Re: what keeps the airway open on expiration?

Post by raisedfist » Sat Jul 06, 2019 11:22 am

On a ventilator you can turn off EPAP if you are using an active circuit with an exhalation valve. EPAP is not an absolute requirement for positive pressure ventilation. A person with a neuromuscular disease could use a Trilogy 100 with an active circuit, non-vented mask, using an IPAP of 18 and an EPAP of 0 just fine, should they not have OSA. If no OSA, no collapse of airway, EPAP or not. On a passive circuit you must have EPAP because the expired air must go somewhere, thus the reason for using a vented mask.

On older bi-level machines the EPAP used to go all the way down to 2cm h2o. Now the lowest I've ever seen is 3.

_________________
Mask
Philips Respironics Trilogy 100
AVAPS-AE Mode
PS Min 6, PS Max 18, EPAP Min 4, EPAP Max 12

User avatar
palerider
Posts: 32299
Joined: Wed Dec 16, 2009 5:43 pm
Location: Dallas(ish).

Re: what keeps the airway open on expiration?

Post by palerider » Sat Jul 06, 2019 12:19 pm

raisedfist wrote:
Sat Jul 06, 2019 11:22 am
On a ventilator you can turn off EPAP if you are using an active circuit with an exhalation valve. EPAP is not an absolute requirement for positive pressure ventilation. A person with a neuromuscular disease could use a Trilogy 100 with an active circuit, non-vented mask, using an IPAP of 18 and an EPAP of 0 just fine, should they not have OSA. If no OSA, no collapse of airway, EPAP or not. On a passive circuit you must have EPAP because the expired air must go somewhere, thus the reason for using a vented mask.

On older bi-level machines the EPAP used to go all the way down to 2cm h2o. Now the lowest I've ever seen is 3.
Well yeah... ventilators, you can do a lot of things. :)

_________________
Mask: Bleep DreamPort CPAP Mask Solution
Additional Comments: S9 VPAP Auto
Get OSCAR

Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.

User avatar
jimbud
Posts: 1474
Joined: Sat Dec 01, 2018 8:19 pm

Re: what keeps the airway open on expiration?

Post by jimbud » Sat Jul 06, 2019 12:26 pm

This is what the OP had when the OP asked question:

Post by tyrinryan » Thu Jul 04, 2019 9:17 am

Machine: AirSense™ 10 AutoSet™ For Her CPAP Machine with HumidAir™ Heated Humidifier

Not a ventilator.
JPB

_________________
Machine: ResMed AirSense™ 10 AutoSet™ CPAP Machine with HumidAir™ Heated Humidifier
Mask: AirFit™ P10 Nasal Pillow CPAP Mask with Headgear
Additional Comments: Sleep on a Buckwheat Hull Pillow.

User avatar
raisedfist
Posts: 1176
Joined: Wed Jun 15, 2016 7:21 am

Re: what keeps the airway open on expiration?

Post by raisedfist » Sat Jul 06, 2019 12:33 pm

party pooper :lol:

_________________
Mask
Philips Respironics Trilogy 100
AVAPS-AE Mode
PS Min 6, PS Max 18, EPAP Min 4, EPAP Max 12

User avatar
jimbud
Posts: 1474
Joined: Sat Dec 01, 2018 8:19 pm

Re: what keeps the airway open on expiration?

Post by jimbud » Sat Jul 06, 2019 1:05 pm

raisedfist wrote:
Sat Jul 06, 2019 12:33 pm
party pooper :lol:
Apologies. :wink:
It had all turned to a general discussion anyway. :)
JPB

_________________
Machine: ResMed AirSense™ 10 AutoSet™ CPAP Machine with HumidAir™ Heated Humidifier
Mask: AirFit™ P10 Nasal Pillow CPAP Mask with Headgear
Additional Comments: Sleep on a Buckwheat Hull Pillow.

HoseCrusher
Posts: 2744
Joined: Tue Oct 12, 2010 6:42 pm

Re: what keeps the airway open on expiration?

Post by HoseCrusher » Sat Jul 06, 2019 5:45 pm

Just to present an alternative perspective...

This is extremely difficult to do so if you are frail don't attempt this.

We normally consider chest expansion and drawing air in as inhalation. Everything relaxes and air goes out during exhalation. Some breathing exercises focus on forcing more air out than what is involved in normal expiration.

If you cycle between forcing air out, then relaxing to draw air in you end up with somewhat of a reverse cycle. Effort to blow air out and relax to breathe in.

This is totally not natural and I don't think you can do it in your sleep, but it is fun to play with breathing. I can only do this for a very short period of time. I find it interesting to observe that either way of breathing, normal or reversed, while awake has no change on my O2 levels as long as I am resting and not engaged in any activity.

_________________
Mask: Brevida™ Nasal Pillow CPAP Mask with Headgear
Additional Comments: Machine is an AirSense 10 AutoSet For Her with Heated Humidifier.
SpO2 96+% and holding...

User avatar
jimbud
Posts: 1474
Joined: Sat Dec 01, 2018 8:19 pm

Re: what keeps the airway open on expiration?

Post by jimbud » Sat Jul 06, 2019 7:14 pm

HoseCrusher wrote:
Sat Jul 06, 2019 5:45 pm
Just to present an alternative perspective...

This is extremely difficult to do so if you are frail don't attempt this.

We normally consider chest expansion and drawing air in as inhalation. Everything relaxes and air goes out during exhalation. Some breathing exercises focus on forcing more air out than what is involved in normal expiration.

If you cycle between forcing air out, then relaxing to draw air in you end up with somewhat of a reverse cycle. Effort to blow air out and relax to breathe in.

This is totally not natural and I don't think you can do it in your sleep, but it is fun to play with breathing. I can only do this for a very short period of time. I find it interesting to observe that either way of breathing, normal or reversed, while awake has no change on my O2 levels as long as I am resting and not engaged in any activity.
If you don't mind I would like to use that in my rotation of party games.
Ring Around the Rosie, Spin the Bottle, Twister and then the coup de gras, Lets All Breath in Reverse(it will slay). :D
Man, Hosecrusher you are one "wild and crazy guy". :lol: :lol: :lol:
You can think up some stuff for sure. :)
JPB

_________________
Machine: ResMed AirSense™ 10 AutoSet™ CPAP Machine with HumidAir™ Heated Humidifier
Mask: AirFit™ P10 Nasal Pillow CPAP Mask with Headgear
Additional Comments: Sleep on a Buckwheat Hull Pillow.

User avatar
katestyles
Posts: 610
Joined: Sun Jan 06, 2019 9:08 am

Re: what keeps the airway open on expiration?

Post by katestyles » Sun Jul 07, 2019 9:14 am

Some thoughts.

1) I play brass instruments. I have spent years training my exhalation to be controlled in both speed and intensity. So that I can play my instrument. Breathing in needs to be swift, and breathing out needs to be significantly slower.
2) In Yoga there is this thing called "square breathing" when you breathe in to a count of, say 4, hold the breath for the same count, breathe out over the same count, and then hold your lungs empty for the same count. I like it. I find that breathing in is more "reflexive" when I am starved of oxygen, and it does not feel like I am working to draw in the air, rather, I am relaxing to allow the air in.
3) My Alexander technique teacher has also taught me to expel ALL the air from my lungs, then wait (which requires muscle action) until I must (reflexively) draw in air.

I think that we all know that muscles come in pairs, and simply relaxing a muscle does not cause a limb to move. It is the contraction of another muscle that causes the movement. If the diaphragm relaxes, gravity will pull the ribcage down, but the intercostal muscles are likely also active.

_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier
Mask: Bleep DreamPort CPAP Mask Solution
Additional Comments: Back up mask - anything in the drawer

User avatar
zonker
Posts: 11369
Joined: Fri Jun 19, 2015 4:36 pm

Re: what keeps the airway open on expiration?

Post by zonker » Sun Jul 07, 2019 11:40 am

katestyles wrote:
Sun Jul 07, 2019 9:14 am
Some thoughts.
hey, there! where you been?

to OP: sorry, i'm just happy to see kate back again.
"Age is not an accomplishment and youth is not a sin"-Robert A. Heinlein
Oscar-Win
https://www.apneaboard.com/OSCAR/OSCAR-1.5.1-Win64.exe
Oscar-Mac
https://www.apneaboard.com/OSCAR/OSCAR-1.5.1.dmg

User avatar
Jas_williams
Posts: 1120
Joined: Tue Dec 19, 2017 2:12 pm
Location: Somerset UK

Re: what keeps the airway open on expiration?

Post by Jas_williams » Sun Jul 07, 2019 12:49 pm

palerider wrote:
Sat Jul 06, 2019 12:19 pm
raisedfist wrote:
Sat Jul 06, 2019 11:22 am
On a ventilator you can turn off EPAP if you are using an active circuit with an exhalation valve. EPAP is not an absolute requirement for positive pressure ventilation. A person with a neuromuscular disease could use a Trilogy 100 with an active circuit, non-vented mask, using an IPAP of 18 and an EPAP of 0 just fine, should they not have OSA. If no OSA, no collapse of airway, EPAP or not. On a passive circuit you must have EPAP because the expired air must go somewhere, thus the reason for using a vented mask.

On older bi-level machines the EPAP used to go all the way down to 2cm h2o. Now the lowest I've ever seen is 3.
Well yeah... ventilators, you can do a lot of things. :)
Sure they can and are a specialist piece of medical equipment to treat many things and not just Obstructive Sleep Apnoea :D

_________________
Mask: Bleep DreamPort CPAP Mask Solution
Additional Comments: Using sleepyhead and a pressure of 6 - 21 Resmed S9 Adapt SV with a Bleep Sleep Mask

User avatar
Jay Aitchsee
Posts: 2936
Joined: Sun May 22, 2011 12:47 pm
Location: Southwest Florida

Re: what keeps the airway open on expiration?

Post by Jay Aitchsee » Sun Jul 07, 2019 1:05 pm

katestyles wrote:
Sun Jul 07, 2019 9:14 am
I think that we all know that muscles come in pairs, and simply relaxing a muscle does not cause a limb to move. It is the contraction of another muscle that causes the movement. If the diaphragm relaxes, gravity will pull the ribcage down, but the intercostal muscles are likely also active.
Don't forget we have atmospheric pressure also playing a part.
It might be helpful to remember that atmospheric pressure at sea level is 14.7 PSI = 29.92 inHg = 1033.5 cmH2O
And 10cmH20 equals approx 0.142PSI

_________________
Mask: AirFit™ P10 Nasal Pillow CPAP Mask with Headgear
Additional Comments: S9 Auto, P10 mask, P=7.0, EPR3, ResScan 5.3, SleepyHead V1.B2, Windows 10, ZEO, CMS50F, Infrared Video

User avatar
zonker
Posts: 11369
Joined: Fri Jun 19, 2015 4:36 pm

Re: what keeps the airway open on expiration?

Post by zonker » Sun Jul 07, 2019 3:01 pm

Jay Aitchsee wrote:
Sun Jul 07, 2019 1:05 pm
katestyles wrote:
Sun Jul 07, 2019 9:14 am
I think that we all know that muscles come in pairs, and simply relaxing a muscle does not cause a limb to move. It is the contraction of another muscle that causes the movement. If the diaphragm relaxes, gravity will pull the ribcage down, but the intercostal muscles are likely also active.
Don't forget we have atmospheric pressure also playing a part.
It might be helpful to remember that atmospheric pressure at sea level is 14.7 PSI = 29.92 inHg = 1033.5 cmH2O
And 10cmH20 equals approx 0.142PSI
i was told there'd be no math!
:lol: :lol: :lol:
"Age is not an accomplishment and youth is not a sin"-Robert A. Heinlein
Oscar-Win
https://www.apneaboard.com/OSCAR/OSCAR-1.5.1-Win64.exe
Oscar-Mac
https://www.apneaboard.com/OSCAR/OSCAR-1.5.1.dmg

User avatar
palerider
Posts: 32299
Joined: Wed Dec 16, 2009 5:43 pm
Location: Dallas(ish).

Re: what keeps the airway open on expiration?

Post by palerider » Sun Jul 07, 2019 3:06 pm

zonker wrote:
Sun Jul 07, 2019 3:01 pm
Jay Aitchsee wrote:
Sun Jul 07, 2019 1:05 pm
katestyles wrote:
Sun Jul 07, 2019 9:14 am
I think that we all know that muscles come in pairs, and simply relaxing a muscle does not cause a limb to move. It is the contraction of another muscle that causes the movement. If the diaphragm relaxes, gravity will pull the ribcage down, but the intercostal muscles are likely also active.
Don't forget we have atmospheric pressure also playing a part.
It might be helpful to remember that atmospheric pressure at sea level is 14.7 PSI = 29.92 inHg = 1033.5 cmH2O
And 10cmH20 equals approx 0.142PSI
i was told there'd be no math!
:lol: :lol: :lol:
Looks like you were lied to.

_________________
Mask: Bleep DreamPort CPAP Mask Solution
Additional Comments: S9 VPAP Auto
Get OSCAR

Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.

User avatar
ChicagoGranny
Posts: 15426
Joined: Sun Jan 29, 2012 1:43 pm
Location: USA

Stop it!

Post by ChicagoGranny » Sun Jul 07, 2019 4:11 pm

1.jpg
"It's not the number of breaths we take, it's the number of moments that take our breath away."

Cuando cuentes cuentos, cuenta cuántas cuentos cuentas.

User avatar
Goofproof
Posts: 16087
Joined: Mon Dec 05, 2005 3:16 pm
Location: Central Indiana, USA

Re: Stop it!

Post by Goofproof » Sun Jul 07, 2019 4:31 pm

ChicagoGranny wrote:
Sun Jul 07, 2019 4:11 pm
1.jpg
When it comes to Hershey, it always comes down to the Math. The Corps all depend on you not being able to do it. A may not use Algebra, but. I can figure out the lowest cost per Oz, for my Fix! Jim
Use data to optimize your xPAP treatment!

"The art of medicine consists in amusing the patient while nature cures the disease." Voltaire