Stomach issues with cpap

General Discussion on any topic relating to CPAP and/or Sleep Apnea.
ElCamello
Posts: 19
Joined: Thu May 28, 2020 10:53 pm

Stomach issues with cpap

Post by ElCamello » Thu May 28, 2020 11:22 pm

Hi all.
I am 45 years old male I was diagnosed with sleep apnea about a year ago. I use a a Philips DreamStation Auto CPAP Machine and a Full face mask form CARA.
The last months were really miserable for me. My sleep is interrupted I wake up exhausted in the morning with nausea. I had episodes of rapid heart rate during the day. I had a 24 hr ECG and blood tests but all came ok and docs are blaming everything on stress.
My CPAP never worked for me and my CPAP was always reporting a high AHI but I think the Situation escalated about 3 months ago, I went back then to my sleep therapist and told him I am constantly waking up at night. He said your CPAP pressure might not be high enough and he adjusted the pressure range form 4-7 to 4-8 range (like that without a sleep study).
After a month I went back to him and told him that I wake up destroyed in the mornings. The doc gave me a sleep study at home and he said my AHI was 4 and My oxygen levels were good and everything seemed ok, I might need to check if my blood pressure is dropping at night. The funny thing that night the CPAP was reporting an AHI of 9 which did not match what the sleep study apparatus reported.
At the time I was using a nose mask, Couple of days later I found out that I wake up with my mouth open and air coming out of my mouth so I switched to a full face mask.
The last weeks I was constantly suffering from nausea, feeling dizzy specially after eating and I feel I have too much air in my stomach that only gets better after a long walks. I also noticed my heart rate jumps up very easily while basically doing no significant effort.
Now if I think of it all my health troubles happened after adjusting the CPAP pressure max level from 7 to 8. I am not sure if that is a coincidence or the root cause.
Am I swallowing too much air when asleep ? Can too much air in the stomach impact the heart? Is the CPAP pressure to blame? What else should I do?

_________________
Machine: DreamStation Auto CPAP Machine
Additional Comments: Mask : CARA Full Face
Last edited by ElCamello on Fri May 29, 2020 4:13 am, edited 1 time in total.

User avatar
Julie
Posts: 20051
Joined: Tue Feb 28, 2006 12:58 pm

Re: Stomach issues with cpap

Post by Julie » Fri May 29, 2020 12:52 am

You can give us more information to go on - like the full name and model of your machine, which mask you have, what your pressure settings are right now, if you take any medication and which. It is possible to have some symptoms of other problems while on Cpap, and I'm not sure your doctor knows enough to be setting your machine. Please read the notes at the top of the page - https://www.sleepfiles.com/OSCAR/
and follow the directions about how to post your results here so the experts can interpret what's going on and help you to reset whatever is not working. The software is free and very comprehensive.

ElCamello
Posts: 19
Joined: Thu May 28, 2020 10:53 pm

Re: Stomach issues with cpap

Post by ElCamello » Fri May 29, 2020 3:52 am

Hi all,
I have uploaded my dream station data to OSCAR. I have attached two screen shots form my yesterday's sleep analysis :
The last weeks I was under a lot of stress and I am currently taking Neurexan to reduce my stress levels.
28_May_2020_v2.png
28_May_2020_v2.png (155.75 KiB) Viewed 2602 times
28_May_2020_v1.png
28_May_2020_v1.png (181.25 KiB) Viewed 2602 times

_________________
Machine: DreamStation Auto CPAP Machine
Additional Comments: Mask : CARA Full Face
Last edited by ElCamello on Fri May 29, 2020 5:58 am, edited 2 times in total.

ElCamello
Posts: 19
Joined: Thu May 28, 2020 10:53 pm

Re: Stomach issues with cpap

Post by ElCamello » Fri May 29, 2020 4:03 am

28_May_2020_overview.png
28_May_2020_overview.png (62.63 KiB) Viewed 2601 times

_________________
Machine: DreamStation Auto CPAP Machine
Additional Comments: Mask : CARA Full Face

User avatar
Julie
Posts: 20051
Joined: Tue Feb 28, 2006 12:58 pm

Re: Stomach issues with cpap

Post by Julie » Fri May 29, 2020 7:25 am

bump

User avatar
LSAT
Posts: 13405
Joined: Sun Nov 16, 2008 10:11 am
Location: SE Wisconsin

Re: Stomach issues with cpap

Post by LSAT » Fri May 29, 2020 7:42 am

Max pressure of 8 is the problem...should be 20. Minimum at 6.... The machine wants to go higher to stop apneas and it can't. It's blocked at 8. Even with 20, the machine will only go as high as necessary to stop apneas...it may never go higher than 10 or 12, but we will never know unless we let it run for a few days.

ElCamello
Posts: 19
Joined: Thu May 28, 2020 10:53 pm

Re: Stomach issues with cpap

Post by ElCamello » Fri May 29, 2020 8:19 am

Thank you for the feedback. My CPAP is locked, only my CPAP provider is allowed to change the pressure. I need to talk to my doctor in order to change the pressure. Could you please let me know how did you get to the conclusion my CPAP is trying to go higher. Is there a specific graph or portion of a graph showing this.

_________________
Machine: DreamStation Auto CPAP Machine
Additional Comments: Mask : CARA Full Face

User avatar
Dog Slobber
Posts: 4330
Joined: Thu Feb 15, 2018 2:05 pm
Location: Ontario, Canada

Re: Stomach issues with cpap

Post by Dog Slobber » Fri May 29, 2020 8:41 am

ElCamello wrote:
Fri May 29, 2020 8:19 am
Thank you for the feedback. My CPAP is locked, only my CPAP provider is allowed to change the pressure. I need to talk to my doctor in order to change the pressure. Could you please let me know how did you get to the conclusion my CPAP is trying to go higher. Is there a specific graph or portion of a graph showing this.
If you take a look at the Pressure graph portion on the second image, you will see the pressure rise to 8 is is stuck for virtually most of the night.

If you take a look during these periods, you will also see a large amount of events occurring that you machine can't properly address because the pressure isn't high enough.

While, I agree with LSAT about changing your pressure to 6-20 (for most people), I would be hesitant to move your maximum to 20 for you just yet.

In your original post you also described symptoms that could be attributed to Aerophagia (bloating from too much air from CPAP). Increasing your pressure so abruptly could increase aerophagia dramatically and well beyond your tolerance.

I would recommend increasing your maximum pressure, but maybe 1 CM at a time over a longer period. Keep an eye on your AHI, but also monitor how your tummy feels. The goal is to eventually get you to 20 CM.

And yes you can change your machine, you just don't know how to get into the secret mode to do so.

Regarding your Doctor. How in the hell could the have let you go so long with such a high machine reported AHI, while seeing that your device was clearly not supplying enough pressure. I wouldn't go anywhere near him again.
Battery Backup: EcoFlow Delta 2

User avatar
Julie
Posts: 20051
Joined: Tue Feb 28, 2006 12:58 pm

Re: Stomach issues with cpap

Post by Julie » Fri May 29, 2020 8:59 am

Your machine is not locked - that's the lie patients are told so doctor continues to have power! We've all been changing our pressures forever, you just need to get the clinician manual from Apneaboard.com - find your machine on the list and follow the instructions which will be mailed to you. It's your therapy and you should be in charge. We'll help you to optimize it.

User avatar
Miss Emerita
Posts: 3783
Joined: Sun Nov 04, 2018 8:07 pm

Re: Stomach issues with cpap

Post by Miss Emerita » Fri May 29, 2020 10:18 am

This is a new low in the annals of sleep-doctoring. I'm so sorry you've had to endure so much trouble without competent medical help. You absolutely should be adjusting your own pressure settings, with the kind of good advice you've already gotten in this thread. To sum the advice up: reset your minimum to 6 and your maximum to 9 and then post a chart.

When you post again, you need to post just the chart with these graphs:

Events
Flow rate
Pressure
Leaks
Snores.

I see on line that Neurexan is an over-the-counter preparation that can help with anxiety and insomnia. Side effects are minimal, BUT it contains lactose. Are you lactose intolerant? If you are, or even might be, I would suggest you work with your doctor to find a different drug. That might help with the belly issues.
Oscar software is available at https://www.sleepfiles.com/OSCAR/

ElCamello
Posts: 19
Joined: Thu May 28, 2020 10:53 pm

Re: Stomach issues with cpap

Post by ElCamello » Fri May 29, 2020 11:33 am

thank you all for your amazing help and advice.
Miss Emerita you are right, I just figured out this morning that Neurexan has lactose. I am lactose intolerant and that explains all the nausea bloating and stomach discomfort I was having.

_________________
Machine: DreamStation Auto CPAP Machine
Additional Comments: Mask : CARA Full Face

User avatar
zonker
Posts: 11458
Joined: Fri Jun 19, 2015 4:36 pm

Re: Stomach issues with cpap

Post by zonker » Fri May 29, 2020 12:11 pm

Miss Emerita wrote:
Fri May 29, 2020 10:18 am

I see on line that Neurexan is an over-the-counter preparation that can help with anxiety and insomnia. Side effects are minimal, BUT it contains lactose. Are you lactose intolerant? If you are, or even might be, I would suggest you work with your doctor to find a different drug. That might help with the belly issues.
remarkable, holmes!

kudos for a very good catch.
"Age is not an accomplishment and youth is not a sin"-Robert A. Heinlein
Oscar-Win
https://www.apneaboard.com/OSCAR/OSCAR-1.5.1-Win64.exe
Oscar-Mac
https://www.apneaboard.com/OSCAR/OSCAR-1.5.1.dmg

User avatar
Miss Emerita
Posts: 3783
Joined: Sun Nov 04, 2018 8:07 pm

Re: Stomach issues with cpap

Post by Miss Emerita » Fri May 29, 2020 12:19 pm

Why thank you, sir. I must look into clutching a pipe as well as a scepter. Scepter in hand, pipe between teeth?

I'm quite lactose-intolerant myself, so I'm always on the alert. For anyone else who might be interested, here's information from the NIH:

"Approximately 65 percent of the human population has a reduced ability to digest lactose after infancy. Lactose intolerance in adulthood is most prevalent in people of East Asian descent, with 70 to 100 percent of people affected in these communities. Lactose intolerance is also very common in people of West African, Arab, Jewish, Greek, and Italian descent.

The prevalence of lactose intolerance is lowest in populations with a long history of dependence on unfermented milk products as an important food source. For example, only about 5 percent of people of Northern European descent are lactose intolerant."

More here: https://ghr.nlm.nih.gov/condition/lactose-intolerance#
Oscar software is available at https://www.sleepfiles.com/OSCAR/

User avatar
Dog Slobber
Posts: 4330
Joined: Thu Feb 15, 2018 2:05 pm
Location: Ontario, Canada

Re: Stomach issues with cpap

Post by Dog Slobber » Fri May 29, 2020 12:32 pm

zonker wrote:
Fri May 29, 2020 12:11 pm
Miss Emerita wrote:
Fri May 29, 2020 10:18 am

I see on line that Neurexan is an over-the-counter preparation that can help with anxiety and insomnia. Side effects are minimal, BUT it contains lactose. Are you lactose intolerant? If you are, or even might be, I would suggest you work with your doctor to find a different drug. That might help with the belly issues.
remarkable, holmes!

kudos for a very good catch.
Absolutely! Nice catch Ms. Miss!
Battery Backup: EcoFlow Delta 2

ElCamello
Posts: 19
Joined: Thu May 28, 2020 10:53 pm

Re: Stomach issues with cpap

Post by ElCamello » Sat May 30, 2020 7:31 am

Last night and this morning were a nightmare.
I did not take Nurexan for the entire day. before going to bed and while brushing my teeth my heart rate was jumping to 150. I had to lay in bed so it calms down.
I woke up at 2 am and could not get back to sleep. My sleep was disrupted the entire night. I was laying in bed and feeling my legs my arms and back all burning form the inside. Like at the bones levels.
I am suspecting too much stress hormone is released in my body. Any ideas ?
In the morning I went to wash my face and my heart rate was jumping to 140. I took Nurexan and a beta blocker that my doc gave me and told me I can take in case I have heart racing again. and than went back to bed.
Now my heart rate is back to normal but emotionally and psychologically I am devastated. Not knowing what really is happening to me.
I am continuing taking Nurexan for the time being I ordered a lactose free variant but will can not get it till Tuesday.
below are my recordings form last night :
29_May_2020_v1.png
29_May_2020_v1.png (189.6 KiB) Viewed 2475 times

_________________
Machine: DreamStation Auto CPAP Machine
Additional Comments: Mask : CARA Full Face