Need Help With OSCAR 1.3.0 Issues

General Discussion on any topic relating to CPAP and/or Sleep Apnea.
User avatar
Pugsy
Posts: 65304
Joined: Thu May 14, 2009 9:31 am
Location: Missouri, USA

Re: Need Help With OSCAR 1.3.0 Issues

Post by Pugsy » Sat Feb 26, 2022 9:40 am

Using the S9 model?? What operating system on your computer?
Did you lock the SD card prior to putting the SD card in your computer?

I assume you already let the machine fix the "corrupted card" by erasing everything but if this should happen in the future there may be a way to fix it so the machine doesn't think the card is corrupted. Just putting this out there in case it happens again.

S9 model machines notoriously don't like it when the SD card doesn't look like it expects the card to look so it wants to get rid of the corruption. Sometimes it's possible to get rid of the "corruption" without totally nuking the entire SD contents.
ProfBrianObvilion wrote:
Sat Feb 26, 2022 9:10 am
Ugh... bad news there... I put my SD card back in my machine last night and it said it was corrupt and wiped it out(reformatted).

_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier
Additional Comments: Mask Bleep Eclipse https://bleepsleep.com/the-eclipse/
I may have to RISE but I refuse to SHINE.

User avatar
BlueDragon
Posts: 554
Joined: Sun May 27, 2018 10:20 am
Location: Sonoran Desert

Re: Need Help With OSCAR 1.3.0 Issues

Post by BlueDragon » Sat Feb 26, 2022 10:05 am

Did you not lock the card before putting it in the computer? Might be a good idea to get a new SD card! Any size up to 32 GB, but not more, will do. A defective SD card might well explain the odd problems you are having.

_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier
Mask: Brevida™ Nasal Pillow CPAP Mask with Headgear
Additional Comments: Dx Mar 2018 (AHI=24, RDI=54; AHI=73 supine). Started APAP June 2018, VAuto Aug 2020.
See OSCAR for the latest release.
OSCAR Lead Developer
ResMed AirCurve 10 VAuto, F&P Brevida.
FlashAir SD and FlashPap for data transfer.

ProfBrianObvilion
Posts: 9
Joined: Wed Feb 23, 2022 11:19 am

Re: Need Help With OSCAR 1.3.0 Issues

Post by ProfBrianObvilion » Sun Feb 27, 2022 12:51 pm

Yes it's a ResMed S9 VPAP (I'll update my profile).

Actually, I just replaced the SD card like 4 months ago because the original, "official" resmed card was doing the same thing from time to time.... thought it was dying because it was old. Maybe I will go ahead and replace it again just to be sure. It's cheap to see if it resolves the problem.

I do quite religiously lock the card the instant I remove it from the machine and before I put it into my computer. So it never happens.


I would like to know how to try and not nuke the sd card and possibly recover without the machine reformatting the whole thing.

User avatar
Pugsy
Posts: 65304
Joined: Thu May 14, 2009 9:31 am
Location: Missouri, USA

Re: Need Help With OSCAR 1.3.0 Issues

Post by Pugsy » Sun Feb 27, 2022 1:04 pm

ProfBrianObvilion wrote:
Sun Feb 27, 2022 12:51 pm
I would like to know how to try and not nuke the sd card and possibly recover without the machine reformatting the whole thing.
Some operating systems put a little text file on the SD card when the card is inserted into the computer.
Now OSCAR doesn't care that this little text file is on the SD card but the S9 model machines care....S9 models are the only models that care...so that's why we advise locking the SD card before it gets inserted into the computer.
Windows 8 and later...and some Mac operating systems will add that little text file.

I don't know how to remove the text file from Macs but Windows is fairly easy.

rmdir "System Volume Information" /s /q

at the contents of the SD card window hold down the shift key and right click mouse...and choose Open command window and paste the above in the black window.

_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier
Additional Comments: Mask Bleep Eclipse https://bleepsleep.com/the-eclipse/
I may have to RISE but I refuse to SHINE.

ProfBrianObvilion
Posts: 9
Joined: Wed Feb 23, 2022 11:19 am

Re: Need Help With OSCAR 1.3.0 Issues

Post by ProfBrianObvilion » Sun Feb 27, 2022 4:49 pm

Cool thanks... in my case, I never put it in my computer without having it locked though. I've gotten burned by that a long time ago :-).

User avatar
BlueDragon
Posts: 554
Joined: Sun May 27, 2018 10:20 am
Location: Sonoran Desert

Re: Need Help With OSCAR 1.3.0 Issues

Post by BlueDragon » Sun Feb 27, 2022 7:18 pm

I can reproduce the missing Flow graph problem using your profile's backup data. I don't know what is causing it but we will look into it and get back to you.

_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier
Mask: Brevida™ Nasal Pillow CPAP Mask with Headgear
Additional Comments: Dx Mar 2018 (AHI=24, RDI=54; AHI=73 supine). Started APAP June 2018, VAuto Aug 2020.
See OSCAR for the latest release.
OSCAR Lead Developer
ResMed AirCurve 10 VAuto, F&P Brevida.
FlashAir SD and FlashPap for data transfer.

ProfBrianObvilion
Posts: 9
Joined: Wed Feb 23, 2022 11:19 am

Re: Need Help With OSCAR 1.3.0 Issues

Post by ProfBrianObvilion » Mon Feb 28, 2022 7:46 am

It happened again last night... Missing the same graphs. I can upload the SDCard and all OSCAR data as previously requested if it will help?

User avatar
BlueDragon
Posts: 554
Joined: Sun May 27, 2018 10:20 am
Location: Sonoran Desert

Re: Need Help With OSCAR 1.3.0 Issues

Post by BlueDragon » Mon Feb 28, 2022 10:16 am

Thank you. Since we can reproduce the failure, we don't need any more data at this time. I'll post a response when we figure out what is going on or need more data.

_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier
Mask: Brevida™ Nasal Pillow CPAP Mask with Headgear
Additional Comments: Dx Mar 2018 (AHI=24, RDI=54; AHI=73 supine). Started APAP June 2018, VAuto Aug 2020.
See OSCAR for the latest release.
OSCAR Lead Developer
ResMed AirCurve 10 VAuto, F&P Brevida.
FlashAir SD and FlashPap for data transfer.