Resmed S9 broken and out of warranty! What now?

General Discussion on any topic relating to CPAP and/or Sleep Apnea.
User avatar
DeeCPAP
Posts: 441
Joined: Mon Dec 14, 2015 11:32 pm

Resmed S9 broken and out of warranty! What now?

Post by DeeCPAP » Sun May 07, 2017 6:52 am

My machine stopped working properly. I'm due a new one in August according to medicare rules.
Instead of showing how many hours slept each time I use it, it's adding up the hours over the few
days. It's reading 23.5 hours slept "last night". The AHI is based on the hours slept so now the
AHI is off. I'm *assuming* the pressure and the other things work properly.

What can I do now other than not buy a Resmed again because it's not trustworthy? Oh, and
my DME hasn't contacted me since I last emailed him about 2 months ago for a minor
issue.

Thanks.

_________________
Machine: ResMed AirSense™ 10 AutoSet™ CPAP Machine with HumidAir™ Heated Humidifier
Mask: DreamWear Nasal CPAP Mask with Headgear
Additional Comments: My headgear varies (STILL!)
Resmed S9 with humidifier and in need of the right mask.

User avatar
LSAT
Posts: 13357
Joined: Sun Nov 16, 2008 10:11 am
Location: SE Wisconsin

Re: Resmed S9 broken and out of warranty! What now?

Post by LSAT » Sun May 07, 2017 7:36 am

Earlier today RogerSC suggested that you re-boot your machine...Did you do that? It was suggested that you download Sleepyhead...Did you do that?

User avatar
DeeCPAP
Posts: 441
Joined: Mon Dec 14, 2015 11:32 pm

Re: Resmed S9 broken and out of warranty! What now?

Post by DeeCPAP » Sun May 07, 2017 7:59 am

LSAT wrote:Earlier today RogerSC suggested that you re-boot your machine...Did you do that? It was suggested that you download Sleepyhead...Did you do that?
Yes, I did. I slept 3 hours last night and it just added 3 hours to the 21 hours it had accumulated. The AHI was a fraction.

I've had SleepyHead for a while. I'll have to look at the stats after noontime because that's when last night's data will probably show up.

_________________
Machine: ResMed AirSense™ 10 AutoSet™ CPAP Machine with HumidAir™ Heated Humidifier
Mask: DreamWear Nasal CPAP Mask with Headgear
Additional Comments: My headgear varies (STILL!)
Resmed S9 with humidifier and in need of the right mask.

User avatar
LSAT
Posts: 13357
Joined: Sun Nov 16, 2008 10:11 am
Location: SE Wisconsin

Re: Resmed S9 broken and out of warranty! What now?

Post by LSAT » Sun May 07, 2017 9:50 am

DeeCPAP wrote:
LSAT wrote:Earlier today RogerSC suggested that you re-boot your machine...Did you do that? It was suggested that you download Sleepyhead...Did you do that?
Yes, I did. I slept 3 hours last night and it just added 3 hours to the 21 hours it had accumulated. The AHI was a fraction.

I've had SleepyHead for a while. I'll have to look at the stats after noontime because that's when last night's data will probably show up.
The S9 has a usage setting that can be day, week, or month. If that is what you are looking at, I'm guessing that the setting is week or month rather than day and the hours are accumulating night after night.

User avatar
Julie
Posts: 20051
Joined: Tue Feb 28, 2006 12:58 pm

Re: Resmed S9 broken and out of warranty! What now?

Post by Julie » Sun May 07, 2017 9:54 am

If you're reading 'last' night's stats after noon today that could explain it - machines count from noon to noon (not 6 am), so that may be the problem.

User avatar
DeeCPAP
Posts: 441
Joined: Mon Dec 14, 2015 11:32 pm

Re: Resmed S9 broken and out of warranty! What now?

Post by DeeCPAP » Sun May 07, 2017 1:04 pm

LSAT wrote:
DeeCPAP wrote:
LSAT wrote:Earlier today RogerSC suggested that you re-boot your machine...Did you do that? It was suggested that you download Sleepyhead...Did you do that?
Yes, I did. I slept 3 hours last night and it just added 3 hours to the 21 hours it had accumulated. The AHI was a fraction.

I've had SleepyHead for a while. I'll have to look at the stats after noontime because that's when last night's data will probably show up.
The S9 has a usage setting that can be day, week, or month. If that is what you are looking at, I'm guessing that the setting is week or month rather than day and the hours are accumulating night after night.
Yes, the day, week, month and year can be accessed by pressing the right buttons. The 23.5 out of 24 shows up under the
heading of DAILY. Today it's back to 0 hours. For the week it shows 7.7 and I'm nowhere near that since I average about 6 or 6.5 on a daily basis. It also shows the pressure for the week is higher than it should be. So, the daily hours of sleep is broken, and that causes the weekly hour. As long as it doesn't effect the other numbers it'll be okay, but I sure wish they'd give a longer warranty.

Are there any other CPAP machine that have a longer warranty? I don't want to have to invest in fixing it when I'm due a new machine from Medicare in August.

_________________
Machine: ResMed AirSense™ 10 AutoSet™ CPAP Machine with HumidAir™ Heated Humidifier
Mask: DreamWear Nasal CPAP Mask with Headgear
Additional Comments: My headgear varies (STILL!)
Resmed S9 with humidifier and in need of the right mask.

User avatar
DeeCPAP
Posts: 441
Joined: Mon Dec 14, 2015 11:32 pm

Re: Resmed S9 broken and out of warranty! What now?

Post by DeeCPAP » Sun May 07, 2017 1:36 pm

Julie wrote:If you're reading 'last' night's stats after noon today that could explain it - machines count from noon to noon (not 6 am), so that may be the problem.
I've used the machine for quite some time, so I know the data turns over at noon. Last night the DAILY HOURS showed I slept for over 20 hours (just for the DAY). In actuality, it was more like 6 or 7 total.

_________________
Machine: ResMed AirSense™ 10 AutoSet™ CPAP Machine with HumidAir™ Heated Humidifier
Mask: DreamWear Nasal CPAP Mask with Headgear
Additional Comments: My headgear varies (STILL!)
Last edited by DeeCPAP on Sun May 07, 2017 3:04 pm, edited 1 time in total.
Resmed S9 with humidifier and in need of the right mask.

User avatar
Uncle_Bob
Posts: 2777
Joined: Tue Feb 24, 2009 12:10 pm
Location: Arizona

Re: Resmed S9 broken and out of warranty! What now?

Post by Uncle_Bob » Sun May 07, 2017 1:48 pm

What the time on the machine? Is it accurate?

User avatar
Wulfman...
Posts: 6688
Joined: Mon Sep 01, 2014 6:41 pm
Location: Nearest fishing spot

Re: Resmed S9 broken and out of warranty! What now?

Post by Wulfman... » Sun May 07, 2017 2:07 pm

Contact http://www.acbio.com

They do certified repair for Philips/Respironics and ResMed (and other) equipment.

They're also a long-time member of this forum under the name "Accurate".

viewtopic/t117938/memberlist.php?mode=v ... ile&u=1002


Den

.
(5) REMstar Autos w/C-Flex & (6) REMstar Pro 2 CPAPs w/C-Flex - Pressure Setting = 14 cm.
"Passover" Humidification - ResMed Ultra Mirage FF - Encore Pro w/Card Reader & MyEncore software - Chiroflow pillow
User since 05/14/05

User avatar
DeeCPAP
Posts: 441
Joined: Mon Dec 14, 2015 11:32 pm

Re: Resmed S9 broken and out of warranty! What now?

Post by DeeCPAP » Sun May 07, 2017 2:59 pm

Uncle_Bob wrote:What the time on the machine? Is it accurate?
I just took 30 minutes rest with the machine on and it registered .5 hours so it's okay today.
I don't see the "time" as in a watch or clock.

I DID notice that the RAMP is set to 20 and I want NO ramp. I don't know whether the tech
did it accidentally or the machine somehow malfunctioned in that area, too.


Thanks.

_________________
Machine: ResMed AirSense™ 10 AutoSet™ CPAP Machine with HumidAir™ Heated Humidifier
Mask: DreamWear Nasal CPAP Mask with Headgear
Additional Comments: My headgear varies (STILL!)
Resmed S9 with humidifier and in need of the right mask.

User avatar
DeeCPAP
Posts: 441
Joined: Mon Dec 14, 2015 11:32 pm

Re: Resmed S9 broken and out of warranty! What now?

Post by DeeCPAP » Sun May 07, 2017 3:00 pm

Julie wrote:If you're reading 'last' night's stats after noon today that could explain it - machines count from noon to noon (not 6 am), so that may be the problem.
Yes, I understand noon to noon, but it said I used the machine 21 out of 24 hours just yesterday alone. That's WAY off.

_________________
Machine: ResMed AirSense™ 10 AutoSet™ CPAP Machine with HumidAir™ Heated Humidifier
Mask: DreamWear Nasal CPAP Mask with Headgear
Additional Comments: My headgear varies (STILL!)
Last edited by DeeCPAP on Sun May 07, 2017 4:49 pm, edited 1 time in total.
Resmed S9 with humidifier and in need of the right mask.

User avatar
DeeCPAP
Posts: 441
Joined: Mon Dec 14, 2015 11:32 pm

Re: Resmed S9 broken and out of warranty! What now?

Post by DeeCPAP » Sun May 07, 2017 3:02 pm

LSAT wrote:
DeeCPAP wrote:
LSAT wrote:Earlier today RogerSC suggested that you re-boot your machine...Did you do that? It was suggested that you download Sleepyhead...Did you do that?
Yes, I did. I slept 3 hours last night and it just added 3 hours to the 21 hours it had accumulated. The AHI was a fraction.

I've had SleepyHead for a while. I'll have to look at the stats after noontime because that's when last night's data will probably show up.
The S9 has a usage setting that can be day, week, or month. If that is what you are looking at, I'm guessing that the setting is week or month rather than day and the hours are accumulating night after night.
Thanks, but I know how to look for day, week, month, etc., so that's not the problem. I made some readjustments
and will now wait a day and see what happens.

_________________
Machine: ResMed AirSense™ 10 AutoSet™ CPAP Machine with HumidAir™ Heated Humidifier
Mask: DreamWear Nasal CPAP Mask with Headgear
Additional Comments: My headgear varies (STILL!)
Last edited by DeeCPAP on Sun May 07, 2017 4:48 pm, edited 1 time in total.
Resmed S9 with humidifier and in need of the right mask.

User avatar
DeeCPAP
Posts: 441
Joined: Mon Dec 14, 2015 11:32 pm

Re: Resmed S9 broken and out of warranty! What now?

Post by DeeCPAP » Sun May 07, 2017 3:09 pm

Wulfman... wrote:Contact http://www.acbio.com

They do certified repair for Philips/Respironics and ResMed (and other) equipment.
They're also a long-time member of this forum under the name "Accurate".
viewtopic/t117938/memberlist.php?mode=v ... ile&u=1002
Den

.
Thanks! I see one of the locations is in Florida. Is it the company or the forum user who
fixes the machines. Have any idea of the price if it's just a problem with the data collection?

_________________
Machine: ResMed AirSense™ 10 AutoSet™ CPAP Machine with HumidAir™ Heated Humidifier
Mask: DreamWear Nasal CPAP Mask with Headgear
Additional Comments: My headgear varies (STILL!)
Resmed S9 with humidifier and in need of the right mask.

User avatar
Wulfman...
Posts: 6688
Joined: Mon Sep 01, 2014 6:41 pm
Location: Nearest fishing spot

Re: Resmed S9 broken and out of warranty! What now?

Post by Wulfman... » Sun May 07, 2017 3:34 pm

DeeCPAP wrote:
Wulfman... wrote:Contact http://www.acbio.com

They do certified repair for Philips/Respironics and ResMed (and other) equipment.
They're also a long-time member of this forum under the name "Accurate".
viewtopic/t117938/memberlist.php?mode=v ... ile&u=1002
Den

.
Thanks! I see one of the locations is in Florida. Is it the company or the forum user who
fixes the machines. Have any idea of the price if it's just a problem with the data collection?
I have no ideas about costs. You'd have to contact them for information.
A number of forum members have had them work on their machines and have been quite happy with their work.
I haven't had them do any work for me, but I did buy some parts from them to fix one of my machines back in 2010.


Den

.
(5) REMstar Autos w/C-Flex & (6) REMstar Pro 2 CPAPs w/C-Flex - Pressure Setting = 14 cm.
"Passover" Humidification - ResMed Ultra Mirage FF - Encore Pro w/Card Reader & MyEncore software - Chiroflow pillow
User since 05/14/05

User avatar
Uncle_Bob
Posts: 2777
Joined: Tue Feb 24, 2009 12:10 pm
Location: Arizona

Re: Resmed S9 broken and out of warranty! What now?

Post by Uncle_Bob » Sun May 07, 2017 3:37 pm

Please check the date/time on your machine ....

http://www.apneaboard.com/wiki/index.ph ... _the_clock

Lean how to change options on your machine including ramp ...

http://osahelp.com/clinician-manuals/S9 ... al-eng.pdf