Auto cpap is..
Auto cpap is..
I am wondering if a auto cpap machine will adjust and ramp downward if it ramped (while asleep) up too high.
I wake up and my machine is almost always running at high setting. You know, if I open my mouth air comes rushing out. This whole this is annoying and I want to fix it.
I pulled it off the night stand one time and I've been wondering every since if I broke something inside. It does it's job but doesn't seem to want to ramp down.
Does anyone have info on this?
Forgot to add my machine: PR System One REMStar 60 Series BiPAP Auto with Bi-Flex
Also have humidifier with it.
Jim
I wake up and my machine is almost always running at high setting. You know, if I open my mouth air comes rushing out. This whole this is annoying and I want to fix it.
I pulled it off the night stand one time and I've been wondering every since if I broke something inside. It does it's job but doesn't seem to want to ramp down.
Does anyone have info on this?
Forgot to add my machine: PR System One REMStar 60 Series BiPAP Auto with Bi-Flex
Also have humidifier with it.
Jim
Jim 
Re: Auto cpap is..
In general, it will (gradually) lower the pressure down to the "floor" as it senses that higher pressures are no longer needed. I'm not sure if it can sense issues specific to "too much" pressure.
- zoocrewphoto
- Posts: 3732
- Joined: Mon Apr 30, 2012 10:34 pm
- Location: Seatac, WA
Re: Auto cpap is..
It sounds like it increased pressure to prevent events, but that level of pressure bothers you enough to wake you up. You can restart it so that it starts down low again. Or wait for it to go back down slowly.
Can you tell us what your pressure settings are? Do you know about sleepyhead? You can use a free program to view your data. And if you post a screenshot of the graphs and the charts on the left side, we can help you make adjustments that will help your therapy and/or your comfort.
Can you tell us what your pressure settings are? Do you know about sleepyhead? You can use a free program to view your data. And if you post a screenshot of the graphs and the charts on the left side, we can help you make adjustments that will help your therapy and/or your comfort.
_________________
| Mask: Quattro™ FX Full Face CPAP Mask with Headgear |
| Humidifier: S9™ Series H5i™ Heated Humidifier with Climate Control |
| Additional Comments: Resmed S9 autoset pressure range 11-17 |
Who would have thought it would be this challenging to sleep and breathe at the same time?
-
HoseCrusher
- Posts: 2744
- Joined: Tue Oct 12, 2010 6:42 pm
Re: Auto cpap is..
I started out with a ResMed S8 Escape Auto. When I left the maximum pressure at 20 it would frequently go to 20 and get stuck there. I would need to shut the machine off and restart it to get out of this maximum pressure situation. My workaround was to set the maximum pressure just above what I needed and then I transitioned to an S8 Autoset that didn't seem to have the same issue.
_________________
| Mask: Brevida™ Nasal Pillow CPAP Mask with Headgear |
| Additional Comments: Machine is an AirSense 10 AutoSet For Her with Heated Humidifier. |
SpO2 96+% and holding...
Re: Auto cpap is..
As soon as the machine stops sensing the events that cause it to raise pressure, it will start dropping pressure back down to the minimum set.
The first step is to pull the data into sleepyhead and post it, like it says in my sig.
unlikely.
Get OSCAR
Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.
Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.
Re: Auto cpap is..
It doesn’t matter what the pressure, if you open your mouth the air is going to go out your mouth instead of into your windpipe. Path of least resistance. So don’t let that be your criteria for too much Pressure.
_________________
| Machine: ResMed AirSense™ 10 AutoSet™ CPAP Machine with HumidAir™ Heated Humidifier |
| Mask: DreamWear Nasal CPAP Mask with Headgear |
What you need to know before you meet your DME http://tinyurl.com/2arffqx
Taming the Mirage Quattro http://tinyurl.com/2ft3lh8
Swift FX Fitting Guide http://tinyurl.com/22ur9ts
Don't Pay that Upcharge! http://tinyurl.com/2ck48rm
Taming the Mirage Quattro http://tinyurl.com/2ft3lh8
Swift FX Fitting Guide http://tinyurl.com/22ur9ts
Don't Pay that Upcharge! http://tinyurl.com/2ck48rm
Re: Auto cpap is..
Okay, is this the page you guys want?
Just to be clear, I've used a machine for close to 10 years and this is my second machine. What I am asking about is having the machine ramp down on it's own automatically - it doesn't seem to do that now. I am aware I can hit the ramp button and it will restart on low but "I want this process to be automatic". It seems to me that it use to be automatic but I am not sure (guess I didn't fully pay attention to that detail but I don't remember having air blasting every morning before.
Here is a screen shot with my pressures. I've (me) limited my upper pressure on auto to 13 (if I remember right) because it used to push air out the corners of my eyes when it was set at 20; stuff would come out of my ears and I was constantly applying cortisone to my ear canals. The high pressure just raised hell with me so I turned it down and I am happy with the pressure that it's on because there are no extra problems from too much air pressure.
Let me know what more/page you need to see if you want more info. I am still thinking this machine use to ramp down on it's own and wouldn't wake me up every morning because of air blasting in my nasal mask.
thank you guys.
Jim
Just to be clear, I've used a machine for close to 10 years and this is my second machine. What I am asking about is having the machine ramp down on it's own automatically - it doesn't seem to do that now. I am aware I can hit the ramp button and it will restart on low but "I want this process to be automatic". It seems to me that it use to be automatic but I am not sure (guess I didn't fully pay attention to that detail but I don't remember having air blasting every morning before.
Here is a screen shot with my pressures. I've (me) limited my upper pressure on auto to 13 (if I remember right) because it used to push air out the corners of my eyes when it was set at 20; stuff would come out of my ears and I was constantly applying cortisone to my ear canals. The high pressure just raised hell with me so I turned it down and I am happy with the pressure that it's on because there are no extra problems from too much air pressure.
Let me know what more/page you need to see if you want more info. I am still thinking this machine use to ramp down on it's own and wouldn't wake me up every morning because of air blasting in my nasal mask.
thank you guys.
Jim
- Attachments
-
- screenshot-20180509-202605.png (121.75 KiB) Viewed 18984 times
Jim 
- zoocrewphoto
- Posts: 3732
- Joined: Mon Apr 30, 2012 10:34 pm
- Location: Seatac, WA
Re: Auto cpap is..
Looks like your machine does ramp down on its own, at least it did a few times. It does take time to go back down. You are probably just waking up while it is high sometimes.
Also, your pressure is set to a very wide range. Much too low to start. If you started at a higher pressure, one that can prevent most of your events, it will be less likely to go as high, chasing events. Your screen shot does show your ahi, but the events chart shows a lot of events. You would probably have more restful sleep and fewer wake ups if you had a narrower range. I would try starting it around 8. That might not be enough, but it helps to see some data after a change and see how it goes.
I also wonder if you are rolling onto your back or if that is REM sleep. It looks like it is low for awhile, and then climbs for awhile, then goes back down for awhile.
Also, your pressure is set to a very wide range. Much too low to start. If you started at a higher pressure, one that can prevent most of your events, it will be less likely to go as high, chasing events. Your screen shot does show your ahi, but the events chart shows a lot of events. You would probably have more restful sleep and fewer wake ups if you had a narrower range. I would try starting it around 8. That might not be enough, but it helps to see some data after a change and see how it goes.
I also wonder if you are rolling onto your back or if that is REM sleep. It looks like it is low for awhile, and then climbs for awhile, then goes back down for awhile.
_________________
| Mask: Quattro™ FX Full Face CPAP Mask with Headgear |
| Humidifier: S9™ Series H5i™ Heated Humidifier with Climate Control |
| Additional Comments: Resmed S9 autoset pressure range 11-17 |
Who would have thought it would be this challenging to sleep and breathe at the same time?
Re: Auto cpap is..
What Zoocrew said.
Get OSCAR
Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.
Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.
Re: Auto cpap is..
Double ditto.
-
HoseCrusher
- Posts: 2744
- Joined: Tue Oct 12, 2010 6:42 pm
Re: Auto cpap is..
Looking at your Pressure graph I count about 18 times the pressure reduced. Some were small and others were larger.
If that were my chart I would raise the minimum from 4 to 5.5 or 6 and expect better results because of fewer pressure transitions.
If you like to play...
You could count the number of events that occurred when the pressure was below 6, normalize those events with the amount of time spent below 6, then compare the number of normalized events that occurred above 6. Theoretically, raising the pressure to 6 would eliminate the events that occurred below that pressure and with less events you should have a more restful sleep.
If that were my chart I would raise the minimum from 4 to 5.5 or 6 and expect better results because of fewer pressure transitions.
If you like to play...
You could count the number of events that occurred when the pressure was below 6, normalize those events with the amount of time spent below 6, then compare the number of normalized events that occurred above 6. Theoretically, raising the pressure to 6 would eliminate the events that occurred below that pressure and with less events you should have a more restful sleep.
_________________
| Mask: Brevida™ Nasal Pillow CPAP Mask with Headgear |
| Additional Comments: Machine is an AirSense 10 AutoSet For Her with Heated Humidifier. |
SpO2 96+% and holding...
Re: Auto cpap is..
Oh okay. At least I know it is doing what I thought it should doLooks like your machine does ramp down on its own, at least it did a few times. It does take time to go back down. You are probably just waking up while it is high sometimes.
Glad I posted now. I set the machine myself thinking that the "auto setting" would adjust. Glad to know I should up the low setting.Also, your pressure is set to a very wide range. Much too low to start. If you started at a higher pressure, one that can prevent most of your events, it will be less likely to go as high, chasing events. Your screen shot does show your ahi, but the events chart shows a lot of events. You would probably have more restful sleep and fewer wake ups if you had a narrower range. I would try starting it around 8. That might not be enough, but it helps to see some data after a change and see how it goes.
What is REM sleep? Can you explain what you mean here?I also wonder if you are rolling onto your back or if that is REM sleep. It looks like it is low for awhile, and then climbs for awhile, then goes back down for awhile.
Thank you guys so much.
Jim
Jim 
Re: Auto cpap is..
Okay, I will try that Hosecrusher. Thank you very muchHoseCrusher wrote: ↑Thu May 10, 2018 9:21 amLooking at your Pressure graph I count about 18 times the pressure reduced. Some were small and others were larger.
If that were my chart I would raise the minimum from 4 to 5.5 or 6 and expect better results because of fewer pressure transitions.
If you like to play...
You could count the number of events that occurred when the pressure was below 6, normalize those events with the amount of time spent below 6, then compare the number of normalized events that occurred above 6. Theoretically, raising the pressure to 6 would eliminate the events that occurred below that pressure and with less events you should have a more restful sleep.
Jim 
Re: Auto cpap is..
Okay, I will try that Hosecrusher. Thank you very muchHoseCrusher wrote: ↑Thu May 10, 2018 9:21 amLooking at your Pressure graph I count about 18 times the pressure reduced. Some were small and others were larger.
If that were my chart I would raise the minimum from 4 to 5.5 or 6 and expect better results because of fewer pressure transitions.
If you like to play...
You could count the number of events that occurred when the pressure was below 6, normalize those events with the amount of time spent below 6, then compare the number of normalized events that occurred above 6. Theoretically, raising the pressure to 6 would eliminate the events that occurred below that pressure and with less events you should have a more restful sleep.
*I am seeing the sleep doctor next week....
Jim 
-
HoseCrusher
- Posts: 2744
- Joined: Tue Oct 12, 2010 6:42 pm
Re: Auto cpap is..
Your pressure graph shows that you came close to your maximum pressure but did not reach it. I think it is OK.
_________________
| Mask: Brevida™ Nasal Pillow CPAP Mask with Headgear |
| Additional Comments: Machine is an AirSense 10 AutoSet For Her with Heated Humidifier. |
SpO2 96+% and holding...