Fix: Respironics System One (50 series) Humidifier Problems

General Discussion on any topic relating to CPAP and/or Sleep Apnea.
nsailsb
Posts: 3
Joined: Mon Jun 27, 2011 9:27 am

Fix: Respironics System One (50 series) Humidifier Problems

Post by nsailsb » Sat Nov 22, 2014 7:37 pm

I want to start out by saying if you try this fix yourself do it at your own risk and I do not guarantee it will fix your machine or if it is even safe to do. Only try this if you are good with soldering electronics.

I have 2 Respironics System One (50 series) that the Humidifiers quit working the same way. At first they seemed to not use as much water and eventually the humidifier light started flashing and the humidifiers didn't work at all. I took them to my local medical equipment supplier and they tested them and said my humidifiers were just fine and that my CPAP machines were not communicating with the humidifiers and that they could fix them for $360 each. Since my second one quit working the same way the first one did and they are both out of warranty I decided to take my older one apart to see if I could figure out what was wrong. As soon as I opened it up I could see the problem and I was able to successfully fix both of them using the method below (please forgive the fact that I repaired them on my poker table lol).

Here's the machine I have:
Image

Take the cover off by doing the following two steps:

Pull the knob off (just pull strait up and it will come off):
Image

Use a torx 15 screwdriver to take the two screws out of the bottom:
Image

If you have the same problem I did you will see the problem right away. Look for the burn in the white plug red wire:
Image

You will need to pry the white plug off because it is melted onto the prong. Here's what the prong looked like on both of my machines:
Image

My theory is that the connection gets corroded and the less contact area the plug has the hotter it gets until it eventually burns the plug and there is no connection left. So I used sandpaper to clean the prong and I used a pair of wire cutters to cut off the burnt end of the white plug and then I cleaned the end of the red wire and soldered it directly to the prong.

Here's what my machines look like after I was done:
Image

Here are the tools I used (I had the soldering iron set at 3 because I didn't want to get it to hot and burn up anything on the circuit board):
Image


Again if you try this do it at your own risk and don't blame me if anything goes wrong.

JimP
Posts: 181
Joined: Thu Oct 02, 2014 2:49 pm

Re: Fix: Respironics System One (50 series) Humidifier Problems

Post by JimP » Sat Nov 22, 2014 7:47 pm

How long did you have these machines before the wire burned?

Could you tell how the moisture got in there?

User avatar
palerider
Posts: 32299
Joined: Wed Dec 16, 2009 5:43 pm
Location: Dallas(ish).

Re: Fix: Respironics System One (50 series) Humidifier Problems

Post by palerider » Sat Nov 22, 2014 7:52 pm

awesome fix for a design flaw!

_________________
Mask: Bleep DreamPort CPAP Mask Solution
Additional Comments: S9 VPAP Auto
Get OSCAR

Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.

nsailsb
Posts: 3
Joined: Mon Jun 27, 2011 9:27 am

Re: Fix: Respironics System One (50 series) Humidifier Problems

Post by nsailsb » Sat Nov 22, 2014 11:25 pm

They each lasted about 2 years and insurance wouldn't buy me a new machine so I paid out of pocket for the second one which was out of warranty when it broke earlier this week. I really do think this is a factory defect and I'm sure they know, but they can get people to pay to fix them without question so why admit the problem.

Janknitz
Posts: 8494
Joined: Sat Mar 20, 2010 1:05 pm
Location: Northern California

Re: Fix: Respironics System One (50 series) Humidifier Problems

Post by Janknitz » Sat Nov 22, 2014 11:40 pm

I'm having the same issue--humidifier isn't working. They gave me a new humidifier and it doesn't work either. I have no patience to work with Crapria on this, but I'm not handy enough to fix my own like that.
What you need to know before you meet your DME http://tinyurl.com/2arffqx
Taming the Mirage Quattro http://tinyurl.com/2ft3lh8
Swift FX Fitting Guide http://tinyurl.com/22ur9ts
Don't Pay that Upcharge! http://tinyurl.com/2ck48rm

User avatar
AlabamaAl
Posts: 128
Joined: Mon Sep 08, 2014 4:21 am

Re: Fix: Respironics System One (50 series) Humidifier Problems

Post by AlabamaAl » Sun Nov 23, 2014 7:59 am

Good fix. Some anti oxidation fluid on the connectors at the time of assembly might have prevented the increased resistance, and thus the heat. Sadly, most manufactures don't think past the end of the warranty date, others design the product to fail during a specific time period. After all they do like to sell new machines.

_________________
Mask: Mirage Quattro™ Full Face CPAP Mask with Headgear
Additional Comments: Sleepyhead software. Pressure 15 low, 20 high

User avatar
palerider
Posts: 32299
Joined: Wed Dec 16, 2009 5:43 pm
Location: Dallas(ish).

Re: Fix: Respironics System One (50 series) Humidifier Problems

Post by palerider » Sun Nov 23, 2014 2:10 pm

Janknitz wrote:I'm having the same issue--humidifier isn't working. They gave me a new humidifier and it doesn't work either. I have no patience to work with Crapria on this, but I'm not handy enough to fix my own like that.
ask around, some of your friends might be handy with a soldering iron... the PRS1s are really easy to pop open...

_________________
Mask: Bleep DreamPort CPAP Mask Solution
Additional Comments: S9 VPAP Auto
Get OSCAR

Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.

User avatar
Sir NoddinOff
Posts: 4190
Joined: Mon May 14, 2012 5:30 pm
Location: California

Re: Fix: Respironics System One (50 series) Humidifier Problems

Post by Sir NoddinOff » Sun Nov 23, 2014 5:30 pm

I'm totally impressed by your discovery, not to mention the GREAT documentation which is certainly of engineering quality. Many thanks because I'm sure I'll confront this problem on one of my PR machines somewhere down the road. <loud and prolonged clapping> Shame on PR for not envisioning this failure, and double shame for not informing their distributors/DMEs.

_________________
Mask: AirFit™ F10 Full Face Mask with Headgear
Additional Comments: Sleepyhead software v.0.9.8.1 Open GL and Encore Pro v2.2.
I like my ResMed AirFit F10 FFM - reasonably low leaks for my ASV therapy. I'm currently using a PR S1 AutoSV 960P Advanced. I also keep a ResMed S9 Adapt as backup. I use a heated Hibernite hose. Still rockin' with Win 7 by using GWX to stop Win 10.

User avatar
squid13
Posts: 2300
Joined: Fri Jul 15, 2011 3:47 pm
Location: Pensacola, FL

Re: Fix: Respironics System One (50 series) Humidifier Problems

Post by squid13 » Sun Nov 23, 2014 6:13 pm

That's a great fix you did, the only thing that I could add is to put a piece of shrink tubing over your solider connection.

_________________
Machine: ResMed AirCurve 10 ASV Machine with Heated Humidifier
Mask: Evora Full Face Mask - Fitpack
Additional Comments: AirCurve 10 ASV, Oscar V1.0.1-r-1
US Navy Retired 1973,AirCurve 10 ASV, Mode: ASV Auto, Min EPAP: 7.2, Max EPAP: 15.0, Min PS:4.0, Max PS: 15.0, Mask ResMed Airtouch F20, Backup: (2) AirCurve 10 ASV

JDS74
Posts: 3397
Joined: Sun Jan 23, 2011 2:57 pm
Location: South Carolina

Re: Fix: Respironics System One (50 series) Humidifier Problems

Post by JDS74 » Fri Dec 05, 2014 7:06 am

Thanks so much for your post.
My Respironics 960 just recently developed the same symptoms.
I called Accurate Biomed Services for a possible repair and they are familiar with this problem.
Since my machine is stiil under warranty, they suggested sending it in and they will handle the warranty.
Since the turnaround will be about four weeks, I ordered a new machine.

The DME is shipping next day air at no extra charge, including an Oracle mask, also at no charge and discounting an additional 10%.

_________________
Mask: Oracle HC452 Oral CPAP Mask
Humidifier: DreamStation Heated Humidifier
Additional Comments: EverFlo Q 3.0 Liters O2 PR DSX900 ASV
Last edited by JDS74 on Fri Jan 09, 2015 9:00 pm, edited 1 time in total.
Oracle 452 Lessons Learned Updated
DSX900 AutoSV with HC150 extra humidifier and Hibernite heated hose
Settings: EPAP Min-10.0, EPAP Max-17, PS Min-3, PS Max-10, Max Pressure-20, Rate-Auto, Biflex-1.
Sleepyhead and Encore Pro 2.21.

User avatar
dadbar1
Posts: 50
Joined: Tue May 13, 2014 1:22 pm
Location: Boston area

Re: Fix: Respironics System One (50 series) Humidifier Problems

Post by dadbar1 » Fri Dec 05, 2014 9:26 am

Ha ha ha...the secret is out. $360 to fix a bad connection. Gotta love that.

My old Remstar unit was also pretty easy to take apart and fix. These boxes are not exactly nuclear warheads.
PR System One Auto A-flex
ResMed Mirage Ultra Mask
Sleepmapper for Android
No humidifier (I take my air shaken, not stirred)

JDS74
Posts: 3397
Joined: Sun Jan 23, 2011 2:57 pm
Location: South Carolina

Re: Fix: Respironics System One (50 series) Humidifier Problems

Post by JDS74 » Mon Dec 29, 2014 11:00 am

Thank you so much for this original post.
As I mentioned before, Accurate Biosystems handles this type of repair.
I sent my Respironics 960 ASV machine to them after experiencing similar symptoms.
They have requested a new replacement from Respironics as mine was still under warranty.

They are great folks. The only cost was my shipping to them and the shipping cost to get it back to me.

_________________
Mask: Oracle HC452 Oral CPAP Mask
Humidifier: DreamStation Heated Humidifier
Additional Comments: EverFlo Q 3.0 Liters O2 PR DSX900 ASV
Last edited by JDS74 on Thu Jan 01, 2015 8:03 pm, edited 1 time in total.
Oracle 452 Lessons Learned Updated
DSX900 AutoSV with HC150 extra humidifier and Hibernite heated hose
Settings: EPAP Min-10.0, EPAP Max-17, PS Min-3, PS Max-10, Max Pressure-20, Rate-Auto, Biflex-1.
Sleepyhead and Encore Pro 2.21.

Suenyo
Posts: 98
Joined: Sun Sep 07, 2014 12:38 am

Re: Fix: Respironics System One (50 series) Humidifier Problems

Post by Suenyo » Thu Jan 01, 2015 4:43 pm

I applaud your handyman skills. One tip if I may offer it to future fixers of this problem: That solder joint looks a bit "cold". If you are worried about overheating, try clamping a tiny pair of ViceGrip pliers on just past the solder point. They will stop the heat from going on down that wire so you can heat the joint till the solder flows completely.
Life is a dream…if I'm wearing my Resmed AirTouch F20 cushion and my Autosense 10 Autoset isn't being hijacked by some dung-eating state security troll.

JDS74
Posts: 3397
Joined: Sun Jan 23, 2011 2:57 pm
Location: South Carolina

Re: Fix: Respironics System One (50 series) Humidifier Problems

Post by JDS74 » Fri Jan 09, 2015 9:05 pm

Just got notified by Accurate Biomed Services that Respironics is replacing my 960P at no charge.
What a great company!!!

_________________
Mask: Oracle HC452 Oral CPAP Mask
Humidifier: DreamStation Heated Humidifier
Additional Comments: EverFlo Q 3.0 Liters O2 PR DSX900 ASV
Oracle 452 Lessons Learned Updated
DSX900 AutoSV with HC150 extra humidifier and Hibernite heated hose
Settings: EPAP Min-10.0, EPAP Max-17, PS Min-3, PS Max-10, Max Pressure-20, Rate-Auto, Biflex-1.
Sleepyhead and Encore Pro 2.21.

User avatar
Sir NoddinOff
Posts: 4190
Joined: Mon May 14, 2012 5:30 pm
Location: California

Re: Fix: Respironics System One (50 series) Humidifier Problems

Post by Sir NoddinOff » Sat Jan 10, 2015 1:09 pm

Suenyo wrote:I applaud your handyman skills. One tip if I may offer it to future fixers of this problem: That solder joint looks a bit "cold". If you are worried about overheating, try clamping a tiny pair of ViceGrip pliers on just past the solder point. They will stop the heat from going on down that wire so you can heat the joint till the solder flows completely.
Thanks Suenyo, that's a great tip for dissipating heat during soldering. I learned that trick back in high school but somehow forgot it along the many years since then. I used to use a hemostat, the same kind pot smokers use for roaches. Somewhat easier to fit in tight spaces.

_________________
Mask: AirFit™ F10 Full Face Mask with Headgear
Additional Comments: Sleepyhead software v.0.9.8.1 Open GL and Encore Pro v2.2.
I like my ResMed AirFit F10 FFM - reasonably low leaks for my ASV therapy. I'm currently using a PR S1 AutoSV 960P Advanced. I also keep a ResMed S9 Adapt as backup. I use a heated Hibernite hose. Still rockin' with Win 7 by using GWX to stop Win 10.