Aerophagia

General Discussion on any topic relating to CPAP and/or Sleep Apnea.
ILoveFlowers
Posts: 159
Joined: Sat Feb 20, 2021 1:58 pm
Location: Pacific Northwest

Aerophagia

Post by ILoveFlowers » Wed Jun 23, 2021 11:35 am

After switching to an Autoset for Her machine, I am having aerophagia so bad that I have to stop using the machine after a few hours. I've tried reducing max pressure, setting EPR to 3, and switching from the "for her" algorithm to apap but these haven't made a difference. I also set the starting pressure as low as I could without feeling air hunger. Two nights ago it went beyond an annoyance to seriously distressing abdominal pain and now I'm afraid to use it again.

I tried to get the data into Oscar so I could have a look at it but there seems to be a problem importing it. Maybe the SD card is too big? 64G was the smallest they had at Walmart. I can view the card in Windows explorer and see there is data on it, but it won't import.

I don't understand why the Airsense is giving me so much more trouble than my System One did? I went back to the System One last night and I realized I did have a small amount of bloating in the morning, maybe I haven't really noticed it before because it's so minor, but for some reason the Airsense is blowing me up like a pufferfish.

Anyway, I'm not really sure where to go from here. I feel like I'm between a rock and a hard place, I can't sleep without my cpap but now I can't sleep with it, either.
AirSense 10 AutoSet For Her
Airfit P10 mask

User avatar
Dog Slobber
Posts: 4336
Joined: Thu Feb 15, 2018 2:05 pm
Location: Ontario, Canada

Re: Aerophagia

Post by Dog Slobber » Wed Jun 23, 2021 12:07 pm

Import the data from the System One into OSCAR, then post your graph, make a not if you experienced Aerophagia.

We can then set the max on your AutoSet to the maximum your System was was achieving on nights where you weren't experiencing problems.

You can also consider setting the Response to Soft to see if that helps.
Battery Backup: EcoFlow Delta 2

ILoveFlowers
Posts: 159
Joined: Sat Feb 20, 2021 1:58 pm
Location: Pacific Northwest

Re: Aerophagia

Post by ILoveFlowers » Wed Jun 23, 2021 4:12 pm

Oh, thanks. I've got to remember now how to post a graph. Are you saying, though, to set the max pressure on the Autoset to the pressure I've been maxing out at on the System One? I'm just looking back at my previous data now and under details it has pressure: min med 95% max. Looks like I never actually go up to the number that the machine's max is set to. Most nights I am maxing at 7.50. I noticed when I was looking at the data on the Autoset screen, it had me at 8.8 when I woke up.
AirSense 10 AutoSet For Her
Airfit P10 mask

User avatar
Dog Slobber
Posts: 4336
Joined: Thu Feb 15, 2018 2:05 pm
Location: Ontario, Canada

Re: Aerophagia

Post by Dog Slobber » Wed Jun 23, 2021 5:02 pm

Yes.

Let's say on your System One, during an evening where you don't experience aerophagia the maximum your pressure get to is 7.5. But on your ResMed at 8.8. your experiencing bad aerophagia, then your ResMed is achieving pressures that is triggering your aerophagia, that your System one wasn't getting to.

7.5 may not be low enough, but it's a good starting point.

Lowering your maximum may sacrifice some therapy, but when one is suffering from aerophagia, sometimes a compromise between addressing apneas and aerophagia must found.
Battery Backup: EcoFlow Delta 2

User avatar
zonker
Posts: 11491
Joined: Fri Jun 19, 2015 4:36 pm

Re: Aerophagia

Post by zonker » Wed Jun 23, 2021 6:03 pm

dogslobber speaks with straight tongue. once we find that pressure combo that won't trigger aerophagia, we can start "inching" you upward. the very best thing about your resmed is that it can be increased in increments of .2 which will help a lot.

in what ds has named "the zonker protocol", we can slowly increase until it all comes together. you'll just have to put up with a higher ahi for awhile.

but let's get you going with no aerophagia first, then we'll dive into the protocol.

edited: corrected spelling
Last edited by zonker on Wed Jun 23, 2021 7:37 pm, edited 1 time in total.
"Age is not an accomplishment and youth is not a sin"-Robert A. Heinlein
Oscar-Win
https://www.apneaboard.com/OSCAR/OSCAR-1.5.1-Win64.exe
Oscar-Mac
https://www.apneaboard.com/OSCAR/OSCAR-1.5.1.dmg

GrumpyHere
Posts: 523
Joined: Sat Feb 13, 2021 2:40 am

Re: Aerophagia

Post by GrumpyHere » Wed Jun 23, 2021 6:59 pm

Follow dogslobber and zonker's advice for translating your System One settings.

ILoveFlowers wrote: ↑
Wed Jun 23, 2021 11:35 am
the SD card is too big? 64G was the smallest they had at Walmart. I can view the card in Windows explorer and see there is data on it, but it won't import.
Regarding the new SD card:
BlueDragon wrote: ↑
Sat Dec 19, 2020 10:25 pm
If I recall correctly, a ResMed machine will simply ignore a 64 GB card and write no data to the card.

The easiest solution is to delete the 64GB partition on the SD card, create a new 31GB partition, and format the partition as FAT-32. On Windows, you can do that from Disk Management.

Sure, you are giving up half the space on the card but it is never going to be used anyway on a CPAP machine. Even a 4GB card will hold years of data.
ResMed Lumis Tx

ILoveFlowers
Posts: 159
Joined: Sat Feb 20, 2021 1:58 pm
Location: Pacific Northwest

Re: Aerophagia

Post by ILoveFlowers » Wed Jun 23, 2021 10:56 pm

Thank you! You guys are amazing.
AirSense 10 AutoSet For Her
Airfit P10 mask

User avatar
Jlfinkels
Posts: 468
Joined: Sun Jun 20, 2021 3:51 pm

Re: Aerophagia

Post by Jlfinkels » Thu Jun 24, 2021 7:27 am

For SD cards Amazon has plenty of 16GB cards for a few dollars and is next day delivery with Prime.
Sometimes it is the very people who no one imagines anything of who do the things that no one imagines

ILoveFlowers
Posts: 159
Joined: Sat Feb 20, 2021 1:58 pm
Location: Pacific Northwest

Re: Aerophagia

Post by ILoveFlowers » Thu Jun 24, 2021 12:08 pm

It worked! That is the best night's sleep I have had since the whole recall thing started. Thank you so much! I was too tired to reformat the SD card last night so I did it today, I'll try it out tonight. If it doesn't work, I'll go the Amazon route. You guys rock!
AirSense 10 AutoSet For Her
Airfit P10 mask

User avatar
zonker
Posts: 11491
Joined: Fri Jun 19, 2015 4:36 pm

Re: Aerophagia

Post by zonker » Thu Jun 24, 2021 3:49 pm

Image
"Age is not an accomplishment and youth is not a sin"-Robert A. Heinlein
Oscar-Win
https://www.apneaboard.com/OSCAR/OSCAR-1.5.1-Win64.exe
Oscar-Mac
https://www.apneaboard.com/OSCAR/OSCAR-1.5.1.dmg

ILoveFlowers
Posts: 159
Joined: Sat Feb 20, 2021 1:58 pm
Location: Pacific Northwest

Re: Aerophagia

Post by ILoveFlowers » Thu Jul 01, 2021 10:41 pm

As I'm working on easing the settings upwards, how long should I give each setting to get a feel for how it is working?
AirSense 10 AutoSet For Her
Airfit P10 mask

User avatar
zonker
Posts: 11491
Joined: Fri Jun 19, 2015 4:36 pm

Re: Aerophagia

Post by zonker » Fri Jul 02, 2021 11:41 am

ILoveFlowers wrote: ↑
Thu Jul 01, 2021 10:41 pm
As I'm working on easing the settings upwards, how long should I give each setting to get a feel for how it is working?
give it a couple of nights to be sure. at least, that's what i did.

now, obviously, if it has gotten WORSE, then feel free to back off of it, then come back to it the next night. this is why this process takes so bloody long.

good luck!
"Age is not an accomplishment and youth is not a sin"-Robert A. Heinlein
Oscar-Win
https://www.apneaboard.com/OSCAR/OSCAR-1.5.1-Win64.exe
Oscar-Mac
https://www.apneaboard.com/OSCAR/OSCAR-1.5.1.dmg

ILoveFlowers
Posts: 159
Joined: Sat Feb 20, 2021 1:58 pm
Location: Pacific Northwest

Re: Aerophagia

Post by ILoveFlowers » Fri Jul 02, 2021 2:29 pm

Thank you, Zonker! Will do.
AirSense 10 AutoSet For Her
Airfit P10 mask

User avatar
zonker
Posts: 11491
Joined: Fri Jun 19, 2015 4:36 pm

Re: Aerophagia

Post by zonker » Fri Jul 02, 2021 4:12 pm

ILoveFlowers wrote: ↑
Fri Jul 02, 2021 2:29 pm
Thank you, Zonker! Will do.
off topic, but it would help me out.

could you start using the quote feature? when you do that, i'll get a notification that you've quoted me and then i'll be able reply to you faster.

thank you.
"Age is not an accomplishment and youth is not a sin"-Robert A. Heinlein
Oscar-Win
https://www.apneaboard.com/OSCAR/OSCAR-1.5.1-Win64.exe
Oscar-Mac
https://www.apneaboard.com/OSCAR/OSCAR-1.5.1.dmg

ILoveFlowers
Posts: 159
Joined: Sat Feb 20, 2021 1:58 pm
Location: Pacific Northwest

Re: Aerophagia

Post by ILoveFlowers » Fri Jul 02, 2021 7:50 pm

zonker wrote: ↑
Fri Jul 02, 2021 4:12 pm
ILoveFlowers wrote: ↑
Fri Jul 02, 2021 2:29 pm
Thank you, Zonker! Will do.
off topic, but it would help me out.

could you start using the quote feature? when you do that, i'll get a notification that you've quoted me and then i'll be able reply to you faster.

thank you.
Okay, I hadn't thought about the notification feature, thanks for pointing that out.
AirSense 10 AutoSet For Her
Airfit P10 mask