CPAP longevity/reliability stats
CPAP longevity/reliability stats
Hey everyone, today seems to be the day where I'm asking a bunch of questions! My curiosity is getting the best of me.
All this learning about sleep apnea and treatment equipment I wanted to ask if anyone here could comment on the equipment brand reliability?
The two big names here in North America seem to be Philips Respironics and ResMed. In Europe it seems like Fisher Paykel and Devilbiss are contenders as well along with a few of the smaller lesser known names in industry.
PR and Resmed that my local sleep clinic sells say they both have the same warranties but the Resmed units seem to have less issues and warranty claims vs. PR. When I had the choice of choosing between the two brands the clinician informed me that both are very good units but the PR had more warranty claims than Resmed did without elaborating on how many or what the cause of failures were against the # of units sold.
That being said I've read that a well cared for CPAP machine can quite easily go for 7-8 trouble free years with basic cleaning and care. Is that actually mostly true? What are the primary modes of failures? Blower? heating elements? circuit boards?
Who is king in the CPAP industry and who truly makes the best machines out there?
Thanks!
All this learning about sleep apnea and treatment equipment I wanted to ask if anyone here could comment on the equipment brand reliability?
The two big names here in North America seem to be Philips Respironics and ResMed. In Europe it seems like Fisher Paykel and Devilbiss are contenders as well along with a few of the smaller lesser known names in industry.
PR and Resmed that my local sleep clinic sells say they both have the same warranties but the Resmed units seem to have less issues and warranty claims vs. PR. When I had the choice of choosing between the two brands the clinician informed me that both are very good units but the PR had more warranty claims than Resmed did without elaborating on how many or what the cause of failures were against the # of units sold.
That being said I've read that a well cared for CPAP machine can quite easily go for 7-8 trouble free years with basic cleaning and care. Is that actually mostly true? What are the primary modes of failures? Blower? heating elements? circuit boards?
Who is king in the CPAP industry and who truly makes the best machines out there?
Thanks!
Re: CPAP longevity/reliability stats
Respironics have a history of under specing the high current connections of their humidifier heating plates, ala: viewtopic/t102275/Fix-Respironics-Syste ... blems.htmlCPAPSteve wrote: ↑Mon Jul 30, 2018 10:57 pmHey everyone, today seems to be the day where I'm asking a bunch of questions! My curiosity is getting the best of me.
All this learning about sleep apnea and treatment equipment I wanted to ask if anyone here could comment on the equipment brand reliability?
The two big names here in North America seem to be Philips Respironics and ResMed. In Europe it seems like Fisher Paykel and Devilbiss are contenders as well along with a few of the smaller lesser known names in industry.
PR and Resmed that my local sleep clinic sells say they both have the same warranties but the Resmed units seem to have less issues and warranty claims vs. PR. When I had the choice of choosing between the two brands the clinician informed me that both are very good units but the PR had more warranty claims than Resmed did without elaborating on how many or what the cause of failures were against the # of units sold.
That being said I've read that a well cared for CPAP machine can quite easily go for 7-8 trouble free years with basic cleaning and care. Is that actually mostly true? What are the primary modes of failures? Blower? heating elements? circuit boards?
Who is king in the CPAP industry and who truly makes the best machines out there?
Thanks!
I'd heard they had problems in the M series before that, and the 60 series after that...I haven't heard about, (or taken apart) a dreamstation to see if they learned their lesson...
I *have* taken apart several Resmeds, (though not the latest) and they're better built, using soldered on, heavier wires for their humidifier, instead of a too small for the current plug.
Get OSCAR
Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.
Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.
Re: CPAP longevity/reliability stats
I had my Philips Respironics DreamStation BiPAP only 4 months when a sensor stopped working, and received a new machine because it was still under warranty. Mind you, that was this year. I would agree that Philips has reliability issues.
_________________
| Machine: DreamStation BiPAP® Auto Machine |
| Mask: ResMed AirFit™ F20 Full Face CPAP Mask with Headgear |
| Humidifier: DreamStation Heated Humidifier |
| Additional Comments: Pressures: IPAP 18/ EPAP 14; Secondary mask - Nuance Pro Nasal Pillow |
Re: CPAP longevity/reliability stats
Interesting news. That isnthebinfo I was seeking! Underspecced connectors and lack of soldering. I was a former quality assurance guy so this is information that piques my interest. Thanks for sharing that link. Will be an interesting read.
- ChicagoGranny
- Posts: 15497
- Joined: Sun Jan 29, 2012 1:43 pm
- Location: USA
Re: CPAP longevity/reliability stats
Filter change? That's all I can think of.ChicagoGranny wrote: ↑Tue Jul 31, 2018 8:07 amWhat would that be? Other than routine dusting of the case?
_________________
| Machine: ResMed AirSense™ 10 AutoSet™ CPAP Machine with HumidAir™ Heated Humidifier |
| Mask: Bleep DreamPort CPAP Mask Solution |
| Additional Comments: Backup machine: AirSense 10 AutoSet with all the fixins |
- chunkyfrog
- Posts: 34544
- Joined: Mon Jul 12, 2010 5:10 pm
- Location: Nowhere special--this year in particular.
Re: CPAP longevity/reliability stats
Avoiding water and impact is a good plan.
No dropping, tossing, dunking, or mallets.
No dropping, tossing, dunking, or mallets.
_________________
| Mask: AirFit™ P10 For Her Nasal Pillow CPAP Mask with Headgear |
| Additional Comments: Airsense 10 Autoset for Her |
Re: CPAP longevity/reliability stats
Routine maintenance to me would be:
1. Wiping down machine
2. Keeping it out of the sun
3. Not dropping it
4. Making sure it's hooked up to "clean power" - surge protection, True sine wave UPS, etc.
5. Filter changes
6. Compressed air dusting IF I somehow neglected to change the filters or if I forgot to install a filter
7. Cleaning the electrical contacts for the heated hose with rubbing alcohol swabs to ensure good contact.
8. Cleaning of hoses/humidifier/mask (even though that technically isn't the CPAP machine itself)
Definitely overkill but I'm OCD that way
1. Wiping down machine
2. Keeping it out of the sun
3. Not dropping it
4. Making sure it's hooked up to "clean power" - surge protection, True sine wave UPS, etc.
5. Filter changes
6. Compressed air dusting IF I somehow neglected to change the filters or if I forgot to install a filter
7. Cleaning the electrical contacts for the heated hose with rubbing alcohol swabs to ensure good contact.
8. Cleaning of hoses/humidifier/mask (even though that technically isn't the CPAP machine itself)
Definitely overkill but I'm OCD that way
Re: CPAP longevity/reliability stats
Here's the inside of a S9: https://imgur.com/a/S0ojG
Get OSCAR
Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.
Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.
Re: CPAP longevity/reliability stats
Very cool stuff! Thanks for sharing.
_________________
| Machine: DreamStation BiPAP® Auto Machine |
| Mask: ResMed AirFit™ F20 Full Face CPAP Mask with Headgear |
| Humidifier: DreamStation Heated Humidifier |
| Additional Comments: Pressures: IPAP 18/ EPAP 14; Secondary mask - Nuance Pro Nasal Pillow |
- ChicagoGranny
- Posts: 15497
- Joined: Sun Jan 29, 2012 1:43 pm
- Location: USA
Re: CPAP longevity/reliability stats
Well, I asked for it.CPAPSteve wrote: ↑Tue Jul 31, 2018 1:41 pm1. Wiping down machine
2. Keeping it out of the sun
3. Not dropping it
4. Making sure it's hooked up to "clean power" - surge protection, True sine wave UPS, etc.
5. Filter changes
6. Compressed air dusting IF I somehow neglected to change the filters or if I forgot to install a filter
7. Cleaning the electrical contacts for the heated hose with rubbing alcohol swabs to ensure good contact.
8. Cleaning of hoses/humidifier/mask (even though that technically isn't the CPAP machine itself)
Re: CPAP longevity/reliability stats
*totally* unnecessary... the switching power supply doesn't care.
Get OSCAR
Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.
Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.
Re: CPAP longevity/reliability stats
I'd agree that it isn't necessary but all my UPSes at home are true sine wave 
Re: CPAP longevity/reliability stats
My S9 has been in a drawer for 4+ years. Never taken out. Filter changed at 6-9 months, second hose in 4 years. At night I open the drawer, pull out the mask connected to the hose, put it on and press the power button. In the AM reverse the order.
_________________
| Machine: ResMed AirSense™ 10 AutoSet™ CPAP Machine with HumidAir™ Heated Humidifier |
| Mask: Fisher & Paykel Vitera Full Face Mask with Headgear (S, M, or L Cushion) |
| Additional Comments: Back up is a new AS10. |
Re: CPAP longevity/reliability stats
ResMed user here. Even my original (2007) brick Escape still works, it's now my hose dryer. No problems with my S8 APAP, still works. Had first S9 APAP replaced under warranty for moldy smell (tossed mask as I could not get rid of smell and sterilized hose). ResMed Tech support was great. My sleep doc handled mailing for me for $50 (not sure why I used him, maybe Tech Support said? can't remember, sorry).
ResMed S9 range 9.8-17, RespCare Hybrid FFM
Never, never, never, never say never.
Never, never, never, never say never.



