ASV Study Results

Post by RestlessinWoolley » Thu Mar 30, 2017 2:10 pm

My Doctor messaged me late yesterday and again today. ASV cleared my issues right up on the dedicated study. my AHI on the ASV unit was a 2.3, I slept 5.2 Hours. I had an excellent Dream Sleep, and entered Deep SLeep for 90 minutes, and my SpO2 didnt drop below a 95%, so I wont need supplemental O2 at this time but wants me to consult with my PCP about having Oxygen on hand for my headaches, as it almost cleared the headache instantly, but hopes after some time on the ASV the headaches stay away.

I am being prescribed a ResMed AirCurve 10 ASV, Humidifier, Heated Hose, and making sure I get an F20 Mask, or something better fitting with the higher pressures Ill have with the ASV. I have to return the DreamStation when I get the word to pick up the new machine, and get a drop off slip, as I am firing Lincare and going with a small Local Company called Sound Oxygen Supply, who wants my business, and knows already how my clinic works.. Machine will be set for full auto mode, but she is finalizing pressures now. She pulled the last week off the rental machine (old system one BiPap) and they were better than the CPAP but the ASV was far better, said only one central showed up. She also said to give it 2-3 weeks as the machine learns my breathing,before losing hope it isnt working..

_________________
MachineMask
Additional Comments: Resmed AirCurve 10 ASV
EPAP Min 6, Max 15, PS Min 6, Max 21, BURR 10, Press Max 25, BPM 11, Ti 1.2

"I hate the word 'Handicapped' and 'Disabled'. They imply that you are less than whole don't see myself that way at all". ~Aimee Mullins

klv329
Posts: 497
Joined: Mon Aug 31, 2015 6:43 pm
Location: Bradenton, Florida

Re: ASV Study Results

Post by klv329 » Thu Mar 30, 2017 6:16 pm

Good luck with the Aircurve 10 ASV. It is a good machine to treat central apneas and obstructive apneas.

I don't think the machine 'learns your breathing' over time and changes settings by itself. Maybe the physician will be modifying the ASVauto settings remotely?

You can use SleepyHead software to monitor your statistics.

With ASVauto wide open, it can be quite a ride.

Maybe the physician will adjust Min EPAP to Med EPAP reported statistics after a day or two and adjust Min PS and Max PS to your needs. The best settings for you will ultimately require your subjective feed back about how you feel during the day and what bothers you about the therapy, if anything.

Post some Sleepyhead charts and maybe some of the ASV experts can help out a bit.

Best of luck!

_________________
Machine: ResMed AirCurve 10 ASV Machine with Heated Humidifier
Additional Comments: Resmed Aircurve 10 ASVAUTO Min Epap 10.4, Max Epap 11.6, PS 1.6-12.0

User avatar
palerider
Posts: 32299
Joined: Wed Dec 16, 2009 5:43 pm
Location: Dallas(ish).

Re: ASV Study Results

Post by palerider » Thu Mar 30, 2017 6:32 pm

klv329 wrote:
RestlessinWoolley wrote:She also said to give it 2-3 weeks as the machine learns my breathing,before losing hope it isnt working..
I don't think the machine 'learns your breathing' over time and changes settings by itself. Maybe the physician will be modifying the ASVauto settings remotely?
you're quite right about that, the machine never learns anything, ever. all it does is respond to immediate circumstances based on it's programming and settings.

every time the machine starts up whether by button press or smart start, it's a whole new beginning.

_________________
Mask: Bleep DreamPort CPAP Mask Solution
Additional Comments: S9 VPAP Auto
Get OSCAR

Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.

User avatar
Pugsy
Posts: 65206
Joined: Thu May 14, 2009 9:31 am
Location: Missouri, USA

Re: ASV Study Results

Post by Pugsy » Thu Mar 30, 2017 6:51 pm

palerider wrote:the machine never learns anything, ever. all it does is respond to immediate circumstances based on it's programming and settings.
But it makes the patient feel better instead of saying "it may take a bit of time for your body to adjust and learn to trust the machine".

_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier
Additional Comments: Mask Bleep Eclipse https://bleepsleep.com/the-eclipse/
I may have to RISE but I refuse to SHINE.

User avatar
palerider
Posts: 32299
Joined: Wed Dec 16, 2009 5:43 pm
Location: Dallas(ish).

Re: ASV Study Results

Post by palerider » Thu Mar 30, 2017 8:05 pm

Pugsy wrote:
palerider wrote:the machine never learns anything, ever. all it does is respond to immediate circumstances based on it's programming and settings.
But it makes the patient feel better instead of saying "it may take a bit of time for your body to adjust and learn to trust the machine".
or, as a book I read recently put it "lies to children"

I'm betting that the next evo/revolution in cpap therapy will be a machine that *does* learn, and adjusts itself over time based on data from previous sessions. Resmed is nibbling at that with the 'for her' mode, where it raises the minimum pressure if there are multiple apneas within a period of time...

_________________
Mask: Bleep DreamPort CPAP Mask Solution
Additional Comments: S9 VPAP Auto
Get OSCAR

Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.

User avatar
RestlessinWoolley
Posts: 158
Joined: Sat Feb 11, 2017 2:03 am
Location: Sedro Woolley, Washington USA

Re: ASV Study Results

Post by RestlessinWoolley » Thu Mar 30, 2017 8:24 pm

Thanks for the clarification. I talked with the new DME saying they should hear tomorrow from the clinic. I asked a few questions and they also clarified saying it wont learn, but they will either make the adjustments or have the clinic make the adjustments during the initial startup month. It depends on how she writes the prescription. The new DME is a Seattle Based operation of 7 stores, and they have a dedicated in house team if Compliance people and "monitors".

Basically saying don't give up in the first few weeks as it may be a wild ride till they figure out where the machine needs set. They have a baseline from my test. I do remember having to cinch up my mask a bit and listen to it fart for a bit if I laid on my side, I feel my firm pillow will make it worse. So the DME is ordering me a Special Pillow as well that will cost me minimal. It's not a cutaway pillow but a pillow filled with corn husk apparently. Stays cool, and won't knock the mask around much. Say they sell a bunch of them.

Already I know in with a better company then Lincare, I am dreading the day I return the DreamStation, but will just say my clinic decided to deal with one local DME. And say my goodbyes as I skip out the door to my rig. The nice part is I can leave the rental Sys 1 at the DME and they will return it to my clinic free.

Hope to have it tomorrow or Monday. Gonna download the manual and see how different this beast is.

_________________
MachineMask
Additional Comments: Resmed AirCurve 10 ASV

User avatar
Pugsy
Posts: 65206
Joined: Thu May 14, 2009 9:31 am
Location: Missouri, USA

Re: ASV Study Results

Post by Pugsy » Thu Mar 30, 2017 9:11 pm

I anticipate that you will likely experience some rather high but brief pressure increases so planning the fitting for leak prevention at those higher pressures may be a bit of a challenge.
Part of "you adjusting to the machine"

That's how it breathes for you when you have a central...big burst of air but it doesn't stay up there very long. Hopefully you will sleep through it.

_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier
Additional Comments: Mask Bleep Eclipse https://bleepsleep.com/the-eclipse/
I may have to RISE but I refuse to SHINE.

User avatar
RestlessinWoolley
Posts: 158
Joined: Sat Feb 11, 2017 2:03 am
Location: Sedro Woolley, Washington USA

Re: ASV Study Results

Post by RestlessinWoolley » Thu Mar 30, 2017 9:23 pm

During the test apparently I had 24 "central" events i slept thru. I do remember getting punched in my face one time. The biggest part was at the beginning a tiny inhale resulted in a lung filling blast of air. But I adjusted quickly and knocked off fairly fast.

The only issue I have currently on the rental BiPap is now I'm falling asleep within 10 minutes but waking up at 430 and not able to get back to sleep without taking the mask off. No SD card in the unit to share or see what's up in Sleepyhead.

_________________
MachineMask
Additional Comments: Resmed AirCurve 10 ASV

User avatar
Pugsy
Posts: 65206
Joined: Thu May 14, 2009 9:31 am
Location: Missouri, USA

Re: ASV Study Results

Post by Pugsy » Thu Mar 30, 2017 9:32 pm

You could have used any SD card if you wanted to. Didn't have to be an official SD card from machine brand.

_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier
Additional Comments: Mask Bleep Eclipse https://bleepsleep.com/the-eclipse/
I may have to RISE but I refuse to SHINE.

User avatar
RestlessinWoolley
Posts: 158
Joined: Sat Feb 11, 2017 2:03 am
Location: Sedro Woolley, Washington USA

Re: ASV Study Results

Post by RestlessinWoolley » Thu Mar 30, 2017 9:53 pm

I never thought of grabbing the one in the Dream station but honestly I have a ton of micro SD cards and no adaptors and the only one SD card I had has about 14k songs on it. Lol

_________________
MachineMask
Additional Comments: Resmed AirCurve 10 ASV
EPAP Min 6, Max 15, PS Min 6, Max 21, BURR 10, Press Max 25, BPM 11, Ti 1.2

"I hate the word 'Handicapped' and 'Disabled'. They imply that you are less than whole don't see myself that way at all". ~Aimee Mullins

User avatar
Pugsy
Posts: 65206
Joined: Thu May 14, 2009 9:31 am
Location: Missouri, USA

Re: ASV Study Results

Post by Pugsy » Thu Mar 30, 2017 10:00 pm

That's okay...no need to worry about it now and you will be getting rid of this AirCurve soon anyway.
We could put a card in now and you would have some summary data available but it would be pretty much what is available on the LCD screen now.

Do remember when you get the AirCurve ASV...the SD card HAS to be in the machine during the night or you won't have any detailed data available.
The only thing it stores on the internal memory is the summary numbers...all the detailed stuff (which is what is needed for all those graphs) has to go directly onto the SD card.
So if you pull the card to use the software but sure to put the card back in the machine before you use it.
Can't tell you how many times I have gone to pull the SD card and realized it was still in the computer and of course it always happened on a night that I really wanted to see the details.

_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier
Additional Comments: Mask Bleep Eclipse https://bleepsleep.com/the-eclipse/
I may have to RISE but I refuse to SHINE.

User avatar
palerider
Posts: 32299
Joined: Wed Dec 16, 2009 5:43 pm
Location: Dallas(ish).

Re: ASV Study Results

Post by palerider » Thu Mar 30, 2017 10:23 pm

Pugsy wrote:Can't tell you how many times I have gone to pull the SD card and realized it was still in the computer and of course it always happened on a night that I really wanted to see the details.
maybe put a big orange dayglow tape flag on the SD card so you'll see that sticking out of the computer...

maybe we should go into business selling CPAP SD cards with "PUT ME BACK IN THE DANG CPAP" flags on them

_________________
Mask: Bleep DreamPort CPAP Mask Solution
Additional Comments: S9 VPAP Auto
Get OSCAR

Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.

User avatar
poppi2
Posts: 590
Joined: Thu Nov 18, 2010 2:54 pm
Location: Houston, near JSC

Re: ASV Study Results

Post by poppi2 » Thu Mar 30, 2017 11:11 pm

RestlessinWoolley wrote:.... but a pillow filled with corn husk apparently. Stays cool, and won't knock the mask around much. ...
I've not familiar with the corn husk filled pillow but love my buckwheat hull filled pillow. When side sleeping, I just move some hulls around to form a hole for my P10.

Please let us know what pillow you get. I'll have to Google the corn husk filled pillow. Always in search for the holy grail. Earl

_________________
Machine: AirSense 11 Autoset
Mask: AirFit™ P10 Nasal Pillow CPAP Mask with Headgear
Additional Comments: PAPCap, 3M Microfoam Surgical Tape, PoliGrip Strip, APAP 10.0 - 14.0 cm

User avatar
zoocrewphoto
Posts: 3732
Joined: Mon Apr 30, 2012 10:34 pm
Location: Seatac, WA

Re: ASV Study Results

Post by zoocrewphoto » Thu Mar 30, 2017 11:35 pm

palerider wrote:
Pugsy wrote:Can't tell you how many times I have gone to pull the SD card and realized it was still in the computer and of course it always happened on a night that I really wanted to see the details.
maybe put a big orange dayglow tape flag on the SD card so you'll see that sticking out of the computer...

maybe we should go into business selling CPAP SD cards with "PUT ME BACK IN THE DANG CPAP" flags on them

After doing this several times, I learned to leave a post-it note on the cpap machine. So, when I get ready for bed, I open the drawer, see the post-it, and retrieve the card from the computer.

_________________
Mask: Quattro™ FX Full Face CPAP Mask with Headgear
Humidifier: S9™ Series H5i™ Heated Humidifier with Climate Control
Additional Comments: Resmed S9 autoset pressure range 11-17
Who would have thought it would be this challenging to sleep and breathe at the same time?

User avatar
palerider
Posts: 32299
Joined: Wed Dec 16, 2009 5:43 pm
Location: Dallas(ish).

Re: ASV Study Results

Post by palerider » Thu Mar 30, 2017 11:41 pm

zoocrewphoto wrote:
palerider wrote:
Pugsy wrote:Can't tell you how many times I have gone to pull the SD card and realized it was still in the computer and of course it always happened on a night that I really wanted to see the details.
maybe put a big orange dayglow tape flag on the SD card so you'll see that sticking out of the computer...

maybe we should go into business selling CPAP SD cards with "PUT ME BACK IN THE DANG CPAP" flags on them

After doing this several times, I learned to leave a post-it note on the cpap machine. So, when I get ready for bed, I open the drawer, see the post-it, and retrieve the card from the computer.
I applaud you, but I don't think I'm organized enough to pull that off

_________________
Mask: Bleep DreamPort CPAP Mask Solution
Additional Comments: S9 VPAP Auto
Get OSCAR

Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.