Things...

General Discussion on any topic relating to CPAP and/or Sleep Apnea.
DMA97701
Posts: 11
Joined: Thu May 14, 2020 6:21 am

Things...

Post by DMA97701 » Thu May 14, 2020 6:57 am

First, both my CPAP machine and humidifier are not listed in the respective drop-down lists, so I am unable to complete my profile.

Now that that is out of the way, I am wondering in anyone might be able to assist with my CPAP settings.

I participated in a sleep study years back and at the end of the session I was diagnosed with "positional sleep apnea," meaning it impacted me the least when sleeping on my side. Now, that might be common enough, but you'll never guess what my prescription was...a tee-shirt with a pocket in the middle of the back, in which a styrofoam tube was inserted to force side sleeping. That was a $2,500 shirt! Is that really a thing?! Needless to say, I have never returned.

Over the years, my apnea has apparently grown more noticeable to the point where my wife says I stop breathing and if I don't wake myself, she wakes me. This is, of course, in addition to the snoring.

So, I decided to pick up a CPAP and give 'er a go. Yup, you guessed it, that's why I'm inquiring about assistance with the settings. I've played around with the clinical settings and tried about every combination of changes I could, but just can't seem to find the right mix to help me make it past that three hour mark. At or around that point, I wake up because it feels like the pressure is way too high, the mask feels as though it has severely tightened due to said pressure, and it's too difficult to exhale. I'm primarily using the "myAir" app on my phone for the results, but have downloaded a couple of other PC-based applications.

I know everyone will probably advise another sleep study, but with the results I received previously, I have no desire to return.

Anyway, my CPAP is a RedMed AirCurve 10 ASV, with heated humidifier and my mask is the Mirage Quattro full face mask with headgear.

Anyone out there care to chime in on possible modifications to the min and max epap settings as well as the min and max ps settings?

Thanx in advance.

User avatar
roadcycler
Posts: 142
Joined: Mon Oct 15, 2018 10:08 am
Location: BC Canada

Re: Things...

Post by roadcycler » Thu May 14, 2020 7:48 am

Well welcome aboard.

First thing you have to do is to download Oscar https://www.sleepfiles.com/OSCAR/. Read on the top posts on what graphs are needed and then post snapshots. Until we can see what is going on giving you any advice is just a shot in the dark blindfolded.

User avatar
Julie
Posts: 20051
Joined: Tue Feb 28, 2006 12:58 pm

Re: Things...

Post by Julie » Thu May 14, 2020 7:51 am

You can't find the listing because of your spelling - It's Resmed, not Redmed.

User avatar
Pugsy
Posts: 65358
Joined: Thu May 14, 2009 9:31 am
Location: Missouri, USA

Re: Things...

Post by Pugsy » Thu May 14, 2020 8:00 am

When you can't find a machine model in the drop down menu we advise people to simply list the model in the comments section.

All equipment menu choices are tied to cpap.com current inventory and if not sold right now....won't be available as a choice.

_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier
Additional Comments: Mask Bleep Eclipse https://bleepsleep.com/the-eclipse/
I may have to RISE but I refuse to SHINE.

User avatar
LSAT
Posts: 13405
Joined: Sun Nov 16, 2008 10:11 am
Location: SE Wisconsin

Re: Things...

Post by LSAT » Thu May 14, 2020 8:09 am

Is there any specific reason for purchasing an ASV machine...That is a very expensive specialized machine for treating Central Apnea. If you don't understand the machine it's difficult to use for standard sleep apnea.

User avatar
Pugsy
Posts: 65358
Joined: Thu May 14, 2009 9:31 am
Location: Missouri, USA

Re: Things...

Post by Pugsy » Thu May 14, 2020 8:18 am

DMA97701 wrote:
Thu May 14, 2020 6:57 am
Anyone out there care to chime in on possible modifications to the min and max epap settings as well as the min and max ps settings?
Gotta see the detailed reports before we can offer any ideas.

Did you download OSCAR?

OSCAR https://www.sleepfiles.com/OSCAR/
http://www.apneaboard.com/wiki/index.ph ... stallation
http://www.apneaboard.com/wiki/index.ph ... rpretation

_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier
Additional Comments: Mask Bleep Eclipse https://bleepsleep.com/the-eclipse/
I may have to RISE but I refuse to SHINE.

DMA97701
Posts: 11
Joined: Thu May 14, 2020 6:21 am

Re: Things...

Post by DMA97701 » Thu May 14, 2020 11:44 am

Julie wrote:
Thu May 14, 2020 7:51 am
You can't find the listing because of your spelling - It's Resmed, not Redmed.
Yeah, dangit. Misspelled it in the post, but still don't see the "exact" product (ResMed AirCurve 10 ASV) listed in the drop-down (see below). Should I just use one of the other AirCurve models listed or just use the suggested spell-it-out-in-the-comment-field choice? Thanx.

Image
Machine: ResMed AirCurve 10 ASV
Humidifier: ResMed HumidAir S
Tubing: ResMed ClimateLineAir Heated Tubing
Mask: ResMed Mirage Quattro Full Face Mask with Headgear
Sanitation: SoClean 2 CPAP Cleaner/Sanitizer
Software: OSCAR 1.1.0

DMA97701
Posts: 11
Joined: Thu May 14, 2020 6:21 am

Re: Things...

Post by DMA97701 » Thu May 14, 2020 11:47 am

Pugsy wrote:
Thu May 14, 2020 8:18 am
DMA97701 wrote:
Thu May 14, 2020 6:57 am
Anyone out there care to chime in on possible modifications to the min and max epap settings as well as the min and max ps settings?
Gotta see the detailed reports before we can offer any ideas.

Did you download OSCAR?

OSCAR https://www.sleepfiles.com/OSCAR/
http://www.apneaboard.com/wiki/index.ph ... stallation
http://www.apneaboard.com/wiki/index.ph ... rpretation

Just downloaded, installed, and configured OSCAR. As soon as I download the data from the card, I'll upload it. Thanx in advance.
Machine: ResMed AirCurve 10 ASV
Humidifier: ResMed HumidAir S
Tubing: ResMed ClimateLineAir Heated Tubing
Mask: ResMed Mirage Quattro Full Face Mask with Headgear
Sanitation: SoClean 2 CPAP Cleaner/Sanitizer
Software: OSCAR 1.1.0

User avatar
chunkyfrog
Posts: 34544
Joined: Mon Jul 12, 2010 5:10 pm
Location: Nowhere special--this year in particular.

Re: Things...

Post by chunkyfrog » Thu May 14, 2020 12:41 pm

Twelve hundred bucks for a T-SHIRT?
Can't blame you for not wanting to go back.

_________________
Mask: AirFit™ P10 For Her Nasal Pillow CPAP Mask with Headgear
Additional Comments: Airsense 10 Autoset for Her

User avatar
Pugsy
Posts: 65358
Joined: Thu May 14, 2009 9:31 am
Location: Missouri, USA

Re: Things...

Post by Pugsy » Thu May 14, 2020 12:50 pm

People should always just add the correct machine model in the comments section manually.

To choose a machine that "is close" will just cause confusion because each machine can do different things.

Leave the machine choice blank and just manually add it to the comments section....this will avoid potential confusion and delay in getting optimal suggestions to best deal with any issues.

_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier
Additional Comments: Mask Bleep Eclipse https://bleepsleep.com/the-eclipse/
I may have to RISE but I refuse to SHINE.

DMA97701
Posts: 11
Joined: Thu May 14, 2020 6:21 am

Re: Things...

Post by DMA97701 » Thu May 14, 2020 1:22 pm

chunkyfrog wrote:
Thu May 14, 2020 12:41 pm
Twelve hundred bucks for a T-SHIRT?
Can't blame you for not wanting to go back.
Double that and add another $100.

:shock:
Machine: ResMed AirCurve 10 ASV
Humidifier: ResMed HumidAir S
Tubing: ResMed ClimateLineAir Heated Tubing
Mask: ResMed Mirage Quattro Full Face Mask with Headgear
Sanitation: SoClean 2 CPAP Cleaner/Sanitizer
Software: OSCAR 1.1.0

DMA97701
Posts: 11
Joined: Thu May 14, 2020 6:21 am

Re: Things...

Post by DMA97701 » Thu May 14, 2020 2:12 pm

LSAT wrote:
Thu May 14, 2020 8:09 am
Is there any specific reason for purchasing an ASV machine...That is a very expensive specialized machine for treating Central Apnea. If you don't understand the machine it's difficult to use for standard sleep apnea.
Got a killer deal on the machine. Wasn't looking for any specific type of machine, just looking for a CPAP in general. Picked this up for around $600.

On another note, I noticed your avatar...GO PACK GO!!!
Machine: ResMed AirCurve 10 ASV
Humidifier: ResMed HumidAir S
Tubing: ResMed ClimateLineAir Heated Tubing
Mask: ResMed Mirage Quattro Full Face Mask with Headgear
Sanitation: SoClean 2 CPAP Cleaner/Sanitizer
Software: OSCAR 1.1.0

DMA97701
Posts: 11
Joined: Thu May 14, 2020 6:21 am

Re: Things...

Post by DMA97701 » Thu May 14, 2020 2:56 pm

Pugsy wrote:
Thu May 14, 2020 8:18 am
DMA97701 wrote:
Thu May 14, 2020 6:57 am
Anyone out there care to chime in on possible modifications to the min and max epap settings as well as the min and max ps settings?
Gotta see the detailed reports before we can offer any ideas.

Did you download OSCAR?

OSCAR https://www.sleepfiles.com/OSCAR/
http://www.apneaboard.com/wiki/index.ph ... stallation
http://www.apneaboard.com/wiki/index.ph ... rpretation
I've provided links to the reports (only 5 days so far) on my site. Hopefully it will give some insight. Thanx.

<links removed>
Last edited by DMA97701 on Thu May 14, 2020 6:07 pm, edited 1 time in total.
Machine: ResMed AirCurve 10 ASV
Humidifier: ResMed HumidAir S
Tubing: ResMed ClimateLineAir Heated Tubing
Mask: ResMed Mirage Quattro Full Face Mask with Headgear
Sanitation: SoClean 2 CPAP Cleaner/Sanitizer
Software: OSCAR 1.1.0

User avatar
Pugsy
Posts: 65358
Joined: Thu May 14, 2009 9:31 am
Location: Missouri, USA

Re: Things...

Post by Pugsy » Thu May 14, 2020 3:14 pm

Your current settings are doing a very decent job of handling your OSA.

I don't know that they need adjusting at this time. This machine is pretty good at dealing with OSA at the default settings (pretty much what you are using) as long as the EPAP is high enough to prevent the airway from collapsing and it appears that your settings are doing that fairly well.

You are having some awfully short sleep sessions though...any special reason why they are so short?
Are you having trouble sleeping with the machine doing its thing with these settings?

Oh....BTW...those sleep reports aren't the best for evaluating things....a detailed report is better.
Explained with examples in this thread.
viewtopic/t158560/How-to-post-images-for-review.html

_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier
Additional Comments: Mask Bleep Eclipse https://bleepsleep.com/the-eclipse/
I may have to RISE but I refuse to SHINE.

DMA97701
Posts: 11
Joined: Thu May 14, 2020 6:21 am

Re: Things...

Post by DMA97701 » Thu May 14, 2020 4:25 pm

Pugsy wrote:
Thu May 14, 2020 3:14 pm
You are having some awfully short sleep sessions though...any special reason why they are so short?
Are you having trouble sleeping with the machine doing its thing with these settings?
Yes, that's exactly why my sleep sessions are so short. I'm having major issues with the machine "doing it's thing." (lol)

I wake up because I can't breathe...due to the pressure being so high that it's difficult to exhale. Therefore, I reach over and hit the power button, remove the mask, and go back to my crappy, constantly interrupted sleep.

How the heck do I get past this?

Thanx again.
Machine: ResMed AirCurve 10 ASV
Humidifier: ResMed HumidAir S
Tubing: ResMed ClimateLineAir Heated Tubing
Mask: ResMed Mirage Quattro Full Face Mask with Headgear
Sanitation: SoClean 2 CPAP Cleaner/Sanitizer
Software: OSCAR 1.1.0