0.33 reading but rubbish sleep?

General Discussion on any topic relating to CPAP and/or Sleep Apnea.
jono_1uk
Posts: 23
Joined: Thu Feb 14, 2019 6:30 am

0.33 reading but rubbish sleep?

Post by jono_1uk » Sun Feb 17, 2019 10:06 am

So last night i supposedly got a reading of 033 but i felt i had a crap night sleep and i had to keep adjust the mask to sit under my nose correctly ..

can anyone please tell me weather that would / has affected the readings?
sleepyhead results are below ..
Capture.PNG
Capture.PNG (122.17 KiB) Viewed 980 times

User avatar
Pugsy
Posts: 65284
Joined: Thu May 14, 2009 9:31 am
Location: Missouri, USA

Re: 0.33 reading but rubbish sleep?

Post by Pugsy » Sun Feb 17, 2019 10:14 am

Yes. All that BND (Breathing Not Detected) pretty much totally invalidates any numbers you see or most likely don't see.
If the machine can't detect breathing it can't detect much of anything. No way to know if not much happened or a lot happened but the machine simply was clueless. I would vote for a clueless machine.

_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier
Additional Comments: Mask Bleep Eclipse https://bleepsleep.com/the-eclipse/
I may have to RISE but I refuse to SHINE.

jono_1uk
Posts: 23
Joined: Thu Feb 14, 2019 6:30 am

Re: 0.33 reading but rubbish sleep?

Post by jono_1uk » Sun Feb 17, 2019 10:55 am

thanks so Wednesday to Thursday i felt i had a brilliant sleep so does this reflect that?
Capture.PNG
Capture.PNG (124.11 KiB) Viewed 966 times

User avatar
Pugsy
Posts: 65284
Joined: Thu May 14, 2009 9:31 am
Location: Missouri, USA

Re: 0.33 reading but rubbish sleep?

Post by Pugsy » Sun Feb 17, 2019 11:09 am

jono_1uk wrote:
Sun Feb 17, 2019 10:55 am
thanks so Wednesday to Thursday i felt i had a brilliant sleep so does this reflect that?
Not particularly. Maybe the first hour and a half if you slept.
Anytime you see BND of this magnitude...the data reported is pretty much meaningless.
Whatever you are doing to create this much BND...you need to fix it.

Also can you review this thread for how to format the images and which graphs to include?
viewtopic/t158560/How-to-post-images-for-review.html
Will be useful once you ever get some reports where there is something to look at that isn't BND.

_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier
Additional Comments: Mask Bleep Eclipse https://bleepsleep.com/the-eclipse/
I may have to RISE but I refuse to SHINE.

jono_1uk
Posts: 23
Joined: Thu Feb 14, 2019 6:30 am

Re: 0.33 reading but rubbish sleep?

Post by jono_1uk » Sun Feb 17, 2019 11:49 pm

Thanks for your help here.... so last night BND was 50% again .. my mask is the open one that sits under the nose ( having swapped from a closed one i really couldnt stand) does it need to press right up against the nose?

User avatar
Okie bipap
Posts: 3567
Joined: Thu Oct 15, 2015 4:14 pm
Location: Central Oklahoma

Re: 0.33 reading but rubbish sleep?

Post by Okie bipap » Mon Feb 18, 2019 10:56 am

The cushion needs to seal against the bottom of the nose in order for it to deliver the necessary air pressure. Have you considered trying the nasal pillows? I used the dreamwear full face mask which seals against the bottom of the nose, but have quit using it because when I got it snug enough to seal around the nose, it caused my nose to be very sensitive and would hurt when I first took the mask off every morning.

_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier
Mask: Evora Full Face Mask - Fitpack
Additional Comments: IPAP 20-25, ps 4, OSCAR software
Growing old is mandatory, but growing up is optional.

User avatar
palerider
Posts: 32299
Joined: Wed Dec 16, 2009 5:43 pm
Location: Dallas(ish).

Re: 0.33 reading but rubbish sleep?

Post by palerider » Mon Feb 18, 2019 11:46 am

jono_1uk wrote:
Sun Feb 17, 2019 11:49 pm
Thanks for your help here.... so last night BND was 50% again .. my mask is the open one that sits under the nose ( having swapped from a closed one i really couldnt stand) does it need to press right up against the nose?
You need to figure out what it is, and put it in your equipment profile.

There are no "open" masks. and yeah, surprisingly enough they have to make a tight seal, or they're worthless.

_________________
Mask: Bleep DreamPort CPAP Mask Solution
Additional Comments: S9 VPAP Auto
Get OSCAR

Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.

jono_1uk
Posts: 23
Joined: Thu Feb 14, 2019 6:30 am

Re: 0.33 reading but rubbish sleep?

Post by jono_1uk » Mon Feb 18, 2019 1:14 pm

OK so equipment posted .. just looked back at sleepyhead results for when i was using the Amara full face ...AHI was 49 BUT BND was only 5% ..
so should i go back to that anf make some changes and if so what?
i swapped it out as i didnt like the air blowing in my face ..so they reduced the max pressure when they swapped out.
would prefer the nasal pillow out of the 2 if i could reduce that BND so any advice please ..

User avatar
palerider
Posts: 32299
Joined: Wed Dec 16, 2009 5:43 pm
Location: Dallas(ish).

Re: 0.33 reading but rubbish sleep?

Post by palerider » Mon Feb 18, 2019 1:51 pm

jono_1uk wrote:
Mon Feb 18, 2019 1:14 pm
OK so equipment posted .. just looked back at sleepyhead results for when i was using the Amara full face ...AHI was 49 BUT BND was only 5% ..
so should i go back to that anf make some changes and if so what?
i swapped it out as i didnt like the air blowing in my face ..so they reduced the max pressure when they swapped out.
would prefer the nasal pillow out of the 2 if i could reduce that BND so any advice please ..
You need to get the mask fitting property, so that it seals well. Lots of people find that it works well for them, but every mask doesn't work for everyone.

Maybe the new Bleep mask would work for you.

_________________
Mask: Bleep DreamPort CPAP Mask Solution
Additional Comments: S9 VPAP Auto
Get OSCAR

Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.

jono_1uk
Posts: 23
Joined: Thu Feb 14, 2019 6:30 am

Re: 0.33 reading but rubbish sleep?

Post by jono_1uk » Tue Feb 19, 2019 12:29 am

ok so made some adjustments and got rid of the BND ... AHI is a bit higher then i would like ...
Capture.PNG
Capture.PNG (21.62 KiB) Viewed 806 times

User avatar
palerider
Posts: 32299
Joined: Wed Dec 16, 2009 5:43 pm
Location: Dallas(ish).

Re: 0.33 reading but rubbish sleep?

Post by palerider » Tue Feb 19, 2019 12:52 am

Considerably higher than it should be.

But, we can't tell anything from a partial screen shot.

See below.

_________________
Mask: Bleep DreamPort CPAP Mask Solution
Additional Comments: S9 VPAP Auto
Get OSCAR

Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.

jono_1uk
Posts: 23
Joined: Thu Feb 14, 2019 6:30 am

Re: 0.33 reading but rubbish sleep?

Post by jono_1uk » Tue Feb 19, 2019 1:22 am

Thanks.. Ill post the graphs up tonight.

User avatar
palerider
Posts: 32299
Joined: Wed Dec 16, 2009 5:43 pm
Location: Dallas(ish).

Re: 0.33 reading but rubbish sleep?

Post by palerider » Tue Feb 19, 2019 1:23 am

jono_1uk wrote:
Tue Feb 19, 2019 1:22 am
Thanks.. Ill post the graphs up tonight.
This may lead us to the cause of your rubbish sleep :)

_________________
Mask: Bleep DreamPort CPAP Mask Solution
Additional Comments: S9 VPAP Auto
Get OSCAR

Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.

jono_1uk
Posts: 23
Joined: Thu Feb 14, 2019 6:30 am

Re: 0.33 reading but rubbish sleep?

Post by jono_1uk » Tue Feb 19, 2019 1:32 am

lol -so can nothing be gained from what i posted then? it needs to be the graphs?

User avatar
palerider
Posts: 32299
Joined: Wed Dec 16, 2009 5:43 pm
Location: Dallas(ish).

Re: 0.33 reading but rubbish sleep?

Post by palerider » Tue Feb 19, 2019 1:46 am

jono_1uk wrote:
Tue Feb 19, 2019 1:32 am
lol -so can nothing be gained from what i posted then? it needs to be the graphs?
There's a reason I created this page, referenced in the sig of every one of my posts:
wiki/index.php/Sleepyhead:Organize

Your last post basically tells us "a thing happened", without any clues about what that thing was, or how it happened.

_________________
Mask: Bleep DreamPort CPAP Mask Solution
Additional Comments: S9 VPAP Auto
Get OSCAR

Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.