..now a new problem with my new RESMED model 11 ELITE
..now a new problem with my new RESMED model 11 ELITE
...now I need help, if anyone has gained some expertise on this new machine....This is my problem....machine is on ready to start therapy, green tiny light on at the big button on the top of the machine.....and when therapy begins it turns white
If I get up to go to the bathroom I press the big button once and machine pauses but continues blowing air at an "almost imperceptible rate.. I come back put mask back on and the machine starts automatically back on, I read that that is a "smarton" option that is set on the machine for when it notices that you begin breathing...I also read that there is a "smartoff" set up so machine stops when no breathing is detected...
Now my problem is in the morning, I slept 8 or 9 hours no problems at all, if I remove the mask the machine continues blowing air, so I press big black button on top of the machine and it stops, but it continues to blow air at an "almost imperceptible rate", a soft blowing of air, also I noticed the symbols on the screen that indicates "cooling of water chamber". All is fine but the problem is that the machine never stops that "almost imperceptible blowing of air". Now out of desperation I unplugged the machine from the wall and THAT stops the "almost imperceptible blowing of air" and the "cooling tank indicator" disappears.
Should I try to check if the "smartoff" function is properly set on? and if so how do I do that.
Thanks for any advice ..greatly appreciated..
If I get up to go to the bathroom I press the big button once and machine pauses but continues blowing air at an "almost imperceptible rate.. I come back put mask back on and the machine starts automatically back on, I read that that is a "smarton" option that is set on the machine for when it notices that you begin breathing...I also read that there is a "smartoff" set up so machine stops when no breathing is detected...
Now my problem is in the morning, I slept 8 or 9 hours no problems at all, if I remove the mask the machine continues blowing air, so I press big black button on top of the machine and it stops, but it continues to blow air at an "almost imperceptible rate", a soft blowing of air, also I noticed the symbols on the screen that indicates "cooling of water chamber". All is fine but the problem is that the machine never stops that "almost imperceptible blowing of air". Now out of desperation I unplugged the machine from the wall and THAT stops the "almost imperceptible blowing of air" and the "cooling tank indicator" disappears.
Should I try to check if the "smartoff" function is properly set on? and if so how do I do that.
Thanks for any advice ..greatly appreciated..
Re: ..now a new problem with my new RESMED model 11 ELITE
Define "never", and please stop making more new threads about your new machine, just stick to one.
Get OSCAR
Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.
Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.
Re: ..now a new problem with my new RESMED model 11 ELITE
Normal cool down mode. Resmed machines have all done this for years and years and various models.
_________________
| Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier |
| Additional Comments: Mask Bleep Eclipse https://bleepsleep.com/the-eclipse/ |
I may have to RISE but I refuse to SHINE.
- chunkyfrog
- Posts: 34544
- Joined: Mon Jul 12, 2010 5:10 pm
- Location: Nowhere special--this year in particular.
Re: ..now a new problem with my new RESMED model 11 ELITE
Cool-down mode fascinates the cat.
First human to channel a cat.
How do you like laser pointers?
Another thing: if you want to carry this all the way,
remember cats do not need toilet paper.
First human to channel a cat.
How do you like laser pointers?
Another thing: if you want to carry this all the way,
remember cats do not need toilet paper.
_________________
| Mask: AirFit™ P10 For Her Nasal Pillow CPAP Mask with Headgear |
| Additional Comments: Airsense 10 Autoset for Her |
Re: ..now a new problem with my new RESMED model 11 ELITE
chunkyfrog wrote: ↑Sun Sep 05, 2021 4:30 pm
Another thing: if you want to carry this all the way,
remember cats do not need toilet paper.

_________________
| Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier |
| Mask: AirFit™ P10 Nasal Pillow CPAP Mask with Headgear |
"Age is not an accomplishment and youth is not a sin"-Robert A. Heinlein
Oscar-Win
https://www.apneaboard.com/OSCAR/OSCAR-1.5.1-Win64.exe
Oscar-Mac
https://www.apneaboard.com/OSCAR/OSCAR-1.5.1.dmg
Oscar-Win
https://www.apneaboard.com/OSCAR/OSCAR-1.5.1-Win64.exe
Oscar-Mac
https://www.apneaboard.com/OSCAR/OSCAR-1.5.1.dmg
Re: ..now a new problem with my new RESMED model 11 ELITE
Sorry palerider, a simple mistake on my part!!......never, as I thought it was happening, meant to me that the small amount of air blowing was not stopping, but this morning I decided to let that condition run as long as it will take (water tank left in the machine) and to my surprise the air stopped blowing and the symbol "cooling off" disappear after about 20 or 30 minutes later (after I had manually stopped the macine/therapy).
I will inquiere with my doctors office on my next visit if that is because the "smartoff" parameter is not turned on ....I can live with the way my resmed 11 is currently operating.
I will inquiere with my doctors office on my next visit if that is because the "smartoff" parameter is not turned on ....I can live with the way my resmed 11 is currently operating.
Re: ..now a new problem with my new RESMED model 11 ELITE
has nothing to do with SmartStart or SmartStop.....it's cool down mode.
Did you miss my reply above???
_________________
| Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier |
| Additional Comments: Mask Bleep Eclipse https://bleepsleep.com/the-eclipse/ |
I may have to RISE but I refuse to SHINE.
- chunkyfrog
- Posts: 34544
- Joined: Mon Jul 12, 2010 5:10 pm
- Location: Nowhere special--this year in particular.
Re: ..now a new problem with my new RESMED model 11 ELITE
This one has a steep learning curve . . .
And low reading comprehension--or none at all.
And low reading comprehension--or none at all.
_________________
| Mask: AirFit™ P10 For Her Nasal Pillow CPAP Mask with Headgear |
| Additional Comments: Airsense 10 Autoset for Her |
- ChicagoGranny
- Posts: 15407
- Joined: Sun Jan 29, 2012 1:43 pm
- Location: USA
Re: ..now a new problem with my new RESMED model 11 ELITE
Seems to be ESL.chunkyfrog wrote: ↑Mon Sep 06, 2021 10:17 amThis one has a steep learning curve . . .
And low reading comprehension--or none at all.
"It's not the number of breaths we take, it's the number of moments that take our breath away."
Cuando cuentes cuentos, cuenta cuántas cuentos cuentas.
Cuando cuentes cuentos, cuenta cuántas cuentos cuentas.
Re: ..now a new problem with my new RESMED model 11 ELITE
Funny, the way that things sometimes work the way the manual says they will.svo wrote: ↑Mon Sep 06, 2021 8:09 amSorry palerider, a simple mistake on my part!!......never, as I thought it was happening, meant to me that the small amount of air blowing was not stopping, but this morning I decided to let that condition run as long as it will take (water tank left in the machine) and to my surprise the air stopped blowing and the symbol "cooling off" disappear after about 20 or 30 minutes later (after I had manually stopped the macine/therapy).
So, turn it on.
Get OSCAR
Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.
Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.
- chunkyfrog
- Posts: 34544
- Joined: Mon Jul 12, 2010 5:10 pm
- Location: Nowhere special--this year in particular.
Re: ..now a new problem with my new RESMED model 11 ELITE
I suppose that is possible; but the bullfrog has English as his FIFTH languageChicagoGranny wrote: ↑Mon Sep 06, 2021 11:10 amSeems to be ESL.chunkyfrog wrote: ↑Mon Sep 06, 2021 10:17 amThis one has a steep learning curve . . .
And low reading comprehension--or none at all.
and does quite well, even to the extent of crosswords and Scrabble.
His mother spoke NINE languages, and her spelling was hilarious.
It perfectly reflected her heavy accent. Loved that woman!
_________________
| Mask: AirFit™ P10 For Her Nasal Pillow CPAP Mask with Headgear |
| Additional Comments: Airsense 10 Autoset for Her |
Re: ..now a new problem with my new RESMED model 11 ELITE
You know palerider, your replies usually amuse me....but now I realized that they are curt and impolite at best! ...sometimes I wish you were more silent!!!
Most of us do not appreciate your tone....the one "who die and made you king or queen"....lol
Most of us do not appreciate your tone....the one "who die and made you king or queen"....lol
- chunkyfrog
- Posts: 34544
- Joined: Mon Jul 12, 2010 5:10 pm
- Location: Nowhere special--this year in particular.
Re: ..now a new problem with my new RESMED model 11 ELITE
Really intelligent people have little patience for the intentionally ignorant.
_________________
| Mask: AirFit™ P10 For Her Nasal Pillow CPAP Mask with Headgear |
| Additional Comments: Airsense 10 Autoset for Her |
Re: ..now a new problem with my new RESMED model 11 ELITE
THAT makes a lot of sense. My machine appears to turn off OK, but I never actually checked to see if it was still slightly on. Will do that tomorrow.
_________________
| Machine: ResMed AirSense™ 10 AutoSet™ CPAP Machine with HumidAir™ Heated Humidifier |
| Mask: AirTouch™ F20 Full Face CPAP Mask with Headgear |
Airsense 10 Autoset
AirTouch and AirFit F10 masks

AirTouch and AirFit F10 masks
Re: ..now a new problem with my new RESMED model 11 ELITE
My cat has NOT THE SLIGHTEST interest in my machine, but he REALLY wants to play with my USB charging cables.chunkyfrog wrote: ↑Sun Sep 05, 2021 4:30 pmCool-down mode fascinates the cat.
First human to channel a cat.
How do you like laser pointers?
![]()
![]()
Another thing: if you want to carry this all the way,
remember cats do not need toilet paper.
_________________
| Machine: ResMed AirSense™ 10 AutoSet™ CPAP Machine with HumidAir™ Heated Humidifier |
| Mask: AirTouch™ F20 Full Face CPAP Mask with Headgear |
Airsense 10 Autoset
AirTouch and AirFit F10 masks

AirTouch and AirFit F10 masks