New user.... trying to solve tiredness - graph added

General Discussion on any topic relating to CPAP and/or Sleep Apnea.
Dan_McD
Posts: 146
Joined: Sat Dec 22, 2018 1:15 pm

Re: 5 months.... Flow Limitations - graph added

Post by Dan_McD » Sat Apr 20, 2019 9:16 am

It's been exactly 5 months. Obviously AHI is not my problem. If I didn't mention earlier, I have been diagnosed with COPD years back. My question is what causes 'Flow Limitations' and do you see anything in my graphs that would indicate a problem/solution? I will have anywhere from 3 to 10 or 11 FL events in a given night and all settings will remain the same. In addition to the 15/15 apap setting, my mask is set at X1 and 'Flex' is at 3 to relieve expiration. I suspect that sometimes my mask slips down and semi restricts my nasal passages. Input would be appreciated. I apologize for the quality of the graph.

Image

_________________
Machine: DreamStation Auto CPAP Machine
Mask: Simplus Full Face CPAP Mask with Headgear
Humidifier: DreamStation Heated Humidifier
Additional Comments: Quattro FX, 1L Everflo Q Concentrator, OSCAR Software - Q-Mirage as back-up

User avatar
Julie
Posts: 20052
Joined: Tue Feb 28, 2006 12:58 pm

Re: New user.... trying to solve tiredness - graph added

Post by Julie » Sat Apr 20, 2019 10:39 am

If you upload to Imgur.com you'll do better at posting graphs... see 1st Announcement main pg.

User avatar
palerider
Posts: 32299
Joined: Wed Dec 16, 2009 5:43 pm
Location: Dallas(ish).

Re: 5 months.... Flow Limitations - graph added

Post by palerider » Sat Apr 20, 2019 3:06 pm

Dan_McD wrote:
Sat Apr 20, 2019 9:16 am
It's been exactly 5 months. Obviously AHI is not my problem. If I didn't mention earlier, I have been diagnosed with COPD years back. My question is what causes 'Flow Limitations' and do you see anything in my graphs that would indicate a problem/solution? I will have anywhere from 3 to 10 or 11 FL events in a given night and all settings will remain the same. In addition to the 15/15 apap setting, my mask is set at X1 and 'Flex' is at 3 to relieve expiration. I suspect that sometimes my mask slips down and semi restricts my nasal passages. Input would be appreciated. I apologize for the quality of the graph.

Image
There's really not much on that graph that tells me anything. if you could review the guidelines in my sig on what is helpful with sleepyhead before your next graph posting, that would be preferable.

Flow limitations are what happens when you've got a narrowing of your airway, and it makes it harder to breathe in, like breathing through a straw, you can still get a full breath (so you're not having a hypopnea), but it takes more work to get that breath, and takes longer.
Image

Here's a video about different sleep disordered breathing issues:
https://www.youtube.com/watch?v=-gie2dhqP2c

_________________
Mask: Bleep DreamPort CPAP Mask Solution
Additional Comments: S9 VPAP Auto
Get OSCAR

Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.

Dan_McD
Posts: 146
Joined: Sat Dec 22, 2018 1:15 pm

Re: 5 months.... Flow Limitations - graph added

Post by Dan_McD » Sun Apr 21, 2019 12:38 pm

palerider wrote:
Sat Apr 20, 2019 3:06 pm
Dan_McD wrote:
Sat Apr 20, 2019 9:16 am
It's been exactly 5 months. Obviously AHI is not my problem. If I didn't mention earlier, I have been diagnosed with COPD years back. My question is what causes 'Flow Limitations' and do you see anything in my graphs that would indicate a problem/solution? I will have anywhere from 3 to 10 or 11 FL events in a given night and all settings will remain the same. In addition to the 15/15 apap setting, my mask is set at X1 and 'Flex' is at 3 to relieve expiration. I suspect that sometimes my mask slips down and semi restricts my nasal passages. Input would be appreciated. I apologize for the quality of the graph.

Image
There's really not much on that graph that tells me anything. if you could review the guidelines in my sig on what is helpful with sleepyhead before your next graph posting, that would be preferable.








Flow limitations are what happens when you've got a narrowing of your airway, and it makes it harder to breathe in, like breathing through a straw, you can still get a full breath (so you're not having a hypopnea), but it takes more work to get that breath, and takes longer.
Image

Here's a video about different sleep disordered breathing issues:
https://www.youtube.com/watch?v=-gie2dhqP2c
I'm working on it.

_________________
Machine: DreamStation Auto CPAP Machine
Mask: Simplus Full Face CPAP Mask with Headgear
Humidifier: DreamStation Heated Humidifier
Additional Comments: Quattro FX, 1L Everflo Q Concentrator, OSCAR Software - Q-Mirage as back-up

Dan_McD
Posts: 146
Joined: Sat Dec 22, 2018 1:15 pm

Re: New user.... trying to solve tiredness - graph added

Post by Dan_McD » Mon Apr 22, 2019 8:37 am

screenshot-20190422-092956.png
screenshot-20190422-092956.png (126.82 KiB) Viewed 3611 times


palerider: I don't know if this is any better. This is last night. More mask leak than the last week and a restless night

_________________
Machine: DreamStation Auto CPAP Machine
Mask: Simplus Full Face CPAP Mask with Headgear
Humidifier: DreamStation Heated Humidifier
Additional Comments: Quattro FX, 1L Everflo Q Concentrator, OSCAR Software - Q-Mirage as back-up

User avatar
Pugsy
Posts: 65131
Joined: Thu May 14, 2009 9:31 am
Location: Missouri, USA

Re: New user.... trying to solve tiredness - graph added

Post by Pugsy » Mon Apr 22, 2019 8:47 am

In your effort to include so many graphs you make the graphs that might help someone so tiny it's hard to evaluate.

Please see this thread for graph format.
viewtopic/t158560/How-to-post-images-for-review.html
Omit the respiration rate and time at pressure graph.
Make your other graphs larger so easier to read.

_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier
Additional Comments: Mask Bleep Eclipse https://bleepsleep.com/the-eclipse/
I may have to RISE but I refuse to SHINE.

User avatar
katestyles
Posts: 610
Joined: Sun Jan 06, 2019 9:08 am

Re: New user.... trying to solve tiredness - graph added

Post by katestyles » Mon Apr 22, 2019 8:48 am

That graph shows the information that the experts normally ask for - and I would love to have your AHI! you have the unicorn score!

You are asking about flow limitations - can you take another screenshot with the Flow limitation graph showing? I suggest you drop the pressure graph, since it is not changing.

_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier
Mask: Bleep DreamPort CPAP Mask Solution
Additional Comments: Back up mask - anything in the drawer

User avatar
Pugsy
Posts: 65131
Joined: Thu May 14, 2009 9:31 am
Location: Missouri, USA

Re: New user.... trying to solve tiredness - graph added

Post by Pugsy » Mon Apr 22, 2019 8:52 am

katestyles wrote:
Mon Apr 22, 2019 8:48 am
can you take another screenshot with the Flow limitation graph showing?
Respironics machines don't do flow limitation graphs. Instead they just flag FLs up on the events graph and treat them like the other flagged stuff.
The only real useful graphs from a Respironics machine are
Events
Flow rate
Pressure....if it varies any
Leak

_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier
Additional Comments: Mask Bleep Eclipse https://bleepsleep.com/the-eclipse/
I may have to RISE but I refuse to SHINE.

Dan_McD
Posts: 146
Joined: Sat Dec 22, 2018 1:15 pm

Re: New user.... trying to solve tiredness - graph added

Post by Dan_McD » Mon Apr 22, 2019 8:53 am

This is from last Thursday. A lot of FL and no mask leak. Very few other events.
*
*
screenshot-20190422-094533.png
screenshot-20190422-094533.png (121.83 KiB) Viewed 3606 times

_________________
Machine: DreamStation Auto CPAP Machine
Mask: Simplus Full Face CPAP Mask with Headgear
Humidifier: DreamStation Heated Humidifier
Additional Comments: Quattro FX, 1L Everflo Q Concentrator, OSCAR Software - Q-Mirage as back-up

Dan_McD
Posts: 146
Joined: Sat Dec 22, 2018 1:15 pm

Re: New user.... trying to solve tiredness - graph added

Post by Dan_McD » Mon Apr 22, 2019 8:56 am

Pugsy wrote:
Mon Apr 22, 2019 8:47 am
In your effort to include so many graphs you make the graphs that might help someone so tiny it's hard to evaluate.

Please see this thread for graph format.
viewtopic/t158560/How-to-post-images-for-review.html
Omit the respiration rate and time at pressure graph.
Make your other graphs larger so easier to read.
My second and 3rd graph is what palerider asked for. I can post whatever you want to see???

_________________
Machine: DreamStation Auto CPAP Machine
Mask: Simplus Full Face CPAP Mask with Headgear
Humidifier: DreamStation Heated Humidifier
Additional Comments: Quattro FX, 1L Everflo Q Concentrator, OSCAR Software - Q-Mirage as back-up

User avatar
katestyles
Posts: 610
Joined: Sun Jan 06, 2019 9:08 am

Re: New user.... trying to solve tiredness - graph added

Post by katestyles » Mon Apr 22, 2019 8:59 am

Pugsy wrote:
Mon Apr 22, 2019 8:52 am
katestyles wrote:
Mon Apr 22, 2019 8:48 am
can you take another screenshot with the Flow limitation graph showing?
Respironics machines don't do flow limitation graphs. Instead they just flag FLs up on the events graph and treat them like the other flagged stuff.
The only real useful graphs from a Respironics machine are
Events
Flow rate
Pressure....if it varies any
Leak
Oops - My bad!

_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier
Mask: Bleep DreamPort CPAP Mask Solution
Additional Comments: Back up mask - anything in the drawer

User avatar
Pugsy
Posts: 65131
Joined: Thu May 14, 2009 9:31 am
Location: Missouri, USA

Re: New user.... trying to solve tiredness - graph added

Post by Pugsy » Mon Apr 22, 2019 8:59 am

Actually what he said was to review the guidelines in his signature and I am pretty sure respiration rate and time at pressure aren't mentioned. I had a hand in those guidelines.

_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier
Additional Comments: Mask Bleep Eclipse https://bleepsleep.com/the-eclipse/
I may have to RISE but I refuse to SHINE.

Dan_McD
Posts: 146
Joined: Sat Dec 22, 2018 1:15 pm

Re: New user.... trying to solve tiredness - graph added

Post by Dan_McD » Mon Apr 22, 2019 9:03 am

Pugsy wrote:
Mon Apr 22, 2019 8:47 am
In your effort to include so many graphs you make the graphs that might help someone so tiny it's hard to evaluate.

Please see this thread for graph format.
viewtopic/t158560/How-to-post-images-for-review.html
Omit the respiration rate and time at pressure graph.
Make your other graphs larger so easier to read.
As you asked.....
*
*
screenshot-20190422-100054.png
screenshot-20190422-100054.png (112.63 KiB) Viewed 3595 times

_________________
Machine: DreamStation Auto CPAP Machine
Mask: Simplus Full Face CPAP Mask with Headgear
Humidifier: DreamStation Heated Humidifier
Additional Comments: Quattro FX, 1L Everflo Q Concentrator, OSCAR Software - Q-Mirage as back-up

Dan_McD
Posts: 146
Joined: Sat Dec 22, 2018 1:15 pm

Re: New user.... trying to solve tiredness - graph added

Post by Dan_McD » Mon Apr 22, 2019 9:19 am

One of the reasons I included the 'Respiration Rate', Inspiration, Expiration,. I thought I/We could find the reason for the 'Flow Limitation' number of events. But, whatever.

_________________
Machine: DreamStation Auto CPAP Machine
Mask: Simplus Full Face CPAP Mask with Headgear
Humidifier: DreamStation Heated Humidifier
Additional Comments: Quattro FX, 1L Everflo Q Concentrator, OSCAR Software - Q-Mirage as back-up

User avatar
Pugsy
Posts: 65131
Joined: Thu May 14, 2009 9:31 am
Location: Missouri, USA

Re: New user.... trying to solve tiredness - graph added

Post by Pugsy » Mon Apr 22, 2019 9:21 am

Fragmented sleep for any reason will make us feel like crap during the day.
I see at least 2 breaks in therapy where we know you were awake and turned the machine off and back on again.
How many more times do you remember being awakened? Any idea?
If very many....any idea why?

I know you are looking for an answer in your graphs for the fatigue but to be honest...I don't see it here.
The FLs...a maybe if they are in clusters but a weak maybe IMHO.
I don't think that tweaks to the machine settings are your answer here. I think that unfortunately you have to look elsewhere for the poor sleep or excessive fatigue. These machines are designed to fix one problem only....airway issues. They totally suck at fixing problems unrelated to the airway.

The leaks are well within the machine's ability to compensate for them...now if they are waking you up then that's a different story as anything (even small leaks) that disturb our sleep is unwanted.

I think maybe you should take a crash course in learning how to distinguish arousal breathing from asleep breathing and take a stroll through your flow rate and look for evidence of arousal breathing.
http://freecpapadvice.com/sleepyhead-free-software
Watch the videos as it explains how to take a stroll through the flow rate and don't just look at flow rate with events flagged...look at all of it to get an idea how much your actual sleep might be impacted by arousals.
Usually I ask people about medication side effects and other health issues....the PLMD could be playing a significant part in arousals if you are having very many.
And arousals may or may not be remembered...so don't think that just because you don't remember something that it didn't happen.

_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier
Additional Comments: Mask Bleep Eclipse https://bleepsleep.com/the-eclipse/
I may have to RISE but I refuse to SHINE.