What's wrong with me?

General Discussion on any topic relating to CPAP and/or Sleep Apnea.
User avatar
palerider
Posts: 32308
Joined: Wed Dec 16, 2009 5:43 pm
Location: Dallas(ish).

Re: What's wrong with me?

Post by palerider » Tue Dec 17, 2019 10:42 am

It says to set your min pressure to 8, and read the post at the top of the forum that says READ THIS FIRST to learn how to format your charts.

Also, it's a Resmed AirSense 10 Autoset. Use the right words when searching and you get better results.

_________________
Mask: Bleep DreamPort CPAP Mask Solution
Additional Comments: S9 VPAP Auto
Get OSCAR

Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.

andrenym00
Posts: 20
Joined: Sun Dec 01, 2019 7:20 pm

Re: What's wrong with me?

Post by andrenym00 » Tue Dec 17, 2019 10:43 am

Pugsy wrote: ↑
Tue Dec 17, 2019 10:38 am
It tells me you have terrible sleep quality and wake up often....why? Any known reason?

Do you take any medications of any kind? If so, what?
I take Synthroid and citalopram in the morning and a small amount of Xanax at night before sleep. I am under a lot of stress lately. I have always woken up for many years to go to the bathroom 4-8 times per night.

User avatar
palerider
Posts: 32308
Joined: Wed Dec 16, 2009 5:43 pm
Location: Dallas(ish).

Re: What's wrong with me?

Post by palerider » Tue Dec 17, 2019 10:49 am

andrenym00 wrote: ↑
Tue Dec 17, 2019 10:43 am
I have always woken up for many years to go to the bathroom 4-8 times per night.
That's a classic side effect of untreated obstructive sleep apnea.

_________________
Mask: Bleep DreamPort CPAP Mask Solution
Additional Comments: S9 VPAP Auto
Get OSCAR

Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.

andrenym00
Posts: 20
Joined: Sun Dec 01, 2019 7:20 pm

Re: What's wrong with me?

Post by andrenym00 » Tue Dec 17, 2019 10:50 am

palerider wrote: ↑
Tue Dec 17, 2019 10:49 am
andrenym00 wrote: ↑
Tue Dec 17, 2019 10:43 am
I have always woken up for many years to go to the bathroom 4-8 times per night.
That's a classic side effect of untreated obstructive sleep apnea.
So you think I should set minimum pressure to 8?

User avatar
palerider
Posts: 32308
Joined: Wed Dec 16, 2009 5:43 pm
Location: Dallas(ish).

Re: What's wrong with me?

Post by palerider » Tue Dec 17, 2019 10:51 am

andrenym00 wrote: ↑
Tue Dec 17, 2019 10:50 am
palerider wrote: ↑
Tue Dec 17, 2019 10:49 am
andrenym00 wrote: ↑
Tue Dec 17, 2019 10:43 am
I have always woken up for many years to go to the bathroom 4-8 times per night.
That's a classic side effect of untreated obstructive sleep apnea.
So you think I should set minimum pressure to 8?
I'm betting on 10, but let's work up to it.

_________________
Mask: Bleep DreamPort CPAP Mask Solution
Additional Comments: S9 VPAP Auto
Get OSCAR

Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.

andrenym00
Posts: 20
Joined: Sun Dec 01, 2019 7:20 pm

Re: What's wrong with me?

Post by andrenym00 » Tue Dec 17, 2019 10:54 am

palerider wrote: ↑
Tue Dec 17, 2019 10:51 am
andrenym00 wrote: ↑
Tue Dec 17, 2019 10:50 am
palerider wrote: ↑
Tue Dec 17, 2019 10:49 am
andrenym00 wrote: ↑
Tue Dec 17, 2019 10:43 am
I have always woken up for many years to go to the bathroom 4-8 times per night.
That's a classic side effect of untreated obstructive sleep apnea.
So you think I should set minimum pressure to 8?
I'm betting on 10, but let's work up to it.
So even though my ahi was under 5 at 2.49, why do you think I'm still waking up multiple times at night? Are other settings correct for now? No ramp?

User avatar
Pugsy
Posts: 65411
Joined: Thu May 14, 2009 9:31 am
Location: Missouri, USA

Re: What's wrong with me?

Post by Pugsy » Tue Dec 17, 2019 10:58 am

Read up on the side effects for your citalopram. It's well known to mess with sleep architecture (those sleep cycles) and even cause insomnia.
It's a SSRI type of medication...google "SSRI and sleep" and start reading. SSRI and SNRI....bad, bad stuff for sleep quality.

So part of your sleep issues could be medication side effects contributing to your problem.
Could also be from the stress that the meds are supposed to treat.

Are you male and of an age that the prostate is maybe a factor in the nocturia? Do you have to pee a lot during the day as well.
While sleep apnea is a common cause of nocturia....so is BPH in the male.

Follow palerider's advice on the pressure settings...might help to smooth out that pressure line a bit.

_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier
Additional Comments: Mask Bleep Eclipse https://bleepsleep.com/the-eclipse/
I may have to RISE but I refuse to SHINE.

User avatar
palerider
Posts: 32308
Joined: Wed Dec 16, 2009 5:43 pm
Location: Dallas(ish).

Re: What's wrong with me?

Post by palerider » Tue Dec 17, 2019 11:11 am

andrenym00 wrote: ↑
Tue Dec 17, 2019 10:54 am
So even though my ahi was under 5 at 2.49, why do you think I'm still waking up multiple times at night? Are other settings correct for now? No ramp?
1) 5 is a crap target, an ahi of 5 means your sleep is being disrupted at least every 12 minutes, on average, all night long. It's good enough for doctors, but not for us. Plus the ahi doesn't count other breathing related sleep disturbances. So no, get the ahi under 2, preferably under 1.5.

2) all those times the pressure line jumps up? Breathing problem, and those disturb your sleep, and raise your BP which makes your bladder fill up, and you have to go pee.

_________________
Mask: Bleep DreamPort CPAP Mask Solution
Additional Comments: S9 VPAP Auto
Last edited by palerider on Fri Jan 17, 2020 10:43 am, edited 1 time in total.
Get OSCAR

Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.

andrenym00
Posts: 20
Joined: Sun Dec 01, 2019 7:20 pm

Re: What's wrong with me?

Post by andrenym00 » Tue Dec 17, 2019 11:15 am

palerider wrote: ↑
Tue Dec 17, 2019 11:11 am
andrenym00 wrote: ↑
Tue Dec 17, 2019 10:54 am
So even though my ahi was under 5 at 2.49, why do you think I'm still waking up multiple times at night? Are other settings correct for now? No ramp?
1) 5 is a crap target, an ahi of 5 means your sleep is being disrupted at least every 12.5 minutes, on average, all night long. It's good enough for doctors, but not for us. Plus the ahi doesn't count other breathing related sleep disturbances. So no, get the ahi under 2, preferably under 1.5.
When you say I have breathing problems and I'm not breathing well, that's caused by the apnea correct?

User avatar
palerider
Posts: 32308
Joined: Wed Dec 16, 2009 5:43 pm
Location: Dallas(ish).

Re: What's wrong with me?

Post by palerider » Tue Dec 17, 2019 11:20 am

andrenym00 wrote: ↑
Tue Dec 17, 2019 11:15 am
When you say I have breathing problems and I'm not breathing well, that's caused by the apnea correct?
Apnea means "not breathing", so, apnea is a type of breathing problem falling under the umbrella of "Sleep Disordered Breathing".

_________________
Mask: Bleep DreamPort CPAP Mask Solution
Additional Comments: S9 VPAP Auto
Get OSCAR

Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.

andrenym00
Posts: 20
Joined: Sun Dec 01, 2019 7:20 pm

Re: What's wrong with me?

Post by andrenym00 » Sat Dec 21, 2019 9:06 am

So I increased minimum pressure to 8. Should I keep it at 8 or increase? Here are last nights results:
Attachments
screenshot-20191221-095418.png
screenshot-20191221-095418.png (76.96 KiB) Viewed 1794 times
screenshot-20191221-095506.png
screenshot-20191221-095506.png (60.15 KiB) Viewed 1794 times
screenshot-20191221-095537.png
screenshot-20191221-095537.png (57.56 KiB) Viewed 1794 times

User avatar
palerider
Posts: 32308
Joined: Wed Dec 16, 2009 5:43 pm
Location: Dallas(ish).

Re: What's wrong with me?

Post by palerider » Sat Dec 21, 2019 10:44 am

Still betting on 10.

Leave it for a couple days, or go to 9 or 10,your choice.

_________________
Mask: Bleep DreamPort CPAP Mask Solution
Additional Comments: S9 VPAP Auto
Get OSCAR

Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.

User avatar
zonker
Posts: 11491
Joined: Fri Jun 19, 2015 4:36 pm

Re: What's wrong with me?

Post by zonker » Sat Dec 21, 2019 12:36 pm

andrenym00 wrote: ↑
Sat Dec 21, 2019 9:06 am
So I increased minimum pressure to 8. Should I keep it at 8 or increase? Here are last nights results:
most important-how did you sleep? how are you feeling when you wake up?
"Age is not an accomplishment and youth is not a sin"-Robert A. Heinlein
Oscar-Win
https://www.apneaboard.com/OSCAR/OSCAR-1.5.1-Win64.exe
Oscar-Mac
https://www.apneaboard.com/OSCAR/OSCAR-1.5.1.dmg

andrenym00
Posts: 20
Joined: Sun Dec 01, 2019 7:20 pm

Re: What's wrong with me?

Post by andrenym00 » Sat Dec 21, 2019 2:25 pm

zonker wrote: ↑
Sat Dec 21, 2019 12:36 pm
andrenym00 wrote: ↑
Sat Dec 21, 2019 9:06 am
So I increased minimum pressure to 8. Should I keep it at 8 or increase? Here are last nights results:
most important-how did you sleep? how are you feeling when you wake up?
Not great and not bad. I haven't been waking up suffocating anymore. Which is good.

User avatar
zonker
Posts: 11491
Joined: Fri Jun 19, 2015 4:36 pm

Re: What's wrong with me?

Post by zonker » Sat Dec 21, 2019 7:22 pm

andrenym00 wrote: ↑
Sat Dec 21, 2019 2:25 pm
zonker wrote: ↑
Sat Dec 21, 2019 12:36 pm
andrenym00 wrote: ↑
Sat Dec 21, 2019 9:06 am
So I increased minimum pressure to 8. Should I keep it at 8 or increase? Here are last nights results:
most important-how did you sleep? how are you feeling when you wake up?
Not great and not bad. I haven't been waking up suffocating anymore. Which is good.
were it me, i'd leave it at 8. give your body time to adjust.

good luck!
"Age is not an accomplishment and youth is not a sin"-Robert A. Heinlein
Oscar-Win
https://www.apneaboard.com/OSCAR/OSCAR-1.5.1-Win64.exe
Oscar-Mac
https://www.apneaboard.com/OSCAR/OSCAR-1.5.1.dmg