New CPAP user! But, AHI is high...

General Discussion on any topic relating to CPAP and/or Sleep Apnea.
jddqr
Posts: 12
Joined: Sun Mar 09, 2008 10:18 am

New CPAP user! But, AHI is high...

Post by jddqr » Sun Mar 09, 2008 10:25 am

Hi everyone! I'm John, a brand new CPAP user. I was diagnosed with mild-to-moderate sleep apnea, and after two sleep studies I was prescribed with an 8 cm CPAP machine. I've been on it for two nights now. The first night I was able to sleep with it for 4 hours, the second night for 6. (I'm using a S8 Elite with a comfort gel mask.)

My concern is that the AHI numbers seem to be really high. For example this is what I had last night.

Avg pressure: 8.2
Leak: 0.10L/s
AHI: 15.6
AI: 6.5
HI: 9.1

The leaks seem (?) to be minimal so I'm wondering what I'm doing wrong.

Any advice? Thanks!!
-John


User avatar
Wulfman
Posts: 12317
Joined: Thu Jul 07, 2005 3:43 pm
Location: Nearest fishing spot

Post by Wulfman » Sun Mar 09, 2008 10:44 am

It would appear that your pressure is too low. AHI should be less than 5.0.

In the first few nights, it's typical that the numbers may be somewhat higher, however, you probably need to bump it up to 10 and see how that works.

Den

(5) REMstar Autos w/C-Flex & (6) REMstar Pro 2 CPAPs w/C-Flex - Pressure Setting = 14 cm.
"Passover" Humidification - ResMed Ultra Mirage FF - Encore Pro w/Card Reader & MyEncore software - Chiroflow pillow
User since 05/14/05

RipVW
Posts: 1684
Joined: Tue Oct 09, 2007 7:29 pm
Location: USA
Contact:

Post by RipVW » Sun Mar 09, 2008 10:56 am

Image
Hello & Welcome, John! Like you, my sleep study (July 2006) set my pressure very low. After more than a year of really not feeling any positive results, I found this forum, learned about data, discovered that my AHI was high (9.5), increased my pressure, and now my AHI averages 0.9. Unlike you, it took me over a year to find the forum. Glad you found it so soon!
Image
"You are getting sleepy . . ."

Check out my chinstrap--> http://cpapchinstraps.com
Image

User avatar
rwalther
Posts: 65
Joined: Fri Dec 28, 2007 11:05 am
Location: Northern California
Contact:

Post by rwalther » Sun Mar 09, 2008 10:59 am

I agree with Wuff-Man !!!

Your pressure is pretty low.

I'm in the same category, my average pressure on my Auto-CPAP fluctuates between 10 and 13 (high is set to 14), and my AHI is around 5.0 over the past month. Depending on a lot of facors, it fluctuate between 1.0 and 8.0.

You're is too high ... IMHO.

Hope you get it sorted out. I'm lucky, in that I'm allowed to change all of my own settings, so I'm not dependant upon anyone else. I watch all of my data, and tweak it on my own. I think I've gotten it to a set of settings that work very well.

Rusty


User avatar
Wulfman
Posts: 12317
Joined: Thu Jul 07, 2005 3:43 pm
Location: Nearest fishing spot

Post by Wulfman » Sun Mar 09, 2008 11:05 am

And, if you want to change your own settings......and don't have the instructions......just ask. They've been posted many, many times on these forum pages.

Den
(5) REMstar Autos w/C-Flex & (6) REMstar Pro 2 CPAPs w/C-Flex - Pressure Setting = 14 cm.
"Passover" Humidification - ResMed Ultra Mirage FF - Encore Pro w/Card Reader & MyEncore software - Chiroflow pillow
User since 05/14/05

jddqr
Posts: 12
Joined: Sun Mar 09, 2008 10:18 am

Thanks for the feedback! But...

Post by jddqr » Sun Mar 09, 2008 4:09 pm

Thanks for the warm welcome and feedback everyone

I took a nap today with a pressure of 10 and still, the same results

Pressure: 10.2
Leak: 0.06L/s
AHI: 15.5
AI: 5.7
HI: 9.8

Gah!!

Please don't tell me I need to push it above 10... I can't believe my prescription would be THAT far off.

Any other advice??

Thanks all!!!


User avatar
Wulfman
Posts: 12317
Joined: Thu Jul 07, 2005 3:43 pm
Location: Nearest fishing spot

Re: Thanks for the feedback! But...

Post by Wulfman » Sun Mar 09, 2008 4:13 pm

jddqr wrote:Thanks for the warm welcome and feedback everyone

I took a nap today with a pressure of 10 and still, the same results

Pressure: 10.2
Leak: 0.06L/s
AHI: 15.5
AI: 5.7
HI: 9.8

Gah!!

Please don't tell me I need to push it above 10... I can't believe my prescription would be THAT far off.

Any other advice??

Thanks all!!!
WHY NOT? I was prescibed a pressure of 18 cm and ended up between 10 and 12. That's a lot farther off than yours may be. I've seen others like that on the forum, too.

It happens.

Den

(5) REMstar Autos w/C-Flex & (6) REMstar Pro 2 CPAPs w/C-Flex - Pressure Setting = 14 cm.
"Passover" Humidification - ResMed Ultra Mirage FF - Encore Pro w/Card Reader & MyEncore software - Chiroflow pillow
User since 05/14/05

User avatar
goose
Posts: 1382
Joined: Sun Mar 11, 2007 7:59 pm
Location: The left coast - CA... If you're not living on the edge, you're taking up too much space!!

Re: Thanks for the feedback! But...

Post by goose » Sun Mar 09, 2008 4:24 pm

jddqr wrote: ....Gah!!

Please don't tell me I need to push it above 10... I can't believe my prescription would be THAT far off.
Mine was the opposite from Wulfman.
I was titrated at 5cm and I'm now at 10. I couldn't breathe at 5cm so set it to 7cm. After a while at that, I asked for and got an Auto CPAP but the range was set from 8-15. The higher pressures caused centrals, so it was backed off. After about 4 months of experimentation with different settings I've settled in at 10 (in CPAP mode) with an average AHI of 2.5 (it's really about 1.2/1.3 over the past month or two, but it's averaged in with my old data).....

So as you can see, the prescriptions are not an exact science. If you change your pressures you should leave it for at least a week (I go for 3-4 weeks) to see what the data shows. If the trend is downward, leave it for another week and check again. It takes some time to zero in on your "magic" number.
Your leak numbers look pretty good....Too many AI's and HI's though....

Fill in your profile so we can see what you're using -- it's easier to help if we know what you've got.
If you don't have one, ask for a data capable machine so you can keep an eye on all the data.

Keep at it -- you'll nail it down in time.
Good luck -- take care
cheers
goose

_________________

CPAPopedia Keywords Contained In This Post (Click For Definition): auto cpap, CPAP, AHI, Prescription


_________________
Humidifier: HC150 Heated Humidifier With Hose, 2 Chambers and Stand
Additional Comments: Also Use ComfortGel (s); Headrest (XL) and a PAP-Cap.
Wars arise from a failure to understand one another's humanness. Instead of summit meetings, why not have families meet for a picnic and get to know each other while the children play together?

-the Dalai Lama

RipVW
Posts: 1684
Joined: Tue Oct 09, 2007 7:29 pm
Location: USA
Contact:

Re: Thanks for the feedback! But...

Post by RipVW » Sun Mar 09, 2008 5:24 pm

jddqr wrote:Please don't tell me I need to push it above 10... I can't believe my prescription would be THAT far off.
My titration pressure was 6.0, though I actually need around 12.0 to get good results.
Image

Check out my chinstrap--> http://cpapchinstraps.com
Image

jddqr
Posts: 12
Joined: Sun Mar 09, 2008 10:18 am

I'll try 12 tonight

Post by jddqr » Sun Mar 09, 2008 7:09 pm

Thanks all, being new to all this is a bit overwhelming, but I'll try a pressure of 12 tonight and see if it makes a difference or not.

Thanks again!

ozij
Posts: 10517
Joined: Fri Mar 18, 2005 11:52 pm

Post by ozij » Mon Mar 10, 2008 4:55 am

You should definitely use the machine and mask for naps - but don't make and decisions based on nap data. Napping is not normal sleep, and can't tell you much about night needs.

Good luck.
O.

_________________
Mask: AirFit™ P10 Nasal Pillow CPAP Mask with Headgear
Additional Comments: Machine: Resmed AirSense10 for Her with Climateline heated hose ; alternating masks.
And now here is my secret, a very simple secret; it is only with the heart that one can see rightly, what is essential is invisible to the eye.
Antoine de Saint-Exupery

Good advice is compromised by missing data
Forum member Dog Slobber Nov. 2023

User avatar
ColinP
Posts: 349
Joined: Wed Nov 21, 2007 7:26 am
Location: South Africa

Post by ColinP » Mon Mar 10, 2008 5:55 am

Personally, I'd be hesitant to make changes based on one night's sleep too - I've been trying CPAP at 10cm for over a mointh now, and have had AHI reading varying from 7.5 to 30.9, but the trend is still slowly coming down - I assume it's as I adjust to things. (the first week the average was 22.1, last week it was 11.6) I'm planning to stay at those settings until the figurs stabilise.

Colin


jddqr
Posts: 12
Joined: Sun Mar 09, 2008 10:18 am

Post by jddqr » Mon Mar 10, 2008 8:52 am

Yep, I think ColinP is right. Last night I was at 12 and guess what? no changes. In fact I had some pain in my chest, presumably from the greater pressure.

I'm going back to 8 and am going to stick with it for a few weeks to see what happens.

Thanks for the advice everyone

User avatar
Wulfman
Posts: 12317
Joined: Thu Jul 07, 2005 3:43 pm
Location: Nearest fishing spot

Post by Wulfman » Mon Mar 10, 2008 8:57 am

jddqr wrote:Yep, I think ColinP is right. Last night I was at 12 and guess what? no changes. In fact I had some pain in my chest, presumably from the greater pressure.

I'm going back to 8 and am going to stick with it for a few weeks to see what happens.

Thanks for the advice everyone
One night may not necessarily show significant results. It's also advantageous to make smller adjustments for longer periods of time. You should monitor the results over a week or longer. Were you looking at one night's data or averages?

Den
(5) REMstar Autos w/C-Flex & (6) REMstar Pro 2 CPAPs w/C-Flex - Pressure Setting = 14 cm.
"Passover" Humidification - ResMed Ultra Mirage FF - Encore Pro w/Card Reader & MyEncore software - Chiroflow pillow
User since 05/14/05

Guest

Post by Guest » Mon Mar 10, 2008 11:37 am

Generally, one night's data.

I'll stick with the 8 for a week and see if the AHI's (especially the AI's) go down or not...