AHI worse with CPAP than without?

General Discussion on any topic relating to CPAP and/or Sleep Apnea.
User avatar
BamaGramma
Posts: 61
Joined: Thu Jun 20, 2019 8:00 am

AHI worse with CPAP than without?

Post by BamaGramma » Fri Jun 28, 2019 12:29 pm

It was reported to me that I had 6.5 episodes of apnea during my sleep study, and therefore I was put on CPAP therapy. I put the DreamMapper app on my phone, and in the 5 nights since I installed the app, my AHI has ranged from 6.3 to 14.9 (in order they were 14.9, 9.5, 6.3, 7.9, 10). The first 3 nights were with a mask that was not fitting properly because the frame was too big; last two nights are with a new frame that helps with the fit. I use a chin strap to keep my mouth closed but still experience episodes where air sort of burps out of my mouth on its own. And I've been waking with a terrible dry mouth, even drier than before the CPAP and chin strap.

It looks to me like I'm worse with the CPAP with regard to AHI. I'm averaging between 4 and 6 hours of sleep before I wake up and can't go back to sleep (but I usually crash in the afternoon for a nap). Is this just the initial "honeymoon" phase with CPAP? Or should I be concerned? I don't see the sleep doctor for follow-up until July 31.

Thanks in advance.

User avatar
chunkyfrog
Posts: 34544
Joined: Mon Jul 12, 2010 5:10 pm
Location: Nowhere special--this year in particular.

Re: AHI worse with CPAP than without?

Post by chunkyfrog » Fri Jun 28, 2019 1:14 pm

Some events are likely being caused by your not being used to the mask, air, sound, etc.
Give it more time.
5 days is not enough for most people.

_________________
Mask: AirFit™ P10 For Her Nasal Pillow CPAP Mask with Headgear
Additional Comments: Airsense 10 Autoset for Her

NoOnesPerfect
Posts: 163
Joined: Thu May 09, 2019 7:29 am
Location: Washington, USA

Re: AHI worse with CPAP than without?

Post by NoOnesPerfect » Fri Jun 28, 2019 1:22 pm

If your sleep test was in-lab, any events that happened while you were awake would have been excluded from AHI, unlike what you are seeing from your machine. If you had the at-home test, it likely is related to just getting used to therapy. Anxiety (some might also say obsession 😁) doesn’t make sleep any easier.

You definitely want to look at your OSCAR charts and see if you find a lot of events occurring while you are not asleep - they account for almost all of mine, and you may find they are a significant portion of yours as well. I fond that turning my machine off and back on quickly (to start the ramp function again) when I wake at night reduced my AHI from 6-10 down to sub 1-3.

User avatar
palerider
Posts: 32299
Joined: Wed Dec 16, 2009 5:43 pm
Location: Dallas(ish).

Re: AHI worse with CPAP than without?

Post by palerider » Fri Jun 28, 2019 1:33 pm

BamaGramma wrote:
Fri Jun 28, 2019 12:29 pm
It looks to me like I'm worse with the CPAP with regard to AHI. I'm averaging between 4 and 6 hours of sleep before I wake up and can't go back to sleep (but I usually crash in the afternoon for a nap). Is this just the initial "honeymoon" phase with CPAP? Or should I be concerned? I don't see the sleep doctor for follow-up until July 31.
Several things, 1, your sleep study may have been inaccurate, (it happens all the time} 2. your CPAP isn't adjusted for your needs, and, 3. as was mentioned, you're just getting adjusted to CPAP.

Post some Oscar charts so we can see what's happening.

_________________
Mask: Bleep DreamPort CPAP Mask Solution
Additional Comments: S9 VPAP Auto
Get OSCAR

Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.

User avatar
BamaGramma
Posts: 61
Joined: Thu Jun 20, 2019 8:00 am

Re: AHI worse with CPAP than without?

Post by BamaGramma » Fri Jun 28, 2019 3:02 pm

I downloaded Oscar but couldn't figure it out and didn't want to risk erasing the SD card. Is there a manual somewhere?

User avatar
LSAT
Posts: 13399
Joined: Sun Nov 16, 2008 10:11 am
Location: SE Wisconsin

Re: AHI worse with CPAP than without?

Post by LSAT » Fri Jun 28, 2019 3:56 pm

You won't erase the SD card..and even if you would, the data is in your machine and it would reload on the card the next time you put it in.

User avatar
BamaGramma
Posts: 61
Joined: Thu Jun 20, 2019 8:00 am

Re: AHI worse with CPAP than without?

Post by BamaGramma » Fri Jun 28, 2019 4:58 pm

I'll try again with Oscar when I have time to pay close attention. But based on everyone's comments, my thought that it might be related to getting used to the machine seems to be on target. Maybe I can stop worrying.

Though... I also have restless leg syndrome, diagnosed nearly 20 years ago via a sleep study. I take 1 mg of Mirapex at bedtime. The sleep doc pulled me off Melatonin because he said it aggravates RLS. Also said antihistamines do too. My legs haven't bothered me since I started the CPAP last week. Let's hope that continues.

User avatar
zonker
Posts: 11433
Joined: Fri Jun 19, 2015 4:36 pm

Re: AHI worse with CPAP than without?

Post by zonker » Fri Jun 28, 2019 5:16 pm

BamaGramma wrote:
Fri Jun 28, 2019 4:58 pm
I'll try again with Oscar when I have time to pay close attention. But based on everyone's comments, my thought that it might be related to getting used to the machine seems to be on target. Maybe I can stop worrying.

Though... I also have restless leg syndrome, diagnosed nearly 20 years ago via a sleep study. I take 1 mg of Mirapex at bedtime. The sleep doc pulled me off Melatonin because he said it aggravates RLS. Also said antihistamines do too. My legs haven't bothered me since I started the CPAP last week. Let's hope that continues.
bammie, when you DO start posting some charts, could you do that in this thread rather than start a new one? in would help to keep all your info together to make it easier for th e experts. thank you!
"Age is not an accomplishment and youth is not a sin"-Robert A. Heinlein
Oscar-Win
https://www.apneaboard.com/OSCAR/OSCAR-1.5.1-Win64.exe
Oscar-Mac
https://www.apneaboard.com/OSCAR/OSCAR-1.5.1.dmg


User avatar
zonker
Posts: 11433
Joined: Fri Jun 19, 2015 4:36 pm

Re: AHI worse with CPAP than without?

Post by zonker » Fri Jun 28, 2019 5:26 pm

BamaGramma wrote:
Fri Jun 28, 2019 5:18 pm
Will do!

hat-tip.gif
"Age is not an accomplishment and youth is not a sin"-Robert A. Heinlein
Oscar-Win
https://www.apneaboard.com/OSCAR/OSCAR-1.5.1-Win64.exe
Oscar-Mac
https://www.apneaboard.com/OSCAR/OSCAR-1.5.1.dmg

User avatar
ChicagoGranny
Posts: 15487
Joined: Sun Jan 29, 2012 1:43 pm
Location: USA

Re: AHI worse with CPAP than without?

Post by ChicagoGranny » Fri Jun 28, 2019 5:37 pm

BamaGramma wrote:
Fri Jun 28, 2019 4:58 pm
The sleep doc pulled me off Melatonin because he said it aggravates RLS.
Interesting. And, it seems to be true. I'll have to look for the study later.

Thanks.

User avatar
BamaGramma
Posts: 61
Joined: Thu Jun 20, 2019 8:00 am

Re: AHI worse with CPAP than without?

Post by BamaGramma » Fri Jun 28, 2019 5:40 pm

zonker wrote:
Fri Jun 28, 2019 5:26 pm
BamaGramma wrote:
Fri Jun 28, 2019 5:18 pm
Will do!


hat-tip.gif
How did you know I have a crush on Sam? :lol:

User avatar
zonker
Posts: 11433
Joined: Fri Jun 19, 2015 4:36 pm

Re: AHI worse with CPAP than without?

Post by zonker » Fri Jun 28, 2019 5:49 pm

BamaGramma wrote:
Fri Jun 28, 2019 5:40 pm
zonker wrote:
Fri Jun 28, 2019 5:26 pm
BamaGramma wrote:
Fri Jun 28, 2019 5:18 pm
Will do!


hat-tip.gif
How did you know I have a crush on Sam? :lol:
who doesn't?
:wink:
"Age is not an accomplishment and youth is not a sin"-Robert A. Heinlein
Oscar-Win
https://www.apneaboard.com/OSCAR/OSCAR-1.5.1-Win64.exe
Oscar-Mac
https://www.apneaboard.com/OSCAR/OSCAR-1.5.1.dmg

User avatar
BamaGramma
Posts: 61
Joined: Thu Jun 20, 2019 8:00 am

Re: AHI worse with CPAP than without?

Post by BamaGramma » Sun Jun 30, 2019 9:26 am

I put Oscar on my laptop but the screen isn't large enough to screenshot the data needed to post here. I downloaded it onto my desktop since the monitor is much larger, but it won't import the SD card. On the laptop I get an auto-message that it has detected a Philips SD card and do I want to import. No such message on the desktop. And when it does appear to import and asks me for a file location, I select Oscar_Date and... nothing.

Help!

User avatar
Goofproof
Posts: 16082
Joined: Mon Dec 05, 2005 3:16 pm
Location: Central Indiana, USA

Re: AHI worse with CPAP than without?

Post by Goofproof » Sun Jun 30, 2019 10:14 am

NoOnesPerfect wrote:
Fri Jun 28, 2019 1:22 pm
I fond that turning my machine off and back on quickly (to start the ramp function again) when I wake at night reduced my AHI from 6-10 down to sub 1-3.
A very good way to insure you machine isn't treating you, resetting to stop treatment. treatment isn't helping until ramping stops.
Use data to optimize your xPAP treatment!

"The art of medicine consists in amusing the patient while nature cures the disease." Voltaire