I'm a dentist treating myself! CPAP Results... Advice please?

General Discussion on any topic relating to CPAP and/or Sleep Apnea.
nee
Posts: 121
Joined: Thu Jan 09, 2020 10:45 am
Location: Southern California

Re: I'm a dentist treating myself! CPAP Results... Advice please?

Post by nee » Fri Feb 14, 2020 11:06 am

musculus wrote:
Fri Feb 14, 2020 10:02 am
Just curious what is your Mallampati score. A large tongue is hard to treat with CPAP even with high pressures.

BTW: I have done both MMA and genioplasty. My personal experience is they don't have any (durable) effect on sleep quality. When totally relaxed and having no muscle tone, a large tongue will always fall down to block the airway (in supine position).
1 or 2 according to Dr. Zaghi. Would myofunctional tongue exercises help for you? Have you tried them?

I bumped the APAP up to 15 last night and I'm still getting things like this:

Screen Shot 2020-02-14 at 9.06.25 AM.png
Screen Shot 2020-02-14 at 9.06.25 AM.png (77.24 KiB) Viewed 2359 times

User avatar
Pugsy
Posts: 65115
Joined: Thu May 14, 2009 9:31 am
Location: Missouri, USA

Re: I'm a dentist treating myself! CPAP Results... Advice please?

Post by Pugsy » Fri Feb 14, 2020 11:19 am

nee wrote:
Fri Feb 14, 2020 11:06 am
I bumped the APAP up to 15 last night and I'm still getting things like this:
I think that OA is arousal related. The breathing pattern prior to the flag doesn't look like asleep breathing.
I am not surprised that more pressure doesn't fix things like this....more pressure can't do a thing for arousal/awake breathing irregularities.
Instead we have to try to figure out why the arousal breathing happened in the first place....which unfortunately isn't nearly as easy as just cranking up the pressure to try to fix something.
Cranking up the pressure only fixes things that can be fixed with more pressure. Not everything can be fixed with more pressure.

_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier
Additional Comments: Mask Bleep Eclipse https://bleepsleep.com/the-eclipse/
I may have to RISE but I refuse to SHINE.

musculus
Posts: 240
Joined: Tue Feb 26, 2013 8:35 am

Re: I'm a dentist treating myself! CPAP Results... Advice please?

Post by musculus » Fri Feb 14, 2020 11:59 am

nee wrote:
Fri Feb 14, 2020 11:06 am
musculus wrote:
Fri Feb 14, 2020 10:02 am
Just curious what is your Mallampati score. A large tongue is hard to treat with CPAP even with high pressures.

BTW: I have done both MMA and genioplasty. My personal experience is they don't have any (durable) effect on sleep quality. When totally relaxed and having no muscle tone, a large tongue will always fall down to block the airway (in supine position).
1 or 2 according to Dr. Zaghi. Would myofunctional tongue exercises help for you? Have you tried them?

I bumped the APAP up to 15 last night and I'm still getting things like this:


Screen Shot 2020-02-14 at 9.06.25 AM.png
This looks to me as you have worse airway obstruction at exhalation than inhalation. note from 6:28:00 to 6:28:30 the exhalation air flow start a little and then quickly diminish to zero for ~1 second

might be caused by the soft palate as it could be pushed by the outgoing air to shut off (soft palate collapse on the posterior pharyngeal wall).

BTW: Tongue exercise didn't work for me. Considering tongue reduction surgery.

_________________
Mask: Quattro™ Air Full Face Mask with Headgear
Additional Comments: sleepyhead

nee
Posts: 121
Joined: Thu Jan 09, 2020 10:45 am
Location: Southern California

Re: I'm a dentist treating myself! CPAP Results... Advice please?

Post by nee » Fri Feb 14, 2020 12:56 pm

musculus wrote:
Fri Feb 14, 2020 11:59 am
nee wrote:
Fri Feb 14, 2020 11:06 am
musculus wrote:
Fri Feb 14, 2020 10:02 am
Just curious what is your Mallampati score. A large tongue is hard to treat with CPAP even with high pressures.

BTW: I have done both MMA and genioplasty. My personal experience is they don't have any (durable) effect on sleep quality. When totally relaxed and having no muscle tone, a large tongue will always fall down to block the airway (in supine position).
1 or 2 according to Dr. Zaghi. Would myofunctional tongue exercises help for you? Have you tried them?

I bumped the APAP up to 15 last night and I'm still getting things like this:


Screen Shot 2020-02-14 at 9.06.25 AM.png
This looks to me as you have worse airway obstruction at exhalation than inhalation. note from 6:28:00 to 6:28:30 the exhalation air flow start a little and then quickly diminish to zero for ~1 second

might be caused by the soft palate as it could be pushed by the outgoing air to shut off (soft palate collapse on the posterior pharyngeal wall).

BTW: Tongue exercise didn't work for me. Considering tongue reduction surgery.

I groan and exert effort to breathe on exhalation, so this makes sense. Could I be shooting myself in the foot here by having my exhalation pressure set at 12 cmH2O? (15 cmH2O with EPR of 3)

musculus
Posts: 240
Joined: Tue Feb 26, 2013 8:35 am

Re: I'm a dentist treating myself! CPAP Results... Advice please?

Post by musculus » Fri Feb 14, 2020 2:25 pm

it could be due to the high EPAP, but do you see similar flow pattern at lower pressures, at usual REM times like 4-6 am?

try lying down and relax jaw and throat muscles, you should be able to induce the blockage by breathing in and out at different velocities. repeat the test by sticking the tongue out to confirm it is the soft palate.

_________________
Mask: Quattro™ Air Full Face Mask with Headgear
Additional Comments: sleepyhead

nee
Posts: 121
Joined: Thu Jan 09, 2020 10:45 am
Location: Southern California

Re: I'm a dentist treating myself! CPAP Results... Advice please?

Post by nee » Fri Feb 14, 2020 3:11 pm

musculus wrote:
Fri Feb 14, 2020 2:25 pm
it could be due to the high EPAP, but do you see similar flow pattern at lower pressures, at usual REM times like 4-6 am?

try lying down and relax jaw and throat muscles, you should be able to induce the blockage by breathing in and out at different velocities. repeat the test by sticking the tongue out to confirm it is the soft palate.
This was when I had EPAP at 7 cmH2O.

Screen Shot 2020-02-14 at 1.10.22 PM.png
Screen Shot 2020-02-14 at 1.10.22 PM.png (88.25 KiB) Viewed 2330 times

It's definitely my soft palate; I have retruded jaws. A MAD didn't do much for me. When I lie down, jut my jaw forward, and relax my throat/tongue, I still feel the blockage.

musculus
Posts: 240
Joined: Tue Feb 26, 2013 8:35 am

Re: I'm a dentist treating myself! CPAP Results... Advice please?

Post by musculus » Fri Feb 14, 2020 3:23 pm

so it happens at lower EPAP 8-9 cm as well. Using moderately low constant pressure might work better for you, with 0 EPR.

A good thing is that a flaggy soft palate is easily treatable with UPPP.

_________________
Mask: Quattro™ Air Full Face Mask with Headgear
Additional Comments: sleepyhead

nee
Posts: 121
Joined: Thu Jan 09, 2020 10:45 am
Location: Southern California

Re: I'm a dentist treating myself! CPAP Results... Advice please?

Post by nee » Wed Feb 19, 2020 11:02 am

I switched to the P10 for the past week because I was tired of the whole Bleep routine (shaving near the nose, witch hazel, extra tape around the Beep to neutralize leakages) and got massive leakages. This is with lansinoh AND cervical collar AND mouth tape AND mandibular advancement device AND the headband thing, the works... and this morning was the most difficult morning I've had in quite a while. Just could not get out of bed. I switched to the Bleep a few nights ago, but still, same leaks. Started at 15 cmH2O and thinking it was too high I went back down to 13 cmH2O last night. No luck.

I have a buckwheat hull pillow coming in today that I can't wait to use.

Screen Shot 2020-02-19 at 8.58.36 AM.png
Screen Shot 2020-02-19 at 8.58.36 AM.png (186.58 KiB) Viewed 2279 times
Screen Shot 2020-02-19 at 8.58.45 AM.png
Screen Shot 2020-02-19 at 8.58.45 AM.png (188.15 KiB) Viewed 2279 times

I'm definitely puzzled by all of this; the leaks are definitely due to mouth breathing but everything I throw at it doesn't seem to stop it :(

Oh and something interesting I noted; it seems like the flow limitations directly precede the mouth breathing.

nee
Posts: 121
Joined: Thu Jan 09, 2020 10:45 am
Location: Southern California

Re: I'm a dentist treating myself! CPAP Results... Advice please?

Post by nee » Wed Feb 19, 2020 11:13 am

FlowLeak.jpg
FlowLeak.jpg (299.09 KiB) Viewed 2275 times

I know correlation does not mean causation, but this just seems like every time I switch back to nasal breathing, some type of blockage is leading back to mouth breathing.

User avatar
Pugsy
Posts: 65115
Joined: Thu May 14, 2009 9:31 am
Location: Missouri, USA

Re: I'm a dentist treating myself! CPAP Results... Advice please?

Post by Pugsy » Wed Feb 19, 2020 11:19 am

I think the absence of FLs on the graph during periods of large leak are just the machine not being able to sense them due to the leaks.
So I think your FLs are likely fairly consistent but just look intermittent.
If those FLs originate in the nose then more pressure is unlikely to fix them.
Have you played around with humidity settings to see if there is much of a change with variations in humidity?

_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier
Additional Comments: Mask Bleep Eclipse https://bleepsleep.com/the-eclipse/
I may have to RISE but I refuse to SHINE.

User avatar
ChicagoGranny
Posts: 15273
Joined: Sun Jan 29, 2012 1:43 pm
Location: USA

Re: I'm a dentist treating myself! CPAP Results... Advice please?

Post by ChicagoGranny » Wed Feb 19, 2020 1:39 pm

Sorry, I haven't followed your thread, so feel free to ignore.
nee wrote:
Wed Feb 19, 2020 11:02 am
everything I throw at it
Have you thrown a FFM at it?
nee wrote:
Wed Feb 19, 2020 11:13 am
some type of blockage is leading back to mouth breathing
Have you had a total airway evaluation by an ENT?

User avatar
jimbud
Posts: 1465
Joined: Sat Dec 01, 2018 8:19 pm

Re: I'm a dentist treating myself! CPAP Results... Advice please?

Post by jimbud » Wed Feb 19, 2020 2:05 pm

Ditto what CGranny said.

I finally had to give up on the P10 and switched to the F30i. I have a beard and now have no leaks.
I also use a Buckwheat pillow (I have two of them).

I fought mouth breathing every way I could think of. Tape, tape with gag, tape with gag and cervical collar.
I was successful, but not comfortable enough for quality sleep.
Thus the F30i. It is working well for me (with cervical collar).

As with all CPAP solutions it works for me. For you...?

I hope you find something.
It can be very frustrating, but you will beat it in the end. :)

Good luck.

JPB

_________________
Machine: ResMed AirSense™ 10 AutoSet™ CPAP Machine with HumidAir™ Heated Humidifier
Mask: AirFit™ P10 Nasal Pillow CPAP Mask with Headgear
Additional Comments: Sleep on a Buckwheat Hull Pillow.

slowriter
Posts: 383
Joined: Sat Jul 20, 2019 4:37 am

Re: I'm a dentist treating myself! CPAP Results... Advice please?

Post by slowriter » Wed Feb 19, 2020 2:44 pm

nee wrote:
Wed Feb 19, 2020 11:02 am
I switched to the P10 for the past week because I was tired of the whole Bleep routine (shaving near the nose, witch hazel, extra tape around the Beep to neutralize leakages) and got massive leakages.
Hmm ... there's only two cases where I have gotten leaks with the Bleep:

First, if I haven't attached the Dreamports quite right, in which case there's a slight leak around the adhesive.

Second, if I haven't attached the mask part to the ports correctly. This sort of leak can be bigger, and can be avoided by testing before attaching to CPAP hose.

_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier
Mask: Bleep DreamPort CPAP Mask Solution
Additional Comments: UARS; VAuto Mode, 7-15, PS 5.8

User avatar
squid13
Posts: 2301
Joined: Fri Jul 15, 2011 3:47 pm
Location: Pensacola, FL

Re: I'm a dentist treating myself! CPAP Results... Advice please?

Post by squid13 » Wed Feb 19, 2020 6:16 pm

I agree with Granny and jimbud, it's about time you try a full face mask. I switched a couple of months ago to the ResMed F20 AirTouch and have zero leaks.

_________________
Machine: ResMed AirCurve 10 ASV Machine with Heated Humidifier
Mask: Evora Full Face Mask - Fitpack
Additional Comments: AirCurve 10 ASV, Oscar V1.0.1-r-1
US Navy Retired 1973,AirCurve 10 ASV, Mode: ASV Auto, Min EPAP: 7.2, Max EPAP: 15.0, Min PS:4.0, Max PS: 15.0, Mask ResMed Airtouch F20, Backup: (2) AirCurve 10 ASV

Dreaming1
Posts: 166
Joined: Fri Aug 31, 2018 9:55 am

Re: I'm a dentist treating myself! CPAP Results... Advice please?

Post by Dreaming1 » Wed Feb 19, 2020 7:26 pm

Do you care so much about the facial hair to avoid a full face mask? If not, shave it all off and get the mouth under control. OTHERWISE, try a full full face for a couple days and see what happens. You're a dentist, you can afford it.