Reducing pressure setting

General Discussion on any topic relating to CPAP and/or Sleep Apnea.
lhpangler
Posts: 59
Joined: Sat Jul 10, 2010 2:13 pm

Reducing pressure setting

Post by lhpangler » Mon Mar 11, 2019 2:14 pm

Hi folks, ive tried different searches and i just cant seem to find what i am looking for. My ahi for the last year is 0.2. Pressure is set to 6 and 10. My friend is in a similar situation. Any thoughts regarding lowering my pressure to see if my AHI creeps back up? Tia

_________________
Mask: Mirage Activa™ LT Nasal CPAP Mask with Headgear
Additional Comments: Resmed S9 VPAP
Last edited by lhpangler on Wed Mar 13, 2019 10:24 am, edited 1 time in total.

User avatar
palerider
Posts: 32299
Joined: Wed Dec 16, 2009 5:43 pm
Location: Dallas(ish).

Re: Reducing pressure setting

Post by palerider » Mon Mar 11, 2019 2:37 pm

lhpangler wrote:
Mon Mar 11, 2019 2:14 pm
Hi folks, ive tried different searches and i just cant seem to find what i am looking for. My ahi for the last year is 0.2. Pressure is set to 6 and 10. My friend is in a similar situation. Any thoughts regarding lowering my pressure to see if i get a betters nights sleep? Tia
Less pressure rarely results in better sleep. There's more to sleep disordered breathing than just the ahi

_________________
Mask: Bleep DreamPort CPAP Mask Solution
Additional Comments: S9 VPAP Auto
Get OSCAR

Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.

User avatar
Dog Slobber
Posts: 4272
Joined: Thu Feb 15, 2018 2:05 pm
Location: Ontario, Canada

Re: Reducing pressure setting

Post by Dog Slobber » Mon Mar 11, 2019 2:41 pm

There are no graphs to properly advise anyone. So you're asking people to guess..
Lowering the pressure is more likely to raise your AHI.

I'm assuming your intent is less pressure will be more comfortable. Perhaps, lowering the minimum could result in more volatile pressure swings with a net result of higher peaks as the machine tries to catch up. Will you really feel the difference between a minimum of 6 and 7.

But I think experimentation, looking for better therapy should always be encouraged.

Make a change, give it some time, note changes and see how you feel........rinse, repeat.
Last edited by Dog Slobber on Mon Mar 11, 2019 3:16 pm, edited 1 time in total.
Battery Backup: EcoFlow Delta 2

lhpangler
Posts: 59
Joined: Sat Jul 10, 2010 2:13 pm

Re: Reducing pressure setting

Post by lhpangler » Mon Mar 11, 2019 3:10 pm

Two great responses so far. Ty

_________________
Mask: Mirage Activa™ LT Nasal CPAP Mask with Headgear
Additional Comments: Resmed S9 VPAP

User avatar
kteague
Posts: 7783
Joined: Tue May 16, 2006 8:30 pm
Location: West and Midwest

Re: Reducing pressure setting

Post by kteague » Mon Mar 11, 2019 4:21 pm

By "6 and 10" are you saying you're using an auto range of 6 to 10? Some things to consider... Having any AHI at all means something measurable (though negligible in the big picture) was able to slip past the machines defenses. If your pressure rises above 6 at all, it is in response to something it senses, and lowering the pressures is likely to cause it to have to make more efforts towards resolving those things. Having a machine taking longer to address issues is for some more sleep disruptive. Unless your starting pressure of 6 causes air in your belly or you are one of the exceptions who develop centrals even at low pressures, I wouldn't think lowering your pressure is of any advantage. Raising your starting pressure may give you perfect zeroes, not that it is necessary. What kind of problems are you having with your sleep? Maybe there's some other things that need to be considered.

_________________
Mask: TAP PAP Nasal Pillow CPAP Mask with Improved Stability Mouthpiece
Humidifier: S9™ Series H5i™ Heated Humidifier with Climate Control
Additional Comments: Bleep/DreamPort for full nights, Tap Pap for shorter sessions

User avatar
palerider
Posts: 32299
Joined: Wed Dec 16, 2009 5:43 pm
Location: Dallas(ish).

Re: Reducing pressure setting

Post by palerider » Mon Mar 11, 2019 4:49 pm

KnightSleeper wrote:
Mon Mar 11, 2019 3:52 pm
lhpangler wrote:
Mon Mar 11, 2019 2:14 pm
Hi folks, ive tried different searches and i just cant seem to find what i am looking for. My ahi for the last year is 0.2. Pressure is set to 6 and 10. My friend is in a similar situation. Any thoughts regarding lowering my pressure to see if i get a betters nights sleep? Tia
lol...this is like someone getting an A+ in a class and saying "can anyone find out how I can get a better grade???" I would think with that low of an ahi, you're sleeping on Cloud-9, no??? :)
You could have a low ahi, but still have FLs and snores that are disturbing sleep...

_________________
Mask: Bleep DreamPort CPAP Mask Solution
Additional Comments: S9 VPAP Auto
Get OSCAR

Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.

User avatar
zonker
Posts: 11369
Joined: Fri Jun 19, 2015 4:36 pm

Re: Reducing pressure setting

Post by zonker » Mon Mar 11, 2019 4:57 pm

lhpangler wrote:
Mon Mar 11, 2019 2:14 pm
Hi folks, ive tried different searches and i just cant seem to find what i am looking for.
howdy. you don't list a machine in your sig, just a mask. i'll assume you have a sleepyhead compatible machine. while you aren't a newbie, perhaps reading here will get you up to speed-

viewtopic/t172378/Sticky--Newbies-PLEAS ... STING.html

once the good boys and girls here see your data, they can make some recommendations.

good luck!
Last edited by zonker on Mon Mar 11, 2019 5:37 pm, edited 1 time in total.
"Age is not an accomplishment and youth is not a sin"-Robert A. Heinlein
Oscar-Win
https://www.apneaboard.com/OSCAR/OSCAR-1.5.1-Win64.exe
Oscar-Mac
https://www.apneaboard.com/OSCAR/OSCAR-1.5.1.dmg

User avatar
palerider
Posts: 32299
Joined: Wed Dec 16, 2009 5:43 pm
Location: Dallas(ish).

Re: Reducing pressure setting

Post by palerider » Mon Mar 11, 2019 6:45 pm

KnightSleeper wrote:
Mon Mar 11, 2019 5:11 pm
palerider wrote:
Mon Mar 11, 2019 4:49 pm
KnightSleeper wrote:
Mon Mar 11, 2019 3:52 pm
lhpangler wrote:
Mon Mar 11, 2019 2:14 pm
Hi folks, ive tried different searches and i just cant seem to find what i am looking for. My ahi for the last year is 0.2. Pressure is set to 6 and 10. My friend is in a similar situation. Any thoughts regarding lowering my pressure to see if i get a betters nights sleep? Tia
lol...this is like someone getting an A+ in a class and saying "can anyone find out how I can get a better grade???" I would think with that low of an ahi, you're sleeping on Cloud-9, no??? :)
You could have a low ahi, but still have FLs and snores that are disturbing sleep...

Hey, all my homies are following me...lol

That's interesting that one could have a real low ahi, ie. .2, yet still have some sleep disturbances! Btw, I surely didn't mean to step-on lhpangler's thread... I was just surprised that one could have a very, very low ahi and still not be sleeping that well. - - I'm always learning something new on this forum! Thanks guys! :D
I think this is relevant. We see it with people saying "my AHI is good (and it is) but I still feel like crap".

And that's when I try to remind people that there's more to sleep disordered breathing than just apneas and hypopneas (AHI)... and, of course, there's more to bad sleep than just breathing problems.

But, yeah, you can have an active flow limitation graph, which will disturb your sleep, but never get bad enough to have apneas or hypos.

_________________
Mask: Bleep DreamPort CPAP Mask Solution
Additional Comments: S9 VPAP Auto
Get OSCAR

Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.

lhpangler
Posts: 59
Joined: Sat Jul 10, 2010 2:13 pm

Re: Reducing pressure setting

Post by lhpangler » Wed Mar 13, 2019 10:23 am

kteague wrote:
Mon Mar 11, 2019 4:21 pm
By "6 and 10" are you saying you're using an auto range of 6 to 10? Some things to consider... Having any AHI at all means something measurable (though negligible in the big picture) was able to slip past the machines defenses. If your pressure rises above 6 at all, it is in response to something it senses, and lowering the pressures is likely to cause it to have to make more efforts towards resolving those things. Having a machine taking longer to address issues is for some more sleep disruptive. Unless your starting pressure of 6 causes air in your belly or you are one of the exceptions who develop centrals even at low pressures, I wouldn't think lowering your pressure is of any advantage. Raising your starting pressure may give you perfect zeroes, not that it is necessary. What kind of problems are you having with your sleep? Maybe there's some other things that need to be considered.
6-10 IPAP/EPAP I have a bipap.
I don't have any sleep problems with the exception of machine related interruptions: mask makes funny noise, exhaust air bouncing off of the pillow, nasal pillow getting dislodged.

So I wonder to what extent my bipap is still necessary. There are conflicting opinions as to whether apnea is permanent. We have seen diabetes and heart disease reverse by diet, why not apnea?

https://www.mayoclinic.org/diseases-con ... c-20352095

_________________
Mask: Mirage Activa™ LT Nasal CPAP Mask with Headgear
Additional Comments: Resmed S9 VPAP

User avatar
Julie
Posts: 20051
Joined: Tue Feb 28, 2006 12:58 pm

Re: Reducing pressure setting

Post by Julie » Wed Mar 13, 2019 11:40 am

Because in the cases of so many here who've lost some weight and then had a new test done, while they might need less pressure to open their airways, they still end up needing Cpap except in rare cases where they e.g. lost 100 lbs, had stomach stapling, or possibly only had 'positional' or very mild apnea, etc. If it were that easy a lot of us would not be on Cpap. Trouble is some people initially feel better after wt loss, but gain it back (sooo often) and/or don't take a new test because they feel a bit better or AHI is down a bit, but end up needing to go back on Cpap anyhow.

User avatar
palerider
Posts: 32299
Joined: Wed Dec 16, 2009 5:43 pm
Location: Dallas(ish).

Re: Reducing pressure setting

Post by palerider » Wed Mar 13, 2019 11:45 am

lhpangler wrote:
Wed Mar 13, 2019 10:23 am
There are conflicting opinions as to whether apnea is permanent. We have seen diabetes and heart disease reverse by diet, why not apnea?
Because diabetes and heart disease aren't a result of a physical structure problem.

_________________
Mask: Bleep DreamPort CPAP Mask Solution
Additional Comments: S9 VPAP Auto
Get OSCAR

Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.

User avatar
Julie
Posts: 20051
Joined: Tue Feb 28, 2006 12:58 pm

Re: Reducing pressure setting

Post by Julie » Wed Mar 13, 2019 1:03 pm

Good point!

lhpangler
Posts: 59
Joined: Sat Jul 10, 2010 2:13 pm

Re: Reducing pressure setting

Post by lhpangler » Wed Mar 13, 2019 1:26 pm

zonker wrote:
Mon Mar 11, 2019 4:57 pm
lhpangler wrote:
Mon Mar 11, 2019 2:14 pm
Hi folks, ive tried different searches and i just cant seem to find what i am looking for.
howdy. you don't list a machine in your sig, just a mask. i'll assume you have a sleepyhead compatible machine. while you aren't a newbie, perhaps reading here will get you up to speed-

viewtopic/t172378/Sticky--Newbies-PLEAS ... STING.html

once the good boys and girls here see your data, they can make some recommendations.

good luck!
I updated my machine. I set the value years ago, not sure where it went. Resmed S9. Machine updated. ty!

_________________
Mask: Mirage Activa™ LT Nasal CPAP Mask with Headgear
Additional Comments: Resmed S9 VPAP

lhpangler
Posts: 59
Joined: Sat Jul 10, 2010 2:13 pm

Re: Reducing pressure setting

Post by lhpangler » Wed Mar 13, 2019 1:31 pm

Julie wrote:
Wed Mar 13, 2019 11:40 am
Because in the cases of so many here who've lost some weight and then had a new test done, while they might need less pressure to open their airways, they still end up needing Cpap except in rare cases where they e.g. lost 100 lbs, had stomach stapling, or possibly only had 'positional' or very mild apnea, etc. If it were that easy a lot of us would not be on Cpap. Trouble is some people initially feel better after wt loss, but gain it back (sooo often) and/or don't take a new test because they feel a bit better or AHI is down a bit, but end up needing to go back on Cpap anyhow.
I am approaching the 100#s down marker and been dropping weight for three years. The mayo article suggests weight loss can help. I understand the issue being structural. I understand that people need to go back or continue to use the bipap. So how can one reconfirm the need to be on the bipap?

_________________
Mask: Mirage Activa™ LT Nasal CPAP Mask with Headgear
Additional Comments: Resmed S9 VPAP

User avatar
Dog Slobber
Posts: 4272
Joined: Thu Feb 15, 2018 2:05 pm
Location: Ontario, Canada

Re: Reducing pressure setting

Post by Dog Slobber » Wed Mar 13, 2019 1:43 pm

lhpangler wrote:
Wed Mar 13, 2019 10:23 am
So I wonder to what extent my bipap is still necessary. There are conflicting opinions as to whether apnea is permanent. We have seen diabetes and heart disease reverse by diet, why not apnea?

https://www.mayoclinic.org/diseases-con ... c-20352095
There are conflicting opinions on whether the earth is flat.