Newbie Self Diagnosis - In Denial - AHI Question

General Discussion on any topic relating to CPAP and/or Sleep Apnea.
marktheshark
Posts: 92
Joined: Mon Sep 17, 2018 10:01 am

Re: Newbie Self Diagnosis - In Denial - AHI Question

Post by marktheshark » Tue Mar 12, 2019 10:33 am

metsfan302 wrote: ↑
Fri Mar 08, 2019 10:38 am
How did ya do with the new machine? Hopefully a good sleep, might take a bit to dial it in, but boy does it really work for you and with you vs other brands n types.
Still dealing with this head cold/sinus infection, so not yet!

marktheshark
Posts: 92
Joined: Mon Sep 17, 2018 10:01 am

Re: Newbie Self Diagnosis - In Denial - AHI Question

Post by marktheshark » Tue Apr 02, 2019 2:57 pm

Howdy folks, I'm back.

The new bi-level machine is SO much more comfortable to fall asleep with and overall feels so much more natural.

But I feel I'm still dealing with the same problem, making up a lot. You can see how many times I restarted the machine.

The one thing I did notice, is the flow limitations seem different, before it was widespread for the whole night, now it seems like there are higher peaks but lower valleys?

I know you all mentioned go to a higher minimum in the beginning of this journey, wondering if I should do that with this new machine, if so can you explain what settings to change?
Attachments
First bi-level sleepyhead data.PNG
First bi-level sleepyhead data.PNG (106.87 KiB) Viewed 1607 times

User avatar
metsfan302
Posts: 687
Joined: Fri Oct 09, 2015 9:44 pm
Location: Dewey Beach, Delaware

Re: Newbie Self Diagnosis - In Denial - AHI Question

Post by metsfan302 » Tue Apr 02, 2019 3:10 pm

I wish my screenshots looked like that! So glad it seems to be working out great for ya and yes the ResMed sure does feels so natural vs fighting against the machine as it seemed to be the case the few times I’ve been on the Philips. I have no knowledgeable input as far as tweaking it further yet. The machine seems much happier with you on the other end lol :lol:

_________________
Mask: AirFit™ P10 Nasal Pillow CPAP Mask with Headgear
Humidifier: DreamStation Heated Humidifier
Additional Comments: Prior - ResMed AirCurve ASV User
This is my Orig Post "The start of my venture to find proper therapy" - http://www.cpaptalk.com/viewtopic.php?f ... &sk=t&sd=a

User avatar
palerider
Posts: 32308
Joined: Wed Dec 16, 2009 5:43 pm
Location: Dallas(ish).

Re: Newbie Self Diagnosis - In Denial - AHI Question

Post by palerider » Tue Apr 02, 2019 3:12 pm

Are you having problems getting back to sleep?

Are you having problems dealing with the pressure?

I see you've got the ramp set for 45 minutes, and starting really low, with all the times you're restarting the machine, that's a lot of the night you're going without therapy.

_________________
Mask: Bleep DreamPort CPAP Mask Solution
Additional Comments: S9 VPAP Auto
Get OSCAR

Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.

marktheshark
Posts: 92
Joined: Mon Sep 17, 2018 10:01 am

Re: Newbie Self Diagnosis - In Denial - AHI Question

Post by marktheshark » Tue Apr 02, 2019 3:18 pm

palerider wrote: ↑
Tue Apr 02, 2019 3:12 pm
Are you having problems getting back to sleep?

Are you having problems dealing with the pressure?

I see you've got the ramp set for 45 minutes, and starting really low, with all the times you're restarting the machine, that's a lot of the night you're going without therapy.
I don't think I'm having trouble getting back to sleep.

I think I'm waking up and the with the higher pressure I can't fall back asleep unless I restart the machine.

I thought I had the ramp at 20 minutes, it says that on the machine itself, is that wrong?

User avatar
palerider
Posts: 32308
Joined: Wed Dec 16, 2009 5:43 pm
Location: Dallas(ish).

Re: Newbie Self Diagnosis - In Denial - AHI Question

Post by palerider » Tue Apr 02, 2019 6:11 pm

marktheshark wrote: ↑
Tue Apr 02, 2019 3:18 pm
palerider wrote: ↑
Tue Apr 02, 2019 3:12 pm
Are you having problems getting back to sleep?

Are you having problems dealing with the pressure?

I see you've got the ramp set for 45 minutes, and starting really low, with all the times you're restarting the machine, that's a lot of the night you're going without therapy.
I don't think I'm having trouble getting back to sleep.

I think I'm waking up and the with the higher pressure I can't fall back asleep unless I restart the machine.

I thought I had the ramp at 20 minutes, it says that on the machine itself, is that wrong?
No, you're right, I screwed up reading the numbers. but still, that's over three hours that you aren't getting treatment during that night you pasted... see where it's not doing anything till it ramps up? If you can handle a shorter ramp, it might be a good idea.

_________________
Mask: Bleep DreamPort CPAP Mask Solution
Additional Comments: S9 VPAP Auto
Get OSCAR

Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.

marktheshark
Posts: 92
Joined: Mon Sep 17, 2018 10:01 am

Re: Newbie Self Diagnosis - In Denial - AHI Question

Post by marktheshark » Mon Apr 22, 2019 7:02 pm

Any other recommendations?

It seems every night looks the same with too many wake ups.

I did see an ENT last week who said I need sinus surgery! Guess I have a deviated septum and other obstructions that I can't remember what he called them. So, not sure how that plays into this. I went to see him because I finally kicked the afrin nose spray 3 months ago and now I am back to being blocked up every night.

I know for certain now that if I can sleep the first 4 hours of the night, I can really tell the next day. I am now wondering if my headgear is the problem, nasal irritation even with Lansinoh cream and head pressure from the chin strap. Its definitely better with the cream, but still wake up in pain.

User avatar
Miss Emerita
Posts: 3783
Joined: Sun Nov 04, 2018 8:07 pm

Re: Newbie Self Diagnosis - In Denial - AHI Question

Post by Miss Emerita » Mon Apr 22, 2019 9:11 pm

Interesting news from the ENT! I will be very curious whether the nasal surgery will resolve your FL problem. Please let us know!

About discomfort from the chin strap: did you ever wind up trying taping? I can recommend Somnifix (though it is pricey).

And abour your nasal discomfort: have you tried a nasal mask (goes over the bottom part of your nose but doesn’t touch those sensitive ends of your nostrils)?
Oscar software is available at https://www.sleepfiles.com/OSCAR/

marktheshark
Posts: 92
Joined: Mon Sep 17, 2018 10:01 am

Re: Newbie Self Diagnosis - In Denial - AHI Question

Post by marktheshark » Thu Apr 25, 2019 7:37 pm

Miss Emerita wrote: ↑
Mon Apr 22, 2019 9:11 pm
Interesting news from the ENT! I will be very curious whether the nasal surgery will resolve your FL problem. Please let us know!

About discomfort from the chin strap: did you ever wind up trying taping? I can recommend Somnifix (though it is pricey).

And abour your nasal discomfort: have you tried a nasal mask (goes over the bottom part of your nose but doesn’t touch those sensitive ends of your nostrils)?
I will check out that over the nose mask. Thanks!

marktheshark
Posts: 92
Joined: Mon Sep 17, 2018 10:01 am

Re: Newbie Self Diagnosis - In Denial - AHI Question

Post by marktheshark » Sun May 05, 2019 3:13 pm

So I am a little confused about settings.

I wanted to try different pressure limits, but when I go into the settings and change MAX EP and MAX PS, it doesn't seem to limit the max pressure to 8.0 like I have it set?

Not sure if you can see my settings from last night?

Maybe I need to be changing the IPAP MAX but I don't see that in the settings?

I feel like this new machine is making a difference and is so much more comfortable, I just feel I need to work on tweaking the pressure as I feel it's still waking me up.
Attachments
5-5-2019 Sleepyhead.PNG
5-5-2019 Sleepyhead.PNG (188.34 KiB) Viewed 1487 times

User avatar
Pugsy
Posts: 65411
Joined: Thu May 14, 2009 9:31 am
Location: Missouri, USA

Re: Newbie Self Diagnosis - In Denial - AHI Question

Post by Pugsy » Sun May 05, 2019 3:29 pm

There is no special separate setting for Max IPAP.
Max IPAP is a function of EPAP plus PS and since PS (2 to 8cm) can vary then IPAP is going to vary.
If you want to really limit IPAP you have to limit EPAP and also limit PS max.

Remember PS gets added on to EPAP to give you IPAP.

_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier
Additional Comments: Mask Bleep Eclipse https://bleepsleep.com/the-eclipse/
I may have to RISE but I refuse to SHINE.

marktheshark
Posts: 92
Joined: Mon Sep 17, 2018 10:01 am

Re: Newbie Self Diagnosis - In Denial - AHI Question

Post by marktheshark » Sun May 05, 2019 3:36 pm

Pugsy wrote: ↑
Sun May 05, 2019 3:29 pm
There is no special separate setting for Max IPAP.
Max IPAP is a function of EPAP plus PS and since PS (2 to 8cm) can vary then IPAP is going to vary.
If you want to really limit IPAP you have to limit EPAP and also limit PS max.

Remember PS gets added on to EPAP to give you IPAP.
Got it, I think lol.

What settings should I am for if I want to try a max pressure (or in the range) of 9.0?

Or am I just assuming 5.0 EPAP plus 5.0 PS = 10.00?

User avatar
Jas_williams
Posts: 1120
Joined: Tue Dec 19, 2017 2:12 pm
Location: Somerset UK

Re: Newbie Self Diagnosis - In Denial - AHI Question

Post by Jas_williams » Sun May 05, 2019 3:44 pm

marktheshark wrote: ↑
Sun May 05, 2019 3:36 pm
Pugsy wrote: ↑
Sun May 05, 2019 3:29 pm
There is no special separate setting for Max IPAP.
Max IPAP is a function of EPAP plus PS and since PS (2 to 8cm) can vary then IPAP is going to vary.
If you want to really limit IPAP you have to limit EPAP and also limit PS max.

Remember PS gets added on to EPAP to give you IPAP.
Got it, I think lol.

What settings should I am for if I want to try a max pressure (or in the range) of 9.0?

Or am I just assuming 5.0 EPAP plus 5.0 PS = 10.00?

Mark you will be seriously cramping the ASV, if you tune it down it won’t be able to do what it needs to do to treat any central apnoea.

_________________
Mask: Bleep DreamPort CPAP Mask Solution
Additional Comments: Using sleepyhead and a pressure of 6 - 21 Resmed S9 Adapt SV with a Bleep Sleep Mask

User avatar
Pugsy
Posts: 65411
Joined: Thu May 14, 2009 9:31 am
Location: Missouri, USA

Re: Newbie Self Diagnosis - In Denial - AHI Question

Post by Pugsy » Sun May 05, 2019 4:00 pm

If you do EPAP 5 and PS of 5 you get IPAP max of 10...and essentially a fixed bilevel setting.
You can add in some variables like min and max EPAP and min and max PS...as long as the max settings don't add up to more than 10 and each has some room to move around a bit...you have some auto adjusting functions available but 5 and 5....that's a fixed setting.

_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier
Additional Comments: Mask Bleep Eclipse https://bleepsleep.com/the-eclipse/
I may have to RISE but I refuse to SHINE.