Respiratory Rate of 7 on ASV - UARS x 13 years

General Discussion on any topic relating to CPAP and/or Sleep Apnea.
dataq1
Posts: 832
Joined: Sat Dec 24, 2005 8:09 am
Location: Northeast Ohio

Re: Respiratory Rate of 7 on ASV - UARS x 13 years

Post by dataq1 » Fri May 27, 2022 1:54 pm

This is a request for clarification on breath flow patterns that was raised earlier in this thread.
It was suggested that a sinusoidal flow rate pattern is characteristic of "awake " breathing, but not of "asleep" breathing.( see viewtopic/t184231/Respiratory-Rate-of-7 ... l#p1410330 and surrounding discussion.
However, Resmed's videos that are trying to discuss (for example) flow limitations appears to begin with a flow-rate pattern (during sleep) that looks like this:
Flow limitation_002.jpg
Flow limitation_002.jpg (48.13 KiB) Viewed 1416 times
This screen grab seems to show a unobstructed "asleep" flow morphology that is sinusoidal.

So what I'm trying to understand is this ideal (and untreated) flow pattern during sleep sinusoidal nor not .
"THE INFORMATION PROVIDED ON CPAPTALK.COM IS NOT INTENDED NOR RECOMMENDED AS A SUBSTITUTE FOR PROFESSIONAL MEDICAL ADVICE."

Tec5
Posts: 173
Joined: Mon Mar 28, 2022 10:32 am

Re: Respiratory Rate of 7 on ASV - UARS x 13 years

Post by Tec5 » Fri May 27, 2022 10:06 pm

I don't have an answer for data, but it's an interesting question.
Particularly because the resmed algorithm uses the relative deformation of flow rate curve as a principle in a determination of flow limitations.

Palerider seemed to be confident that the balanced (sinusoidal) flow rate breaths is only characteristic of breathing while awake, and not expected while sleeping. Maybe that's not what he meant, but he seemed to be asserting that, and no one disputed it.

Another question might be: what is the limiting factor (tongue muscles relaxing, uvula flapping, mucus plugs etc) that shows up only on the inspiration side of the breath cycle ? In other words, why wouldn't the expiration flowrate be a mirror image of the inspiration flowrate?
I am neither a physician nor a lawyer, so DO NOT rely on me for professional medical or legal advice.

User avatar
palerider
Posts: 32299
Joined: Wed Dec 16, 2009 5:43 pm
Location: Dallas(ish).

Re: Respiratory Rate of 7 on ASV - UARS x 13 years

Post by palerider » Fri May 27, 2022 10:25 pm

Tec5 wrote:
Fri May 27, 2022 10:06 pm
Another question might be: what is the limiting factor (tongue muscles relaxing, uvula flapping, mucus plugs etc) that shows up only on the inspiration side of the breath cycle ? In other words, why wouldn't the expiration flowrate be a mirror image of the inspiration flowrate?
Ever tried sucking air through a tubular balloon?

Try it, get one of those balloons that they make balloon animals with, cut the end off, try sucking air through it, report back.

Compare and contrast that with blowing through the same one.

Report your findings.

_________________
Mask: Bleep DreamPort CPAP Mask Solution
Additional Comments: S9 VPAP Auto
Get OSCAR

Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.

User avatar
Rubicon
Posts: 1756
Joined: Sun Feb 20, 2022 6:59 am

Re: Respiratory Rate of 7 on ASV - UARS x 13 years

Post by Rubicon » Sat May 28, 2022 2:39 am

Tec5 wrote:
Fri May 27, 2022 10:06 pm
In other words, why wouldn't the expiration flowrate be a mirror image of the inspiration flowrate?
Because normally, inspiration is active while expiration is passive, resulting in an Inspiratory: Expiratory (I:E) Ratio of ~1.0 : 2.0.
Freeze this moment a little bit longer.
Make each sensation a little bit stronger.
Experience slips away.

User avatar
ozij
Posts: 10451
Joined: Fri Mar 18, 2005 11:52 pm

Re: Respiratory Rate of 7 on ASV - UARS x 13 years

Post by ozij » Sat May 28, 2022 3:33 am

Rubicon wrote:
Sat May 28, 2022 2:39 am
Because normally, inspiration is active while expiration is passive, resulting in an Inspiratory: Expiratory (I:E) Ratio of ~1.0 : 2.0.
Ummm ....

I:E Ratio of what measure? Flowrate? and if so, I'd be grateful for an explanation of what is "flow" and what is "rate" in that measure.

_________________
Mask: AirFit™ P10 Nasal Pillow CPAP Mask with Headgear
Additional Comments: Machine: Resmed AirSense10 for Her with Climateline heated hose ; alternating masks.
And now here is my secret, a very simple secret; it is only with the heart that one can see rightly, what is essential is invisible to the eye.
Antoine de Saint-Exupery

Good advice is compromised by missing data
Forum member Dog Slobber Nov. 2023

User avatar
Rubicon
Posts: 1756
Joined: Sun Feb 20, 2022 6:59 am

Re: Respiratory Rate of 7 on ASV - UARS x 13 years

Post by Rubicon » Sat May 28, 2022 3:40 am

ozij wrote:
Sat May 28, 2022 3:33 am
I:E Ratio of what measure? Flowrate?
Time.

Consequently, as volume is constant, flow rate becomes variable.
Freeze this moment a little bit longer.
Make each sensation a little bit stronger.
Experience slips away.

User avatar
ozij
Posts: 10451
Joined: Fri Mar 18, 2005 11:52 pm

Re: Respiratory Rate of 7 on ASV - UARS x 13 years

Post by ozij » Sat May 28, 2022 3:45 am

Thanks.

_________________
Mask: AirFit™ P10 Nasal Pillow CPAP Mask with Headgear
Additional Comments: Machine: Resmed AirSense10 for Her with Climateline heated hose ; alternating masks.
And now here is my secret, a very simple secret; it is only with the heart that one can see rightly, what is essential is invisible to the eye.
Antoine de Saint-Exupery

Good advice is compromised by missing data
Forum member Dog Slobber Nov. 2023

User avatar
ozij
Posts: 10451
Joined: Fri Mar 18, 2005 11:52 pm

Re: Respiratory Rate of 7 on ASV - UARS x 13 years

Post by ozij » Sat May 28, 2022 3:48 am

Ignore this if I'm wrong. I hear some bells ringing.
Do you cross the Rubicon every 1000 posts?

_________________
Mask: AirFit™ P10 Nasal Pillow CPAP Mask with Headgear
Additional Comments: Machine: Resmed AirSense10 for Her with Climateline heated hose ; alternating masks.
And now here is my secret, a very simple secret; it is only with the heart that one can see rightly, what is essential is invisible to the eye.
Antoine de Saint-Exupery

Good advice is compromised by missing data
Forum member Dog Slobber Nov. 2023

Tec5
Posts: 173
Joined: Mon Mar 28, 2022 10:32 am

Re: Respiratory Rate of 7 on ASV - UARS x 13 years

Post by Tec5 » Sat May 28, 2022 5:53 am

No one has an answer to data’s question? Is the representation in the Resmed video incorrect, sinusoidal pattern is not characteristic of normal ( non-limited) sleep ?
I am neither a physician nor a lawyer, so DO NOT rely on me for professional medical or legal advice.

User avatar
ozij
Posts: 10451
Joined: Fri Mar 18, 2005 11:52 pm

Re: Respiratory Rate of 7 on ASV - UARS x 13 years

Post by ozij » Sat May 28, 2022 7:25 am

Tec5 wrote:
Sat May 28, 2022 5:53 am
No one has an answer to data’s question? Is the representation in the Resmed video incorrect, sinusoidal pattern is not characteristic of normal ( non-limited) sleep ?
Rubicon did indeed reply.
  1. Inspiration and expiration are not identical
  2. Inspiration (let me add: in humans) requires effort expiration does not, its passive
  3. "resulting in an Inspiratory: Expiratory (I:E) time Ratio of ~1.0 : 2.0."
Rubicon wrote:
Sat May 28, 2022 3:40 am
ozij wrote:
Sat May 28, 2022 3:33 am
I:E Ratio of what measure? Flowrate?
Time.

Consequently, as volume is constant, flow rate becomes variable.
My emphasis.
It takes twice as along to let the air out as it take to pull it in. So the flow can't be perfectly sinusoidal.

There is no reason to assume Resmed's explanatory video is showing real time flowrate graphs, any more that there is reason to assume it's showing a real person.
Both are graphic simplifications for the sake of explaining an issue.

_________________
Mask: AirFit™ P10 Nasal Pillow CPAP Mask with Headgear
Additional Comments: Machine: Resmed AirSense10 for Her with Climateline heated hose ; alternating masks.
And now here is my secret, a very simple secret; it is only with the heart that one can see rightly, what is essential is invisible to the eye.
Antoine de Saint-Exupery

Good advice is compromised by missing data
Forum member Dog Slobber Nov. 2023

User avatar
Miss Emerita
Posts: 3732
Joined: Sun Nov 04, 2018 8:07 pm

Re: Respiratory Rate of 7 on ASV - UARS x 13 years

Post by Miss Emerita » Sat May 28, 2022 11:58 am

The idealized flow trace inside the O in the Oscar emblem is much closer to reality than the trace in the video.
Oscar software is available at https://www.sleepfiles.com/OSCAR/

Tec5
Posts: 173
Joined: Mon Mar 28, 2022 10:32 am

Re: Respiratory Rate of 7 on ASV - UARS x 13 years

Post by Tec5 » Sat May 28, 2022 12:54 pm

@ozij,
I could be mistaken, but I "think" that data1 was questioning the difference between "asleep" and "awake' breathing patterns. Previously the suggestion was made that a sinusoidal pattern was more characteristic of an "awake" breathing pattern.

But that can be clarified when he or she returns.

On the other area that Rubicon brought up, the matter of the expiration time being twice as long as the inspiration time, but maintaining the same volume. That would suggest that Resmed's stylized plot ought to look more like this:
Stylized breath curve.jpg
Stylized breath curve.jpg (30.32 KiB) Viewed 1332 times
(half the amplitude but twice the duration)
I am neither a physician nor a lawyer, so DO NOT rely on me for professional medical or legal advice.

dataq1
Posts: 832
Joined: Sat Dec 24, 2005 8:09 am
Location: Northeast Ohio

Re: Respiratory Rate of 7 on ASV - UARS x 13 years

Post by dataq1 » Sat May 28, 2022 9:04 pm

Tec5 wrote:
Sat May 28, 2022 12:54 pm
@ozij,
I could be mistaken, but I "think" that data1 was questioning the difference between "asleep" and "awake' breathing patterns. Previously the suggestion was made that a sinusoidal pattern was more characteristic of an "awake" breathing pattern.
Teck5 is correct, my question (above) stated: " It was suggested that a sinusoidal flow rate pattern is characteristic of "awake " breathing, but not of "asleep" breathing.( see viewtopic/t184231/Respiratory-Rate-of-7 ... l#p1410330 and surrounding discussion."

Is the "sinusoidal" flow pattern representative of awake or asleep breathing or not (or neither) ?

BTW, it would be very odd indeed that non-limited breathing pattern be a perfect sine wave. So Teck5's example, where the inspiration phase is a half-sine, and the expiration phase is a different half sine (reduced amplitude and longer time duration (thereby maintaining the volume), while still a stylized representation is a sine shaped pattern.
"THE INFORMATION PROVIDED ON CPAPTALK.COM IS NOT INTENDED NOR RECOMMENDED AS A SUBSTITUTE FOR PROFESSIONAL MEDICAL ADVICE."

User avatar
Applecheeks
Posts: 86
Joined: Fri Dec 24, 2021 10:07 am

Re: Respiratory Rate of 7 on ASV - UARS x 13 years

Post by Applecheeks » Sat May 28, 2022 9:56 pm

Miss Emerita wrote:
Sat May 28, 2022 11:58 am
The idealized flow trace inside the O in the Oscar emblem is much closer to reality than the trace in the video.
This is typical of my asleep pattern with no flow limitations, @5 cm pressure, 1 cm EPR, no snore, no leaks. Doesn't seem at all like the OSCAR logo.
typical breathing patterns.jpg

User avatar
Rubicon
Posts: 1756
Joined: Sun Feb 20, 2022 6:59 am

Re: Respiratory Rate of 7 on ASV - UARS x 13 years

Post by Rubicon » Sun May 29, 2022 4:12 am

I think I would be looking at something like this for "normal" waveform (from a post by JayHC);

Image

As that little blip on the expiratory limb appears in the same spot every breath, I don't think that's CB artifact, more like a palatal something.

Expiration starts out strong and tapers off.
Freeze this moment a little bit longer.