EPR. to use or not to use?

General Discussion on any topic relating to CPAP and/or Sleep Apnea.
JBarizona
Posts: 66
Joined: Mon Oct 29, 2018 2:18 pm

EPR. to use or not to use?

Post by JBarizona » Mon Oct 29, 2018 3:04 pm

I've had my epr setting at 3 since I started treatment about 6 weeks ago.... My ahi is around 7 or 8...can the epr setting affect your AHI or is it merely just a tool for more comfortable and tolerable therapy?

User avatar
chunkyfrog
Posts: 34544
Joined: Mon Jul 12, 2010 5:10 pm
Location: Nowhere special--this year in particular.

Re: EPR. to use or not to use?

Post by chunkyfrog » Mon Oct 29, 2018 3:17 pm

Whether or not it affects therapy significantly is still open to debate, even among experts.
It is a feature intended to allow an easier transition to therapy.
Since everyone is different, you may try different settings, for a few nights at a time.
Your are free to choose what is most comfortable.
Only your results should determine your choice.

_________________
Mask: AirFit™ P10 For Her Nasal Pillow CPAP Mask with Headgear
Additional Comments: Airsense 10 Autoset for Her

User avatar
LSAT
Posts: 13399
Joined: Sun Nov 16, 2008 10:11 am
Location: SE Wisconsin

Re: EPR. to use or not to use?

Post by LSAT » Mon Oct 29, 2018 3:42 pm

It's a comfort tool. If it makes your breathing easier...use it. It has very little effect on AHI..if any. Everyone is different. I have used EPR 2 for a long time. On a few occasions I tried EPR 1 and EPR off. In both occasions , my breathing was uncomfortable and my AHI went up (From the breathing issues). If the machine listed in your profile is correct, it supplies no meaningful data. I would try increasing your pressure to see if the AHI comes down. \

NOTE...If your pressure is 8/15, you have an Autoset and the machine you listed is incorrect. If this is the case..increase your minimum pressure a bit.

Stom
Posts: 383
Joined: Sat Jul 21, 2018 9:44 am

Re: EPR. to use or not to use?

Post by Stom » Mon Oct 29, 2018 4:31 pm

Whether EPR affects your therapy may depend on what kind of apnea you have. For some people, EPR can increase treatment emergent central apneas.
Official Foe™ on @Palerider's public list of foes. ;-)

Rules to live by: Be wary of anyone boasting their behavior is like that of a fictional sociopath.

JBarizona
Posts: 66
Joined: Mon Oct 29, 2018 2:18 pm

Re: EPR. to use or not to use?

Post by JBarizona » Mon Oct 29, 2018 6:03 pm

LSAT wrote:
Mon Oct 29, 2018 3:42 pm
It's a comfort tool. If it makes your breathing easier...use it. It has very little effect on AHI..if any. Everyone is different. I have used EPR 2 for a long time. On a few occasions I tried EPR 1 and EPR off. In both occasions , my breathing was uncomfortable and my AHI went up (From the breathing issues). If the machine listed in your profile is correct, it supplies no meaningful data. I would try increasing your pressure to see if the AHI comes down. \

NOTE...If your pressure is 8/15, you have an Autoset and the machine you listed is incorrect. If this is the case..increase your minimum pressure a bit.
I'm not sure about Auto sets or type of machine. It's a ResMed airsense 10 set to APAP... So does that mean I have an autoset machine like you said? Also tell me why you think raising the lower pressure would be helpful as opposed to raising the higher pressure..

User avatar
Pugsy
Posts: 65294
Joined: Thu May 14, 2009 9:31 am
Location: Missouri, USA

Re: EPR. to use or not to use?

Post by Pugsy » Mon Oct 29, 2018 6:07 pm

I wouldn't go increasing any pressure until I knew exactly what that AHI is composed of if it were me.

_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier
Additional Comments: Mask Bleep Eclipse https://bleepsleep.com/the-eclipse/
I may have to RISE but I refuse to SHINE.

JBarizona
Posts: 66
Joined: Mon Oct 29, 2018 2:18 pm

Re: EPR. to use or not to use?

Post by JBarizona » Mon Oct 29, 2018 6:09 pm

Stom wrote:
Mon Oct 29, 2018 4:31 pm
Whether EPR affects your therapy may depend on what kind of apnea you have. For some people, EPR can increase treatment emergent central apneas.
I've definitely had an emerging Central apneas but I've heard that this is common and overtime dissipates... I do have obstructive sleep apnea...

JBarizona
Posts: 66
Joined: Mon Oct 29, 2018 2:18 pm

Re: EPR. to use or not to use?

Post by JBarizona » Mon Oct 29, 2018 6:11 pm

Pugsy wrote:
Mon Oct 29, 2018 6:07 pm
I wouldn't go increasing any pressure until I knew exactly what that AHI is composed of if it were me.
Typically my AHi is around 7... Approximately three out of seven obstructive three out of Seven Central and 1 out of 7 hypopnea

User avatar
Pugsy
Posts: 65294
Joined: Thu May 14, 2009 9:31 am
Location: Missouri, USA

Re: EPR. to use or not to use?

Post by Pugsy » Mon Oct 29, 2018 6:14 pm

More pressure won't fix the centrals.
Small chance more pressure might make the centrals worse.

Before advising anything I prefer to see the actual detailed graphs so I know exactly what is going on.
Need to use the software to see the details.
For all we know the majority of the flagged events are right smack after a known awake event and not even real.

_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier
Additional Comments: Mask Bleep Eclipse https://bleepsleep.com/the-eclipse/
I may have to RISE but I refuse to SHINE.

JBarizona
Posts: 66
Joined: Mon Oct 29, 2018 2:18 pm

Re: EPR. to use or not to use?

Post by JBarizona » Mon Oct 29, 2018 6:29 pm

Pugsy wrote:
Mon Oct 29, 2018 6:14 pm
More pressure won't fix the centrals.
Small chance more pressure might make the centrals worse.

Before advising anything I prefer to see the actual detailed graphs so I know exactly what is going on.
Need to use the software to see the details.
For all we know the majority of the flagged events are right smack after a known awake event and not even real.
Yes I'm trying to fix the obstructive not the centrals...I will do my best to post a few nights of sleepyhead day to here shortly and appreciate any feedback

User avatar
palerider
Posts: 32299
Joined: Wed Dec 16, 2009 5:43 pm
Location: Dallas(ish).

Re: EPR. to use or not to use?

Post by palerider » Mon Oct 29, 2018 6:40 pm

JBarizona wrote:
Mon Oct 29, 2018 3:04 pm
I've had my epr setting at 3 since I started treatment about 6 weeks ago.... My ahi is around 7 or 8...can the epr setting affect your AHI or is it merely just a tool for more comfortable and tolerable therapy?
Yes, it absolutely can, because EPR drops the pressure, and if it drops it too much, you'll have obstructive events.

EPR also increases ventilation, and if you're sensitive to that, it can cause central events.

Post some charts.

_________________
Mask: Bleep DreamPort CPAP Mask Solution
Additional Comments: S9 VPAP Auto
Get OSCAR

Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.

JBarizona
Posts: 66
Joined: Mon Oct 29, 2018 2:18 pm

Re: EPR. to use or not to use?

Post by JBarizona » Mon Oct 29, 2018 6:45 pm

palerider wrote:
Mon Oct 29, 2018 6:40 pm
JBarizona wrote:
Mon Oct 29, 2018 3:04 pm
I've had my epr setting at 3 since I started treatment about 6 weeks ago.... My ahi is around 7 or 8...can the epr setting affect your AHI or is it merely just a tool for more comfortable and tolerable therapy?
Yes, it absolutely can, because EPR drops the pressure, and if it drops it too much, you'll have obstructive events.

EPR also increases ventilation, and if you're sensitive to that, it can cause central events.

Post some charts.
I'm having trouble figuring out how to get the charts posted

JBarizona
Posts: 66
Joined: Mon Oct 29, 2018 2:18 pm

Re: EPR. to use or not to use?

Post by JBarizona » Mon Oct 29, 2018 6:47 pm

palerider wrote:
Mon Oct 29, 2018 6:40 pm
JBarizona wrote:
Mon Oct 29, 2018 3:04 pm
I've had my epr setting at 3 since I started treatment about 6 weeks ago.... My ahi is around 7 or 8...can the epr setting affect your AHI or is it merely just a tool for more comfortable and tolerable therapy?
Yes, it absolutely can, because EPR drops the pressure, and if it drops it too much, you'll have obstructive events.

EPR also increases ventilation, and if you
're sensitive to that, it can cause central events.

Post some charts.
okay this is strange I toggled my keyboard in the sleepyhead software and now I can't seem to close out of the sleepyhead software or do anything in it. It's like a jammed up my computer or something anybody going to know what's going on?

User avatar
palerider
Posts: 32299
Joined: Wed Dec 16, 2009 5:43 pm
Location: Dallas(ish).

Re: EPR. to use or not to use?

Post by palerider » Mon Oct 29, 2018 6:49 pm

JBarizona wrote:
Mon Oct 29, 2018 6:47 pm
palerider wrote:
Mon Oct 29, 2018 6:40 pm
JBarizona wrote:
Mon Oct 29, 2018 3:04 pm
I've had my epr setting at 3 since I started treatment about 6 weeks ago.... My ahi is around 7 or 8...can the epr setting affect your AHI or is it merely just a tool for more comfortable and tolerable therapy?
Yes, it absolutely can, because EPR drops the pressure, and if it drops it too much, you'll have obstructive events.

EPR also increases ventilation, and if you
're sensitive to that, it can cause central events.

Post some charts.
okay this is strange I toggled my keyboard in the sleepyhead software and now I can't seem to close out of the sleepyhead software or do anything in it. It's like a jammed up my computer or something anybody going to know what's going on?
Try hitting F11

_________________
Mask: Bleep DreamPort CPAP Mask Solution
Additional Comments: S9 VPAP Auto
Get OSCAR

Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.

JBarizona
Posts: 66
Joined: Mon Oct 29, 2018 2:18 pm

Re: EPR. to use or not to use?

Post by JBarizona » Mon Oct 29, 2018 6:53 pm

palerider wrote:
Mon Oct 29, 2018 6:49 pm
JBarizona wrote:
Mon Oct 29, 2018 6:47 pm
palerider wrote:
Mon Oct 29, 2018 6:40 pm
JBarizona wrote:
Mon Oct 29, 2018 3:04 pm
I've had my epr setting at 3 since I started treatment about 6 weeks ago.... My ahi is around 7 or 8...can the epr setting affect your AHI or is it merely just a tool for more comfortable and tolerable therapy?
Yes, it absolutely can, because EPR drops the pressure, and if it drops it too much, you'll have obstructive events.

EPR also increases ventilation, and if you
're sensitive to that, it can cause central events.

Post some charts.
okay this is strange I toggled my keyboard in the sleepyhead software and now I can't seem to close out of the sleepyhead software or do anything in it. It's like a jammed up my computer or something anybody going to know what's going on?
Try hitting F11
That didn't work. I can navigate on the charts but up at the top where there's file and View and data I can't do anything with it and I can't close down the program without restarting my computer and then when I restart it again the same thing happens. This is messed up