CPAP Humidifier Pushing Water Drops Through Hose

General Discussion on any topic relating to CPAP and/or Sleep Apnea.
JeffNC
Posts: 19
Joined: Wed Apr 06, 2022 5:57 am

CPAP Humidifier Pushing Water Drops Through Hose

Post by JeffNC » Sun Aug 17, 2025 7:29 am

I have a Resmed AirSense 10 CPAP with a humidifier; I need and depend upon this feature.
Problem: the humidifier has now started passing water droplets through the hose (hence into my nose) while I am sleeping (or TRYING to sleep).
Has anyone else eve had this problem? If so, is it "solvable" at home, rather than having to send the CPAP out for repair or having to replace it?

User avatar
LSAT
Posts: 13395
Joined: Sun Nov 16, 2008 10:11 am
Location: SE Wisconsin

Re: CPAP Humidifier Pushing Water Drops Through Hose

Post by LSAT » Sun Aug 17, 2025 8:14 am

You may have a problem with condensation...It's called "rainout". Condensation is formed when your room temperature is lower than the air in the hose, If you are using a heated hose, set it to auto. Check out this web site... https://www.youtube.com/watch?v=ctw73sdPU6k.

User avatar
zonker
Posts: 11431
Joined: Fri Jun 19, 2015 4:36 pm

Re: CPAP Humidifier Pushing Water Drops Through Hose

Post by zonker » Sun Aug 17, 2025 5:45 pm

unfortunately this is not a matter of sending it off and having it repaired. it's a very individual thing. I went though this last year.

what it boils down to is you having to change settings on your machine. you know how to do that, right?
"Age is not an accomplishment and youth is not a sin"-Robert A. Heinlein
Oscar-Win
https://www.apneaboard.com/OSCAR/OSCAR-1.5.1-Win64.exe
Oscar-Mac
https://www.apneaboard.com/OSCAR/OSCAR-1.5.1.dmg

JeffNC
Posts: 19
Joined: Wed Apr 06, 2022 5:57 am

Re: CPAP Humidifier Pushing Water Drops Through Hose

Post by JeffNC » Mon Aug 18, 2025 8:28 am

Last might, I tried not filling the water reservoir to capacity; really, I put very little in at all. Guess what? NO water got into the hose or the mask, I slept for 9 peaceful hours and there was STILL a little water left in it this morning! Whether it "should" operate this way or not, at least I've learned "the trick"!

Thanks for the suggestions!

Jeff

User avatar
ChicagoGranny
Posts: 15471
Joined: Sun Jan 29, 2012 1:43 pm
Location: USA

Re: CPAP Humidifier Pushing Water Drops Through Hose

Post by ChicagoGranny » Mon Aug 18, 2025 11:54 am

JeffNC wrote:
Mon Aug 18, 2025 8:28 am
at least I've learned "the trick"!
Trick? You found something that worked for ONE night for YOU.

I hope it continues to work for you throughout your life.

I would not recommend this to anyone else.
"It's not the number of breaths we take, it's the number of moments that take our breath away."

Cuando cuentes cuentos, cuenta cuántas cuentos cuentas.

User avatar
Dog Slobber
Posts: 4320
Joined: Thu Feb 15, 2018 2:05 pm
Location: Ontario, Canada

Re: CPAP Humidifier Pushing Water Drops Through Hose

Post by Dog Slobber » Mon Aug 18, 2025 1:37 pm

Last night I had pizza for dinner.

I did not experience any rain-out. In fact, I can say with 100% authority, I have never experienced rain-out, on nights that I have had pizza.

I like my trick better than yours.
JeffNC wrote:
Mon Aug 18, 2025 8:28 am
Last might, I tried not filling the water reservoir to capacity; really, I put very little in at all. Guess what? NO water got into the hose or the mask, I slept for 9 peaceful hours and there was STILL a little water left in it this morning! Whether it "should" operate this way or not, at least I've learned "the trick"!
https://en.wikipedia.org/wiki/Human_sacrifice
Battery Backup: EcoFlow Delta 2

User avatar
zonker
Posts: 11431
Joined: Fri Jun 19, 2015 4:36 pm

Re: CPAP Humidifier Pushing Water Drops Through Hose

Post by zonker » Mon Aug 18, 2025 4:21 pm

Dog Slobber wrote:
Mon Aug 18, 2025 1:37 pm
Last night I had pizza for dinner.

I did not experience any rain-out. In fact, I can say with 100% authority, I have never experienced rain-out, on nights that I have had pizza.

I like my trick better than yours.

I don't. yer trick costs more money!
"Age is not an accomplishment and youth is not a sin"-Robert A. Heinlein
Oscar-Win
https://www.apneaboard.com/OSCAR/OSCAR-1.5.1-Win64.exe
Oscar-Mac
https://www.apneaboard.com/OSCAR/OSCAR-1.5.1.dmg

User avatar
ChicagoGranny
Posts: 15471
Joined: Sun Jan 29, 2012 1:43 pm
Location: USA

Re: CPAP Humidifier Pushing Water Drops Through Hose

Post by ChicagoGranny » Mon Aug 18, 2025 4:54 pm

Make mine a thin-crust margarita.


Image
"It's not the number of breaths we take, it's the number of moments that take our breath away."

Cuando cuentes cuentos, cuenta cuántas cuentos cuentas.

Aretino
Posts: 109
Joined: Sun Mar 15, 2020 2:07 pm

Re: CPAP Humidifier Pushing Water Drops Through Hose

Post by Aretino » Tue Aug 19, 2025 8:51 am

Put your CPAP machine lower than your head.

_________________
Machine: ResMed AirCurve 10 ASV Machine with Heated Humidifier
Mask: AirFit™ P10 Nasal Pillow CPAP Mask with Headgear
Additional Comments: Started 2005 with a Remstar machine. Then a 2020 Dreamstation 1. Now a Resmed Aircurve ST

User avatar
zonker
Posts: 11431
Joined: Fri Jun 19, 2015 4:36 pm

Re: CPAP Humidifier Pushing Water Drops Through Hose

Post by zonker » Tue Aug 19, 2025 12:54 pm

Aretino wrote:
Tue Aug 19, 2025 8:51 am
Put your CPAP machine lower than your head.
that has been found to be ineffective.
"Age is not an accomplishment and youth is not a sin"-Robert A. Heinlein
Oscar-Win
https://www.apneaboard.com/OSCAR/OSCAR-1.5.1-Win64.exe
Oscar-Mac
https://www.apneaboard.com/OSCAR/OSCAR-1.5.1.dmg