Can't fall asleep with cpap anymore. Please help.
Can't fall asleep with cpap anymore. Please help.
Hey.
In the last few weeks i have been experiencing a difficulty in falling asleep with my cpap device. I basically spend 15-50 minutes on bed and never get past the initial sleep phase. When i go to my cpap device it always shows apnea ratings such as 9-20 apneas per hour. When i go to my pc to have my sd card read by sleepyhead it always show that all those apneas are Clear Airways (followed by RERAs sometimes) . I already use the lowest pressure setting (4) and don't know what to do anymore. I've never had a very severe apnea, and recently i've been wondering if the Cpap device is a bit "overkill" for my case. The thing is that the nature of my apnea is a bit ellusive and may be related to muscle weakness rather than airway collapse (unconfirmed).
Anyways, can someone show a light on how to stop these Clear Airways apneas from happening? There are days that they are completely absent and i can initiate sleep fine, but the days that the opposite happen i end up sleeping 3-4 AM and can't function properly in the other day. I am afraid of not sleeping with Cpap and end up being worse, but not being able to fall asleep because i feel suffocated because of those Clear airways apneas is also very awful.
Device-Resmed S10 (with that heating tube)
Mask-Airfit P10 Resmed
Pressure: 4 or 5
In the last few weeks i have been experiencing a difficulty in falling asleep with my cpap device. I basically spend 15-50 minutes on bed and never get past the initial sleep phase. When i go to my cpap device it always shows apnea ratings such as 9-20 apneas per hour. When i go to my pc to have my sd card read by sleepyhead it always show that all those apneas are Clear Airways (followed by RERAs sometimes) . I already use the lowest pressure setting (4) and don't know what to do anymore. I've never had a very severe apnea, and recently i've been wondering if the Cpap device is a bit "overkill" for my case. The thing is that the nature of my apnea is a bit ellusive and may be related to muscle weakness rather than airway collapse (unconfirmed).
Anyways, can someone show a light on how to stop these Clear Airways apneas from happening? There are days that they are completely absent and i can initiate sleep fine, but the days that the opposite happen i end up sleeping 3-4 AM and can't function properly in the other day. I am afraid of not sleeping with Cpap and end up being worse, but not being able to fall asleep because i feel suffocated because of those Clear airways apneas is also very awful.
Device-Resmed S10 (with that heating tube)
Mask-Airfit P10 Resmed
Pressure: 4 or 5
- Attachments
-
- apnea3.png (215.15 KiB) Viewed 2885 times
-
- apnea2.png (213.53 KiB) Viewed 2885 times
-
- apnea1.png (221.57 KiB) Viewed 2885 times
- ChicagoGranny
- Posts: 15183
- Joined: Sun Jan 29, 2012 1:43 pm
- Location: USA
Re: Can't fall asleep with cpap anymore. Please help.
What medications do you take? How is your sleep hygiene, diet and exercise? When and how much caffeine do you consume? Don't overlook smaller amounts like in chocolate.
Re: Can't fall asleep with cpap anymore. Please help.
Hello.ChicagoGranny wrote: ↑Thu Oct 18, 2018 11:12 amWhat medications do you take? How is your sleep hygiene, diet and exercise? When and how much caffeine do you consume? Don't overlook smaller amounts like in chocolate.
I take levothyroxine and a few supplements for exercises. My diet is fine, i try to eat as healthy as possible. No caffeine. During late afternoon and night i rarely ever eat anything with sugar or chocolate.
The thing is, i actually feel suffocated when i try to sleep.
Here is last night:
After trying to sleep 3-4 times i gave up and slept without the cpap.
- Attachments
-
- apnea4.png (198.09 KiB) Viewed 2848 times
- Jas_williams
- Posts: 1120
- Joined: Tue Dec 19, 2017 2:12 pm
- Location: Somerset UK
Re: Can't fall asleep with cpap anymore. Please help.
Somebody has been messing Your machine is not providing enough air pressure it is currently in CPAP mode with a fixed pressure of 4
It’s no wonder you feel starved of air/suffocated
What was your prescription I see in an earlier graph the machine was set to a fixed pressure of 5
I suggest setting it to apap mode with a minimum pressure of 6 and a max pressure of 20 and sleep with it for a night and let’s see what happens. Post up a graph when you’ve done that.
You can’t count the CA’s before you go to sleep as those are not true CA’s as our awake breathing is just too irregular
It’s no wonder you feel starved of air/suffocated
What was your prescription I see in an earlier graph the machine was set to a fixed pressure of 5
I suggest setting it to apap mode with a minimum pressure of 6 and a max pressure of 20 and sleep with it for a night and let’s see what happens. Post up a graph when you’ve done that.
You can’t count the CA’s before you go to sleep as those are not true CA’s as our awake breathing is just too irregular
_________________
Mask: Bleep DreamPort CPAP Mask Solution |
Additional Comments: Using sleepyhead and a pressure of 6 - 21 Resmed S9 Adapt SV with a Bleep Sleep Mask |
Re: Can't fall asleep with cpap anymore. Please help.
I agree...trying to breathe with a pressure of 4 is ridiculous. Change your setting to Auto with a min pressure of 6 and max pressure of 12 and see what happens. The machine can't stop the CAs from happening, but with a higher pressure you may be able to fall asleep faster which will eliminate most of them.
_________________
Machine: ResMed AirSense™ 10 AutoSet™ CPAP Machine with HumidAir™ Heated Humidifier |
Mask: Fisher & Paykel Vitera Full Face Mask with Headgear (S, M, or L Cushion) |
Additional Comments: Back up is a new AS10. |
Re: Can't fall asleep with cpap anymore. Please help.
I am the one that changed the settings lol. I used to have a automatic pressure, but because the device would always put me in the highest pressure i would wake up bloated and with stomach pain so i decided to put a fixed pressure and slowly reduce the pressure until i found the most adequate setting for me.
Turns out that no matter how i changed the settings the ahi would stay the same so i decided to lower to the lowest setting (4). Another reason for changing the pressure to a lower setting is that when i did a sleep exam with a Cpap it would also show a very high number of central apneas that would be possibly linked to the cpap use (60% of them were central apneas).
Also, my apologies if i can't bring any more relevant info, i am struggling to keep myself awake at the moment.
Turns out that no matter how i changed the settings the ahi would stay the same so i decided to lower to the lowest setting (4). Another reason for changing the pressure to a lower setting is that when i did a sleep exam with a Cpap it would also show a very high number of central apneas that would be possibly linked to the cpap use (60% of them were central apneas).
Also, my apologies if i can't bring any more relevant info, i am struggling to keep myself awake at the moment.
- zoocrewphoto
- Posts: 3732
- Joined: Mon Apr 30, 2012 10:34 pm
- Location: Seatac, WA
Re: Can't fall asleep with cpap anymore. Please help.
The lower pressure is why you are currently feeling suffocated. You should raise it to at least 6 and see if you can get to sleep better. The centrals you are having on those sleepless nights are probably sleep/wake/junk as you toss and turn. They are all at the beginning and it is normal if we are not sleeping.user20398 wrote: ↑Thu Oct 18, 2018 11:29 pmI am the one that changed the settings lol. I used to have a automatic pressure, but because the device would always put me in the highest pressure i would wake up bloated and with stomach pain so i decided to put a fixed pressure and slowly reduce the pressure until i found the most adequate setting for me.
Turns out that no matter how i changed the settings the ahi would stay the same so i decided to lower to the lowest setting (4). Another reason for changing the pressure to a lower setting is that when i did a sleep exam with a Cpap it would also show a very high number of central apneas that would be possibly linked to the cpap use (60% of them were central apneas).
Also, my apologies if i can't bring any more relevant info, i am struggling to keep myself awake at the moment.
What was your original pressure range? Perhaps starting with a smaller range that doesn't go as high would help you sleep better and see what pressures you actually need. Start at 6 and maybe go up to 9 or 10 and see if you have any issues.
_________________
Mask: Quattro™ FX Full Face CPAP Mask with Headgear |
Humidifier: S9™ Series H5i™ Heated Humidifier with Climate Control |
Additional Comments: Resmed S9 autoset pressure range 11-17 |
Who would have thought it would be this challenging to sleep and breathe at the same time?
- Midwest_non_sleeper
- Posts: 436
- Joined: Wed Jun 20, 2018 1:03 pm
Re: Can't fall asleep with cpap anymore. Please help.
Those clear airway events look like sleep-onset events to me, or just sleep-wake-junk from being roused from sleep by something. I get a couple of minor batches of them per night usually.
Put your machine in Auto mode (APAP) and set the minimum to 6 or 7 and the max to 15 or 20. 4cmH20 is way too low for most people to feel like they're getting enough airflow. That could be what's preventing you from getting to sleep or waking up once you're finally exhausted enough to sleep.
I could be wrong, as I often am - and readily admit it.
Put your machine in Auto mode (APAP) and set the minimum to 6 or 7 and the max to 15 or 20. 4cmH20 is way too low for most people to feel like they're getting enough airflow. That could be what's preventing you from getting to sleep or waking up once you're finally exhausted enough to sleep.
I could be wrong, as I often am - and readily admit it.
_________________
Machine: ResMed AirSense™ 10 AutoSet™ CPAP Machine with HumidAir™ Heated Humidifier |
Mask: DreamWear Gel Nasal Pillow CPAP Mask with Headgear - Fit Pack (All Nasal Pillows with Medium Frame) |
Additional Comments: 9.2cmH20 / 20cmH20 pressure |
Re: Can't fall asleep with cpap anymore. Please help.
I was able to sleep with an automatic pressure of 5-8. First time in a while that i was able to get a decent night of sleep. Still a bit sleepy, though lol.
Also, thanks everyone here for the help. You people are very kind.
Also, thanks everyone here for the help. You people are very kind.

- Midwest_non_sleeper
- Posts: 436
- Joined: Wed Jun 20, 2018 1:03 pm
Re: Can't fall asleep with cpap anymore. Please help.
Do you mean 5cmH20 minimum and 8cmH20 maximum? Or between 5-8cmH20 minimum only?
_________________
Machine: ResMed AirSense™ 10 AutoSet™ CPAP Machine with HumidAir™ Heated Humidifier |
Mask: DreamWear Gel Nasal Pillow CPAP Mask with Headgear - Fit Pack (All Nasal Pillows with Medium Frame) |
Additional Comments: 9.2cmH20 / 20cmH20 pressure |
- Jas_williams
- Posts: 1120
- Joined: Tue Dec 19, 2017 2:12 pm
- Location: Somerset UK
Re: Can't fall asleep with cpap anymore. Please help.
Post a graph and let’s see how it looks, max pressure of 8 is quite low but if you did not get any aerophagia then great.. hopefully you will feel better with a few nights of sleep
_________________
Mask: Bleep DreamPort CPAP Mask Solution |
Additional Comments: Using sleepyhead and a pressure of 6 - 21 Resmed S9 Adapt SV with a Bleep Sleep Mask |
Re: Can't fall asleep with cpap anymore. Please help.
5cmH2O minimum and 8cmH2O maximum.Midwest_non_sleeper wrote: ↑Fri Oct 19, 2018 8:50 amDo you mean 5cmH20 minimum and 8cmH20 maximum? Or between 5-8cmH20 minimum only?
*7cmH2O maximum to be exact.
Re: Can't fall asleep with cpap anymore. Please help.
Here it is.Jas_williams wrote: ↑Fri Oct 19, 2018 10:58 amPost a graph and let’s see how it looks, max pressure of 8 is quite low but if you did not get any aerophagia then great.. hopefully you will feel better with a few nights of sleep
It was actually a max pressure of 7 not 8

- Attachments
-
- apnea5.png (218.16 KiB) Viewed 2693 times
Re: Can't fall asleep with cpap anymore. Please help.
And that's why your AHI is higher than it needs to be.
Get OSCAR
Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.
Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.
- Jas_williams
- Posts: 1120
- Joined: Tue Dec 19, 2017 2:12 pm
- Location: Somerset UK
Re: Can't fall asleep with cpap anymore. Please help.
Can you remove the mask pressure graph and add in the flow limitation graph, also I would increase the max pressure another 1cm to 8 to help treat the Hypopnoea
_________________
Mask: Bleep DreamPort CPAP Mask Solution |
Additional Comments: Using sleepyhead and a pressure of 6 - 21 Resmed S9 Adapt SV with a Bleep Sleep Mask |