SD Card Error - Resmed S9

General Discussion on any topic relating to CPAP and/or Sleep Apnea.
FatBaldOldMan
Posts: 2
Joined: Tue Dec 15, 2015 11:43 pm
Location: Abingdon, Maryland

SD Card Error - Resmed S9

Post by FatBaldOldMan » Mon Feb 25, 2019 10:20 pm

I have an issue with my Resmed S9. Up until recently, the SD card was working properly, but now I'm receiving the error, "SD Card error, remove and press start to begin therapy." Initially, I thought that the SD card may have become corrupted, so I replaced the card with a new one today, but the error again reoccurred. I thought it may have been a formatting issue; so I reformated the card to no avail. I tried the card in another PAP device, and it worked properly. After it worked properly in the other PAP, I placed it in the original PAP, and the error persisted.

Is this an issue with my S9, or does anyone have any other suggestions?

User avatar
palerider
Posts: 32299
Joined: Wed Dec 16, 2009 5:43 pm
Location: Dallas(ish).

Re: SD Card Error - Resmed S9

Post by palerider » Mon Feb 25, 2019 11:10 pm

FatBaldOldMan wrote:
Mon Feb 25, 2019 10:20 pm
I have an issue with my Resmed S9. Up until recently, the SD card was working properly, but now I'm receiving the error, "SD Card error, remove and press start to begin therapy." Initially, I thought that the SD card may have become corrupted, so I replaced the card with a new one today, but the error again reoccurred. I thought it may have been a formatting issue; so I reformated the card to no avail. I tried the card in another PAP device, and it worked properly. After it worked properly in the other PAP, I placed it in the original PAP, and the error persisted.

Is this an issue with my S9, or does anyone have any other suggestions?
It *might* be dust/cruft in the sd card slot that's preventing it making good contact. I've seen that before.

_________________
Mask: Bleep DreamPort CPAP Mask Solution
Additional Comments: S9 VPAP Auto
Get OSCAR

Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.

User avatar
zonker
Posts: 11399
Joined: Fri Jun 19, 2015 4:36 pm

Re: SD Card Error - Resmed S9

Post by zonker » Thu Apr 25, 2019 10:49 am

palerider wrote:
Mon Feb 25, 2019 11:10 pm
FatBaldOldMan wrote:
Mon Feb 25, 2019 10:20 pm
I have an issue with my Resmed S9. Up until recently, the SD card was working properly, but now I'm receiving the error, "SD Card error, remove and press start to begin therapy." Initially, I thought that the SD card may have become corrupted, so I replaced the card with a new one today, but the error again reoccurred. I thought it may have been a formatting issue; so I reformated the card to no avail. I tried the card in another PAP device, and it worked properly. After it worked properly in the other PAP, I placed it in the original PAP, and the error persisted.

Is this an issue with my S9, or does anyone have any other suggestions?
It *might* be dust/cruft in the sd card slot that's preventing it making good contact. I've seen that before.
this has started happening to me intermittently. i have some compressed air or canned air or whatever it's called. do you a squirt of it into the card slot would help?
"Age is not an accomplishment and youth is not a sin"-Robert A. Heinlein
Oscar-Win
https://www.apneaboard.com/OSCAR/OSCAR-1.5.1-Win64.exe
Oscar-Mac
https://www.apneaboard.com/OSCAR/OSCAR-1.5.1.dmg

User avatar
grayghost4
Posts: 1554
Joined: Wed Aug 20, 2014 6:52 pm
Location: Norther Illinois
Contact:

Re: SD Card Error - Resmed S9

Post by grayghost4 » Thu Apr 25, 2019 1:37 pm

there are also electronic cleaners in spray cans

https://www.amazon.com/CRC-05103-Electr ... way&sr=8-2

also available at auto supply stores or home depot
If you're not part of the solution you're just scumming up the bottom of the beaker!

Get the Clinicians manual here : http://apneaboard.com/adjust-cpap-press ... tup-manual

User avatar
zonker
Posts: 11399
Joined: Fri Jun 19, 2015 4:36 pm

Re: SD Card Error - Resmed S9

Post by zonker » Thu Apr 25, 2019 1:44 pm

grayghost4 wrote:
Thu Apr 25, 2019 1:37 pm
there are also electronic cleaners in spray cans

https://www.amazon.com/CRC-05103-Electr ... way&sr=8-2

also available at auto supply stores or home depot
thank you. i'm somewhat concerned by this-

"This item is not for sale in Catalina Island"

:!: :roll:
"Age is not an accomplishment and youth is not a sin"-Robert A. Heinlein
Oscar-Win
https://www.apneaboard.com/OSCAR/OSCAR-1.5.1-Win64.exe
Oscar-Mac
https://www.apneaboard.com/OSCAR/OSCAR-1.5.1.dmg

User avatar
grayghost4
Posts: 1554
Joined: Wed Aug 20, 2014 6:52 pm
Location: Norther Illinois
Contact:

Re: SD Card Error - Resmed S9

Post by grayghost4 » Thu Apr 25, 2019 2:00 pm

that is not where to OP is located ... amazon may only have air shipment to that Island.

Location: Abingdon, Maryland
If you're not part of the solution you're just scumming up the bottom of the beaker!

Get the Clinicians manual here : http://apneaboard.com/adjust-cpap-press ... tup-manual

User avatar
palerider
Posts: 32299
Joined: Wed Dec 16, 2009 5:43 pm
Location: Dallas(ish).

Re: SD Card Error - Resmed S9

Post by palerider » Thu Apr 25, 2019 4:35 pm

zonker wrote:
Thu Apr 25, 2019 1:44 pm
grayghost4 wrote:
Thu Apr 25, 2019 1:37 pm
there are also electronic cleaners in spray cans

https://www.amazon.com/CRC-05103-Electr ... way&sr=8-2

also available at auto supply stores or home depot
thank you. i'm somewhat concerned by this-

"This item is not for sale in Catalina Island"

:!: :roll:
Apparently, it's cuz there's no ground shipping to there:
https://www.reddit.com/r/LosAngeles/com ... _for_this/

_________________
Mask: Bleep DreamPort CPAP Mask Solution
Additional Comments: S9 VPAP Auto
Get OSCAR

Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.

User avatar
zonker
Posts: 11399
Joined: Fri Jun 19, 2015 4:36 pm

Re: SD Card Error - Resmed S9

Post by zonker » Thu Apr 25, 2019 5:14 pm

palerider wrote:
Thu Apr 25, 2019 4:35 pm
zonker wrote:
Thu Apr 25, 2019 1:44 pm
grayghost4 wrote:
Thu Apr 25, 2019 1:37 pm
there are also electronic cleaners in spray cans

https://www.amazon.com/CRC-05103-Electr ... way&sr=8-2

also available at auto supply stores or home depot
thank you. i'm somewhat concerned by this-

"This item is not for sale in Catalina Island"

:!: :roll:
Apparently, it's cuz there's no ground shipping to there:
https://www.reddit.com/r/LosAngeles/com ... _for_this/
well, that's what they get for living on an island, i guess.

without regard to that, would i potentially cause damage to my machine's card reader if i blew air in there?
"Age is not an accomplishment and youth is not a sin"-Robert A. Heinlein
Oscar-Win
https://www.apneaboard.com/OSCAR/OSCAR-1.5.1-Win64.exe
Oscar-Mac
https://www.apneaboard.com/OSCAR/OSCAR-1.5.1.dmg

User avatar
palerider
Posts: 32299
Joined: Wed Dec 16, 2009 5:43 pm
Location: Dallas(ish).

Re: SD Card Error - Resmed S9

Post by palerider » Thu Apr 25, 2019 6:01 pm

zonker wrote:
Thu Apr 25, 2019 5:14 pm
without regard to that, would i potentially cause damage to my machine's card reader if i blew air in there?
I wouldn't think so, just don't blast it with a big compressor or anything :)

_________________
Mask: Bleep DreamPort CPAP Mask Solution
Additional Comments: S9 VPAP Auto
Get OSCAR

Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.

User avatar
zonker
Posts: 11399
Joined: Fri Jun 19, 2015 4:36 pm

Re: SD Card Error - Resmed S9

Post by zonker » Thu Apr 25, 2019 6:02 pm

palerider wrote:
Thu Apr 25, 2019 6:01 pm
zonker wrote:
Thu Apr 25, 2019 5:14 pm
without regard to that, would i potentially cause damage to my machine's card reader if i blew air in there?
I wouldn't think so, just don't blast it with a big compressor or anything :)
damn! was gonna take out to the garage and strap it down to my workbench.

you're no fun at all! :lol:

nope just the canned air.
"Age is not an accomplishment and youth is not a sin"-Robert A. Heinlein
Oscar-Win
https://www.apneaboard.com/OSCAR/OSCAR-1.5.1-Win64.exe
Oscar-Mac
https://www.apneaboard.com/OSCAR/OSCAR-1.5.1.dmg

User avatar
palerider
Posts: 32299
Joined: Wed Dec 16, 2009 5:43 pm
Location: Dallas(ish).

Re: SD Card Error - Resmed S9

Post by palerider » Thu Apr 25, 2019 6:18 pm

zonker wrote:
Thu Apr 25, 2019 6:02 pm
palerider wrote:
Thu Apr 25, 2019 6:01 pm
zonker wrote:
Thu Apr 25, 2019 5:14 pm
without regard to that, would i potentially cause damage to my machine's card reader if i blew air in there?
I wouldn't think so, just don't blast it with a big compressor or anything :)
damn! was gonna take out to the garage and strap it down to my workbench.

you're no fun at all! :lol:

nope just the canned air.
Just be gentle when you poke the tube in, so you don't hook and possibly bend any of the little fingers that make contact to the sd card.

_________________
Mask: Bleep DreamPort CPAP Mask Solution
Additional Comments: S9 VPAP Auto
Get OSCAR

Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.

User avatar
zonker
Posts: 11399
Joined: Fri Jun 19, 2015 4:36 pm

Re: SD Card Error - Resmed S9

Post by zonker » Thu Apr 25, 2019 6:22 pm

palerider wrote:
Thu Apr 25, 2019 6:18 pm
zonker wrote:
Thu Apr 25, 2019 6:02 pm
palerider wrote:
Thu Apr 25, 2019 6:01 pm
zonker wrote:
Thu Apr 25, 2019 5:14 pm
without regard to that, would i potentially cause damage to my machine's card reader if i blew air in there?
I wouldn't think so, just don't blast it with a big compressor or anything :)
damn! was gonna take out to the garage and strap it down to my workbench.

you're no fun at all! :lol:

nope just the canned air.
Just be gentle when you poke the tube in, so you don't hook and possibly bend any of the little fingers that make contact to the sd card.
ended up chasing the dog with the canned air. i'm easily distracted!

sprayed sd card and sd reader input slot, but got the same error message. tried a different, but not new, sd card. once it "prepared" the sd card, saw no error message.

will use this card tonight and see what tomorrow brings. if nothing else, it's a troubleshooting step.
"Age is not an accomplishment and youth is not a sin"-Robert A. Heinlein
Oscar-Win
https://www.apneaboard.com/OSCAR/OSCAR-1.5.1-Win64.exe
Oscar-Mac
https://www.apneaboard.com/OSCAR/OSCAR-1.5.1.dmg

FatBaldOldMan
Posts: 2
Joined: Tue Dec 15, 2015 11:43 pm
Location: Abingdon, Maryland

Re: SD Card Error - Resmed S9

Post by FatBaldOldMan » Thu Apr 25, 2019 9:28 pm

Thank you to all those who contributed. As it turns out, my BIPAP was malfunctioning and it's been replaced. I'm now using an auto-BIPAP, also know as an ASV device. Again thanks to those who contributed.

Sent from my SM-G973U using Tapatalk


User avatar
zonker
Posts: 11399
Joined: Fri Jun 19, 2015 4:36 pm

Re: SD Card Error - Resmed S9

Post by zonker » Thu Apr 25, 2019 9:49 pm

FatBaldOldMan wrote:
Thu Apr 25, 2019 9:28 pm
Thank you to all those who contributed. As it turns out, my BIPAP was malfunctioning and it's been replaced. I'm now using an auto-BIPAP, also know as an ASV device. Again thanks to those who contributed.

Sent from my SM-G973U using Tapatalk
glad you got it fixed!

i just re-read your first post. i hope that changing my sd card works!!
"Age is not an accomplishment and youth is not a sin"-Robert A. Heinlein
Oscar-Win
https://www.apneaboard.com/OSCAR/OSCAR-1.5.1-Win64.exe
Oscar-Mac
https://www.apneaboard.com/OSCAR/OSCAR-1.5.1.dmg

User avatar
palerider
Posts: 32299
Joined: Wed Dec 16, 2009 5:43 pm
Location: Dallas(ish).

Re: SD Card Error - Resmed S9

Post by palerider » Thu Apr 25, 2019 11:38 pm

FatBaldOldMan wrote:
Thu Apr 25, 2019 9:28 pm
Thank you to all those who contributed. As it turns out, my BIPAP was malfunctioning and it's been replaced. I'm now using an auto-BIPAP, also know as an ASV device. Again thanks to those who contributed.
In order to avoid confusion to people that may be reading this:

An "auto bilevel" machine is one that acts much like an auto cpap, but with separate settable pressures for inhale and exhale. It is a type of the generic category of bilevel machines.

An ASV (Adaptive Servo Ventilator) is a highly specialized type of bilevel machine, and is not an 'auto' machine. Auto machines respond slowly, over many minutes of time, whereas ASVs respond almost instantly, changing pressure breath by breath. And ASV will force air into you, making you breathe, and auto will not.

_________________
Mask: Bleep DreamPort CPAP Mask Solution
Additional Comments: S9 VPAP Auto
Get OSCAR

Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.