Water Resevoir stuck
-
Crystll222
- Posts: 1
- Joined: Fri Jan 28, 2022 10:45 pm
Water Resevoir stuck
Hi All,
I have a ResMed AirSense 10. The water resevoir is totally stuck in the machine. As you probably know, it's supposed to slide into and out of the main device easily. And it has, until now. Now it's stuck so tight, it seems like it's glued into the main part. Has anyone ever had this problem before? Any idea how to separate the resevoir from the body?
Thanks.
Crystll222
I have a ResMed AirSense 10. The water resevoir is totally stuck in the machine. As you probably know, it's supposed to slide into and out of the main device easily. And it has, until now. Now it's stuck so tight, it seems like it's glued into the main part. Has anyone ever had this problem before? Any idea how to separate the resevoir from the body?
Thanks.
Crystll222
Re: Water Resevoir stuck
Be careful, I damaged the back of my thumb from trying to pull out the tray. I have had the same machine for nearly four years and have numbness and tingling, surgery was recommended but it is feeling better for now. I also use my left hand to pull it out now. I have a garden tool which helps remove the tray, it's a asparagus fork/knife, a weeder. I got mine at Lowe's, it's ~$5.
https://www.lowes.com/pd/Yardsmith-7-in ... 1000383861
The steel fork part might be a little thick if the tray cannot be wedged away a bit first. Others can be found on Amazon. Having the handle helps with needed strength.
You are absolutely not the first to have problems pulling out the tray on the ResMed machine. I think it is a very good machine otherwise. I also have arthritis at the base of my thumb which is probably contributory to fraying the nerve fibers. Hope you find a solution that works for you.
Zabeth
https://www.lowes.com/pd/Yardsmith-7-in ... 1000383861
The steel fork part might be a little thick if the tray cannot be wedged away a bit first. Others can be found on Amazon. Having the handle helps with needed strength.
You are absolutely not the first to have problems pulling out the tray on the ResMed machine. I think it is a very good machine otherwise. I also have arthritis at the base of my thumb which is probably contributory to fraying the nerve fibers. Hope you find a solution that works for you.
Zabeth
_________________
| Machine: ResMed AirSense™ 10 AutoSet™ CPAP Machine with HumidAir™ Heated Humidifier |
| Mask: DreamWear Gel Nasal Pillow CPAP Mask with Headgear - Fit Pack (All Nasal Pillows with Medium Frame) |
| Additional Comments: Dx'd mild apnea Feb 2018 CPAP ordered for AFib |
Re: Water Resevoir stuck
Press down on the white tab with your thumb and pull out, or give it a little twist. There have been messages before from people having trouble, but it does come out, https://www.youtube.com/watch?v=4E6w6fCwxrU
_________________
| Machine: ResMed AirSense™ 10 AutoSet™ CPAP Machine with HumidAir™ Heated Humidifier |
| Mask: Fisher & Paykel Vitera Full Face Mask with Headgear (S, M, or L Cushion) |
| Additional Comments: Back up is a new AS10. |
- Dog Slobber
- Posts: 4312
- Joined: Thu Feb 15, 2018 2:05 pm
- Location: Ontario, Canada
Re: Water Resevoir stuck
Sometime stuff on the bottom of humidifier can fuse to the heater plate.
Try inserting a butter knife under the tank to separate them. Careful you don't dump any remaining water into the machine.
Try inserting a butter knife under the tank to separate them. Careful you don't dump any remaining water into the machine.
_________________
| Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier |
| Mask: AirFit™ P30i Nasal Pillow CPAP Mask with Headgear Starter Pack |
| Additional Comments: Min EPAP: 8.2, Max IPAP: 25, PS:4 |
Battery Backup: EcoFlow Delta 2
- chunkyfrog
- Posts: 34544
- Joined: Mon Jul 12, 2010 5:10 pm
- Location: Nowhere special--this year in particular.
Re: Water Resevoir stuck
Reduced hand strength seems to be an issue.
No idea if Resmed has addressed it with the Airsense 11..
No idea if Resmed has addressed it with the Airsense 11..
_________________
| Mask: AirFit™ P10 For Her Nasal Pillow CPAP Mask with Headgear |
| Additional Comments: Airsense 10 Autoset for Her |
Re: Water Resevoir stuck
I've not had the problem but from what I've read the problem comes from when the gasket cover is closed while the gasket is out of place making the cover stick up just a bit more. So this requires more force to remove it.Crystll222 wrote: ↑Fri Jan 28, 2022 11:33 pmHas anyone ever had this problem before? Any idea how to separate the resevoir from the body?
One way around it, after correcting the seal placement, is to leave the cover closed and use the two hole for adding more water. Just be careful not to overfill.
IF you're using distilled water you won't have to open this cover for cleaning (as often).
HTH
_________________
| Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier |
| Mask: AirFit™ P10 Nasal Pillow CPAP Mask with Headgear |
| Humidifier: HC150 Heated Humidifier With Hose, 2 Chambers and Stand |
Happiness is being on Dog Slobber pr & zonkers foe list
CrankyGranny is Whale Road + many other ids
They are here to help.
CrankyGranny is Whale Road + many other ids
They are here to help.
zonkers + palerider aka GrumpyHere wrote: ↑What exactly do you think you're adding to this thread?
Re: Water Resevoir stuck
The heater plate doesn't get THAT hot that something would 'fuse' the two together. Well ok, maybe if you had candle wax or melted crayons all over the thing then they might stick together. If that's the case, just allow the thing to warm up and then remove the reservoir.Dog Slobber wrote: ↑Sat Jan 29, 2022 7:58 amSometime stuff on the bottom of humidifier can fuse to the heater plate.
Try inserting a butter knife under the tank to separate them. Careful you don't dump any remaining water into the machine.
_________________
| Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier |
| Mask: AirFit™ P10 Nasal Pillow CPAP Mask with Headgear |
ResMed AirCurve 10 Vauto Swift FX
Do not regret growing older. It is a privilege denied to many...
It takes less time to do a thing right than to explain why you did it wrong.
Do not regret growing older. It is a privilege denied to many...
It takes less time to do a thing right than to explain why you did it wrong.
- Dog Slobber
- Posts: 4312
- Joined: Thu Feb 15, 2018 2:05 pm
- Location: Ontario, Canada
Re: Water Resevoir stuck
Conrad wrote: ↑Sat Jan 29, 2022 9:06 amThe heater plate doesn't get THAT hot that something would 'fuse' the two together. Well ok, maybe if you had candle wax or melted crayons all over the thing then they might stick together. If that's the case, just allow the thing to warm up and then remove the reservoir.Dog Slobber wrote: ↑Sat Jan 29, 2022 7:58 amSometime stuff on the bottom of humidifier can fuse to the heater plate.
Try inserting a butter knife under the tank to separate them. Careful you don't dump any remaining water into the machine.
Ash6555 wrote: ↑Thu Jan 27, 2022 10:22 amMy reservoir was stuck in place. It took me two hours to wiggle, tug and ply it out. It was stuck on the bottom not at the top. On inspection, once i finally wrestled it out, I noticed that there was a sticky substance on the plate in the machine. Once I cleaned that out everything was fine. Now I check regularly.
Tom Magarian wrote: ↑Sun Dec 23, 2018 10:28 amToday no matter how hard I yanked on the tank, I could not remove it. Upon investigating it further, I found that the stainless steel bottom of the tank had cemented itself to the stainless steel heating element.
Guest wrote: ↑Wed Apr 12, 2017 11:48 pmI had the same problem. Finally got it loose. I think when i refilled the tank last night I set it down in something slightly sticky and the two metal elements baked themselves together. It seemed like the tank was glued in and not pinned by a part. Try wiggling it side to side. I guess they mean it when they say clean the tank regularly!
_________________
| Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier |
| Mask: AirFit™ P30i Nasal Pillow CPAP Mask with Headgear Starter Pack |
| Additional Comments: Min EPAP: 8.2, Max IPAP: 25, PS:4 |
Battery Backup: EcoFlow Delta 2
Re: Water Resevoir stuck

_________________
| Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier |
| Mask: AirFit™ P10 Nasal Pillow CPAP Mask with Headgear |
"Age is not an accomplishment and youth is not a sin"-Robert A. Heinlein
Oscar-Win
https://www.apneaboard.com/OSCAR/OSCAR-1.5.1-Win64.exe
Oscar-Mac
https://www.apneaboard.com/OSCAR/OSCAR-1.5.1.dmg
Oscar-Win
https://www.apneaboard.com/OSCAR/OSCAR-1.5.1-Win64.exe
Oscar-Mac
https://www.apneaboard.com/OSCAR/OSCAR-1.5.1.dmg
Re: Water Resevoir stuck
Wouldn't it be nice if posters came back to tell us if the problem was resolved and what worked?
_________________
| Machine: ResMed AirSense™ 10 AutoSet™ CPAP Machine with HumidAir™ Heated Humidifier |
| Mask: Fisher & Paykel Vitera Full Face Mask with Headgear (S, M, or L Cushion) |
| Additional Comments: Back up is a new AS10. |
- Respirator99
- Posts: 402
- Joined: Mon Jul 05, 2021 12:39 am
- Location: Australia
Re: Water Resevoir stuck
Fair go Larry! She only joined yesterday. Not everybody is glued to their device 24/7.
_________________
| Machine: ResMed AirCurve 10 ASV Machine with Heated Humidifier |
| Mask: ResMed AirFit™ F20 Full Face CPAP Mask with Headgear |
* Download Oscar
* Oscar help
* An alternative to Oscar - try SleepHQ
I have no medical training or qualifications. Take my advice for what it's worth.
* Oscar help
* An alternative to Oscar - try SleepHQ
I have no medical training or qualifications. Take my advice for what it's worth.
Re: Water Resevoir stuck
Most are only glued to it 8x7.Respirator99 wrote: ↑Sat Jan 29, 2022 8:59 pmFair go Larry! She only joined yesterday. Not everybody is glued to their device 24/7.
However, you wanna take bets on whether either of the first time posters in the thread come back?
Get OSCAR
Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.
Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.
Re: Water Resevoir stuck
What do all of these (except one) above have in common?Dog Slobber wrote: ↑Sat Jan 29, 2022 9:58 amConrad wrote: ↑Sat Jan 29, 2022 9:06 amThe heater plate doesn't get THAT hot that something would 'fuse' the two together. Well ok, maybe if you had candle wax or melted crayons all over the thing then they might stick together. If that's the case, just allow the thing to warm up and then remove the reservoir.Dog Slobber wrote: ↑Sat Jan 29, 2022 7:58 amSometime stuff on the bottom of humidifier can fuse to the heater plate.
Try inserting a butter knife under the tank to separate them. Careful you don't dump any remaining water into the machine.Ash6555 wrote: ↑Thu Jan 27, 2022 10:22 amMy reservoir was stuck in place. It took me two hours to wiggle, tug and ply it out. It was stuck on the bottom not at the top. On inspection, once i finally wrestled it out, I noticed that there was a sticky substance on the plate in the machine. Once I cleaned that out everything was fine. Now I check regularly.Tom Magarian wrote: ↑Sun Dec 23, 2018 10:28 amToday no matter how hard I yanked on the tank, I could not remove it. Upon investigating it further, I found that the stainless steel bottom of the tank had cemented itself to the stainless steel heating element.Guest wrote: ↑Wed Apr 12, 2017 11:48 pmI had the same problem. Finally got it loose. I think when i refilled the tank last night I set it down in something slightly sticky and the two metal elements baked themselves together. It seemed like the tank was glued in and not pinned by a part. Try wiggling it side to side. I guess they mean it when they say clean the tank regularly!
"I noticed that there was a sticky substance on the plate in the machine. "
"I think when i refilled the tank last night I set it down in something slightly sticky and the two metal elements baked themselves together."
Stainless steel does not "fuse together" at the temps the heating plate gets up to. I'm sure that the one case posted above also had some substance on the bottom of the reservoir that, when heated, glued the two together.
How hot do you think the heating plate gets up to? At what temp do you think that two stainless steel plates would have to be heated to in order for them to fuse together?
_________________
| Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier |
| Mask: AirFit™ P10 Nasal Pillow CPAP Mask with Headgear |
ResMed AirCurve 10 Vauto Swift FX
Do not regret growing older. It is a privilege denied to many...
It takes less time to do a thing right than to explain why you did it wrong.
Do not regret growing older. It is a privilege denied to many...
It takes less time to do a thing right than to explain why you did it wrong.
- Dog Slobber
- Posts: 4312
- Joined: Thu Feb 15, 2018 2:05 pm
- Location: Ontario, Canada
Re: Water Resevoir stuck
Conrad wrote: ↑Sun Jan 30, 2022 9:16 amWhat do all of these (except one) above have in common?Dog Slobber wrote: ↑Sat Jan 29, 2022 9:58 amConrad wrote: ↑Sat Jan 29, 2022 9:06 amThe heater plate doesn't get THAT hot that something would 'fuse' the two together. Well ok, maybe if you had candle wax or melted crayons all over the thing then they might stick together. If that's the case, just allow the thing to warm up and then remove the reservoir.Dog Slobber wrote: ↑Sat Jan 29, 2022 7:58 amSometime stuff on the bottom of humidifier can fuse to the heater plate.
Try inserting a butter knife under the tank to separate them. Careful you don't dump any remaining water into the machine.Ash6555 wrote: ↑Thu Jan 27, 2022 10:22 amMy reservoir was stuck in place. It took me two hours to wiggle, tug and ply it out. It was stuck on the bottom not at the top. On inspection, once i finally wrestled it out, I noticed that there was a sticky substance on the plate in the machine. Once I cleaned that out everything was fine. Now I check regularly.Tom Magarian wrote: ↑Sun Dec 23, 2018 10:28 amToday no matter how hard I yanked on the tank, I could not remove it. Upon investigating it further, I found that the stainless steel bottom of the tank had cemented itself to the stainless steel heating element.Guest wrote: ↑Wed Apr 12, 2017 11:48 pmI had the same problem. Finally got it loose. I think when i refilled the tank last night I set it down in something slightly sticky and the two metal elements baked themselves together. It seemed like the tank was glued in and not pinned by a part. Try wiggling it side to side. I guess they mean it when they say clean the tank regularly!
"I noticed that there was a sticky substance on the plate in the machine. "
"I think when i refilled the tank last night I set it down in something slightly sticky and the two metal elements baked themselves together."
Stainless steel does not "fuse together" at the temps the heating plate gets up to. I'm sure that the one case posted above also had some substance on the bottom of the reservoir that, when heated, glued the two together.
How hot do you think the heating plate gets up to? At what temp do you think that two stainless steel plates would have to be heated to in order for them to fuse together?
I never said that the two stainless steel plates would fuse together.
I said "stuff on the bottom of humidifier can fuse to the heater plate".
Those examples I presented are exactly what I described.
Why you incorrectly, interpreted my description of "stuff on the humidifier fusing" as "two stainless steel plates" fusing together is beyond me?
_________________
| Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier |
| Mask: AirFit™ P30i Nasal Pillow CPAP Mask with Headgear Starter Pack |
| Additional Comments: Min EPAP: 8.2, Max IPAP: 25, PS:4 |
Battery Backup: EcoFlow Delta 2
Re: Water Resevoir stuck
You're right and I'm wrong, my bad. I guess that I should practice what I preach about reading comprehension...Dog Slobber wrote: ↑Sun Jan 30, 2022 9:26 am
I never said that the two stainless steel plates would fuse together.
I said "stuff on the bottom of humidifier can fuse to the heater plate".
Those examples I presented are exactly what I described.
Why you incorrectly, interpreted my description of "stuff on the humidifier fusing" as "two stainless steel plates" fusing together is beyond me?
_________________
| Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier |
| Mask: AirFit™ P10 Nasal Pillow CPAP Mask with Headgear |
ResMed AirCurve 10 Vauto Swift FX
Do not regret growing older. It is a privilege denied to many...
It takes less time to do a thing right than to explain why you did it wrong.
Do not regret growing older. It is a privilege denied to many...
It takes less time to do a thing right than to explain why you did it wrong.