Ok - I was going to restrict the max pressure just to start with - I have this irrational fear that the machine will shoot straight up to 25cms and hang out there. I was trying to feel as though I had some head room for a day or two.Pugsy wrote: ↑Fri May 17, 2019 6:35 pmPS of 4 is what the labs commonly start out with and most people come out of a bilevel titration with PS of 4 or 5.
On a rare occasion a little more for some reason or other.
My suggestion is start with PS of 4
use auto mode...
Minimum EPAP 17
PS of 4
Maximum IPAP 25
See how much a little more IPAP helps with those FLs before adding more EPAP....might help out just enough.
Kate's Therapy
- katestyles
- Posts: 610
- Joined: Sun Jan 06, 2019 9:08 am
Re: Kate's Therapy
_________________
| Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier |
| Mask: Bleep DreamPort CPAP Mask Solution |
| Additional Comments: Back up mask - anything in the drawer |
Re: Kate's Therapy
Oh...it also will have some other comfort settings.....use the factory defaults first and see how you do with them.
Most people won't need to change those but if you have trouble getting in sync with the machine rate...then we can change them.
When I was using the VAuto S9 I did just fine with those default settings.
You know to go to the apneaboard and get the clinical manual????
It's a combo manual.
https://www.apneaboard.com/adjust-cpap- ... tup-manual
If you want to restrict the max...try 22 IPAP max...tiny room to roam.
Most people won't need to change those but if you have trouble getting in sync with the machine rate...then we can change them.
When I was using the VAuto S9 I did just fine with those default settings.
You know to go to the apneaboard and get the clinical manual????
It's a combo manual.
https://www.apneaboard.com/adjust-cpap- ... tup-manual
If you want to restrict the max...try 22 IPAP max...tiny room to roam.
_________________
| Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier |
| Additional Comments: Mask Bleep Eclipse https://bleepsleep.com/the-eclipse/ |
I may have to RISE but I refuse to SHINE.
- katestyles
- Posts: 610
- Joined: Sun Jan 06, 2019 9:08 am
Re: Kate's Therapy
I have the clinical manual - I have been reading it avidly for days!Pugsy wrote: ↑Fri May 17, 2019 6:39 pmOh...it also will have some other comfort settings.....use the factory defaults first and see how you do with them.
Most people won't need to change those but if you have trouble getting in sync with the machine rate...then we can change them.
When I was using the VAuto S9 I did just fine with those default settings.
You know to go to the apneaboard and get the clinical manual????
It's a combo manual.
https://www.apneaboard.com/adjust-cpap- ... tup-manual
If you want to restrict the max...try 22 IPAP max...tiny room to roam.
_________________
| Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier |
| Mask: Bleep DreamPort CPAP Mask Solution |
| Additional Comments: Back up mask - anything in the drawer |
Re: Kate's Therapy
I predict you will like the machine.
I see no sense in using PS of 3...that's what your EPR is doing. Make use of the machine's special abilities.
First time I ever used a bilevel machine...was a fixed bilevel machine and I used 4 PS...and it took me about 5 minutes to say to myself..."you gotta get one of these"
It was a used machine that I bought from a forum member for a friend who needed bilevel because her pressure needs were like yours and she had some COPD and was struggling.
She wanted me to use it first to make sure it was working like it should and set it up for her.
So I did.
When I sent it to her I told her that if she didn't like it that I would gladly buy it back....she got the machine and used it one night and the next morning the only message I got was "over my dead body"
I see no sense in using PS of 3...that's what your EPR is doing. Make use of the machine's special abilities.
First time I ever used a bilevel machine...was a fixed bilevel machine and I used 4 PS...and it took me about 5 minutes to say to myself..."you gotta get one of these"
It was a used machine that I bought from a forum member for a friend who needed bilevel because her pressure needs were like yours and she had some COPD and was struggling.
She wanted me to use it first to make sure it was working like it should and set it up for her.
So I did.
When I sent it to her I told her that if she didn't like it that I would gladly buy it back....she got the machine and used it one night and the next morning the only message I got was "over my dead body"
_________________
| Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier |
| Additional Comments: Mask Bleep Eclipse https://bleepsleep.com/the-eclipse/ |
I may have to RISE but I refuse to SHINE.
- katestyles
- Posts: 610
- Joined: Sun Jan 06, 2019 9:08 am
Re: Kate's Therapy
lol;Pugsy wrote: ↑Fri May 17, 2019 6:48 pmI predict you will like the machine.
I see no sense in using PS of 3...that's what your EPR is doing. Make use of the machine's special abilities.
First time I ever used a bilevel machine...was a fixed bilevel machine and I used 4 PS...and it took me about 5 minutes to say to myself..."you gotta get one of these"![]()
It was a used machine that I bought from a forum member for a friend who needed bilevel because her pressure needs were like yours and she had some COPD and was struggling.
She wanted me to use it first to make sure it was working like it should and set it up for her.
So I did.
When I sent it to her I told her that if she didn't like it that I would gladly buy it back....she got the machine and used it one night and the next morning the only message I got was "over my dead body"![]()
I am just excited not to be maxxed out at 20
_________________
| Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier |
| Mask: Bleep DreamPort CPAP Mask Solution |
| Additional Comments: Back up mask - anything in the drawer |
Re: Kate's Therapy
What I would suggest is:katestyles wrote: ↑Fri May 17, 2019 6:17 pmSo - trying hard not to be overexcited, but thanks to LSAT, I have a tracking number that says I will be the proud owner of a Resmed Vauto tommorrow by 8pm!!
I am planning to start at comparable settings to my current APAP which is running at min 20, max 20, EPR 3.
Can someone check my math please?
1) Should I be putting Epap 17, Ipap 20?
2) All the titration protocols say PS should be at 4. Is 3 OK?
3) Should I be doing any other kind of settings?
4) What is the next change to make? I'm trying to limit FL.
MinEPAP 17
MaxIPAP 25
PS 4 (or 3)
and see what it wants to do.
The auto doesn't have epap and ipap, it has minepap and maxipap, and PS to set the pressures, and ranges.
If you go with a PS of 3, it'll *start* at what you've got now, and go up if it needs to.. if you go with the 4 that Pugsy recommends, then it'd be like a pressure of 21, EPR 4 (which, of course, you can't do.. .but I figured if I said it that way, it'd make more sense since that's the nomenclature you're used to at this point.
Get OSCAR
Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.
Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.
Re: Kate's Therapy
Please don't be one of those "I want to lower the max because 'reasons'" people.katestyles wrote: ↑Fri May 17, 2019 6:39 pmOk - I was going to restrict the max pressure just to start with - I have this irrational fear that the machine will shoot straight up to 25cms and hang out there. I was trying to feel as though I had some head room for a day or two.Pugsy wrote: ↑Fri May 17, 2019 6:35 pmPS of 4 is what the labs commonly start out with and most people come out of a bilevel titration with PS of 4 or 5.
On a rare occasion a little more for some reason or other.
My suggestion is start with PS of 4
use auto mode...
Minimum EPAP 17
PS of 4
Maximum IPAP 25
See how much a little more IPAP helps with those FLs before adding more EPAP....might help out just enough.
Put maxipap on 25, and *trust the machine*.
Get OSCAR
Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.
Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.
- katestyles
- Posts: 610
- Joined: Sun Jan 06, 2019 9:08 am
Re: Kate's Therapy
I am not one of those - I am one of the ones who let the max be 20, and found the machine bumping against the top of the graphs.palerider wrote: ↑Fri May 17, 2019 6:53 pmPlease don't be one of those "I want to lower the max because 'reasons'" people.katestyles wrote: ↑Fri May 17, 2019 6:39 pmOk - I was going to restrict the max pressure just to start with - I have this irrational fear that the machine will shoot straight up to 25cms and hang out there. I was trying to feel as though I had some head room for a day or two.Pugsy wrote: ↑Fri May 17, 2019 6:35 pmPS of 4 is what the labs commonly start out with and most people come out of a bilevel titration with PS of 4 or 5.
On a rare occasion a little more for some reason or other.
My suggestion is start with PS of 4
use auto mode...
Minimum EPAP 17
PS of 4
Maximum IPAP 25
See how much a little more IPAP helps with those FLs before adding more EPAP....might help out just enough.
Put maxipap on 25, and *trust the machine*.
Just for a few days I wanted daylight above the max pressure line.....
_________________
| Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier |
| Mask: Bleep DreamPort CPAP Mask Solution |
| Additional Comments: Back up mask - anything in the drawer |
- katestyles
- Posts: 610
- Joined: Sun Jan 06, 2019 9:08 am
Re: Kate's Therapy
OK.palerider wrote: ↑Fri May 17, 2019 6:52 pmWhat I would suggest is:katestyles wrote: ↑Fri May 17, 2019 6:17 pmSo - trying hard not to be overexcited, but thanks to LSAT, I have a tracking number that says I will be the proud owner of a Resmed Vauto tommorrow by 8pm!!
I am planning to start at comparable settings to my current APAP which is running at min 20, max 20, EPR 3.
Can someone check my math please?
1) Should I be putting Epap 17, Ipap 20?
2) All the titration protocols say PS should be at 4. Is 3 OK?
3) Should I be doing any other kind of settings?
4) What is the next change to make? I'm trying to limit FL.
MinEPAP 17
MaxIPAP 25
PS 4 (or 3)
and see what it wants to do.
The auto doesn't have epap and ipap, it has minepap and maxipap, and PS to set the pressures, and ranges.
If you go with a PS of 3, it'll *start* at what you've got now, and go up if it needs to.. if you go with the 4 that Pugsy recommends, then it'd be like a pressure of 21, EPR 4 (which, of course, you can't do.. .but I figured if I said it that way, it'd make more sense since that's the nomenclature you're used to at this point.
MinEPAP 17
MaxIPAP 25
PS 4 (or 3)
_________________
| Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier |
| Mask: Bleep DreamPort CPAP Mask Solution |
| Additional Comments: Back up mask - anything in the drawer |
- babydinosnoreless
- Posts: 2373
- Joined: Fri Nov 02, 2018 2:53 pm
Re: Kate's Therapy
Try ps 4. You will like it. I tried 3 and it sucked.katestyles wrote: ↑Fri May 17, 2019 8:12 pmOK.palerider wrote: ↑Fri May 17, 2019 6:52 pmWhat I would suggest is:katestyles wrote: ↑Fri May 17, 2019 6:17 pmSo - trying hard not to be overexcited, but thanks to LSAT, I have a tracking number that says I will be the proud owner of a Resmed Vauto tommorrow by 8pm!!
I am planning to start at comparable settings to my current APAP which is running at min 20, max 20, EPR 3.
Can someone check my math please?
1) Should I be putting Epap 17, Ipap 20?
2) All the titration protocols say PS should be at 4. Is 3 OK?
3) Should I be doing any other kind of settings?
4) What is the next change to make? I'm trying to limit FL.
MinEPAP 17
MaxIPAP 25
PS 4 (or 3)
and see what it wants to do.
The auto doesn't have epap and ipap, it has minepap and maxipap, and PS to set the pressures, and ranges.
If you go with a PS of 3, it'll *start* at what you've got now, and go up if it needs to.. if you go with the 4 that Pugsy recommends, then it'd be like a pressure of 21, EPR 4 (which, of course, you can't do.. .but I figured if I said it that way, it'd make more sense since that's the nomenclature you're used to at this point.
MinEPAP 17
MaxIPAP 25
PS 4 (or 3)
you can always back it down if you don't like it. It doesn't seem like it would make a huge difference in comfort but it does.
_________________
| Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier |
| Mask: AirFit™ N20 Nasal CPAP Mask with Headgear |
Re: Kate's Therapy
Whatever you set it at - best of luck to you! I'm late getting in here, but congrats!
Heck, I'm excited for you. Looking forward to seeing how it goes.
Heck, I'm excited for you. Looking forward to seeing how it goes.
_________________
| Machine: ResMed AirSense™ 10 AutoSet™ CPAP Machine with HumidAir™ Heated Humidifier |
| Mask: Bleep DreamPort CPAP Mask Solution |
| Additional Comments: Backup machine: AirSense 10 AutoSet with all the fixins |
- Dog Slobber
- Posts: 4336
- Joined: Thu Feb 15, 2018 2:05 pm
- Location: Ontario, Canada
Re: Kate's Therapy
Who can't wait for katestyles to get her new bilevel and improve her therapy?

And while we're at it, three cheers for LSAT!


And while we're at it, three cheers for LSAT!

_________________
| Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier |
| Mask: AirFit™ P30i Nasal Pillow CPAP Mask with Headgear Starter Pack |
| Additional Comments: Min EPAP: 8.2, Max IPAP: 25, PS:4 |
Battery Backup: EcoFlow Delta 2
Re: Kate's Therapy
Raises hand, and yells "HOORAY!"
You're awesome DogSlobber. Your gif's sum up my feelings better than my words do!
You're awesome DogSlobber. Your gif's sum up my feelings better than my words do!
_________________
| Machine: ResMed AirSense™ 10 AutoSet™ CPAP Machine with HumidAir™ Heated Humidifier |
| Mask: Bleep DreamPort CPAP Mask Solution |
| Additional Comments: Backup machine: AirSense 10 AutoSet with all the fixins |
- Dog Slobber
- Posts: 4336
- Joined: Thu Feb 15, 2018 2:05 pm
- Location: Ontario, Canada
Re: Kate's Therapy
Now that our buddy Zonk is back, I have to step up my friendliness and support game. Hard to compete with that guy.
_________________
| Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier |
| Mask: AirFit™ P30i Nasal Pillow CPAP Mask with Headgear Starter Pack |
| Additional Comments: Min EPAP: 8.2, Max IPAP: 25, PS:4 |
Battery Backup: EcoFlow Delta 2
- katestyles
- Posts: 610
- Joined: Sun Jan 06, 2019 9:08 am
Re: Kate's Therapy
It is so beautiful!
_________________
| Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier |
| Mask: Bleep DreamPort CPAP Mask Solution |
| Additional Comments: Back up mask - anything in the drawer |
- Attachments
-
- IMG_1673.JPG (113.27 KiB) Viewed 3185 times