pressure difference with new mask?

General Discussion on any topic relating to CPAP and/or Sleep Apnea.
Redstormsju
Posts: 28
Joined: Fri Oct 08, 2021 3:13 am

Re: pressure difference with new mask?

Post by Redstormsju » Mon Oct 18, 2021 2:47 am

[/quote]
Maybe they need to be adjusted now. :D Are you reviewing your daily AHI and leak level that shows on the machine display?
[/quote]
I am reviewing them daily. I am getting good numbers generally…AHI between 2 and 4, leaks usually less than 12L/min. I am going to install OSCAR today and hopefully get a lot more data than my airsense 11 gives me….I only get time used, leaks, AHI, Total AI and CI. No other report settings with this machine…not even in the clinical menu.

What do you suggest in terms of adjustments? I was thinking of starting off with changing the AutoSet to Standard versus Soft.
Mask: Airfit F30i
Machine: Airsense 11
New York

User avatar
Julie
Posts: 20051
Joined: Tue Feb 28, 2006 12:58 pm

Re: pressure difference with new mask?

Post by Julie » Mon Oct 18, 2021 5:07 am

Post Oscar, then we can see what's what.

User avatar
ChicagoGranny
Posts: 15534
Joined: Sun Jan 29, 2012 1:43 pm
Location: USA

Re: pressure difference with new mask?

Post by ChicagoGranny » Mon Oct 18, 2021 11:28 am

Redstormsju wrote: ↑
Mon Oct 18, 2021 2:47 am
I was thinking of starting off with changing the AutoSet to Standard versus Soft.
If you mean changing this setting when you switch to a nasal mask, I wouldn't do it. Standard practice is to change only one variable at a time. If two are changed and the results are different, you likely won't know which of the two variables caused the different results.

I would leave the machine settings the same when starting the new mask. The settings can be changed later if they prove to be suboptimal.

BTW, it is sometimes said that a higher pressure is needed with a FFM versus a nasal mask. Actually, not a change in mask, but mouth-breathing may result in higher pressure needs. In humans, when the mouth opens, the mandible moves backward slightly. This tends to narrow the airway.
Redstormsju wrote: ↑
Mon Oct 18, 2021 2:47 am
I am getting good numbers generally…AHI between 2 and 4, leaks usually less than 12L/min.
These are good results for a newbie (and many oldies :D ). I wouldn't change any settings at this point. OSCAR may tell us more.

Redstormsju
Posts: 28
Joined: Fri Oct 08, 2021 3:13 am

Re: pressure difference with new mask?

Post by Redstormsju » Mon Oct 18, 2021 1:39 pm

Redstormsju wrote: ↑
Mon Oct 18, 2021 2:47 am
I am getting good numbers generally…AHI between 2 and 4, leaks usually less than 12L/min.
These are good results for a newbie (and many oldies :D ). I wouldn't change any settings at this point. OSCAR may tell us more.
[/quote]

Thank you so much.
Mask: Airfit F30i
Machine: Airsense 11
New York

Redstormsju
Posts: 28
Joined: Fri Oct 08, 2021 3:13 am

Re: pressure difference with new mask?

Post by Redstormsju » Tue Oct 19, 2021 6:39 pm

These are good results for a newbie (and many oldies :D ). I wouldn't change any settings at this point. OSCAR may tell us more.
What graphs should I have up on OSCAR for posting on this forum? I see there are many that I can scroll through on OSCAR.
Mask: Airfit F30i
Machine: Airsense 11
New York

User avatar
Pugsy
Posts: 65411
Joined: Thu May 14, 2009 9:31 am
Location: Missouri, USA

Re: pressure difference with new mask?

Post by Pugsy » Tue Oct 19, 2021 6:54 pm

Redstormsju wrote: ↑
Tue Oct 19, 2021 6:39 pm
What graphs should I have up on OSCAR for posting on this forum?
Take a look here. Start with the basics.

viewtopic/t158560/How-to-post-images-for-review.html

_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier
Additional Comments: Mask Bleep Eclipse https://bleepsleep.com/the-eclipse/
I may have to RISE but I refuse to SHINE.

User avatar
zonker
Posts: 11491
Joined: Fri Jun 19, 2015 4:36 pm

Re: pressure difference with new mask?

Post by zonker » Tue Oct 19, 2021 6:57 pm

Redstormsju wrote: ↑
Tue Oct 19, 2021 6:39 pm
These are good results for a newbie (and many oldies :D ). I wouldn't change any settings at this point. OSCAR may tell us more.
What graphs should I have up on OSCAR for posting on this forum? I see there are many that I can scroll through on OSCAR.
forum friend grumpy here has this to say-

https://www.cpaptalk.com/viewtopic.php? ... 9#p1390059
"Age is not an accomplishment and youth is not a sin"-Robert A. Heinlein
Oscar-Win
https://www.apneaboard.com/OSCAR/OSCAR-1.5.1-Win64.exe
Oscar-Mac
https://www.apneaboard.com/OSCAR/OSCAR-1.5.1.dmg

Redstormsju
Posts: 28
Joined: Fri Oct 08, 2021 3:13 am

Re: pressure difference with new mask?

Post by Redstormsju » Sat Oct 23, 2021 8:36 am

Hello all, these screenshots are from last nights sleep. I see the numbers are good, but I am not quite confident in how to read the graphs. I’ve been reading from the help page but still am not quite sure how to interpret the graphs. Any insight is appreciated.
Attachments
A663A53D-18ED-465D-8A13-ACEBE8D39C25.png
A663A53D-18ED-465D-8A13-ACEBE8D39C25.png (229.62 KiB) Viewed 2619 times
CD70FF8B-B3FA-4607-9F57-77BC2E12731F.png
CD70FF8B-B3FA-4607-9F57-77BC2E12731F.png (237.98 KiB) Viewed 2619 times
8EB56541-FCD5-43D4-B915-687693CF4D6B.png
8EB56541-FCD5-43D4-B915-687693CF4D6B.png (228.35 KiB) Viewed 2619 times
Mask: Airfit F30i
Machine: Airsense 11
New York

User avatar
ChicagoGranny
Posts: 15534
Joined: Sun Jan 29, 2012 1:43 pm
Location: USA

Re: pressure difference with new mask?

Post by ChicagoGranny » Sat Oct 23, 2021 8:42 am

Redstormsju wrote: ↑
Sat Oct 23, 2021 8:36 am
but still am not quite sure how to interpret the graphs. Any insight is appreciated.
Can you be more specific about what you don't understand?

User avatar
ChicagoGranny
Posts: 15534
Joined: Sun Jan 29, 2012 1:43 pm
Location: USA

Re: pressure difference with new mask?

Post by ChicagoGranny » Sat Oct 23, 2021 8:45 am

Redstormsju wrote: ↑
Sat Oct 23, 2021 8:36 am
these screenshots are from last nights sleep.
These are not in the recommended format.

Redstormsju
Posts: 28
Joined: Fri Oct 08, 2021 3:13 am

Re: pressure difference with new mask?

Post by Redstormsju » Sat Oct 23, 2021 9:01 am

Can you be more specific about what you don't understand?
I’m not sure how to read/interpret the AHI graph, Minute vent, Insp. Time, Exp. Time.

And the pressure graph, I assume inhale and exhale pressures?
Mask: Airfit F30i
Machine: Airsense 11
New York

Redstormsju
Posts: 28
Joined: Fri Oct 08, 2021 3:13 am

Re: pressure difference with new mask?

Post by Redstormsju » Sat Oct 23, 2021 9:04 am


These are not in the recommended format.
What is the recommended format?
Mask: Airfit F30i
Machine: Airsense 11
New York

User avatar
palerider
Posts: 32308
Joined: Wed Dec 16, 2009 5:43 pm
Location: Dallas(ish).

Re: pressure difference with new mask?

Post by palerider » Sat Oct 23, 2021 9:46 am

Redstormsju wrote: ↑
Sat Oct 23, 2021 9:01 am
Can you be more specific about what you don't understand?
I’m not sure how to read/interpret the AHI graph, Minute vent, Insp. Time, Exp. Time.
The AHI graph is useless,
MV is your average total respiration volume over a minute.
Exp time, how long your exhalation is

Ignore all of them.
Redstormsju wrote: ↑
Sat Oct 23, 2021 9:01 am
And the pressure graph, I assume inhale and exhale pressures?
Yes.

As explained in the "READ THIS BEFORE POSTING" post at the top of the forum, we only need six graphs 99% of the time.

flags
flow rate
pressure (not mask pressure)
leaks
snores
flow limitation.

_________________
Mask: Bleep DreamPort CPAP Mask Solution
Additional Comments: S9 VPAP Auto
Get OSCAR

Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.

User avatar
palerider
Posts: 32308
Joined: Wed Dec 16, 2009 5:43 pm
Location: Dallas(ish).

Re: pressure difference with new mask?

Post by palerider » Sat Oct 23, 2021 9:48 am

Redstormsju wrote: ↑
Sat Oct 23, 2021 9:04 am

These are not in the recommended format.
What is the recommended format?
https://www.cpaptalk.com/wiki/index.php/Oscar:organize

_________________
Mask: Bleep DreamPort CPAP Mask Solution
Additional Comments: S9 VPAP Auto
Get OSCAR

Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.

Redstormsju
Posts: 28
Joined: Fri Oct 08, 2021 3:13 am

Re: pressure difference with new mask?

Post by Redstormsju » Sat Oct 23, 2021 10:26 am

Not sure how my attachments are not the recommended format. According what I read on the link provided, the format is png image file and that is what my images are. I took screenshots directly from the Take Screenshot option on Oscar, I saved the image to my desktop and just changed the file name to the date of the data. I then used the attachment option in the forum page to attach the pictures to my post. The only thing I did notice was that when they attached to the post, they attached as one large image versus the three individual images I attached while creating g the post.
Mask: Airfit F30i
Machine: Airsense 11
New York