High AHI numbers

General Discussion on any topic relating to CPAP and/or Sleep Apnea.
User avatar
Miss Emerita
Posts: 3783
Joined: Sun Nov 04, 2018 8:07 pm

Re: High AHI numbers

Post by Miss Emerita » Fri Jun 21, 2019 6:00 pm

Hey, no quitting now, just when you made your wife’s life better!

I’m curious what the experts think about the PS of 4. I know zip about the VAuto machine, but I do wonder whether higher pressures on inhalation, to knock down the OAs, would be more tolerable if the PS were greater.
Oscar software is available at https://www.sleepfiles.com/OSCAR/

User avatar
palerider
Posts: 32299
Joined: Wed Dec 16, 2009 5:43 pm
Location: Dallas(ish).

Re: High AHI numbers

Post by palerider » Fri Jun 21, 2019 7:16 pm

Miss Emerita wrote:
Fri Jun 21, 2019 6:00 pm
I do wonder whether higher pressures on inhalation, to knock down the OAs
It doesn't work that way.

OAs are 'knocked down' *between* inhalations. ie, it's the EPAP that prevents OA..

OP needs higher min epap.

Now, more PS can make that higher EPAP 'feel better' because of the pressure drop at the end of inhalation... but too much PS might exacerbate the CAs

_________________
Mask: Bleep DreamPort CPAP Mask Solution
Additional Comments: S9 VPAP Auto
Get OSCAR

Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.

User avatar
Pugsy
Posts: 65327
Joined: Thu May 14, 2009 9:31 am
Location: Missouri, USA

Re: High AHI numbers

Post by Pugsy » Fri Jun 21, 2019 8:16 pm

TCW wrote:
Fri Jun 21, 2019 1:30 pm
EPAP set at 10 did bring down the AHI to 15.
What is the breakdown into each category of that AHI?
How much of it was the clear airway/central index?
How much of it was OA index?
How much of it was hyponea index?

_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier
Additional Comments: Mask Bleep Eclipse https://bleepsleep.com/the-eclipse/
I may have to RISE but I refuse to SHINE.

ajack
Posts: 977
Joined: Thu Mar 02, 2017 2:54 am
Location: australia

Re: High AHI numbers

Post by ajack » Sat Jun 22, 2019 6:56 am

To me it looks like you have some positional apnea as well and my benefit from a $10 foam cervical collar to see if that helps your airway. Hopfully you will get your treatment pressures lower.

On that chart I would have the min epap at 16 and may need more.

this titration guide may help
https://www.resmed.com/us/dam/documents ... er_eng.pdf

_________________
Mask: ResMed AirFit™ F20 Full Face CPAP Mask with Headgear
Humidifier: S9™ Series H5i™ Heated Humidifier with Climate Control
Additional Comments: S9 ST-A iVAPS and adapt ASV

User avatar
Pugsy
Posts: 65327
Joined: Thu May 14, 2009 9:31 am
Location: Missouri, USA

Re: High AHI numbers

Post by Pugsy » Sat Jun 22, 2019 7:09 am

I figured that going higher than 10 cm minimum EPAP would be a big enough challenge and your trying 12 cm and having problems confirmed it.

First of all...we need to know what kind of events you are having...and if the bulk of those events can be fixed with more pressure anyway...and if they can the we go slower with the increases in EPAP that will probably be needed.
That's why I asked about what category of events compose that AHI of 15.
They are probably obstructive in nature but I don't like to go on probably....I want to base plans of attack on facts.
If the bulk of the AHI is central now...we do something different....so really important to know what kind of events we are fighting.

If you do end up needing more minimum EPAP....we go with smaller increases and give your body a chance to get used to them.

_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier
Additional Comments: Mask Bleep Eclipse https://bleepsleep.com/the-eclipse/
I may have to RISE but I refuse to SHINE.

User avatar
Miss Emerita
Posts: 3783
Joined: Sun Nov 04, 2018 8:07 pm

Re: High AHI numbers

Post by Miss Emerita » Sat Jun 22, 2019 9:41 am

palerider, thanks for your help with my continuing education.
Oscar software is available at https://www.sleepfiles.com/OSCAR/

User avatar
palerider
Posts: 32299
Joined: Wed Dec 16, 2009 5:43 pm
Location: Dallas(ish).

Re: High AHI numbers

Post by palerider » Sat Jun 22, 2019 1:25 pm

Miss Emerita wrote:
Sat Jun 22, 2019 9:41 am
palerider, thanks for your help with my continuing education.
Think about it like this... most machines (except for the ventilators like ST, VAPS and ASV) are triggered to go to a higher pressure (even counting EPR/*flex) by your starting an inhalation.

If you don't have enough pressure *between breaths* (EPAP) to hold your airway open, then you can't take that next breath, and the machine never goes to the higher pressure.

Thus, you eliminate OA with EPAP, and twiddle with PS for a variety of other things that it can help.

_________________
Mask: Bleep DreamPort CPAP Mask Solution
Additional Comments: S9 VPAP Auto
Get OSCAR

Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.

User avatar
Miss Emerita
Posts: 3783
Joined: Sun Nov 04, 2018 8:07 pm

Re: High AHI numbers

Post by Miss Emerita » Sat Jun 22, 2019 5:53 pm

Got it! (Now to REMEMBER it....)
Oscar software is available at https://www.sleepfiles.com/OSCAR/

TCW
Posts: 10
Joined: Sun May 19, 2019 3:46 pm
Location: Maryville, TN

Re: High AHI numbers

Post by TCW » Sun Jun 23, 2019 12:51 pm

Here's what happened last night.

AHI 12.81
Obstructive index 9.26
Hypopnea 1.65
Clear airway index 1.90
Toms Stats ver 2.jpg
Toms Stats ver 2.jpg (814.32 KiB) Viewed 2187 times
Many thanks to all of you!

Tom

_________________
MachineMask

User avatar
palerider
Posts: 32299
Joined: Wed Dec 16, 2009 5:43 pm
Location: Dallas(ish).

Re: High AHI numbers

Post by palerider » Sun Jun 23, 2019 1:10 pm

TCW wrote:
Sun Jun 23, 2019 12:51 pm
Here's what happened last night.
Please review the formatting directions for sleepyhead/oscar:
wiki/index.php/Sleepyhead:Organize
So that you don't waste your time posting things that aren't useful.

_________________
Mask: Bleep DreamPort CPAP Mask Solution
Additional Comments: S9 VPAP Auto
Last edited by palerider on Sun Jun 23, 2019 4:32 pm, edited 1 time in total.
Get OSCAR

Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.

User avatar
raisedfist
Posts: 1176
Joined: Wed Jun 15, 2016 7:21 am

Re: High AHI numbers

Post by raisedfist » Sun Jun 23, 2019 1:14 pm

You spend like no time at your starting pressures which means they are inadequate

_________________
Mask
Philips Respironics Trilogy 100
AVAPS-AE Mode
PS Min 6, PS Max 18, EPAP Min 4, EPAP Max 12

User avatar
Pugsy
Posts: 65327
Joined: Thu May 14, 2009 9:31 am
Location: Missouri, USA

Re: High AHI numbers

Post by Pugsy » Sun Jun 23, 2019 1:24 pm

What were the settings last night? EPAP and IPAP?
Looks like EPAP was lower than 10 or 11...is that correct?
If so, did you do that because you just couldn't manage 10/11 EPAP and 14 IPAP ????

If you would do the screen shots in the usual way we could see the settings and stuff on the left side and the events graph and the other graphs on the right side...and get a better idea what happened.

_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier
Additional Comments: Mask Bleep Eclipse https://bleepsleep.com/the-eclipse/
I may have to RISE but I refuse to SHINE.

TCW
Posts: 10
Joined: Sun May 19, 2019 3:46 pm
Location: Maryville, TN

Re: High AHI numbers

Post by TCW » Sun Jun 23, 2019 2:13 pm

Thanks again to all of you who tried to help.

I have also developed swelling in my eyes and cheeks. My right eye is almost swollen shut.

I tried. I really did. I'm done. I'm gone.

Tom

_________________
MachineMask

User avatar
zonker
Posts: 11450
Joined: Fri Jun 19, 2015 4:36 pm

Re: High AHI numbers

Post by zonker » Sun Jun 23, 2019 3:22 pm

TCW wrote:
Sun Jun 23, 2019 2:13 pm
Thanks again to all of you who tried to help.

I have also developed swelling in my eyes and cheeks. My right eye is almost swollen shut.

I tried. I really did. I'm done. I'm gone.

Tom
well, i hope that you really aren't gone and that you came back. i know it seems like a long time in hell, but you haven't given it long enough.

you can do this. stick with it. keep on keepin' on, as the kids today say.

(NARRATOR: kids today don't say that.)
"Age is not an accomplishment and youth is not a sin"-Robert A. Heinlein
Oscar-Win
https://www.apneaboard.com/OSCAR/OSCAR-1.5.1-Win64.exe
Oscar-Mac
https://www.apneaboard.com/OSCAR/OSCAR-1.5.1.dmg

HoseCrusher
Posts: 2744
Joined: Tue Oct 12, 2010 6:42 pm

Re: High AHI numbers

Post by HoseCrusher » Sun Jun 23, 2019 4:36 pm

Tom, let's back up a moment...

The machine is working to keep your airway open. Is it possible that the swelling in the eyes and cheeks are being caused by something else?

I totally agree that you tried... I am sorry I was unable to help you succeed.

_________________
Mask: Brevida™ Nasal Pillow CPAP Mask with Headgear
Additional Comments: Machine is an AirSense 10 AutoSet For Her with Heated Humidifier.
SpO2 96+% and holding...