HIGH NUMBER OF EVENTS AT START OF SLEEP
Hello,
I have been noticing a high number of obstructive apnea events occurring in rapid succession. It usually stops during the night and the apneas reduce to zero events. It has happened a few times now so I am wondering what is the cause. Please have a look at the attached screen capture of my last night's rest. Please have a look and let me know what are your thoughts.
Baje
I have been noticing a high number of obstructive apnea events occurring in rapid succession. It usually stops during the night and the apneas reduce to zero events. It has happened a few times now so I am wondering what is the cause. Please have a look at the attached screen capture of my last night's rest. Please have a look and let me know what are your thoughts.
Baje
_________________
| Machine: DreamStation Auto CPAP Machine |
| Mask: ResMed AirFit™ F30 Full Face CPAP Mask with Headgear |
| Humidifier: DreamStation Heated Humidifier |
| Additional Comments: Using OSCAR......Using apap mode minimum pressure = 14 cmH20, Maximum 19.5 cmH2O |
Re: HIGH NUMBER OF EVENTS AT START OF SLEEP
Are you actually asleep when those events are flagged?
Can you zoom in on a few sections?
Watch the videos here.
http://freecpapadvice.com/sleepyhead-free-software
Please include the stuff on the left side of the detailed report....it's important.
See this
viewtopic/t158560/How-to-post-images-for-review.html
Can you zoom in on a few sections?
Watch the videos here.
http://freecpapadvice.com/sleepyhead-free-software
Please include the stuff on the left side of the detailed report....it's important.
See this
viewtopic/t158560/How-to-post-images-for-review.html
_________________
| Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier |
| Additional Comments: Mask Bleep Eclipse https://bleepsleep.com/the-eclipse/ |
I may have to RISE but I refuse to SHINE.
-
bajeiceman
- Posts: 20
- Joined: Sun Mar 03, 2019 2:06 pm
- Location: Barbados
Re: HIGH NUMBER OF EVENTS AT START OF SLEEP
I have added some more screenshots for you. i hope they give the information you need.
What I did realise after your question was that there seemed to be two sessions. I must have awoken and went back to sleep...….really don't remember that
and here is what the charts look like when they are separated
1st Hour
What I did realise after your question was that there seemed to be two sessions. I must have awoken and went back to sleep...….really don't remember that
1st Hour
_________________
| Machine: DreamStation Auto CPAP Machine |
| Mask: ResMed AirFit™ F30 Full Face CPAP Mask with Headgear |
| Humidifier: DreamStation Heated Humidifier |
| Additional Comments: Using OSCAR......Using apap mode minimum pressure = 14 cmH20, Maximum 19.5 cmH2O |
- Attachments
-
- Annotation 2020-02-02 191143_1st Hour sidebar.png (57.2 KiB) Viewed 1745 times
-
bajeiceman
- Posts: 20
- Joined: Sun Mar 03, 2019 2:06 pm
- Location: Barbados
Re: HIGH NUMBER OF EVENTS AT START OF SLEEP
Here is the 2nd Sleep session
_________________
| Machine: DreamStation Auto CPAP Machine |
| Mask: ResMed AirFit™ F30 Full Face CPAP Mask with Headgear |
| Humidifier: DreamStation Heated Humidifier |
| Additional Comments: Using OSCAR......Using apap mode minimum pressure = 14 cmH20, Maximum 19.5 cmH2O |
- Attachments
-
- Annotation 2020-02-02 191143_2nd Sleep side bar.png (58.93 KiB) Viewed 1745 times
Re: HIGH NUMBER OF EVENTS AT START OF SLEEP
You've not zoomed in enough. Did you watch those videos? They show the level of zooming in needed to maybe decide if awake or asleep.
If you aren't asleep when all those flagged events are noted....they don't count and have to be ignored.
You have to be asleep for anything flagged to be worth worrying about.
Did you read the thread I linked to? It shows how we like to see the detailed reports.
Overall image format. Level of zooming in format needed for flow rate

If you aren't asleep when all those flagged events are noted....they don't count and have to be ignored.
You have to be asleep for anything flagged to be worth worrying about.
Did you read the thread I linked to? It shows how we like to see the detailed reports.
Overall image format. Level of zooming in format needed for flow rate

_________________
| Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier |
| Additional Comments: Mask Bleep Eclipse https://bleepsleep.com/the-eclipse/ |
I may have to RISE but I refuse to SHINE.
-
bajeiceman
- Posts: 20
- Joined: Sun Mar 03, 2019 2:06 pm
- Location: Barbados
Re: HIGH NUMBER OF EVENTS AT START OF SLEEP
How are these
I am pretty sure I was sleeping when those events occurred.
I am pretty sure I was sleeping when those events occurred.
_________________
| Machine: DreamStation Auto CPAP Machine |
| Mask: ResMed AirFit™ F30 Full Face CPAP Mask with Headgear |
| Humidifier: DreamStation Heated Humidifier |
| Additional Comments: Using OSCAR......Using apap mode minimum pressure = 14 cmH20, Maximum 19.5 cmH2O |
Re: HIGH NUMBER OF EVENTS AT START OF SLEEP
Your breathing doesn't look like asleep breathing.
You've zoomed in a little too close this time but at least the breaths can be seen.
Asleep breathing flagged events here. Note the extremely rhythmic pattern and no irregularities.

Asleep breathing.

Asleep breathing circled in red. The rest of it is awake/arousal related even the part with the flags.

You've zoomed in a little too close this time but at least the breaths can be seen.
Asleep breathing flagged events here. Note the extremely rhythmic pattern and no irregularities.

Asleep breathing.

Asleep breathing circled in red. The rest of it is awake/arousal related even the part with the flags.

_________________
| Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier |
| Additional Comments: Mask Bleep Eclipse https://bleepsleep.com/the-eclipse/ |
I may have to RISE but I refuse to SHINE.
-
bajeiceman
- Posts: 20
- Joined: Sun Mar 03, 2019 2:06 pm
- Location: Barbados
Re: HIGH NUMBER OF EVENTS AT START OF SLEEP
Ok. Not sure what the take away is here. Did some checking of the machine when I got in from work this evening and found that the dehumidifier was not pushed all of the way in (must have occurred during reassembly after one of my recent travels). Anyway, let's see what happens tonight. Thanks for the thoughts though.
_________________
| Machine: DreamStation Auto CPAP Machine |
| Mask: ResMed AirFit™ F30 Full Face CPAP Mask with Headgear |
| Humidifier: DreamStation Heated Humidifier |
| Additional Comments: Using OSCAR......Using apap mode minimum pressure = 14 cmH20, Maximum 19.5 cmH2O |
Re: HIGH NUMBER OF EVENTS AT START OF SLEEP
*humidifier*.
Cpap's can add moisture, not take it away
Get OSCAR
Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.
Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.
-
bajeiceman
- Posts: 20
- Joined: Sun Mar 03, 2019 2:06 pm
- Location: Barbados
Re: HIGH NUMBER OF EVENTS AT START OF SLEEP
OOps!!! I meant humidifier! Thanks for pointing out the error.
_________________
| Machine: DreamStation Auto CPAP Machine |
| Mask: ResMed AirFit™ F30 Full Face CPAP Mask with Headgear |
| Humidifier: DreamStation Heated Humidifier |
| Additional Comments: Using OSCAR......Using apap mode minimum pressure = 14 cmH20, Maximum 19.5 cmH2O |
Re: HIGH NUMBER OF EVENTS AT START OF SLEEP
I point out things like that because they will confuse new people.bajeiceman wrote: ↑Tue Feb 04, 2020 9:35 amOOps!!! I meant humidifier! Thanks for pointing out the error.
Hope that getting the humidifier seated properly fixes your issue!
Get OSCAR
Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.
Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.
Re: HIGH NUMBER OF EVENTS AT START OF SLEEP
I have a similar issue. I suggest that you consider your meds; those that you take in the evening.
For me, I believe that the cause is a common prostate relaxant....and even sal pametto might cause the same slowed breathing.
After an hour or two, the AHI declines (improves) to about 8. This at a starting pressure of #16.
For now, I have started taking the meds in the AM.
A female sleep tech suggested the medication idea, but no one has suggested the best way to keep
my scores from sometimes starting in the mid 20's to 50's (AHI).
I am beginning to think that maybe a bipap will help, but my HMO is
slow to go in that direction, due to costs. More patient pressure is indicated.
For me, I believe that the cause is a common prostate relaxant....and even sal pametto might cause the same slowed breathing.
After an hour or two, the AHI declines (improves) to about 8. This at a starting pressure of #16.
For now, I have started taking the meds in the AM.
A female sleep tech suggested the medication idea, but no one has suggested the best way to keep
my scores from sometimes starting in the mid 20's to 50's (AHI).
I am beginning to think that maybe a bipap will help, but my HMO is
slow to go in that direction, due to costs. More patient pressure is indicated.
Re: HIGH NUMBER OF EVENTS AT START OF SLEEP
hello, lobo. if you want some advice on those pressure settings, please start your own post and let the experts here help you.Lobo7 wrote: ↑Sat Feb 08, 2020 2:10 pmI have a similar issue. I suggest that you consider your meds; those that you take in the evening.
For me, I believe that the cause is a common prostate relaxant....and even sal pametto might cause the same slowed breathing.
After an hour or two, the AHI declines (improves) to about 8. This at a starting pressure of #16.
For now, I have started taking the meds in the AM.
A female sleep tech suggested the medication idea, but no one has suggested the best way to keep
my scores from sometimes starting in the mid 20's to 50's (AHI).
I am beginning to think that maybe a bipap will help, but my HMO is
slow to go in that direction, due to costs. More patient pressure is indicated.
good luck!
_________________
| Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier |
| Mask: AirFit™ P10 Nasal Pillow CPAP Mask with Headgear |