New To CPAP - Looking for Advice
New To CPAP - Looking for Advice
Hello,
Was diagnosed with moderate sleep apnea from a home test at the end of December with a 15.1 AHI. Got a machine about a week ago, Dreamstation auto CPAP and Amara full head gear mask. Had success the first 3 nights with little issues sleeping and AHI around 4 but had issues with the mask digging into the bridge of my nose and forehead. Tried loosening it more to no avail so ordered the Amara view and tried it last night for the first time. Was like night and day in terms of comfort, slept very soundly and only woke up once in the night which is typical. I would say I feel more rested than usual this morning but still residual sleepiness which I know is to be expected.
Attaching my sleepyhead data from last night. Just looking for any tips to improve my setup, lower my AHI, or anything else helpful.
Thanks,
Paul
Was diagnosed with moderate sleep apnea from a home test at the end of December with a 15.1 AHI. Got a machine about a week ago, Dreamstation auto CPAP and Amara full head gear mask. Had success the first 3 nights with little issues sleeping and AHI around 4 but had issues with the mask digging into the bridge of my nose and forehead. Tried loosening it more to no avail so ordered the Amara view and tried it last night for the first time. Was like night and day in terms of comfort, slept very soundly and only woke up once in the night which is typical. I would say I feel more rested than usual this morning but still residual sleepiness which I know is to be expected.
Attaching my sleepyhead data from last night. Just looking for any tips to improve my setup, lower my AHI, or anything else helpful.
Thanks,
Paul
_________________
| Machine: DreamStation Auto CPAP Machine |
| Mask: Amara View Full Face CPAP Mask with Headgear |
| Humidifier: DreamStation Heated Humidifier |
- Attachments
-
- screenshot-20190129-155758.png (115.18 KiB) Viewed 1054 times
Re: New To CPAP - Looking for Advice
I just adore newbies that post a perfect screenshot right out of the gate. Well done.
I'd bump your min pressure a cm or two... That should help smooth out some of the breathing issues. There is room for improvement, but over all, you're doing pretty good.
I'd bump your min pressure a cm or two... That should help smooth out some of the breathing issues. There is room for improvement, but over all, you're doing pretty good.
Get OSCAR
Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.
Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.
Re: New To CPAP - Looking for Advice
Thanks, appreciate it. I'll try bumping it up and report back.
_________________
| Machine: DreamStation Auto CPAP Machine |
| Mask: Amara View Full Face CPAP Mask with Headgear |
| Humidifier: DreamStation Heated Humidifier |
Re: New To CPAP - Looking for Advice
welcome to the zoo. it really looks like you've got a good handle on things, right out of the gate. i agree with palerider (i'd be foolish not to!), raise the minimum pressure and see how it goes.
look forward to seeing how you progress!
_________________
| Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier |
| Mask: AirFit™ P10 Nasal Pillow CPAP Mask with Headgear |
"Age is not an accomplishment and youth is not a sin"-Robert A. Heinlein
Oscar-Win
https://www.apneaboard.com/OSCAR/OSCAR-1.5.1-Win64.exe
Oscar-Mac
https://www.apneaboard.com/OSCAR/OSCAR-1.5.1.dmg
Oscar-Win
https://www.apneaboard.com/OSCAR/OSCAR-1.5.1-Win64.exe
Oscar-Mac
https://www.apneaboard.com/OSCAR/OSCAR-1.5.1.dmg
Re: New To CPAP - Looking for Advice
YOU are lucky! When I was diagnosed last year, it was because I was getting O2 numbers in the law 70s when I was asleep during a hospital stay. They gave me a device to check my sleep, and after using and returning it, I had over 58 events per hour!! Yes, they REALLY wanted me on a machine ASAP! Took almost 6 months, and open heart surgery to get me comfortable with the machine, but it has really helped me feel better.pweiric7 wrote: ↑Tue Jan 29, 2019 3:09 pmHello,
Was diagnosed with moderate sleep apnea from a home test at the end of December with a 15.1 AHI. Got a machine about a week ago, Dreamstation auto CPAP and Amara full head gear mask. Had success the first 3 nights with little issues sleeping and AHI around 4 but had issues with the mask digging into the bridge of my nose and forehead. Tried loosening it more to no avail so ordered the Amara view and tried it last night for the first time. Was like night and day in terms of comfort, slept very soundly and only woke up once in the night which is typical. I would say I feel more rested than usual this morning but still residual sleepiness which I know is to be expected.
Attaching my sleepyhead data from last night. Just looking for any tips to improve my setup, lower my AHI, or anything else helpful.
Thanks,
Paul
_________________
| Machine: ResMed AirSense™ 10 AutoSet™ CPAP Machine with HumidAir™ Heated Humidifier |
| Mask: AirTouch™ F20 Full Face CPAP Mask with Headgear |
Airsense 10 Autoset
AirTouch and AirFit F10 masks

AirTouch and AirFit F10 masks
Re: New To CPAP - Looking for Advice
* Note, open heart surgery not required to successfully use CPAP.
Get OSCAR
Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.
Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.
Re: New To CPAP - Looking for Advice
Glad to hear it helped eventually for you! I certainly know I’m fortunate with my start after reading lots of posts here. Hoping it translates into feeling less tired and lethargic over time.SDBud wrote: ↑Tue Jan 29, 2019 9:02 pm
YOU are lucky! When I was diagnosed last year, it was because I was getting O2 numbers in the law 70s when I was asleep during a hospital stay. They gave me a device to check my sleep, and after using and returning it, I had over 58 events per hour!! Yes, they REALLY wanted me on a machine ASAP! Took almost 6 months, and open heart surgery to get me comfortable with the machine, but it has really helped me feel better.
_________________
| Machine: DreamStation Auto CPAP Machine |
| Mask: Amara View Full Face CPAP Mask with Headgear |
| Humidifier: DreamStation Heated Humidifier |
Re: New To CPAP - Looking for Advice
it absolutely WILL. wish i could tell you exactly when. wish i could tell you it will all come at once.
but can't lie. if you're like most people, it will get better and better by degrees.
but it will come!
_________________
| Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier |
| Mask: AirFit™ P10 Nasal Pillow CPAP Mask with Headgear |
"Age is not an accomplishment and youth is not a sin"-Robert A. Heinlein
Oscar-Win
https://www.apneaboard.com/OSCAR/OSCAR-1.5.1-Win64.exe
Oscar-Mac
https://www.apneaboard.com/OSCAR/OSCAR-1.5.1.dmg
Oscar-Win
https://www.apneaboard.com/OSCAR/OSCAR-1.5.1-Win64.exe
Oscar-Mac
https://www.apneaboard.com/OSCAR/OSCAR-1.5.1.dmg
Re: New To CPAP - Looking for Advice
Here's last nights screenshot after raising the minimum pressure to 9.0. Seems like a lot more going on compared to my last chart.
_________________
| Machine: DreamStation Auto CPAP Machine |
| Mask: Amara View Full Face CPAP Mask with Headgear |
| Humidifier: DreamStation Heated Humidifier |
- Attachments
-
- screenshot-20190130-160316.png (108.82 KiB) Viewed 951 times
- Jas_williams
- Posts: 1120
- Joined: Tue Dec 19, 2017 2:12 pm
- Location: Somerset UK
Re: New To CPAP - Looking for Advice
Not really very different no more than normal change night to night but I would suggest going to 10cm
_________________
| Mask: Bleep DreamPort CPAP Mask Solution |
| Additional Comments: Using sleepyhead and a pressure of 6 - 21 Resmed S9 Adapt SV with a Bleep Sleep Mask |
Re: New To CPAP - Looking for Advice
*Points* what he said.Jas_williams wrote: ↑Wed Jan 30, 2019 4:05 pmNot really very different no more than normal change night to night but I would suggest going to 10cm
Get OSCAR
Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.
Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.
Re: New To CPAP - Looking for Advice
Thanks both of you. I figured I was overthinking it after I posted what I said.palerider wrote: ↑Wed Jan 30, 2019 4:24 pm*Points* what he said.Jas_williams wrote: ↑Wed Jan 30, 2019 4:05 pmNot really very different no more than normal change night to night but I would suggest going to 10cm
_________________
| Machine: DreamStation Auto CPAP Machine |
| Mask: Amara View Full Face CPAP Mask with Headgear |
| Humidifier: DreamStation Heated Humidifier |
Re: New To CPAP - Looking for Advice
Here's last night's data after upping the minimum pressure to 10.
_________________
| Machine: DreamStation Auto CPAP Machine |
| Mask: Amara View Full Face CPAP Mask with Headgear |
| Humidifier: DreamStation Heated Humidifier |
- Attachments
-
- screenshot-20190131-083910.png (109.41 KiB) Viewed 889 times
Re: New To CPAP - Looking for Advice
Give it a few nights, and then consider 11.
Get OSCAR
Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.
Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.
Re: New To CPAP - Looking for Advice
what palerider said. but i just want to add, look how your ahi has dropped already! good job.
_________________
| Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier |
| Mask: AirFit™ P10 Nasal Pillow CPAP Mask with Headgear |
"Age is not an accomplishment and youth is not a sin"-Robert A. Heinlein
Oscar-Win
https://www.apneaboard.com/OSCAR/OSCAR-1.5.1-Win64.exe
Oscar-Mac
https://www.apneaboard.com/OSCAR/OSCAR-1.5.1.dmg
Oscar-Win
https://www.apneaboard.com/OSCAR/OSCAR-1.5.1-Win64.exe
Oscar-Mac
https://www.apneaboard.com/OSCAR/OSCAR-1.5.1.dmg