Flow Limit

General Discussion on any topic relating to CPAP and/or Sleep Apnea.
sleepbetter
Posts: 11
Joined: Thu Jul 26, 2018 9:48 am

Flow Limit

Post by sleepbetter » Tue Sep 14, 2021 4:02 pm

I replaced my Philips Respironics with an AirSense 10, 83 days ago. Looking at my statistics on Oscar, my FL has gone to 0.0 since. That seems odd. Does anyone have any insight on why? The 525 days before on the Philips it was 1.12.
Thanks in advance.
AirSense 10 Auto-Set with P-10 Nasal Mask

User avatar
Julie
Posts: 20051
Joined: Tue Feb 28, 2006 12:58 pm

Re: Flow Limit

Post by Julie » Tue Sep 14, 2021 4:10 pm

Post some Oscar results - we're not that good at guessing. :D

sleepbetter
Posts: 11
Joined: Thu Jul 26, 2018 9:48 am

Re: Flow Limit

Post by sleepbetter » Tue Sep 14, 2021 4:26 pm

Here you go.
Screen Shot 2021-09-14 at 5.22.39 PM.png
Screen Shot 2021-09-14 at 5.22.39 PM.png (480.17 KiB) Viewed 2232 times
Screen Shot 2021-09-14 at 5.23.22 PM.png
Screen Shot 2021-09-14 at 5.23.22 PM.png (525.22 KiB) Viewed 2232 times
Screen Shot 2021-09-14 at 5.24.09 PM.png
Screen Shot 2021-09-14 at 5.24.09 PM.png (514.47 KiB) Viewed 2231 times
AirSense 10 Auto-Set with P-10 Nasal Mask

User avatar
zonker
Posts: 11509
Joined: Fri Jun 19, 2015 4:36 pm

Re: Flow Limit

Post by zonker » Tue Sep 14, 2021 4:32 pm

sleepbetter wrote: ↑
Tue Sep 14, 2021 4:02 pm
I replaced my Philips Respironics with an AirSense 10, 83 days ago. Looking at my statistics on Oscar, my FL has gone to 0.0 since. That seems odd. Does anyone have any insight on why? The 525 days before on the Philips it was 1.12.
Thanks in advance.
pugsy said something about this some time ago-

https://www.cpaptalk.com/viewtopic.php?t=155472

and bluedragon chips in here-

https://www.cpaptalk.com/viewtopic.php?t=181218

in case you are unaware, bluedragon is one of the developers of oscar, so it's good to have his input.
"Age is not an accomplishment and youth is not a sin"-Robert A. Heinlein
Oscar-Win
https://www.apneaboard.com/OSCAR/OSCAR-1.5.1-Win64.exe
Oscar-Mac
https://www.apneaboard.com/OSCAR/OSCAR-1.5.1.dmg

User avatar
palerider
Posts: 32308
Joined: Wed Dec 16, 2009 5:43 pm
Location: Dallas(ish).

Re: Flow Limit

Post by palerider » Tue Sep 14, 2021 5:34 pm

sleepbetter wrote: ↑
Tue Sep 14, 2021 4:02 pm
I replaced my Philips Respironics with an AirSense 10, 83 days ago. Looking at my statistics on Oscar, my FL has gone to 0.0 since. That seems odd. Does anyone have any insight on why? The 525 days before on the Philips it was 1.12.
Thanks in advance.
Because 1) the machines report things different, and (most importantly) 2) the Resmeds just work better

Your max pressure should be turned up (preferably to 20)
Get OSCAR

Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.

User avatar
Miss Emerita
Posts: 3783
Joined: Sun Nov 04, 2018 8:07 pm

Re: Flow Limit

Post by Miss Emerita » Tue Sep 14, 2021 7:59 pm

Great links from Zonker! As you can see from your daily charts, you do continue to have flow limitations. They are sometimes pretty heavy. This may mean nothing for your health or for the quality of your sleep. But if you’d like to do an experiment, you could try increasing your EPR from 2 to 3. That might reduce the FLs some. Because you have so few obstructive events, I don’t think you need to raise your minimum pressure to compensate for the extra 1 cm. drop in pressure when you exhale.
Oscar software is available at https://www.sleepfiles.com/OSCAR/

User avatar
zonker
Posts: 11509
Joined: Fri Jun 19, 2015 4:36 pm

Re: Flow Limit

Post by zonker » Tue Sep 14, 2021 10:08 pm

Miss Emerita wrote: ↑
Tue Sep 14, 2021 7:59 pm
Great links from Zonker!
Image

i'm trying to emulate forum friend, grumpyhere, who is spectacular at hunting up links.

one trick is to type "www.cpaptalk.com search criteria" into a google search and pick and choose. the forum search box is pretty useless so one has to get imaginative!
"Age is not an accomplishment and youth is not a sin"-Robert A. Heinlein
Oscar-Win
https://www.apneaboard.com/OSCAR/OSCAR-1.5.1-Win64.exe
Oscar-Mac
https://www.apneaboard.com/OSCAR/OSCAR-1.5.1.dmg

User avatar
Miss Emerita
Posts: 3783
Joined: Sun Nov 04, 2018 8:07 pm

Re: Flow Limit

Post by Miss Emerita » Wed Sep 15, 2021 10:13 am

Oh, cool. I would never have thought of that. Thanks, Zonker!
Oscar software is available at https://www.sleepfiles.com/OSCAR/

User avatar
zonker
Posts: 11509
Joined: Fri Jun 19, 2015 4:36 pm

Re: Flow Limit

Post by zonker » Wed Sep 15, 2021 10:57 am

Miss Emerita wrote: ↑
Wed Sep 15, 2021 10:13 am
Oh, cool. I would never have thought of that. Thanks, Zonker!