Respiratory Rate of 7 on ASV - UARS x 13 years
- Miss Emerita
- Posts: 3766
- Joined: Sun Nov 04, 2018 8:07 pm
Re: Respiratory Rate of 7 on ASV - UARS x 13 years
Applecheeks, I'm so interested to see your graph. I don't think I've seen that particular pattern before. (Doesn't mean there's anything wrong with it!)
Here's what my usual looks like. I do think the oscillations between breaths are cardioballistic artifacts. It makes sense to me that they'd telegraph only when there wasn't any airflow through the airway.
Here's what my usual looks like. I do think the oscillations between breaths are cardioballistic artifacts. It makes sense to me that they'd telegraph only when there wasn't any airflow through the airway.
_________________
| Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier |
| Mask: Bleep DreamPort CPAP Mask Solution |
Oscar software is available at https://www.sleepfiles.com/OSCAR/
Re: Respiratory Rate of 7 on ASV - UARS x 13 years
Because what you're showing there is not normal sleep breathing.Applecheeks wrote: ↑Sat May 28, 2022 9:56 pmThis is typical of my asleep pattern with no flow limitations, @5 cm pressure, 1 cm EPR, no snore, no leaks. Doesn't seem at all like the OSCAR logo.Miss Emerita wrote: ↑Sat May 28, 2022 11:58 amThe idealized flow trace inside the O in the Oscar emblem is much closer to reality than the trace in the video.
typical breathing patterns.jpg
Get OSCAR
Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.
Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.
Re: Respiratory Rate of 7 on ASV - UARS x 13 years
Yes, THAT is an example of normal sleep breathing.Miss Emerita wrote: ↑Sun May 29, 2022 12:09 pmApplecheeks, I'm so interested to see your graph. I don't think I've seen that particular pattern before. (Doesn't mean there's anything wrong with it!)
Here's what my usual looks like. I do think the oscillations between breaths are cardioballistic artifacts. It makes sense to me that they'd telegraph only when there wasn't any airflow through the airway.
Normal flow trace.png
Get OSCAR
Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.
Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.
- Applecheeks
- Posts: 86
- Joined: Fri Dec 24, 2021 10:07 am
Re: Respiratory Rate of 7 on ASV - UARS x 13 years
It would be helpful if you would describe the distinction that makes my pattern abnormal as compared with Miss E.
By the way, I notice that we are using two different machines, and I have no idea of Miss E’s pressures. Does that make any difference?
_________________
| Machine: AirSense 11 Autoset |
| Mask: AirFit™ F20 For Her Full Face CPAP Mask with Headgear |
- Miss Emerita
- Posts: 3766
- Joined: Sun Nov 04, 2018 8:07 pm
Re: Respiratory Rate of 7 on ASV - UARS x 13 years
My pressure when I inhale is 10; when I exhale, 5.
_________________
| Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier |
| Mask: Bleep DreamPort CPAP Mask Solution |
Oscar software is available at https://www.sleepfiles.com/OSCAR/
- Applecheeks
- Posts: 86
- Joined: Fri Dec 24, 2021 10:07 am
Re: Respiratory Rate of 7 on ASV - UARS x 13 years
Hi Miss E,
And you are using a bipap (I’m not knowledgeable with the machine in your profile)
My machine is a Resmed Airsense 11 autoset elite in Apap mode with EPR set for 1 full time.
And you are using a bipap (I’m not knowledgeable with the machine in your profile)
My machine is a Resmed Airsense 11 autoset elite in Apap mode with EPR set for 1 full time.
_________________
| Machine: AirSense 11 Autoset |
| Mask: AirFit™ F20 For Her Full Face CPAP Mask with Headgear |
Re: Respiratory Rate of 7 on ASV - UARS x 13 years
Me neitherMiss Emerita wrote: ↑Sun May 29, 2022 12:09 pmApplecheeks, I'm so interested to see your graph. I don't think I've seen that particular pattern before. (Doesn't mean there's anything wrong with it!)
The pressure and machine type do not make a difference.Applecheeks wrote: ↑Sun May 29, 2022 4:38 pmIt would be helpful if you would describe the distinction that makes my pattern abnormal as compared with Miss E.
By the way, I notice that we are using two different machines, and I have no idea of Miss E’s pressures. Does that make any difference?
The exceptional thing about your pattern is the long, shallow exhale.
In human beings, inhaling is an effort, exhaling is simple a "letting go" - so quite a bit of air rushes out immediately, which is the pointed dip you can see in Miss E.'s graph, and in the Oscar logo. Your exhale has no dip, is seems relatively shallow compared to most graphs.
_________________
| Mask: AirFit™ P10 Nasal Pillow CPAP Mask with Headgear |
| Additional Comments: Machine: Resmed AirSense10 for Her with Climateline heated hose ; alternating masks. |
And now here is my secret, a very simple secret; it is only with the heart that one can see rightly, what is essential is invisible to the eye.
Antoine de Saint-Exupery
Good advice is compromised by missing data
Forum member Dog Slobber Nov. 2023
Antoine de Saint-Exupery
Good advice is compromised by missing data
Forum member Dog Slobber Nov. 2023
Re: Respiratory Rate of 7 on ASV - UARS x 13 years
If you can't *SEE* the difference then I'm quite sure nobody can describe it to you.Applecheeks wrote: ↑Sun May 29, 2022 4:38 pmIt would be helpful if you would describe the distinction that makes my pattern abnormal as compared with Miss E.
No.Applecheeks wrote: ↑Sun May 29, 2022 4:38 pmBy the way, I notice that we are using two different machines, and I have no idea of Miss E’s pressures. Does that make any difference?
Get OSCAR
Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.
Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.
- Applecheeks
- Posts: 86
- Joined: Fri Dec 24, 2021 10:07 am
Re: Respiratory Rate of 7 on ASV - UARS x 13 years
So what does the non-typical exhalation curve mean and what might cause this:ozij wrote: ↑Sun May 29, 2022 10:24 pmThe exceptional thing about your pattern is the long, shallow exhale.
In human beings, inhaling is an effort, exhaling is simple a "letting go" - so quite a bit of air rushes out immediately, which is the pointed dip you can see in Miss E.'s graph, and in the Oscar logo. Your exhale has no dip, is seems relatively shallow compared to most graphs.
1) I'm not human? (haha)
2) there is some restriction in the airway that is preventing the sudden out rush of air (but that restriction is not seen on inhalation)
3) For some reason I am not "letting go", but controlling my chest muscles rate of relaxation.
Your speculation?
_________________
| Machine: AirSense 11 Autoset |
| Mask: AirFit™ F20 For Her Full Face CPAP Mask with Headgear |
- Applecheeks
- Posts: 86
- Joined: Fri Dec 24, 2021 10:07 am
Re: Respiratory Rate of 7 on ASV - UARS x 13 years
I SEE the difference alright, and @ozij did a good job in describing the non-typical part and why it is non-typical. His answer was helpful.
On the other hand, your answer implies that I am stupid because I can't "SEE". Being cranky and dismissive seems to be part of your sthick, your performance. That undermines your credibility - do you realize that? No need to reply to me.
_________________
| Machine: AirSense 11 Autoset |
| Mask: AirFit™ F20 For Her Full Face CPAP Mask with Headgear |
Re: Respiratory Rate of 7 on ASV - UARS x 13 years
Applecheeks....how about starting a new separate thread with your questions about your flow rate/breathing?
That one little snippet doesn't look like my asleep breathing flow rate either but I haven't had time to think on it very much to figure out what the deal is or if it is even a problem or not.
Is that the same pattern that you see all night long when you are asleep or does it ever change?
I would like to keep your problem separate in this situation for various reasons. You could link back to your posts here as part of the first post.
When you do a separate post how about including a typical night's overall detailed report in the traditional format and then include the snippet.
Do you know the "normal" routine? If not I show examples here.
http://www.cpaptalk.com/viewtopic/t1585 ... eview.html
That one little snippet doesn't look like my asleep breathing flow rate either but I haven't had time to think on it very much to figure out what the deal is or if it is even a problem or not.
Is that the same pattern that you see all night long when you are asleep or does it ever change?
I would like to keep your problem separate in this situation for various reasons. You could link back to your posts here as part of the first post.
When you do a separate post how about including a typical night's overall detailed report in the traditional format and then include the snippet.
Do you know the "normal" routine? If not I show examples here.
http://www.cpaptalk.com/viewtopic/t1585 ... eview.html
_________________
| Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier |
| Additional Comments: Mask Bleep Eclipse https://bleepsleep.com/the-eclipse/ |
I may have to RISE but I refuse to SHINE.
Re: Respiratory Rate of 7 on ASV - UARS x 13 years
I'm don't have the knowledge to make a reasonable speculation.Applecheeks wrote: ↑Mon May 30, 2022 6:50 amSo what does the non-typical exhalation curve mean and what might cause this:ozij wrote: ↑Sun May 29, 2022 10:24 pmThe exceptional thing about your pattern is the long, shallow exhale.
In human beings, inhaling is an effort, exhaling is simple a "letting go" - so quite a bit of air rushes out immediately, which is the pointed dip you can see in Miss E.'s graph, and in the Oscar logo. Your exhale has no dip, is seems relatively shallow compared to most graphs.
1) I'm not human? (haha)
2) there is some restriction in the airway that is preventing the sudden out rush of air (but that restriction is not seen on inhalation)
3) For some reason I am not "letting go", but controlling my chest muscles rate of relaxation.
Your speculation?
Here's what I can say:
You stated that's your typical "sleeping" breathing pattern, at a pressure of 5 and EPR of 1.
Could it be you were not sleeping? Do you ever see the more typical patterns?
When I saw your breathing pattern, I searched your messages to see if you have other health conditions. You did mention some. I don't know if and how what you mentioned is related to that breathing pattern. But I wonder: Do you actually use the APAP to treat sleep apnea i.e. sleep induced breathing stops? Or do you have another condition for which continuous positive air pressure may be necessary to help your breathing / oxygen exchange at night?
I know some medications change our breathing - I certainly don't know which do what and how whatever they do looks on the flow rate graph.
That should be "her answer"....
_________________
| Mask: AirFit™ P10 Nasal Pillow CPAP Mask with Headgear |
| Additional Comments: Machine: Resmed AirSense10 for Her with Climateline heated hose ; alternating masks. |
And now here is my secret, a very simple secret; it is only with the heart that one can see rightly, what is essential is invisible to the eye.
Antoine de Saint-Exupery
Good advice is compromised by missing data
Forum member Dog Slobber Nov. 2023
Antoine de Saint-Exupery
Good advice is compromised by missing data
Forum member Dog Slobber Nov. 2023
Re: Respiratory Rate of 7 on ASV - UARS x 13 years
At this point I don't have enough knowledge either which is why I asked for a new thread so we can isolate this question and maybe get attention of someone else who might have the knowledge.
I also don't know if maybe the machine not being able to go below 4 cm is somehow showing us exhale to that 4 cm and simply can't show less.....and 5 with EPR 1 does mean 4 cm.
I am wondering what the flow rate would look like if there weren't those restrictions????? If there were more room available for movement.
Where's SWS when we want him?
_________________
| Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier |
| Additional Comments: Mask Bleep Eclipse https://bleepsleep.com/the-eclipse/ |
I may have to RISE but I refuse to SHINE.
- Applecheeks
- Posts: 86
- Joined: Fri Dec 24, 2021 10:07 am
Re: Respiratory Rate of 7 on ASV - UARS x 13 years
If I define sleeping as "in my partner has to shake me to get me to respond" - yes I was clearly sleeping.
Yes, I do, but only at higher pressures. Those higher pressures having been initiated by "flow limitations" (distorted flow curves).Do you ever see the more typical patterns?
Longish story, I was sent for a home sleep study at the insistence of my electrophysiologist (EP is treating me for AFIB). Pulmonologist who interpreted the home study initially recommended NO treatment in spite of knowing that I had occasional AFIB - therefore no PAP was ordered. The EP guy told the Pul guy that Apnea causes AFIB and insisted that the Pul guy order APAP. That's how I've been on APAP for the last six months or so.Do you actually use the APAP to treat sleep apnea i.e. sleep induced breathing stops?
By the way, the frequency of AFIB events (1 to 2 a month) has not changed in the past six months. For whatever it's worth, my OSCAR average for the past six months is : AHI 2.4, OA 0.37, CA 0.22, and HI is 1.75 (so my bias seems to be toward shallow breathing). I also don't see any abnormal sleep days just prior to an AFIB episode, and almost all of my AFIB events are occurring mid-day so it appears to my way of thinking that any apneas do not (for me) trigger an AFIB event.
So the short answer is: I'm not sure why I'm "actually" using APAP.
_________________
| Machine: AirSense 11 Autoset |
| Mask: AirFit™ F20 For Her Full Face CPAP Mask with Headgear |
- Applecheeks
- Posts: 86
- Joined: Fri Dec 24, 2021 10:07 am
Re: Respiratory Rate of 7 on ASV - UARS x 13 years
Just saw you request, I was off typing a reply to @ozij.
I'd be glad to start a new thread - I'll begin with my breath pattern and add (by link) the conversation that follows. Will that do it?
_________________
| Machine: AirSense 11 Autoset |
| Mask: AirFit™ F20 For Her Full Face CPAP Mask with Headgear |