20M trying to figure out if I inherited CSA from my dad.

General Discussion on any topic relating to CPAP and/or Sleep Apnea.
WeightsNoDates
Posts: 4
Joined: Sat Aug 10, 2019 8:33 pm

20M trying to figure out if I inherited CSA from my dad.

Post by WeightsNoDates » Sat Aug 10, 2019 9:22 pm

Hey all, I've suspected I've had some form of sleep disorder for a while now, I'm a 20 year old male in good shape, I go to the gym for at least an hour 6 days a week and run 5 of those days. My BMI is 23.6 and bodyfat % is below 20%, so I probably don't fit the
normal profile for sleep apnea. Sleep apnea runs in my family however and my dad specifically has central sleep apnea. He got a new CPAP so he gave me his Resmed Airsense 10 Autoset, which I've been using for about a year. I've decided to take this seriously
and started to record sleep data with an SD card the last few months. I've also seen my doctor about this and am now waiting on the call to schedule a full sleep test. I had an at-home sleep test about a year ago, where I was said to have symptoms of hypersomnia,
mild-moderate snoring and my blood oxygen saturation got down to 86%, which he seemed to not think was significant but I continued to push the issue recently. I wanted to share some examples of events I have seen with my CPAP lately.

The following events were taken on a night where I set the maximum and minimum pressure as low as I could, so both were set to 4.0, I wanted to do this to get as close to a baseline reading as possible. As you can see there are two examples of events that appear to last a little over 3 minutes in which I take very few short breaths.
3MinuteEvent2.PNG
3MinuteEvent2.PNG (96.89 KiB) Viewed 940 times
3MinuteEvent.PNG
3MinuteEvent.PNG (95.72 KiB) Viewed 940 times
I also appear to go through periods where my breathing cycles from flatlining, to tiny breaths, to normal breathing, and anywhere in-between.
IrregularBreathing.PNG
IrregularBreathing.PNG (99.88 KiB) Viewed 940 times
I also see plenty of regular ole apneas where I don't breathe for 8-20 seconds. Does this look like central apnea? I don't exhibit symptoms like waking up with extreme headaches or weird drowning dreams, but I do experience daily fatigue, and can fall asleep basically anytime I want. I also sleep like a rock and never, ever have trouble falling asleep, in fact I usually fall asleep within 10 or so minutes I'd wager. Thanks for any help

User avatar
Jas_williams
Posts: 1120
Joined: Tue Dec 19, 2017 2:12 pm
Location: Somerset UK

Re: 20M trying to figure out if I inherited CSA from my dad.

Post by Jas_williams » Sun Aug 11, 2019 12:09 am

WeightsNoDates wrote:
Sat Aug 10, 2019 9:22 pm
Hey all, I've suspected I've had some form of sleep disorder for a while now, I'm a 20 year old male in good shape, I go to the gym for at least an hour 6 days a week and run 5 of those days. My BMI is 23.6 and bodyfat % is below 20%, so I probably don't fit the
normal profile for sleep apnea. Sleep apnea runs in my family however and my dad specifically has central sleep apnea. He got a new CPAP so he gave me his Resmed Airsense 10 Autoset, which I've been using for about a year. I've decided to take this seriously
and started to record sleep data with an SD card the last few months. I've also seen my doctor about this and am now waiting on the call to schedule a full sleep test. I had an at-home sleep test about a year ago, where I was said to have symptoms of hypersomnia,
mild-moderate snoring and my blood oxygen saturation got down to 86%, which he seemed to not think was significant but I continued to push the issue recently. I wanted to share some examples of events I have seen with my CPAP lately.

The following events were taken on a night where I set the maximum and minimum pressure as low as I could, so both were set to 4.0, I wanted to do this to get as close to a baseline reading as possible. As you can see there are two examples of events that appear to last a little over 3 minutes in which I take very few short breaths.

3MinuteEvent2.PNG
3MinuteEvent.PNG

I also appear to go through periods where my breathing cycles from flatlining, to tiny breaths, to normal breathing, and anywhere in-between.

IrregularBreathing.PNG

I also see plenty of regular ole apneas where I don't breathe for 8-20 seconds. Does this look like central apnea? I don't exhibit symptoms like waking up with extreme headaches or weird drowning dreams, but I do experience daily fatigue, and can fall asleep basically anytime I want. I also sleep like a rock and never, ever have trouble falling asleep, in fact I usually fall asleep within 10 or so minutes I'd wager. Thanks for any help
To me it does not look like you have sleep apnoea yet as your machine did not raise pressure

_________________
Mask: Bleep DreamPort CPAP Mask Solution
Additional Comments: Using sleepyhead and a pressure of 6 - 21 Resmed S9 Adapt SV with a Bleep Sleep Mask

User avatar
palerider
Posts: 32299
Joined: Wed Dec 16, 2009 5:43 pm
Location: Dallas(ish).

Re: 20M trying to figure out if I inherited CSA from my dad.

Post by palerider » Sun Aug 11, 2019 12:38 am

Jas_williams wrote:
Sun Aug 11, 2019 12:09 am
WeightsNoDates wrote:
Sat Aug 10, 2019 9:22 pm
The following events were taken on a night where I set the maximum and minimum pressure as low as I could, so both were set to 4.0,
To me it does not look like you have sleep apnoea yet as your machine did not raise pressure
Not a valid test of that.

_________________
Mask: Bleep DreamPort CPAP Mask Solution
Additional Comments: S9 VPAP Auto
Get OSCAR

Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.

User avatar
LSAT
Posts: 13353
Joined: Sun Nov 16, 2008 10:11 am
Location: SE Wisconsin

Re: 20M trying to figure out if I inherited CSA from my dad.

Post by LSAT » Sun Aug 11, 2019 5:45 am

Personally, I see nothing abnormal.

idlewire
Posts: 35
Joined: Sun May 26, 2019 6:07 pm

Re: 20M trying to figure out if I inherited CSA from my dad.

Post by idlewire » Sun Aug 11, 2019 6:36 am

Jas_williams wrote:
Sun Aug 11, 2019 12:09 am
To me it does not look like you have sleep apnoea yet as your machine did not raise pressure
Also you missed the part where he set the min and max pressure both to 4.

User avatar
Miss Emerita
Posts: 3758
Joined: Sun Nov 04, 2018 8:07 pm

Re: 20M trying to figure out if I inherited CSA from my dad.

Post by Miss Emerita » Sun Aug 11, 2019 8:54 am

What was your AHI when you had the home sleep test?
Oscar software is available at https://www.sleepfiles.com/OSCAR/

User avatar
Goofproof
Posts: 16087
Joined: Mon Dec 05, 2005 3:16 pm
Location: Central Indiana, USA

Re: 20M trying to figure out if I inherited CSA from my dad.

Post by Goofproof » Sun Aug 11, 2019 9:09 am

Be it MOM or Dad or Mail Man, if you have Sleep Apnea you need to treat it, anything else is a waste of time and energy and effort!

Life is easier if we spend less time worrying about problems than using solutions. :wink: Jim
Use data to optimize your xPAP treatment!

"The art of medicine consists in amusing the patient while nature cures the disease." Voltaire

D.H.
Posts: 3532
Joined: Sun Sep 04, 2016 7:07 pm

Re: 20M trying to figure out if I inherited CSA from my dad.

Post by D.H. » Sun Aug 11, 2019 10:11 am

First of all, you should include the stuff on the left when posting.

Form what I see, I don't think that you have either variety of SA.

Lastly, stop trying to place blame!

_________________
MachineMask
Additional Comments: Auto PAP; 13.5 cmH2O min - 20 cmH2O max

WeightsNoDates
Posts: 4
Joined: Sat Aug 10, 2019 8:33 pm

Re: 20M trying to figure out if I inherited CSA from my dad.

Post by WeightsNoDates » Sun Aug 11, 2019 11:11 am

palerider wrote:
Sun Aug 11, 2019 12:38 am
Jas_williams wrote:
Sun Aug 11, 2019 12:09 am
WeightsNoDates wrote:
Sat Aug 10, 2019 9:22 pm
The following events were taken on a night where I set the maximum and minimum pressure as low as I could, so both were set to 4.0,
To me it does not look like you have sleep apnoea yet as your machine did not raise pressure
Not a valid test of that.
What's the best way to collect data with these machines? I slept with my min/max setting at 4/4 for one night, and spent a while skimming the entire night for events like this. A majority of my time sleeping with it my pressure is at 6/14 and recently 7/15.
I see events like the following all the time. And those look like standard apneas to me. I incorrectly guessed that those 3-minute long events were more severe and possibly more indicative of sleep apnea, which is why I put those in the OP.
Event2.PNG
Event2.PNG (99.49 KiB) Viewed 860 times

WeightsNoDates
Posts: 4
Joined: Sat Aug 10, 2019 8:33 pm

Re: 20M trying to figure out if I inherited CSA from my dad.

Post by WeightsNoDates » Sun Aug 11, 2019 11:20 am

Miss Emerita wrote:
Sun Aug 11, 2019 8:54 am
What was your AHI when you had the home sleep test?
He made no mention of AHI and at the time I did not know what it was to ask in the first place. From what I gathered it seems that the at-home test neither confirmed or denied that I have a sleep disorder. He did bring up my oxygen saturation getting down to 86% which he said wasn’t a concern, but from what I’ve read on this site you basically never want to get below 88-90%. Reading that is what originally lead me to take this more seriously and gather my own data.

User avatar
Pugsy
Posts: 65130
Joined: Thu May 14, 2009 9:31 am
Location: Missouri, USA

Re: 20M trying to figure out if I inherited CSA from my dad.

Post by Pugsy » Sun Aug 11, 2019 11:21 am

Have you seen the videos here?
http://freecpapadvice.com/sleepyhead-free-software

I am thinking that some flagged events you are seeing (and you sure aren't having many anyway) are arousal/awake related and thus not related to the airway or sleep apnea.
If you aren't asleep...it isn't sleep apnea. Have to be asleep for it to matter other than maybe being a symptom of poor sleep quality instead of the cause of poor sleep quality.

_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier
Additional Comments: Mask Bleep Eclipse https://bleepsleep.com/the-eclipse/
I may have to RISE but I refuse to SHINE.

User avatar
Jas_williams
Posts: 1120
Joined: Tue Dec 19, 2017 2:12 pm
Location: Somerset UK

Re: 20M trying to figure out if I inherited CSA from my dad.

Post by Jas_williams » Sun Aug 11, 2019 11:49 am

palerider wrote:
Sun Aug 11, 2019 12:38 am
Jas_williams wrote:
Sun Aug 11, 2019 12:09 am
WeightsNoDates wrote:
Sat Aug 10, 2019 9:22 pm
The following events were taken on a night where I set the maximum and minimum pressure as low as I could, so both were set to 4.0,
To me it does not look like you have sleep apnoea yet as your machine did not raise pressure
Not a valid test of that.
Oops good point if set to basically a fixed pressure then pressure won’t raise



There is insufficient information trying to use an APAP machine to test for obstructive Apnoea nor CSA. Based on the number of events I see no reason to continue using the machine. But I stand by my statement there was too few events so it does not look like you have CSA. But a CPAP machine is not a diagnostic tool for testing for CSA or OSA. An overnight Oximetry test without any machine may be a better test

_________________
Mask: Bleep DreamPort CPAP Mask Solution
Additional Comments: Using sleepyhead and a pressure of 6 - 21 Resmed S9 Adapt SV with a Bleep Sleep Mask

User avatar
palerider
Posts: 32299
Joined: Wed Dec 16, 2009 5:43 pm
Location: Dallas(ish).

Re: 20M trying to figure out if I inherited CSA from my dad.

Post by palerider » Sun Aug 11, 2019 12:15 pm

WeightsNoDates wrote:
Sun Aug 11, 2019 11:11 am

What's the best way to collect data with these machines?
Set it for 4-20, and post the images properly
wiki/index.php/Sleepyhead:Organize

_________________
Mask: Bleep DreamPort CPAP Mask Solution
Additional Comments: S9 VPAP Auto
Get OSCAR

Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.

User avatar
Miss Emerita
Posts: 3758
Joined: Sun Nov 04, 2018 8:07 pm

Re: 20M trying to figure out if I inherited CSA from my dad.

Post by Miss Emerita » Sun Aug 11, 2019 2:22 pm

WeightsNoDates wrote:
Sun Aug 11, 2019 11:20 am
Miss Emerita wrote:
Sun Aug 11, 2019 8:54 am
What was your AHI when you had the home sleep test?
He made no mention of AHI and at the time I did not know what it was to ask in the first place. From what I gathered it seems that the at-home test neither confirmed or denied that I have a sleep disorder. He did bring up my oxygen saturation getting down to 86% which he said wasn’t a concern, but from what I’ve read on this site you basically never want to get below 88-90%. Reading that is what originally lead me to take this more seriously and gather my own data.
You have a right to a copy of your sleep study report. I’d advise you to get one.
Oscar software is available at https://www.sleepfiles.com/OSCAR/

WeightsNoDates
Posts: 4
Joined: Sat Aug 10, 2019 8:33 pm

Re: 20M trying to figure out if I inherited CSA from my dad.

Post by WeightsNoDates » Sun Aug 11, 2019 6:43 pm

palerider wrote:
Sun Aug 11, 2019 12:15 pm
WeightsNoDates wrote:
Sun Aug 11, 2019 11:11 am

What's the best way to collect data with these machines?
Set it for 4-20, and post the images properly
wiki/index.php/Sleepyhead:Organize
Here's data from last night when it was set to 7/15. I have it set to 4/20 for tonight.
screenshot-20190811-183816.png
screenshot-20190811-183816.png (149.67 KiB) Viewed 781 times
screenshot-20190811-183734.png
screenshot-20190811-183734.png (151.66 KiB) Viewed 781 times