Does my Respironics catch OA and H?
- SleepySwiss
- Posts: 25
- Joined: Mon Feb 11, 2019 11:52 pm
Does my Respironics catch OA and H?
<r>I have a Respironics machine with humidifier since about 2 years. I have a heated hose and use a F&P Eson nasal mask. I love the machine and have no problem with breathing etc. However, at anything above an AHI of 2 I do not feel well rested, below 1.5 I feel 'fully functional.' The first year+ I was fluctuating a lot between ~2 and just below 5. I then started to take Ashwaganda and I dropped to 1-3AHI. While using Sleepyhead for a while to follow my AHI, I never paid much attention to anything else than the AHI. Just recently, I found this board and am intrigued by the offer of support. When I look at my events, it seems the machine responds to FL but not to either OA or H. I post two pngs: one from last night with AHI 2.37 and one from a couple of days ago with 1.1. Today I don't feel so hot, a little headache and concentration issues. Last Friday I was about ready for a marathon. I appreciate that many on this board would be happy with either number. My doc tells me that anything under 5 is great (I think that's BS!). And I also appreciate that there might be many extrinsic factors, like stress etc, BUT if I could set my treatment into a more effective regime I would be thrilled silly - and make the world a better place by being more focused.<br/>
Any thoughts?
<ATTACHMENT filename="screenshot-20190226-141952.png" index="0"><s> </e></ATTACHMENT>
<ATTACHMENT filename="screenshot-20190226-141805.png" index="1"><s> </e></ATTACHMENT></r>
Any thoughts?
<ATTACHMENT filename="screenshot-20190226-141952.png" index="0"><s> </e></ATTACHMENT>
<ATTACHMENT filename="screenshot-20190226-141805.png" index="1"><s> </e></ATTACHMENT></r>
_________________
Machine: AirSense™ 10 CPAP Machine with HumidAir™ Heated Humidifier |
Mask: Nuance & Nuance Pro Nasal Pillow CPAP Mask with Gel Nasal Pillows |
Re: Does my Respironics catch OA and H?
welcome to the forum. unfortunately, i can't see your charts. not sure if this is something wrong at my end or yours. let's see if others can see and advise.
and yes, your machine certainly should respond to both oa and h. so i'm assuming your therapy will need some tweaking.
and yes, your machine certainly should respond to both oa and h. so i'm assuming your therapy will need some tweaking.
_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier |
Mask: AirFit™ P10 Nasal Pillow CPAP Mask with Headgear |
people say i'm self absorbed.
but that's enough about them.
Oscar-Win
https://www.apneaboard.com/OSCAR/OSCAR-1.5.1-Win64.exe
Oscar-Mac
https://www.apneaboard.com/OSCAR/OSCAR-1.5.1.dmg
but that's enough about them.
Oscar-Win
https://www.apneaboard.com/OSCAR/OSCAR-1.5.1-Win64.exe
Oscar-Mac
https://www.apneaboard.com/OSCAR/OSCAR-1.5.1.dmg
Re: Does my Respironics catch OA and H?
They're posting with the wrong kind of markup language (that stuff between < and > ) that's not valid for this kind of forum.
As to respironics responding to OA and H... they will, but they're really slow to do so:
https://www.youtube.com/watch?v=lzCCgNLya_g
Get OSCAR
Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.
Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.
Re: Does my Respironics catch OA and H?
okay. maybe they'll come back and do it proper for this forum.palerider wrote: ↑Tue Feb 26, 2019 6:33 pmThey're posting with the wrong kind of markup language (that stuff between < and > ) that's not valid for this kind of forum.
As to respironics responding to OA and H... they will, but they're really slow to do so:
https://www.youtube.com/watch?v=lzCCgNLya_g
_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier |
Mask: AirFit™ P10 Nasal Pillow CPAP Mask with Headgear |
people say i'm self absorbed.
but that's enough about them.
Oscar-Win
https://www.apneaboard.com/OSCAR/OSCAR-1.5.1-Win64.exe
Oscar-Mac
https://www.apneaboard.com/OSCAR/OSCAR-1.5.1.dmg
but that's enough about them.
Oscar-Win
https://www.apneaboard.com/OSCAR/OSCAR-1.5.1-Win64.exe
Oscar-Mac
https://www.apneaboard.com/OSCAR/OSCAR-1.5.1.dmg
- ChicagoGranny
- Posts: 15286
- Joined: Sun Jan 29, 2012 1:43 pm
- Location: USA
Re: Does my Respironics catch OA and H?
You can attach files by going to the bottom of the Reply box and clicking on the Attachments tab. Then, click Place In Line.
- SleepySwiss
- Posts: 25
- Joined: Mon Feb 11, 2019 11:52 pm
Re: Does my Respironics catch OA and H?
thanks a ton!
Felix
_________________
Machine: AirSense™ 10 CPAP Machine with HumidAir™ Heated Humidifier |
Mask: Nuance & Nuance Pro Nasal Pillow CPAP Mask with Gel Nasal Pillows |
Re: Does my Respironics catch OA and H?
I'd suggest starting by bumping the min pressure up to 7. and seeing how that does over a few days time.SleepySwiss wrote: ↑Wed Feb 27, 2019 12:39 amSo I am looking forward to ideas on what to adjust.
thanks a ton!
Felix
Get OSCAR
Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.
Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.
- SleepySwiss
- Posts: 25
- Joined: Mon Feb 11, 2019 11:52 pm
Re: Does my Respironics catch OA and H?
Thank you, Palerider, I will try that and report back in a few days.
I thought given that it stays at 6 most of the time that would be enough, but you make the good point - or at least your comment implies - that this is exactly my problem, the machine is 'stuck' at 6!
I thought given that it stays at 6 most of the time that would be enough, but you make the good point - or at least your comment implies - that this is exactly my problem, the machine is 'stuck' at 6!
_________________
Machine: AirSense™ 10 CPAP Machine with HumidAir™ Heated Humidifier |
Mask: Nuance & Nuance Pro Nasal Pillow CPAP Mask with Gel Nasal Pillows |
Re: Does my Respironics catch OA and H?
Your brand of machine is very slow to respond to changing needs.SleepySwiss wrote: ↑Wed Feb 27, 2019 10:08 amThank you, Palerider, I will try that and report back in a few days.
I thought given that it stays at 6 most of the time that would be enough, but you make the good point - or at least your comment implies - that this is exactly my problem, the machine is 'stuck' at 6!
https://www.youtube.com/watch?v=lzCCgNLya_g
Thus, it's very important to make sure that the minimum pressure is high enough to take care of the little things that happen... a little more pressure can hold open you airway better, and hopefully smooth out your breathing.
The more of those pressure spikes you have, the better.
Get OSCAR
Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.
Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.
- SleepySwiss
- Posts: 25
- Joined: Mon Feb 11, 2019 11:52 pm
Re: Does my Respironics catch OA and H?
Thank you, Palerider! Do you think the Dreamstation would have a similar response to the System One? Also there is a claim that the video is made by ResMed?
I am one night in on minimal pressure 7. Not the immediate remedy I was wishing for, but I'll keep trying.
My other thought was Central vs Obstructive. I would expect the central apneas to be less responsive?
I am one night in on minimal pressure 7. Not the immediate remedy I was wishing for, but I'll keep trying.
My other thought was Central vs Obstructive. I would expect the central apneas to be less responsive?
_________________
Machine: AirSense™ 10 CPAP Machine with HumidAir™ Heated Humidifier |
Mask: Nuance & Nuance Pro Nasal Pillow CPAP Mask with Gel Nasal Pillows |
Re: Does my Respironics catch OA and H?
Yes, the algorithms are fundamentally unchanged between the PRS1 and the Dreamstation, as well as the Resmed S9 and Airsense 10 lines.SleepySwiss wrote: ↑Thu Feb 28, 2019 11:07 amThank you, Palerider! Do you think the Dreamstation would have a similar response to the System One? Also there is a claim that the video is made by ResMed?
I am one night in on minimal pressure 7. Not the immediate remedy I was wishing for, but I'll keep trying.
My other thought was Central vs Obstructive. I would expect the central apneas to be less responsive?
Yes, the video was done by Resmed, but it's accurate, you can see it in your graphs, where you're having hypopnes, and obstructives, and the machine does nothing. Resmeds respond to single events, Respironics do not.
What that means is that people end up needing a higher minimum pressure on the Respironics, to prevent those single events.
Get OSCAR
Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.
Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.
- SleepySwiss
- Posts: 25
- Joined: Mon Feb 11, 2019 11:52 pm
Re: Does my Respironics catch OA and H?
Thank you Palerider! I wish my sleep Dr would be so knowledgeable!
I had my second night on a 7 base pressure, not great yet, but I remain hopeful
I had my second night on a 7 base pressure, not great yet, but I remain hopeful
_________________
Machine: AirSense™ 10 CPAP Machine with HumidAir™ Heated Humidifier |
Mask: Nuance & Nuance Pro Nasal Pillow CPAP Mask with Gel Nasal Pillows |
Re: Does my Respironics catch OA and H?
how do you feel when you wake up? are you feeling better with the minimum raised?SleepySwiss wrote: ↑Fri Mar 01, 2019 10:47 amThank you Palerider! I wish my sleep Dr would be so knowledgeable!
I had my second night on a 7 base pressure, not great yet, but I remain hopeful
_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier |
Mask: AirFit™ P10 Nasal Pillow CPAP Mask with Headgear |
people say i'm self absorbed.
but that's enough about them.
Oscar-Win
https://www.apneaboard.com/OSCAR/OSCAR-1.5.1-Win64.exe
Oscar-Mac
https://www.apneaboard.com/OSCAR/OSCAR-1.5.1.dmg
but that's enough about them.
Oscar-Win
https://www.apneaboard.com/OSCAR/OSCAR-1.5.1-Win64.exe
Oscar-Mac
https://www.apneaboard.com/OSCAR/OSCAR-1.5.1.dmg
Re: Does my Respironics catch OA and H?
I think you're going to need more pressure, If 7 is bothering you, then wait a few more days, 7 isn't bothering you, then bump it up to 8 and let's see if we can get rid of some of those hypopneas.SleepySwiss wrote: ↑Fri Mar 01, 2019 10:47 amThank you Palerider! I wish my sleep Dr would be so knowledgeable!
I had my second night on a 7 base pressure, not great yet, but I remain hopeful
March1.jpg
Get OSCAR
Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.
Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.
- SleepySwiss
- Posts: 25
- Joined: Mon Feb 11, 2019 11:52 pm
Re: Does my Respironics catch OA and H?
I am sleeping fine with the pressure raised, but not feeling any better in the morning. In fact today I do have a headache which is a bit unusual, but does happen.zonker wrote: ↑Fri Mar 01, 2019 10:51 amhow do you feel when you wake up? are you feeling better with the minimum raised?SleepySwiss wrote: ↑Fri Mar 01, 2019 10:47 amThank you Palerider! I wish my sleep Dr would be so knowledgeable!
I had my second night on a 7 base pressure, not great yet, but I remain hopeful
_________________
Machine: AirSense™ 10 CPAP Machine with HumidAir™ Heated Humidifier |
Mask: Nuance & Nuance Pro Nasal Pillow CPAP Mask with Gel Nasal Pillows |