ResMed Airsense 10 Setting Date Problem

General Discussion on any topic relating to CPAP and/or Sleep Apnea.
User avatar
Reuven Gruber
Posts: 39
Joined: Tue Sep 16, 2014 3:29 am
Location: Israel

ResMed Airsense 10 Setting Date Problem

Post by Reuven Gruber » Tue Nov 15, 2016 11:06 am

For some reason the date in my machine got set to tomorrow's date. When I try to set it back to today's date it gives me a message that says that I cant set a date in the past

HELP!! please

_________________
Machine: ResMed AirSense™ 10 AutoSet™ CPAP Machine with HumidAir™ Heated Humidifier
Mask: AirFit™ P10 Nasal Pillow CPAP Mask with Headgear
Additional Comments: Pressure limits 4.0 to 8.0, ResScan 5.7 Software, Ruby Chin Strap. Untreated AHI 32; With CPAP 1

User avatar
DreamStalker
Posts: 7503
Joined: Mon Aug 07, 2006 9:58 am
Location: Nowhere & Everywhere At Once

Re: ResMed Airsense 10 Setting Date Problem

Post by DreamStalker » Tue Nov 15, 2016 11:35 am

I'm sure someone more knowledgeable will jump in to help you get it all fixed right.

You may need to re-insert the SD card or possibly reformat or insert a new one.
President-pretender, J. Biden, said "the DNC has built the largest voter fraud organization in US history". Too bad they didn’t build the smartest voter fraud organization and got caught.

User avatar
JimW159
Posts: 804
Joined: Fri Jul 08, 2016 10:26 am
Location: Tarpon Springs, FL

Re: ResMed Airsense 10 Setting Date Problem

Post by JimW159 » Tue Nov 15, 2016 12:34 pm

Reuven Gruber wrote:For some reason the date in my machine got set to tomorrow's date. When I try to set it back to today's date it gives me a message that says that I cant set a date in the past
The way I understand it, once data is recorded one cannot reset the machine to a point before the time or date of that data. i.e. if you record a session that actually begins at 10PM on November 10, 2016 and ends at 6AM on November 11, 2016 but the machine reports the session as spanning November 11 to November 12 you cannot tell the machine to record a session that its record says occurred at the same or conflicting time the next day. The easy solution - if you have either a backup or travel machine - is to use that machine for a couple of days and reset your primary machine at a time when the change will not conflict with data already recorded - give it at least one full day to make sure of no conflict.

_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier
Mask: Brevida™ Nasal Pillow CPAP Mask with Headgear
Additional Comments: BU Mach AirSense 10 AutoSet - Mask = F&P Brevida BU ResMed P10 - Pressure = 10-17 On CPAP since 12/05/2008 Prior ID on CPAPTalk.com = JimW203

User avatar
palerider
Posts: 32299
Joined: Wed Dec 16, 2009 5:43 pm
Location: Dallas(ish).

Re: ResMed Airsense 10 Setting Date Problem

Post by palerider » Tue Nov 15, 2016 12:37 pm

either that, or erase the data then set the clock.

_________________
Mask: Bleep DreamPort CPAP Mask Solution
Additional Comments: S9 VPAP Auto
Get OSCAR

Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.

User avatar
JimW159
Posts: 804
Joined: Fri Jul 08, 2016 10:26 am
Location: Tarpon Springs, FL

Re: ResMed Airsense 10 Setting Date Problem

Post by JimW159 » Tue Nov 15, 2016 12:44 pm

palerider wrote:either that, or erase the data then set the clock.
As usual, a possibly simpler solution. In thinking about either suggestion raises a separate question: assuming the data has been read into SleepyHead, how will it deal with the conflicting report dates? As I see it, it may see two sessions spanning the same time period with the same dates.

_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier
Mask: Brevida™ Nasal Pillow CPAP Mask with Headgear
Additional Comments: BU Mach AirSense 10 AutoSet - Mask = F&P Brevida BU ResMed P10 - Pressure = 10-17 On CPAP since 12/05/2008 Prior ID on CPAPTalk.com = JimW203

User avatar
palerider
Posts: 32299
Joined: Wed Dec 16, 2009 5:43 pm
Location: Dallas(ish).

Re: ResMed Airsense 10 Setting Date Problem

Post by palerider » Tue Nov 15, 2016 12:55 pm

JimW159 wrote:
palerider wrote:either that, or erase the data then set the clock.
As usual, a possibly simpler solution. In thinking about either suggestion raises a separate question: assuming the data has been read into SleepyHead, how will it deal with the conflicting report dates? As I see it, it may see two sessions spanning the same time period with the same dates.
that's a good question, and I don't have a good answer for it...

_________________
Mask: Bleep DreamPort CPAP Mask Solution
Additional Comments: S9 VPAP Auto
Get OSCAR

Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.

User avatar
Oltremare
Posts: 187
Joined: Tue Dec 29, 2015 7:01 pm
Location: On the other side of the sea - Italy

Re: ResMed Airsense 10 Setting Date Problem

Post by Oltremare » Tue Nov 15, 2016 4:26 pm

But... what happens to our AIRSENSE 10?
How is this possible?
In a few days this is the second case!
viewtopic/t114257/Please-help-me-Wrong- ... toset.html

I had the same problem.
I held off the CPAP for the day but it was not enough.
I had to reset the CPAP and then I was able to correct the date.
I advise you not to waste time. You make a copy of the sd-card then go with reset.
Good luck

Oltremare

_________________
Machine: ResMed AirSense™ 10 AutoSet™ CPAP Machine with HumidAir™ Heated Humidifier
Mask: ResMed AirFit™ F20 Full Face CPAP Mask with Headgear
Additional Comments: Quattro FX FF mask > CMS50D+ pulse oximeter > Climate line > Settings: auto-CPAP pressure Min 11.00 Max 18.0 EPR 3
I hope you will forgive my language errors. To write in English I use a translator online. Thank you

User avatar
palerider
Posts: 32299
Joined: Wed Dec 16, 2009 5:43 pm
Location: Dallas(ish).

Re: ResMed Airsense 10 Setting Date Problem

Post by palerider » Tue Nov 15, 2016 4:31 pm

Oltremare wrote:But... what happens to our AIRSENSE 10?
How is this possible?
In a few days this is the second case!
mistakes get made in clock setting, that has nothing to do with the machine itself.

_________________
Mask: Bleep DreamPort CPAP Mask Solution
Additional Comments: S9 VPAP Auto
Get OSCAR

Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.

User avatar
Pugsy
Posts: 65332
Joined: Thu May 14, 2009 9:31 am
Location: Missouri, USA

Re: ResMed Airsense 10 Setting Date Problem

Post by Pugsy » Tue Nov 15, 2016 7:45 pm

How does this happen??? Human error...as much as I would like to blame it on the aliens...human error is the only way it happens.
Usually when we go to reset the clock for DST changes...we accidentally change the calendar too.
It's the only way it just up and happens.
I have done it twice....both with DST changes...and that's why I don't bother changing the clock anymore with the DST spring and fall changes.

I just erased the machine then reset the calendar.
I forget what happened to the SH data for the "duplicate" night...Don't know that I even looked.
You can always create a second profile to use after the erasing of the machine and calendar reset.

_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier
Additional Comments: Mask Bleep Eclipse https://bleepsleep.com/the-eclipse/
I may have to RISE but I refuse to SHINE.

User avatar
Oltremare
Posts: 187
Joined: Tue Dec 29, 2015 7:01 pm
Location: On the other side of the sea - Italy

Re: ResMed Airsense 10 Setting Date Problem

Post by Oltremare » Wed Nov 16, 2016 12:24 am

As always thank you for your expertise. I remember correct daylight saving time and surely I made a mistake.
There is no other possibility!
Thanks to you, I continue to learn and to grow.
Pale rider and Pugsy, a virtual hug to you.
Thank you so much.

Oltremare

_________________
Machine: ResMed AirSense™ 10 AutoSet™ CPAP Machine with HumidAir™ Heated Humidifier
Mask: ResMed AirFit™ F20 Full Face CPAP Mask with Headgear
Additional Comments: Quattro FX FF mask > CMS50D+ pulse oximeter > Climate line > Settings: auto-CPAP pressure Min 11.00 Max 18.0 EPR 3
I hope you will forgive my language errors. To write in English I use a translator online. Thank you

User avatar
Reuven Gruber
Posts: 39
Joined: Tue Sep 16, 2014 3:29 am
Location: Israel

Re: ResMed Airsense 10 Setting Date Problem

Post by Reuven Gruber » Wed Nov 16, 2016 1:22 pm

Thanks for the advice - Here is what I did:

With the SD card in the machine I entered the clinical menu and selected "Erase Data" and then confirmed that I actually wanted to do it. The machine erased the data and reformatted the SD card. I was then able to enter the correct date and time
I am pleased to report that this operation did not erase my settings (pressures, units of measure etc)

I did not loose historical data since I had been uploading data from the SD card in my machine to the ResScan application on my laptop (ResScan is the ResMed equivalent of SleepyHead)

_________________
Machine: ResMed AirSense™ 10 AutoSet™ CPAP Machine with HumidAir™ Heated Humidifier
Mask: AirFit™ P10 Nasal Pillow CPAP Mask with Headgear
Additional Comments: Pressure limits 4.0 to 8.0, ResScan 5.7 Software, Ruby Chin Strap. Untreated AHI 32; With CPAP 1

WindRider
Posts: 1
Joined: Fri Jan 24, 2020 9:22 pm

Re: ResMed Airsense 10 Setting Date Problem

Post by WindRider » Fri Jan 24, 2020 9:51 pm

I know this thread is old but I felt the need to chime in here since this thread is one of the only ones I found with information about this issue and others may look here for a solution.

There is no need to reset or reset or erase all your CPAP data to set the date back! I just successfully set my date back on my ResMed Airsense 10 by using a different recommended approach.

You just need to keep setting the time of the machine back as much as the machine will allow you before going to bed. You will end up with 2 sleep sessions in one day. You will eventually reach a point where the machine will allow you to set the time back to midnight (00:00). At that point you will be able to set the time to midnight or the correct time afterwards, with the dates aligned and the clocks aligned again. Minor hour adjustments can always be made after noon and before using the machine for the night moving forwards.

This might sound really complicated but to keep things simple, just keep moving your machine clock back progressively each day until you reach a point at which you can align the clocks on the same day. It might take 2-3 days but trust that you will get to a point where you will be able to set the clock correctly and you won't have to reset or erase your data. This won't cause any issues for SleepyHead or Oscar. There will be a day with 2 sessions but that's no big deal. You get extra credit for lots of sleep that day and all the other days will be ok.

Hope this helps some people,
WindRider

User avatar
zorki1c
Posts: 253
Joined: Mon Jun 10, 2013 5:24 pm
Location: Intermountain West

Re: ResMed Airsense 10 Setting Date Problem

Post by zorki1c » Fri Jan 24, 2020 11:04 pm

Made the mistake of resetting the time on the machine when I flew from the Mainland to Hawaii. Then I couldn’t change back. Won’t do that again.

_________________
MachineMask
Additional Comments: Use the P30i in the cooler months and the P10 the rest of the time.