Turn up lower pressure setting on APAP?

General Discussion on any topic relating to CPAP and/or Sleep Apnea.
User avatar
Geoff33
Posts: 13
Joined: Wed Apr 17, 2019 9:44 am

Turn up lower pressure setting on APAP?

Post by Geoff33 » Tue Apr 23, 2019 8:30 am

In an effort to lower my AHI; I would be most thankful for any suggestions as to what I should turn my lower pressure setting to. Thanks in advance!
Attachments
screenshot-20190423-082028.png
screenshot-20190423-082028.png (238.32 KiB) Viewed 3639 times

User avatar
Julie
Posts: 20051
Joined: Tue Feb 28, 2006 12:58 pm

Re: Turn up lower pressure setting on APAP?

Post by Julie » Tue Apr 23, 2019 8:38 am

Can you please add this one to your other thread(s) so we can see what's gone before?

User avatar
LSAT
Posts: 13403
Joined: Sun Nov 16, 2008 10:11 am
Location: SE Wisconsin

Re: Turn up lower pressure setting on APAP?

Post by LSAT » Tue Apr 23, 2019 8:58 am

Raise min pressure back to 9-10 and watch your OAs drop. How are you sleeping?
Last edited by LSAT on Tue Apr 23, 2019 9:06 am, edited 2 times in total.

User avatar
Pugsy
Posts: 65327
Joined: Thu May 14, 2009 9:31 am
Location: Missouri, USA

Re: Turn up lower pressure setting on APAP?

Post by Pugsy » Tue Apr 23, 2019 9:00 am

Previous thread and one and only post prior to this one today.
viewtopic.php?f=1&t=175569&p=1301236#p1301236

_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier
Additional Comments: Mask Bleep Eclipse https://bleepsleep.com/the-eclipse/
I may have to RISE but I refuse to SHINE.

User avatar
Geoff33
Posts: 13
Joined: Wed Apr 17, 2019 9:44 am

Re: Turn up lower pressure setting on APAP?

Post by Geoff33 » Tue Apr 23, 2019 9:22 am

LSAT wrote:
Tue Apr 23, 2019 8:58 am
Raise min pressure back to 9-10 and watch your OAs drop. How are you sleeping?
sleep is better, thanks for the suggestion!

User avatar
Geoff33
Posts: 13
Joined: Wed Apr 17, 2019 9:44 am

Re: Turn up lower pressure setting on APAP?

Post by Geoff33 » Tue Apr 23, 2019 9:24 am

Pugsy wrote:
Tue Apr 23, 2019 9:00 am
Previous thread and one and only post prior to this one today.
viewtopic.php?f=1&t=175569&p=1301236#p1301236
Thank you, appreciate the help!

User avatar
palerider
Posts: 32299
Joined: Wed Dec 16, 2009 5:43 pm
Location: Dallas(ish).

Re: Turn up lower pressure setting on APAP?

Post by palerider » Tue Apr 23, 2019 12:46 pm

Geoff33 wrote:
Tue Apr 23, 2019 8:30 am
In an effort to lower my AHI; I would be most thankful for any suggestions as to what I should turn my lower pressure setting to. Thanks in advance!
Keep all your posts in this thread from now on.

Raise your minimum pressure to 10 and let's see how things go from there.

_________________
Mask: Bleep DreamPort CPAP Mask Solution
Additional Comments: S9 VPAP Auto
Get OSCAR

Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.

User avatar
Geoff33
Posts: 13
Joined: Wed Apr 17, 2019 9:44 am

Re: Turn up lower pressure setting on APAP?

Post by Geoff33 » Tue Apr 23, 2019 1:02 pm

palerider wrote:
Tue Apr 23, 2019 12:46 pm
Geoff33 wrote:
Tue Apr 23, 2019 8:30 am
In an effort to lower my AHI; I would be most thankful for any suggestions as to what I should turn my lower pressure setting to. Thanks in advance!
Keep all your posts in this thread from now on.

Raise your minimum pressure to 10 and let's see how things go from there.
Great, appreciate your input! Will make those adjustments and follow up on this thread!

User avatar
Tino2You
Posts: 635
Joined: Tue May 01, 2012 7:18 am
Location: Wilmington DE

Re: Turn up lower pressure setting on APAP?

Post by Tino2You » Tue Apr 23, 2019 2:28 pm

after a few days with the min pressure bumped, you might want to tighten up the high pressure to about 13.6. That is your current 95 percentile pressure.
-tino

_________________
Machine: AirSense™ 10 CPAP Machine with HumidAir™ Heated Humidifier
Mask: AirFit™ P10 Nasal Pillow CPAP Mask with Headgear
Additional Comments: Autoset Pressure 9-15.6, EPR 1, no ramp
Tino

User avatar
palerider
Posts: 32299
Joined: Wed Dec 16, 2009 5:43 pm
Location: Dallas(ish).

Re: Turn up lower pressure setting on APAP?

Post by palerider » Tue Apr 23, 2019 2:34 pm

Tino2You wrote:
Tue Apr 23, 2019 2:28 pm
after a few days with the min pressure bumped, you might want to tighten up the high pressure to about 13.6. That is your current 95 percentile pressure.
-tino
There is *absolutely no reason* to change the max pressure.

_________________
Mask: Bleep DreamPort CPAP Mask Solution
Additional Comments: S9 VPAP Auto
Get OSCAR

Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.

User avatar
zonker
Posts: 11450
Joined: Fri Jun 19, 2015 4:36 pm

Re: Turn up lower pressure setting on APAP?

Post by zonker » Tue Apr 23, 2019 3:01 pm

Tino2You wrote:
Tue Apr 23, 2019 2:28 pm
after a few days with the min pressure bumped, you might want to tighten up the high pressure to about 13.6. That is your current 95 percentile pressure.
-tino
why do you suggest this?
"Age is not an accomplishment and youth is not a sin"-Robert A. Heinlein
Oscar-Win
https://www.apneaboard.com/OSCAR/OSCAR-1.5.1-Win64.exe
Oscar-Mac
https://www.apneaboard.com/OSCAR/OSCAR-1.5.1.dmg

User avatar
palerider
Posts: 32299
Joined: Wed Dec 16, 2009 5:43 pm
Location: Dallas(ish).

Re: Turn up lower pressure setting on APAP?

Post by palerider » Tue Apr 23, 2019 3:50 pm

zonker wrote:
Tue Apr 23, 2019 3:01 pm
Tino2You wrote:
Tue Apr 23, 2019 2:28 pm
after a few days with the min pressure bumped, you might want to tighten up the high pressure to about 13.6. That is your current 95 percentile pressure.
-tino
why do you suggest this?
I can't imagine where such spectacularly bad 'advice' comes from.

_________________
Mask: Bleep DreamPort CPAP Mask Solution
Additional Comments: S9 VPAP Auto
Get OSCAR

Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.

User avatar
Geoff33
Posts: 13
Joined: Wed Apr 17, 2019 9:44 am

Re: Turn up lower pressure setting on APAP?

Post by Geoff33 » Wed Apr 24, 2019 3:21 pm

Last night I increased my lower pressure to 10 and my AHI was the best ever at 3.5. I will post my screenshot (SH) when I get home tonight. At any rate, some background info (forgive me if some of this is redundant). I started APAP with my Resmed 10 autosense approximately 12 days ago. I have tried both the Resmed F30 mask and P10 nasal pillow and prefer the nasal pillow. I am using the auto ramp function and it seems to help (didn't have it on initially). I feel like at first I was bouncing around with my settings and I don't want to do that, so I appreciate all input and feedback. I still don't feel totally rested, but the sleep is better than when my AHI=35 (sleep study). So, I had a couple nights at low pressure setting =8, then last night=10. Should I take the lower pressure up to 11? Or give my body some time with 10? I know it may be hard to say without my screenshot, so stay tuned and I will post it, just wanted to get a head start with some background info. Thanks for the help!

User avatar
palerider
Posts: 32299
Joined: Wed Dec 16, 2009 5:43 pm
Location: Dallas(ish).

Re: Turn up lower pressure setting on APAP?

Post by palerider » Wed Apr 24, 2019 3:27 pm

You have the wrong machine in your profile.

There's no such thing as an "autosense", what you have is the Resmed AirSense 10 AutoSet.

Please fix your profile equipment entry to avoid confusion in the future.

_________________
Mask: Bleep DreamPort CPAP Mask Solution
Additional Comments: S9 VPAP Auto
Get OSCAR

Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.

User avatar
Geoff33
Posts: 13
Joined: Wed Apr 17, 2019 9:44 am

Re: Turn up lower pressure setting on APAP?

Post by Geoff33 » Wed Apr 24, 2019 9:08 pm

here is the screenshot from last night. Low pressure of 11? Or should I continue with 10 again and let my body continue to get used to it? Thanks!
Attachments
screenshot-20190424-210515.png
screenshot-20190424-210515.png (220.63 KiB) Viewed 3468 times