DRONE wrote:Don't worry, DSM. I'm a smart guy. Many gray cells still functioning despite the apnea. And I know this is just a forum. If I had the provider manual, I could read it and answer most of these questions myself. Where can I get one?dsm wrote:Drone,
Whoa - be careful, some of the advice posted is partially correct but enough to get you embarrassed if you try to debate with the DME. I have the provider manual & can give accurate detail of the settings - so don't be afraid to ask before a sit down with the DME. We here at cpaptalk do our best to advise but repeatedly we can get details wrong & confused. We always mean well but we aren't RTs. Make sure you have a lot of input before assuming any one of us is the authority. You will get enough good input quickly to be confident.
Rested Gal uses a Bipap Auto & she will be very helpful in explaining the way it works. She has used hers a lot more than I used mine.
Good luck DSM
Regarding RT's, to tell you the truth, I really don't have a high opinion of the RT's I've met. How can I say this in a charitable way? Let's see -- how about this -- The RT's I've met seem to be under-trained. How's that? Not too harsh? Maybe in your area they are experienced and well-trained, but I haven't met any around here. I've been dealing with my sleep apnea for 5 years (I told you there was a long back-story) and I just can't leave these decisions to other people anymore. They've been mis-handling my therapy for five years now. I've decided that I have to take control if I want to live to see 60. I was pre-Med at Yale until I flunked Calculus, so I figure I have enough smarts to get on top of this situation and make certain I get the therapy I need.
I am hoping that Rested Gal will chime in here, but I only started this thread this afternoon. Hopefully she'll see it sometime in the next few days and give us her thoughts.
It's after midnight here, so I'm going to log off for now. Thanks so much for caring. Really. I'm not saying that to be cute. It's great that you take the time and interest to get involved in these discussions. Thanks. But I'm done for tonight!
Newbie with questions about Respironics BiPAP Auto M
xPAP and Quattro std mask (plus a pad-a-cheek anti-leak strap)
- rested gal
- Posts: 12880
- Joined: Thu Sep 09, 2004 10:14 pm
- Location: Tennessee
Hey, Drone...
Welcome to the board. I didn't see your thread until just now.
If it were me, and I'd been using a regular bilevel machine at 10/7... I'd try the M series BiPAP Auto this way:
auto bilevel (Auto-titrating bipap mode)
Max IPAP 20.0
Min EPAP 7.0
Max Press Sup 8.0
Bi-Flex Setting 3
Autoramp time 0:00
Mask Alert Feature OFF
Auto Off Feature OFF
Split Night Time - Off
Show AHI/Leak Feature ON
Patient Reminder OFF
The machine will start out each time giving you an inhale pressure 2 cms more than the exhale pressure (EPAP setting.)
So, with those settings:
Each time you turn it on, it will begin blowing 9 for inhaling and 7 for exhaling. It will vary both those pressures as needed throughout the night.
It will never let the inhale/exhale move closer together than 2 cms, and will never let them move farther apart from each other than 8 cms (the "max pres sup" setting.)
______________________________
Here are links to my explanation (as I understand it) of the Maximum Pressure Support setting -- like two dogs on a leash.
That setting is called "Max Press Sup" in the M series BiPAP Auto.
It's called "PS" in the older non-M series BiPAP Auto.
Same thing.
viewtopic.php?t=22099
Jul 14, 2007 subject: What is 'Max Press Sup'
viewtopic.php?t=15666
Dec 08, 2006 subject: Question for BiPap users - UPDATED 12/14/2006
________________________________
Explanation of "Bi-Flex":
Mar 08, 2007 subject: BiPAP Bi-Flex Feature
viewtopic.php?p=153809
(That discussion was about a pre-M series BiPAP machine.)
Bi-Flex can be turned on when these BiPAP Auto machines are in the listed operating modes (through the therapy setup menu on the machine) --
BiPAP Auto with Bi-Flex M series
bilevel
auto bilevel
BiPAP Auto with Bi-Flex, older non-M series
BFLE
AbFLE
Welcome to the board. I didn't see your thread until just now.
Send me your email address via a Private Message (the "PM me" button at the bottom of my message here) and you shall have it.DRONE wrote:If I had the provider manual, I could read it and answer most of these questions myself. Where can I get one?
If it were me, and I'd been using a regular bilevel machine at 10/7... I'd try the M series BiPAP Auto this way:
auto bilevel (Auto-titrating bipap mode)
Max IPAP 20.0
Min EPAP 7.0
Max Press Sup 8.0
Bi-Flex Setting 3
Autoramp time 0:00
Mask Alert Feature OFF
Auto Off Feature OFF
Split Night Time - Off
Show AHI/Leak Feature ON
Patient Reminder OFF
The machine will start out each time giving you an inhale pressure 2 cms more than the exhale pressure (EPAP setting.)
So, with those settings:
Each time you turn it on, it will begin blowing 9 for inhaling and 7 for exhaling. It will vary both those pressures as needed throughout the night.
It will never let the inhale/exhale move closer together than 2 cms, and will never let them move farther apart from each other than 8 cms (the "max pres sup" setting.)
______________________________
Here are links to my explanation (as I understand it) of the Maximum Pressure Support setting -- like two dogs on a leash.
That setting is called "Max Press Sup" in the M series BiPAP Auto.
It's called "PS" in the older non-M series BiPAP Auto.
Same thing.
viewtopic.php?t=22099
Jul 14, 2007 subject: What is 'Max Press Sup'
viewtopic.php?t=15666
Dec 08, 2006 subject: Question for BiPap users - UPDATED 12/14/2006
________________________________
Explanation of "Bi-Flex":
Mar 08, 2007 subject: BiPAP Bi-Flex Feature
viewtopic.php?p=153809
(That discussion was about a pre-M series BiPAP machine.)
Bi-Flex can be turned on when these BiPAP Auto machines are in the listed operating modes (through the therapy setup menu on the machine) --
BiPAP Auto with Bi-Flex M series
bilevel
auto bilevel
BiPAP Auto with Bi-Flex, older non-M series
BFLE
AbFLE
ResMed S9 VPAP Auto (ASV)
Humidifier: Integrated + Climate Control hose
Mask: Aeiomed Headrest (deconstructed, with homemade straps
3M painters tape over mouth
ALL LINKS by rested gal:
viewtopic.php?t=17435
Humidifier: Integrated + Climate Control hose
Mask: Aeiomed Headrest (deconstructed, with homemade straps
3M painters tape over mouth
ALL LINKS by rested gal:
viewtopic.php?t=17435
Now you are cooking drone! I bought the Bi Pap auto M out of pocket,set it up with the good help of RG(all blessed by my Doctor) and had my provider pick up their Plus M non data straight crap . You will want the auto blower on a permanent basis. Regards
_________________
| Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier |
| Mask: ResMed AirFit™ F20 Full Face CPAP Mask with Headgear |
-
Guest
Drone,
You've received input here from the best. Sorry if my attempt at explaining how the machine works led you astray.
I really like my BiPap auto. It will probably be a battle, but I too suggest that you try to get one on a permanent basis - you can always de-activate the auto feature if you decide you don't like it.
Off-topic, but my guess is that your battle will be with your DME provider, as they get reimbursed a fixed amount for a Bipap, be it auto or not (and so they are economically incented to provide you with the lowest cost unit you are willing to accept). In my area I have a choice of DME providers. Once I mentioned that I knew that I had freedom of choice, and that if they would not provide me with an auto unit I would find a DME provider that would, they became much more accomodating.
You might also want to visit billmyinsurance.com (affiliated with cpaptalk.com's sponsor cpap.com) to see if your insurance plan would allow you to purchase through them.
Best,
Al
You've received input here from the best. Sorry if my attempt at explaining how the machine works led you astray.
I really like my BiPap auto. It will probably be a battle, but I too suggest that you try to get one on a permanent basis - you can always de-activate the auto feature if you decide you don't like it.
Off-topic, but my guess is that your battle will be with your DME provider, as they get reimbursed a fixed amount for a Bipap, be it auto or not (and so they are economically incented to provide you with the lowest cost unit you are willing to accept). In my area I have a choice of DME providers. Once I mentioned that I knew that I had freedom of choice, and that if they would not provide me with an auto unit I would find a DME provider that would, they became much more accomodating.
You might also want to visit billmyinsurance.com (affiliated with cpaptalk.com's sponsor cpap.com) to see if your insurance plan would allow you to purchase through them.
Best,
Al
Hi, fellas! You too, Rested Gal! Things are kinda slow here in the office this afternoon so I'm back online and just finished reading your comments. Thank you one and all.
Well surprise, surprise, the folks at Respironics actually got back to me with an "800" number that put me through to one of their staff Respiratory Therapists. Had a nice chat. He confirmed that the initial IPAP/EPAP settings of the BiPAP Auto M cannot be greater than 2cm apart, unless you disable the "Auto" mode and instead run it as a simple BiPAP machine. I asked him why the engineers at Respironics designed it this way and he admitted that the RT's at Respironics don't know. They don't think 2cm is a magic number and expect that the next generation of machines will have a little more flexibility.
For forum members who don't quite see why this is such a concern, imagine it this way--pretend you're under water and have to breathe through a straw. If you have a jumbo straw - say an inch in diameter - you can breathe easily in AND out with little effort. You could even sleep with that jumbo straw. But then they take away your jumbo straw and give you instead one of those itty-bitty straws. If you suck real hard, then also breathe out real hard, you'll be OK for a while but eventually your gonna need some air. What you need is a machine that will attach to the other end of the straw and push air through it to your lungs, then turn around and suck it back out again. I'm not talking about sleep apnea here, I'm talking about regular breathing while awake. You're gonna need a "push-pull" air pump.
So when I put on my mask, 10 cm of pressure is powerful enough to drive the air through my itty-bitty nose to fill my lungs. But then I need to exhale! For me to try to breathe out through my nose is hard enough with no mask on, so it's really hard if my CPAP machine is fighting me with 8 cm of back pressure. I've learned to tolerate 6 cm of EPAP but what I could really use is an EPAP setting of 0 cm! I hope this helps you to visualize the problem.
Where does this leave me? If I start at a setting of 8/6, I'll turn blue from not getting enough IPAP air, and if I start with a setting of 10/8, I'll turn blue because I won't be able to empty my lungs between breaths. So it looks like I'm going to have to disable the Auto titration feature and use the machine as a straight BiPAP machine with initial settings of 10/6, then just keep tweaking the knobs until I optimize my results.
I thank all of you for your help. Tomorrow I'm off to the DME to get my machine. It will be the BiPAP Auto M, simply because the DME has a boatload of them to get rid of and my RT has already gotten the insurance company's approval, so why not?
But, hey, here's another question. Does Resmed or one of the other outfits make a machine that'll do what I want it to do?
Well surprise, surprise, the folks at Respironics actually got back to me with an "800" number that put me through to one of their staff Respiratory Therapists. Had a nice chat. He confirmed that the initial IPAP/EPAP settings of the BiPAP Auto M cannot be greater than 2cm apart, unless you disable the "Auto" mode and instead run it as a simple BiPAP machine. I asked him why the engineers at Respironics designed it this way and he admitted that the RT's at Respironics don't know. They don't think 2cm is a magic number and expect that the next generation of machines will have a little more flexibility.
For forum members who don't quite see why this is such a concern, imagine it this way--pretend you're under water and have to breathe through a straw. If you have a jumbo straw - say an inch in diameter - you can breathe easily in AND out with little effort. You could even sleep with that jumbo straw. But then they take away your jumbo straw and give you instead one of those itty-bitty straws. If you suck real hard, then also breathe out real hard, you'll be OK for a while but eventually your gonna need some air. What you need is a machine that will attach to the other end of the straw and push air through it to your lungs, then turn around and suck it back out again. I'm not talking about sleep apnea here, I'm talking about regular breathing while awake. You're gonna need a "push-pull" air pump.
So when I put on my mask, 10 cm of pressure is powerful enough to drive the air through my itty-bitty nose to fill my lungs. But then I need to exhale! For me to try to breathe out through my nose is hard enough with no mask on, so it's really hard if my CPAP machine is fighting me with 8 cm of back pressure. I've learned to tolerate 6 cm of EPAP but what I could really use is an EPAP setting of 0 cm! I hope this helps you to visualize the problem.
Where does this leave me? If I start at a setting of 8/6, I'll turn blue from not getting enough IPAP air, and if I start with a setting of 10/8, I'll turn blue because I won't be able to empty my lungs between breaths. So it looks like I'm going to have to disable the Auto titration feature and use the machine as a straight BiPAP machine with initial settings of 10/6, then just keep tweaking the knobs until I optimize my results.
I thank all of you for your help. Tomorrow I'm off to the DME to get my machine. It will be the BiPAP Auto M, simply because the DME has a boatload of them to get rid of and my RT has already gotten the insurance company's approval, so why not?
But, hey, here's another question. Does Resmed or one of the other outfits make a machine that'll do what I want it to do?
- 05Renegade
- Posts: 10
- Joined: Thu Feb 28, 2008 11:21 pm
- Location: South Bend, IN
I've just started BiPAP at 18/12...what auto settings could you suggest and can I request a copy of that provider's guide also?rested gal wrote:Hey, Drone...
Welcome to the board. I didn't see your thread until just now.
Send me your email address via a Private Message (the "PM me" button at the bottom of my message here) and you shall have it.DRONE wrote:If I had the provider manual, I could read it and answer most of these questions myself. Where can I get one?
If it were me, and I'd been using a regular bilevel machine at 10/7... I'd try the M series BiPAP Auto this way:
auto bilevel (Auto-titrating bipap mode)
Max IPAP 20.0
Min EPAP 7.0
Max Press Sup 8.0
Bi-Flex Setting 3
Autoramp time 0:00
Mask Alert Feature OFF
Auto Off Feature OFF
Split Night Time - Off
Show AHI/Leak Feature ON
Patient Reminder OFF
The machine will start out each time giving you an inhale pressure 2 cms more than the exhale pressure (EPAP setting.)
So, with those settings:
Each time you turn it on, it will begin blowing 9 for inhaling and 7 for exhaling. It will vary both those pressures as needed throughout the night.
It will never let the inhale/exhale move closer together than 2 cms, and will never let them move farther apart from each other than 8 cms (the "max pres sup" setting.)
_________________
| Mask: Opus 360 Nasal Pillow CPAP Mask with Headgear |
| Additional Comments: Other masks: Respironics Optilife, F&P FlexiFit HC432 Full Face |
Hey, 05Renegade, I see you're a newbie like me. But maybe your new to forums in general. Here's the deal when in a forum. If you have a new problem or a question to ask, you should go ahead and start a new topic of your own. The people can reply directly to you instead of having two conversations going inside of the same thread. When you change the topic of a thread, it's called "hijacking the thread" and people lose track of who is saying what to whom. Not good.
So your question is a good one, you just need to start a new topic to ask it.
_________________
CPAPopedia Keywords Contained In This Post (Click For Definition): newbie
So your question is a good one, you just need to start a new topic to ask it.
_________________
CPAPopedia Keywords Contained In This Post (Click For Definition): newbie
- 05Renegade
- Posts: 10
- Joined: Thu Feb 28, 2008 11:21 pm
- Location: South Bend, IN
The "hijacking" wasnt intentional...I've been around since before there was an Internet (TheSource and QuantumLink anyone?), just a newb to CPAP and OSA.
Since the thread is titled "Newbie with questions about Respironics BiPAP Auto M", which I am and which I have, that it would be more appropriate to ask here than starting a new and/or duplicate thread.
Sorry to offend...I'll get my answers elsewhere.
_________________
CPAPopedia Keywords Contained In This Post (Click For Definition): respironics, bipap, CPAP, auto
Since the thread is titled "Newbie with questions about Respironics BiPAP Auto M", which I am and which I have, that it would be more appropriate to ask here than starting a new and/or duplicate thread.
Sorry to offend...I'll get my answers elsewhere.
_________________
CPAPopedia Keywords Contained In This Post (Click For Definition): respironics, bipap, CPAP, auto
_________________
| Mask: Opus 360 Nasal Pillow CPAP Mask with Headgear |
| Additional Comments: Other masks: Respironics Optilife, F&P FlexiFit HC432 Full Face |
Not to worry about hijacking a thread here - happens all the time and it's my impression that nobody worries about it!
We get some really good conversations going that started out on one topic and evolved from there.
Mindy
We get some really good conversations going that started out on one topic and evolved from there.
Mindy
_________________
| Mask: Swift™ FX Bella Nasal Pillow CPAP Mask with Headgears |
| Humidifier: S9™ Series H5i™ Heated Humidifier with Climate Control |
| Additional Comments: Pressure 7-11. Padacheek |
"Life isn't about waiting for the storm to pass, it's about learning how to dance in the rain."
--- Author unknown
--- Author unknown
- rested gal
- Posts: 12880
- Joined: Thu Sep 09, 2004 10:14 pm
- Location: Tennessee
05Renegade, it was perfectly fine for you to ask your question here in this thread. I'll be happy to send you the provider manual for the M BiPAP Auto. Just PM me your email address.
I'm not a doctor. But if it were me, and I were prescribed bipap at 18 IPAP / 12 EPAP, I'd set the bipap auto exactly as described in my post to Drone except I'd set it for these I/E pressures:
Max IPAP 20
Min EPAP 12
I'd check online cpap stores to find the lowest price for the Encore Viewer software. If I found it priced lower than at cpap.com, I'd give cpap.com a chance to price match it, which I think they will.
I'd buy the necessary card reader from this Hong Kong store (for about $31 including shipping to the U.S.):
http://www.two-factor.com/shop/index.php?productID=115
__________
Drone, I sent a reply to your PM, but I don't think you've seen it yet. You might want to check your PMs (button at top right of this webpage.)
I'm not a doctor. But if it were me, and I were prescribed bipap at 18 IPAP / 12 EPAP, I'd set the bipap auto exactly as described in my post to Drone except I'd set it for these I/E pressures:
Max IPAP 20
Min EPAP 12
I'd check online cpap stores to find the lowest price for the Encore Viewer software. If I found it priced lower than at cpap.com, I'd give cpap.com a chance to price match it, which I think they will.
I'd buy the necessary card reader from this Hong Kong store (for about $31 including shipping to the U.S.):
http://www.two-factor.com/shop/index.php?productID=115
__________
Drone, I sent a reply to your PM, but I don't think you've seen it yet. You might want to check your PMs (button at top right of this webpage.)
ResMed S9 VPAP Auto (ASV)
Humidifier: Integrated + Climate Control hose
Mask: Aeiomed Headrest (deconstructed, with homemade straps
3M painters tape over mouth
ALL LINKS by rested gal:
viewtopic.php?t=17435
Humidifier: Integrated + Climate Control hose
Mask: Aeiomed Headrest (deconstructed, with homemade straps
3M painters tape over mouth
ALL LINKS by rested gal:
viewtopic.php?t=17435
- 05Renegade
- Posts: 10
- Joined: Thu Feb 28, 2008 11:21 pm
- Location: South Bend, IN
Done, and thanks for the help!rested gal wrote:05Renegade, it was perfectly fine for you to ask your question here in this thread. I'll be happy to send you the provider manual for the M BiPAP Auto. Just PM me your email address.
Right now everything is going well but I'd like to try going auto and see if things get better.
Also, thanks to everyone else for the encouragement and kind comments.
_________________
CPAPopedia Keywords Contained In This Post (Click For Definition): bipap, auto
_________________
| Mask: Opus 360 Nasal Pillow CPAP Mask with Headgear |
| Additional Comments: Other masks: Respironics Optilife, F&P FlexiFit HC432 Full Face |
Last edited by 05Renegade on Thu Mar 06, 2008 12:21 pm, edited 1 time in total.