Please... help with my numbers!

General Discussion on any topic relating to CPAP and/or Sleep Apnea.
wakedog
Posts: 6
Joined: Fri Jun 13, 2008 5:18 am

Please... help with my numbers!

Post by wakedog » Thu Jun 19, 2008 11:47 am

I have been using CPAP for a few weeks now. Below is an average nightly event report. I am still more tired than before the therapy. How do my numbers look and do I just need more time? Thanks.

Image


ozij
Posts: 10544
Joined: Fri Mar 18, 2005 11:52 pm

Post by ozij » Thu Jun 19, 2008 11:53 am

It's not the numbers - it's the squiggles.

You're leaking much too much - possibly mouth leaking - and then therapy simply doesn't happen.

You may find other masks more comfortable and less leaky.

O.

_________________
Mask: AirFit™ P10 Nasal Pillow CPAP Mask with Headgear
Additional Comments: Machine: Resmed AirSense10 for Her with Climateline heated hose ; alternating masks.
And now here is my secret, a very simple secret; it is only with the heart that one can see rightly, what is essential is invisible to the eye.
Antoine de Saint-Exupery

Good advice is compromised by missing data
Forum member Dog Slobber Nov. 2023

User avatar
tangents
Posts: 750
Joined: Thu Mar 08, 2007 11:03 am
Location: Cleveland, Ohio

Post by tangents » Thu Jun 19, 2008 12:16 pm

I agree with Ozij.

Good luck, let us know how you make out.

Cathy

User avatar
Snoredog
Posts: 6397
Joined: Sun Mar 19, 2006 5:09 pm

Post by Snoredog » Thu Jun 19, 2008 12:18 pm

me too, SNORING indicates MOUTH breathing.
someday science will catch up to what I'm saying...

Sleeprider
Posts: 1562
Joined: Tue May 06, 2008 5:57 pm
Location: Murrysville, PA

Post by Sleeprider » Thu Jun 19, 2008 12:27 pm

You have a machine capable of clearing events by automatically adjusting pressure. You are running at a steady 9 cmH2O in CPAP mode. Have you ever tried it in Auto mode? Used the Cflex? Anyway, your AHI is in what I understand to be an acceptable range. What was you PSG result for AHI as a comparison?

As far as leaks, I wouldn't consider your numbers to be unusual for that particular mask. Leaks between 50 and 100 L/Min are more consistent with loss of pressure, and they can appear as individual spikes from a dislodged mask to extended peaks called "large leaks". Your leak rate doesn't exceed 50. It may be lower with a different type of mask, like nasal pillows. My leaks averaged 34 to 43 in ComfortClassic nasal mask a few weeks ago before switching to an Swift LT. My current average leak is around 23 to 27.

Disclaimer: I don't know any more than you do.



_________________

CPAPopedia Keywords Contained In This Post (Click For Definition): cflex, swift, CPAP, AHI, Nasal Pillows, auto


_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier
Mask: AirFit™ P10 Nasal Pillow CPAP Mask with Headgear
Additional Comments: Sleepyhead software. Just changed from PRS1 BiPAP Auto DS760TS

alnhwrd
Posts: 731
Joined: Tue Nov 20, 2007 3:58 pm
Location: Hood River, Oregon

Post by alnhwrd » Fri Jun 20, 2008 1:47 am

Your AHI is just 2.0, which is pretty good, but I agree with oz that you are leaking a lot at times. If you get that under control. you could maybe get your numbers down even lower. Best of luck.


wakedog
Posts: 6
Joined: Fri Jun 13, 2008 5:18 am

Post by wakedog » Fri Jun 20, 2008 6:46 am

Thanks for the relies and advice. From my PSG, the sleep center set the pressure at 9 cmH2O with CPAP and C-FLEX=3. I have been using these settings. I did try Auto mode twice early on but I slept very poorly and thought I just needed to get used to the mask and whole CPAP therapy first. As I recall from my PSG, my AHI was 23.0.

I did have a nasal pillow mask at first and just could not sleep with it. I switched to the comfort gel and was able to sleep much better. I do roll around alot and sometimes wakeup with what I do know to be leaks towards my eyes. I have tried tightening the mask, but it becomes too uncomfortable to sleep. I will keep at it though.