Auto cpap is..

Post by carojae » Wed May 09, 2018 4:50 pm

I am wondering if a auto cpap machine will adjust and ramp downward if it ramped (while asleep) up too high.
I wake up and my machine is almost always running at high setting. You know, if I open my mouth air comes rushing out. This whole this is annoying and I want to fix it.
I pulled it off the night stand one time and I've been wondering every since if I broke something inside. It does it's job but doesn't seem to want to ramp down.
Does anyone have info on this?

Forgot to add my machine: PR System One REMStar 60 Series BiPAP Auto with Bi-Flex
Also have humidifier with it.

Jim

_________________
MachineMask
Jim :-)

D.H.
Posts: 3535
Joined: Sun Sep 04, 2016 7:07 pm

Re: Auto cpap is..

Post by D.H. » Wed May 09, 2018 5:01 pm

In general, it will (gradually) lower the pressure down to the "floor" as it senses that higher pressures are no longer needed. I'm not sure if it can sense issues specific to "too much" pressure.

_________________
MachineMask
Additional Comments: Auto PAP; 13.5 cmH2O min - 20 cmH2O max

User avatar
zoocrewphoto
Posts: 3732
Joined: Mon Apr 30, 2012 10:34 pm
Location: Seatac, WA

Re: Auto cpap is..

Post by zoocrewphoto » Wed May 09, 2018 5:03 pm

It sounds like it increased pressure to prevent events, but that level of pressure bothers you enough to wake you up. You can restart it so that it starts down low again. Or wait for it to go back down slowly.

Can you tell us what your pressure settings are? Do you know about sleepyhead? You can use a free program to view your data. And if you post a screenshot of the graphs and the charts on the left side, we can help you make adjustments that will help your therapy and/or your comfort.

_________________
Mask: Quattro™ FX Full Face CPAP Mask with Headgear
Humidifier: S9™ Series H5i™ Heated Humidifier with Climate Control
Additional Comments: Resmed S9 autoset pressure range 11-17
Who would have thought it would be this challenging to sleep and breathe at the same time?

HoseCrusher
Posts: 2744
Joined: Tue Oct 12, 2010 6:42 pm

Re: Auto cpap is..

Post by HoseCrusher » Wed May 09, 2018 7:36 pm

I started out with a ResMed S8 Escape Auto. When I left the maximum pressure at 20 it would frequently go to 20 and get stuck there. I would need to shut the machine off and restart it to get out of this maximum pressure situation. My workaround was to set the maximum pressure just above what I needed and then I transitioned to an S8 Autoset that didn't seem to have the same issue.

_________________
Mask: Brevida™ Nasal Pillow CPAP Mask with Headgear
Additional Comments: Machine is an AirSense 10 AutoSet For Her with Heated Humidifier.
SpO2 96+% and holding...

User avatar
palerider
Posts: 32308
Joined: Wed Dec 16, 2009 5:43 pm
Location: Dallas(ish).

Re: Auto cpap is..

Post by palerider » Wed May 09, 2018 8:31 pm

carojae wrote: ↑
Wed May 09, 2018 4:50 pm
I am wondering if a auto cpap machine will adjust and ramp downward if it ramped (while asleep) up too high.
As soon as the machine stops sensing the events that cause it to raise pressure, it will start dropping pressure back down to the minimum set.
carojae wrote: ↑
Wed May 09, 2018 4:50 pm
I wake up and my machine is almost always running at high setting. You know, if I open my mouth air comes rushing out. This whole this is annoying and I want to fix it.
The first step is to pull the data into sleepyhead and post it, like it says in my sig.
carojae wrote: ↑
Wed May 09, 2018 4:50 pm
I pulled it off the night stand one time and I've been wondering every since if I broke something inside. It does it's job but doesn't seem to want to ramp down.
unlikely.

_________________
Mask: Bleep DreamPort CPAP Mask Solution
Additional Comments: S9 VPAP Auto
Get OSCAR

Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.

Janknitz
Posts: 8518
Joined: Sat Mar 20, 2010 1:05 pm
Location: Northern California

Re: Auto cpap is..

Post by Janknitz » Wed May 09, 2018 10:06 pm

It doesn’t matter what the pressure, if you open your mouth the air is going to go out your mouth instead of into your windpipe. Path of least resistance. So don’t let that be your criteria for too much Pressure.
What you need to know before you meet your DME http://tinyurl.com/2arffqx
Taming the Mirage Quattro http://tinyurl.com/2ft3lh8
Swift FX Fitting Guide http://tinyurl.com/22ur9ts
Don't Pay that Upcharge! http://tinyurl.com/2ck48rm

User avatar
carojae
Posts: 63
Joined: Wed May 19, 2010 11:12 am
Location: Alaska

Re: Auto cpap is..

Post by carojae » Wed May 09, 2018 10:44 pm

Okay, is this the page you guys want?
Just to be clear, I've used a machine for close to 10 years and this is my second machine. What I am asking about is having the machine ramp down on it's own automatically - it doesn't seem to do that now. I am aware I can hit the ramp button and it will restart on low but "I want this process to be automatic". It seems to me that it use to be automatic but I am not sure (guess I didn't fully pay attention to that detail but I don't remember having air blasting every morning before.

Here is a screen shot with my pressures. I've (me) limited my upper pressure on auto to 13 (if I remember right) because it used to push air out the corners of my eyes when it was set at 20; stuff would come out of my ears and I was constantly applying cortisone to my ear canals. The high pressure just raised hell with me so I turned it down and I am happy with the pressure that it's on because there are no extra problems from too much air pressure.

Let me know what more/page you need to see if you want more info. I am still thinking this machine use to ramp down on it's own and wouldn't wake me up every morning because of air blasting in my nasal mask.

thank you guys.

Jim

_________________
MachineMask
Attachments
screenshot-20180509-202605.png
screenshot-20180509-202605.png (121.75 KiB) Viewed 19743 times
Jim :-)

User avatar
zoocrewphoto
Posts: 3732
Joined: Mon Apr 30, 2012 10:34 pm
Location: Seatac, WA

Re: Auto cpap is..

Post by zoocrewphoto » Wed May 09, 2018 11:24 pm

Looks like your machine does ramp down on its own, at least it did a few times. It does take time to go back down. You are probably just waking up while it is high sometimes.

Also, your pressure is set to a very wide range. Much too low to start. If you started at a higher pressure, one that can prevent most of your events, it will be less likely to go as high, chasing events. Your screen shot does show your ahi, but the events chart shows a lot of events. You would probably have more restful sleep and fewer wake ups if you had a narrower range. I would try starting it around 8. That might not be enough, but it helps to see some data after a change and see how it goes.

I also wonder if you are rolling onto your back or if that is REM sleep. It looks like it is low for awhile, and then climbs for awhile, then goes back down for awhile.

_________________
Mask: Quattro™ FX Full Face CPAP Mask with Headgear
Humidifier: S9™ Series H5i™ Heated Humidifier with Climate Control
Additional Comments: Resmed S9 autoset pressure range 11-17
Who would have thought it would be this challenging to sleep and breathe at the same time?

User avatar
palerider
Posts: 32308
Joined: Wed Dec 16, 2009 5:43 pm
Location: Dallas(ish).

Re: Auto cpap is..

Post by palerider » Thu May 10, 2018 12:11 am

What Zoocrew said.

_________________
Mask: Bleep DreamPort CPAP Mask Solution
Additional Comments: S9 VPAP Auto
Get OSCAR

Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.

User avatar
Julie
Posts: 20051
Joined: Tue Feb 28, 2006 12:58 pm

Re: Auto cpap is..

Post by Julie » Thu May 10, 2018 3:19 am

Double ditto.

HoseCrusher
Posts: 2744
Joined: Tue Oct 12, 2010 6:42 pm

Re: Auto cpap is..

Post by HoseCrusher » Thu May 10, 2018 9:21 am

Looking at your Pressure graph I count about 18 times the pressure reduced. Some were small and others were larger.

If that were my chart I would raise the minimum from 4 to 5.5 or 6 and expect better results because of fewer pressure transitions.

If you like to play...

You could count the number of events that occurred when the pressure was below 6, normalize those events with the amount of time spent below 6, then compare the number of normalized events that occurred above 6. Theoretically, raising the pressure to 6 would eliminate the events that occurred below that pressure and with less events you should have a more restful sleep.

_________________
Mask: Brevida™ Nasal Pillow CPAP Mask with Headgear
Additional Comments: Machine is an AirSense 10 AutoSet For Her with Heated Humidifier.
SpO2 96+% and holding...

User avatar
carojae
Posts: 63
Joined: Wed May 19, 2010 11:12 am
Location: Alaska

Re: Auto cpap is..

Post by carojae » Thu May 10, 2018 9:54 am

Looks like your machine does ramp down on its own, at least it did a few times. It does take time to go back down. You are probably just waking up while it is high sometimes.
Oh okay. At least I know it is doing what I thought it should do :-) Waking up with a high pressure on most days made me think the machine wasn't working but I was wrong.
Also, your pressure is set to a very wide range. Much too low to start. If you started at a higher pressure, one that can prevent most of your events, it will be less likely to go as high, chasing events. Your screen shot does show your ahi, but the events chart shows a lot of events. You would probably have more restful sleep and fewer wake ups if you had a narrower range. I would try starting it around 8. That might not be enough, but it helps to see some data after a change and see how it goes.
Glad I posted now. I set the machine myself thinking that the "auto setting" would adjust. Glad to know I should up the low setting.
I also wonder if you are rolling onto your back or if that is REM sleep. It looks like it is low for awhile, and then climbs for awhile, then goes back down for awhile.
What is REM sleep? Can you explain what you mean here?

Thank you guys so much.

Jim

_________________
MachineMask
Jim :-)

User avatar
carojae
Posts: 63
Joined: Wed May 19, 2010 11:12 am
Location: Alaska

Re: Auto cpap is..

Post by carojae » Thu May 10, 2018 10:01 am

HoseCrusher wrote: ↑
Thu May 10, 2018 9:21 am
Looking at your Pressure graph I count about 18 times the pressure reduced. Some were small and others were larger.

If that were my chart I would raise the minimum from 4 to 5.5 or 6 and expect better results because of fewer pressure transitions.

If you like to play...

You could count the number of events that occurred when the pressure was below 6, normalize those events with the amount of time spent below 6, then compare the number of normalized events that occurred above 6. Theoretically, raising the pressure to 6 would eliminate the events that occurred below that pressure and with less events you should have a more restful sleep.
Okay, I will try that Hosecrusher. Thank you very much :-)

_________________
MachineMask
Jim :-)

User avatar
carojae
Posts: 63
Joined: Wed May 19, 2010 11:12 am
Location: Alaska

Re: Auto cpap is..

Post by carojae » Thu May 10, 2018 10:04 am

HoseCrusher wrote: ↑
Thu May 10, 2018 9:21 am
Looking at your Pressure graph I count about 18 times the pressure reduced. Some were small and others were larger.

If that were my chart I would raise the minimum from 4 to 5.5 or 6 and expect better results because of fewer pressure transitions.

If you like to play...

You could count the number of events that occurred when the pressure was below 6, normalize those events with the amount of time spent below 6, then compare the number of normalized events that occurred above 6. Theoretically, raising the pressure to 6 would eliminate the events that occurred below that pressure and with less events you should have a more restful sleep.
Okay, I will try that Hosecrusher. Thank you very much :-) Also, is the high pressure okay?
*I am seeing the sleep doctor next week....

_________________
MachineMask
Jim :-)

HoseCrusher
Posts: 2744
Joined: Tue Oct 12, 2010 6:42 pm

Re: Auto cpap is..

Post by HoseCrusher » Thu May 10, 2018 10:59 am

Your pressure graph shows that you came close to your maximum pressure but did not reach it. I think it is OK.

_________________
Mask: Brevida™ Nasal Pillow CPAP Mask with Headgear
Additional Comments: Machine is an AirSense 10 AutoSet For Her with Heated Humidifier.