Airsense 10 vs Dream Station Leak Rates

General Discussion on any topic relating to CPAP and/or Sleep Apnea.
DJOS
Posts: 5
Joined: Sat Jun 11, 2016 10:07 am

Airsense 10 vs Dream Station Leak Rates

Post by DJOS » Fri Jul 12, 2019 1:06 pm

I've used Resmed APAP machines for many years with no leak rate problems but this last year has been horrible for leak rates. My leak rate was consistently at 40-42 L/min for about 6 months but now it is at 53 L/min (for the last 3 months now). I've tried at least 5 different masks and none of them improved my leak rate. I'm considering changing to a different APAP machine now to see if i can improve my leak rate and I'm wondering if any of you have changed from Resmed Airsense 10 for Her to another APAP and found that your leak rate improved. I would really appreciate your feedback. Thanks in advance.

User avatar
palerider
Posts: 32299
Joined: Wed Dec 16, 2009 5:43 pm
Location: Dallas(ish).

Re: Airsense 10 vs Dream Station Leak Rates

Post by palerider » Fri Jul 12, 2019 1:24 pm

DJOS wrote:
Fri Jul 12, 2019 1:06 pm
I've used Resmed APAP machines for many years with no leak rate problems but this last year has been horrible for leak rates. My leak rate was consistently at 40-42 L/min for about 6 months but now it is at 53 L/min (for the last 3 months now). I've tried at least 5 different masks and none of them improved my leak rate. I'm considering changing to a different APAP machine now to see if i can improve my leak rate and I'm wondering if any of you have changed from Resmed Airsense 10 for Her to another APAP and found that your leak rate improved. I would really appreciate your feedback. Thanks in advance.
That leak rate is ludicrously high.

Changing *machines* won't make any difference in leak amount, since that is all in the hose and mask.

_________________
Mask: Bleep DreamPort CPAP Mask Solution
Additional Comments: S9 VPAP Auto
Get OSCAR

Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.

User avatar
LSAT
Posts: 13376
Joined: Sun Nov 16, 2008 10:11 am
Location: SE Wisconsin

Re: Airsense 10 vs Dream Station Leak Rates

Post by LSAT » Fri Jul 12, 2019 1:54 pm

I agree..machine will not make a difference. Are you mouth breathing? What masks have you tried?

User avatar
zonker
Posts: 11399
Joined: Fri Jun 19, 2015 4:36 pm

Re: Airsense 10 vs Dream Station Leak Rates

Post by zonker » Fri Jul 12, 2019 2:22 pm

DJOS wrote:
Fri Jul 12, 2019 1:06 pm
I've used Resmed APAP machines for many years with no leak rate problems but this last year has been horrible for leak rates. My leak rate was consistently at 40-42 L/min for about 6 months but now it is at 53 L/min (for the last 3 months now). I've tried at least 5 different masks and none of them improved my leak rate. I'm considering changing to a different APAP machine now to see if i can improve my leak rate and I'm wondering if any of you have changed from Resmed Airsense 10 for Her to another APAP and found that your leak rate improved. I would really appreciate your feedback. Thanks in advance.
nope. have changed from resperonics to resmed with no change in leak rates. (i know that's not what you asked.)

have you tried watching any mask fitting videos on youtube? i don't have direct experience with this, i'm just parroting what i've heard.

good luck!
"Age is not an accomplishment and youth is not a sin"-Robert A. Heinlein
Oscar-Win
https://www.apneaboard.com/OSCAR/OSCAR-1.5.1-Win64.exe
Oscar-Mac
https://www.apneaboard.com/OSCAR/OSCAR-1.5.1.dmg

BScpap20147
Posts: 1
Joined: Fri Jul 12, 2019 6:08 pm

Re: Airsense 10 vs Dream Station Leak Rates

Post by BScpap20147 » Fri Jul 12, 2019 6:48 pm

Hi,

I also have a new autosense auto set by resumed. Using a P-10 pillow mask. Same problem! High leak rates but good sleep times, low events, AHI. But with high leak rates higher then 50. I can do a mask fit check ok with a green smiley face at start and as long as I keep it on fit check. In morning no smiley face though. I hear no big leaks while sleeping 4 then 3 more hours after break. I think it may be low seal leaks in the humidifier throughout the night. I do feel rested in the am.

Will contact provider soon

DJOS
Posts: 5
Joined: Sat Jun 11, 2016 10:07 am

Re: Airsense 10 vs Dream Station Leak Rates

Post by DJOS » Fri Jul 12, 2019 6:53 pm

Thanks so much for the feedback. I used the Swift FX for Her for many years with no leak problems. In the past year I've tried Phillips DreamWear, AirfitP10 and a full face mask. I cannot wear a full face mask - it makes my AHI go sky high and didn't change my leak rate. The other masks made no difference in my leak rate. I've used mouth tape successfully until this year and have also recently tried a chin strap and other types of tape for mouth breathers. Nothing has worked, including the tape I've used for years. I change my hose monthly and have checked for leaks. There's no problem with my hoses. If changing machines won't work, the only other thing I can think to do is try some more masks. I have watched the videos for the masks I've tried and don't think I'm doing anything wrong. I do have narrow nostrils and I'm wondering if there's a mask that might be a better fit for a narrow nose profile. I'd love to solve this leak rate problem but I'm out of ideas at this point. I've used an APAP faithfully since 2006 so I will keep using it and keep working on trying to find a solution to my leak rate. I'd appreciate any of your ideas! Thank you!

User avatar
zonker
Posts: 11399
Joined: Fri Jun 19, 2015 4:36 pm

Re: Airsense 10 vs Dream Station Leak Rates

Post by zonker » Fri Jul 12, 2019 7:00 pm

DJOS wrote:
Fri Jul 12, 2019 6:53 pm
Thanks so much for the feedback. I used the Swift FX for Her for many years with no leak problems.
have you been using the same machine since 2006?
"Age is not an accomplishment and youth is not a sin"-Robert A. Heinlein
Oscar-Win
https://www.apneaboard.com/OSCAR/OSCAR-1.5.1-Win64.exe
Oscar-Mac
https://www.apneaboard.com/OSCAR/OSCAR-1.5.1.dmg

DJOS
Posts: 5
Joined: Sat Jun 11, 2016 10:07 am

Re: Airsense 10 vs Dream Station Leak Rates

Post by DJOS » Sat Jul 13, 2019 8:05 am

Thanks for the question about how long I've used my Resmed. I bought a new Resmed machine a year ago. It was defective and I received a replacement so my current machine is quite new.

User avatar
palerider
Posts: 32299
Joined: Wed Dec 16, 2009 5:43 pm
Location: Dallas(ish).

Re: Airsense 10 vs Dream Station Leak Rates

Post by palerider » Sat Jul 13, 2019 8:34 am

Let's see some Oscar screenshots, so we can see what the leak looks like.

_________________
Mask: Bleep DreamPort CPAP Mask Solution
Additional Comments: S9 VPAP Auto
Get OSCAR

Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.

jsmit86
Posts: 265
Joined: Wed Mar 20, 2019 2:43 pm
Location: Conroe, TX (Suburban Houston TX)

Re: Airsense 10 vs Dream Station Leak Rates

Post by jsmit86 » Sat Jul 13, 2019 8:14 pm

You may also want to double check your hose to verify no leaks.
Sometimes you can get a tear near the end that will seal unless you flex the hose. It can be intermittent. You can always try another hose and see if that helps.