
It's labled MI or MT something. Looks like a mic to me...? What am I missing?
these machines (pretty much all of them) measure at a minimum, 2 main characteristics: Volume FLOW and PRESSURE, they are two different things. For that they use 2 separate circuits found on nearly all xpaps.dsm wrote:Snores,
Take a look at these photos, these are not microphones. They are pressure




well pull the hose off that thermal device and see what happens to the pressure.dsm wrote:PS - Snores
The thermistor used in the fan housing is there as a temperature overload device. It serves no other purpose other than to shut down the xPAP should the motor fail in a way that createss heat. Thermistor means thermal resistor - varies resistance based on temperature.
So if the motor were to overheat (electronic failure, fan jam etc:) the machine *does not want to* allow heated air to find its way to the user. It shuts down. *If* it has a thermistor protection circuit.
D
trust me, I will FIND the actual Resmed document where they refer to the "MIC" and pressure transducer being of the same thing and I'll put it on-line, I only have about a thousand documents to search through but I read it a few weeks back.dsm wrote:Snores,
Again, we are heading down the same track
Actually, most modern mics (except some very expensive models) use a piezo electric crystal to convert sound to electrical energy. As you pointed out - using the principles of a speaker cone (but more flattened as in a diaphram).
Pressure sensors also use piezoelectric devices but constructed in a quite different way from a sound detector.
Here is one example of heavy duty piezoelectric pressure sensors. But these bear little resemblance to a mic.
http://www.davidson.com.au/products/pre ... theory.asp
I may find a diagram of the actual chips used in the xPAPs shown.
D
The Flattening Index is a measure of silent inspiratory airflow limitation. Flow limitation with loud snoring is handled by the snore detector. When a patient snores, sound is generated and the inspiratory flow/time curve is distorted by the frequency of the sound.
Snoredog wrote:trust me, I will FIND the actual Resmed document where they refer to the "MIC" and pressure transducer being of the same thing and I'll put it on-line, I only have about a thousand documents to search through but I read it a few weeks back.dsm wrote:Snores,
Again, we are heading down the same track
Actually, most modern mics (except some very expensive models) use a piezo electric crystal to convert sound to electrical energy. As you pointed out - using the principles of a speaker cone (but more flattened as in a diaphram).
Pressure sensors also use piezoelectric devices but constructed in a quite different way from a sound detector.
Here is one example of heavy duty piezoelectric pressure sensors. But these bear little resemblance to a mic.
http://www.davidson.com.au/products/pre ... theory.asp
I may find a diagram of the actual chips used in the xPAPs shown.
D
Here is what another statement from the Resmed Clinical Manual, 30846, page 9:The Flattening Index is a measure of silent inspiratory airflow limitation. Flow limitation with loud snoring is handled by the snore detector. When a patient snores, sound is generated and the inspiratory flow/time curve is distorted by the frequency of the sound.
Ok, we're hair splitting, but at least now I see why. Mics don't measure anything. They convert pressure modulations to electricity. The computer or DSP or whatever the mic is plugged into can measure all it wants, but that's why we're going down this path.dsm wrote:Mics are devices that measure sound in decibels (not pressure).
that is because the airflow is looked at (by the machine algorithm) as a sine wave like you would see on an oscilloscope. The breath or flow is monitored by assigning numeric values to those waveforms (like shown in the center blue line in the various forms of SDB events below).blarg wrote:DSM, you seem to be saying that pressure sensors are different from mics because they measure outside the typical frequencies we consider sound. That's all fine and dandy.
I guess I'm just confused as to why they'd need to measure outside sound frequencies when detecting snores, because we hear them....
