Supplemental O2 ?

General Discussion on any topic relating to CPAP and/or Sleep Apnea.
russjpaul
Posts: 10
Joined: Fri Jan 10, 2020 5:46 pm

Re: Supplemental O2 ?

Post by russjpaul » Mon Jan 20, 2020 4:53 pm

Hmmm. not seeing any improvement so far. Think I just need to get used to the new settings ?
Attachments
oximetrygraphic012020.png
oximetrygraphic012020.png (22.97 KiB) Viewed 6907 times
oximetry012020.png
oximetry012020.png (37.81 KiB) Viewed 6907 times
Russ012020.png
Russ012020.png (112.06 KiB) Viewed 6907 times

User avatar
palerider
Posts: 32308
Joined: Wed Dec 16, 2009 5:43 pm
Location: Dallas(ish).

Re: Supplemental O2 ?

Post by palerider » Mon Jan 20, 2020 5:33 pm

Go to a minimum of 13, you're still struggling for air a lot of the time.
Get OSCAR

Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.

User avatar
zonker
Posts: 11509
Joined: Fri Jun 19, 2015 4:36 pm

Re: Supplemental O2 ?

Post by zonker » Mon Jan 20, 2020 6:36 pm

palerider wrote: ↑
Mon Jan 20, 2020 5:33 pm
Go to a minimum of 13, you're still struggling for air a lot of the time.
to enlighten me, could you point out where s/he is struggling for air? i'm not doubting it, i just don't know what to look for.

thanks!
"Age is not an accomplishment and youth is not a sin"-Robert A. Heinlein
Oscar-Win
https://www.apneaboard.com/OSCAR/OSCAR-1.5.1-Win64.exe
Oscar-Mac
https://www.apneaboard.com/OSCAR/OSCAR-1.5.1.dmg

User avatar
palerider
Posts: 32308
Joined: Wed Dec 16, 2009 5:43 pm
Location: Dallas(ish).

Re: Supplemental O2 ?

Post by palerider » Mon Jan 20, 2020 10:47 pm

zonker wrote: ↑
Mon Jan 20, 2020 6:36 pm
palerider wrote: ↑
Mon Jan 20, 2020 5:33 pm
Go to a minimum of 13, you're still struggling for air a lot of the time.
to enlighten me, could you point out where s/he is struggling for air? i'm not doubting it, i just don't know what to look for.

thanks!
The Flow Limitation graph. evidence of 'breathing through a straw' syndrome. :D
Get OSCAR

Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.

User avatar
zonker
Posts: 11509
Joined: Fri Jun 19, 2015 4:36 pm

Re: Supplemental O2 ?

Post by zonker » Mon Jan 20, 2020 10:57 pm

palerider wrote: ↑
Mon Jan 20, 2020 10:47 pm
zonker wrote: ↑
Mon Jan 20, 2020 6:36 pm
palerider wrote: ↑
Mon Jan 20, 2020 5:33 pm
Go to a minimum of 13, you're still struggling for air a lot of the time.
to enlighten me, could you point out where s/he is struggling for air? i'm not doubting it, i just don't know what to look for.

thanks!
The Flow Limitation graph. evidence of 'breathing through a straw' syndrome. :D
at the beginning, where it spikes so much?
"Age is not an accomplishment and youth is not a sin"-Robert A. Heinlein
Oscar-Win
https://www.apneaboard.com/OSCAR/OSCAR-1.5.1-Win64.exe
Oscar-Mac
https://www.apneaboard.com/OSCAR/OSCAR-1.5.1.dmg

User avatar
palerider
Posts: 32308
Joined: Wed Dec 16, 2009 5:43 pm
Location: Dallas(ish).

Re: Supplemental O2 ?

Post by palerider » Tue Jan 21, 2020 9:17 am

zonker wrote: ↑
Mon Jan 20, 2020 10:57 pm
palerider wrote: ↑
Mon Jan 20, 2020 10:47 pm
zonker wrote: ↑
Mon Jan 20, 2020 6:36 pm
palerider wrote: ↑
Mon Jan 20, 2020 5:33 pm
Go to a minimum of 13, you're still struggling for air a lot of the time.
to enlighten me, could you point out where s/he is struggling for air? i'm not doubting it, i just don't know what to look for.

thanks!
The Flow Limitation graph. evidence of 'breathing through a straw' syndrome. :D
at the beginning, where it spikes so much?
Well, the whole night, FLs seem to be what's mainly driving pressure increases, It's not apneas, hypos or snores.
Get OSCAR

Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.

User avatar
raisedfist
Posts: 1176
Joined: Wed Jun 15, 2016 7:21 am

Re: Supplemental O2 ?

Post by raisedfist » Tue Jan 21, 2020 11:07 am

The increases in minimum pressure seem to be helping even if they are thus far subtle. There is less sustained clumping of steep desaturations in your last SpO2 graph until about the last hour of the recording where it gets a bit nasty. You have justification to keep increasing the pressure since you have an ugly flow limitation graph and still some breathing events. The increases in pressure are still being done by the book so to speak :).
Philips Respironics Trilogy 100
AVAPS-AE Mode
PS Min 6, PS Max 18, EPAP Min 4, EPAP Max 12

User avatar
palerider
Posts: 32308
Joined: Wed Dec 16, 2009 5:43 pm
Location: Dallas(ish).

Re: Supplemental O2 ?

Post by palerider » Tue Jan 21, 2020 11:18 am

And keep in mind that with classic bilevel therapy in a clinical setting, EPAP is where oxygenation happens, and PS is for ventilation, aka removal of CO2. So if SpO2 is low, they raise EPAP.
Get OSCAR

Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.

russjpaul
Posts: 10
Joined: Fri Jan 10, 2020 5:46 pm

Re: Supplemental O2 ?

Post by russjpaul » Tue Jan 21, 2020 10:01 pm

No real imporovements to speak of yet guys....is it something that will improve as i get used to the new settings ?

Thanks for the help

R.
Attachments
oximetrygraphic012020.png
oximetrygraphic012020.png (22.97 KiB) Viewed 6864 times
oximetry012020.png
oximetry012020.png (37.81 KiB) Viewed 6864 times
Russ012020.png
Russ012020.png (112.06 KiB) Viewed 6864 times

User avatar
raisedfist
Posts: 1176
Joined: Wed Jun 15, 2016 7:21 am

Re: Supplemental O2 ?

Post by raisedfist » Wed Jan 22, 2020 9:14 pm

Did you accidentally post the same old images?
Philips Respironics Trilogy 100
AVAPS-AE Mode
PS Min 6, PS Max 18, EPAP Min 4, EPAP Max 12

russjpaul
Posts: 10
Joined: Fri Jan 10, 2020 5:46 pm

Re: Supplemental O2 ?

Post by russjpaul » Thu Jan 23, 2020 9:28 am

Looks like it - must have labelled the OSCAR file incorrectly. Here's last night's.....
Attachments
OSCAR012220.png
OSCAR012220.png (121.59 KiB) Viewed 6837 times

User avatar
raisedfist
Posts: 1176
Joined: Wed Jun 15, 2016 7:21 am

Re: Supplemental O2 ?

Post by raisedfist » Thu Jan 23, 2020 10:00 pm

Do you have a new corresponding SpO2 chart? The machine still wants to go higher, it doesn't spend much time at 13. I would keep raising the min pressure or drop down the EPR so there is less of a drop in effective pressure. A drop in EPR may have an effect on some of the data statistics but I am afraid it's probably just an experiment. And once you reach a min of 15 or so it may be worth looking at bi-level if it becomes uncomfortable.
Philips Respironics Trilogy 100
AVAPS-AE Mode
PS Min 6, PS Max 18, EPAP Min 4, EPAP Max 12

russjpaul
Posts: 10
Joined: Fri Jan 10, 2020 5:46 pm

Re: Supplemental O2 ?

Post by russjpaul » Sun Jan 26, 2020 1:21 pm

I think I see some definite improvements. Feel good too !!
Attachments
OSCAR012520.png
OSCAR012520.png (111.73 KiB) Viewed 6783 times
Oximetry012520.png
Oximetry012520.png (51.63 KiB) Viewed 6783 times
Oximetry Graphic 012520.png
Oximetry Graphic 012520.png (23.05 KiB) Viewed 6783 times

User avatar
zonker
Posts: 11509
Joined: Fri Jun 19, 2015 4:36 pm

Re: Supplemental O2 ?

Post by zonker » Sun Jan 26, 2020 1:47 pm

russjpaul wrote: ↑
Sun Jan 26, 2020 1:21 pm
I think I see some definite improvements. Feel good too !!

vincentpriceclapping.gif
most excellent news!!
"Age is not an accomplishment and youth is not a sin"-Robert A. Heinlein
Oscar-Win
https://www.apneaboard.com/OSCAR/OSCAR-1.5.1-Win64.exe
Oscar-Mac
https://www.apneaboard.com/OSCAR/OSCAR-1.5.1.dmg

User avatar
raisedfist
Posts: 1176
Joined: Wed Jun 15, 2016 7:21 am

Re: Supplemental O2 ?

Post by raisedfist » Sun Jan 26, 2020 8:03 pm

the dips below 88% SpO2 are better for sure
Philips Respironics Trilogy 100
AVAPS-AE Mode
PS Min 6, PS Max 18, EPAP Min 4, EPAP Max 12