Sleepy55 wrote:I raised my EPAP to 11, leaving IPAP at 16 and BPM at 8. For the first time in my three weeks or so of tweaking, I finally saw a reduction of 50% in my Apnea Count!
Very good!
Sleepy55 wrote:I will raise EPAP tonight to 12, leaving all else alone for now, and see what my report looks like tomorrow.
Will be interesting.
Sleepy55 wrote: It could be, as previously suggested, that I don't even need a backup rate.
Yup.
Sleepy55 wrote:When I'm done tweaking downward my apnea count, I'll turn off S/T for one night, turning my machine into a plain Bipap and see what the numbers look like.
I think that's a good idea.
Sleepy55 wrote:Yes, I do feel better on the S/T than I did with my old bipap. I feel better in the A.M. and have less daytime sleepiness in the afternoons.
I may be wrong, but I think it might not be the "S/T" factor that is giving you better results compared to your old bipap. I think raising the EPAP is what's doing that...plain and simple. I think it's the fact that you're working the EPAP up to where it actually
is accomplishing what EPAP is supposed to do...prevent apneas.
If anything, I think the timed back up rate...that BPM setting... is just getting in your way. That is...
if you don't have a significant "centrals" problem, which I still don't think your diagnostic and titration sleep studies would have missed if they really were there. But I could be wrong. I'm sure not a doctor!
When EPAP is high enough to keep the throat open
enough that it's possible for you to actually get another breath
started, that triggers the higher inhale pressure to keep your throat well and truly open for breathing in. Novel idea, isn't it? Actually being able to breathe while we sleep! Getting EPAP right is the most important setting, imho, when it comes to bilevel treatment.
If you still happen to have your old BiPAP Plus around, another interesting experiment would be to see if you just
might feel the same good effects you're enjoying now (including O2 staying up) by using the old machine at, say IPAP 16 / EPAP 12 ( or IPAP 17/ EPAP 13) instead of the EPAP 8 it was set at before. If you try that, I'd turn on bi-flex at 3 or 2..whatever felt smooth to you.
Even better, but unfortunately you don't have one, would be trying a BiPAP Pro II (the old model) or M series BiPAP Auto set for bilevel mode using the new higher EPAP. With either of those machines, a download of the data with Encore Pro or Encore Viewer would show AHI data.
Sounds like you're doing very well, Sleepy55. You may have to change your nickname, soon!