Lots of events at max pressure

General Discussion on any topic relating to CPAP and/or Sleep Apnea.
zzrrrp
Posts: 17
Joined: Fri May 15, 2015 6:59 am

Re: Lots of events at max pressure

Post by zzrrrp » Sat May 16, 2015 3:21 pm

Wulfman... wrote:Do you have a copy of your sleep study? Did it specify an optimal pressure setting?
Was there any mention of Central Apneas in it?
How did you end up with a range of pressures?
Your minimum pressure does seem a bit low to me, but depending on what your sleep study (titration) says, and how you happened to get that pressure range prescribed is what I have questions about.


Den

.
I do have a copy of the study. Due to a bit of stage fright, my total sleep time at the study was 3.2 hours, 13 minutes of which were REM. I did not have a split study; no titration was performed. The basis of my diagnosis was an extrapolated AHI of 46.2 during REM (an effective AHI of roughly 11, assuming 25% REM sleep). No recorded central apneas.
Machine: ResMed AirSense 10 AutoSet, APAP
Mask: Quattro Air full face mask

User avatar
Wulfman...
Posts: 6688
Joined: Mon Sep 01, 2014 6:41 pm
Location: Nearest fishing spot

Re: Lots of events at max pressure

Post by Wulfman... » Sat May 16, 2015 3:32 pm

zzrrrp wrote:
Wulfman... wrote:Do you have a copy of your sleep study? Did it specify an optimal pressure setting?
Was there any mention of Central Apneas in it?
How did you end up with a range of pressures?
Your minimum pressure does seem a bit low to me, but depending on what your sleep study (titration) says, and how you happened to get that pressure range prescribed is what I have questions about.


Den

.
I do have a copy of the study. Due to a bit of stage fright, my total sleep time at the study was 3.2 hours, 13 minutes of which were REM. I did not have a split study; no titration was performed. The basis of my diagnosis was an extrapolated AHI of 46.2 during REM (an effective AHI of roughly 11, assuming 25% REM sleep). No recorded central apneas.
OK.
Due to what I see in your nightly report(s), I would recommend switching to straight pressure (somewhere between 8 an 10 for now) and see how that goes. Your Flow Limitations are driving your pressures erratically during the night and that could be very disrupting to your sleep (quality).


Den

.
(5) REMstar Autos w/C-Flex & (6) REMstar Pro 2 CPAPs w/C-Flex - Pressure Setting = 14 cm.
"Passover" Humidification - ResMed Ultra Mirage FF - Encore Pro w/Card Reader & MyEncore software - Chiroflow pillow
User since 05/14/05

User avatar
VVV
Posts: 532
Joined: Mon Mar 14, 2011 11:28 am

Re: Lots of events at max pressure

Post by VVV » Sat May 16, 2015 3:50 pm

zzrrrp wrote:13 minutes of which were REM
Typically, pressure requirement in REM are higher than in the other sleep stages. 13 minutes is not a lot of time to do a proper titration that accounts for REM periods. This suggests that they titrated your pressure settings too low.

I think if you set your pressure at min 8, max 20, you will start to get a better idea of your pressure needs.
Wulfman... wrote:I would recommend switching to straight pressure
That's exactly bassackwards Wulfman. When you are trying to find an ideal pressure setting at home with your CPAP, you start with a min/max range. Trying to do this with straight pressures will take more, maybe many more, nights of experimenting. You may actually never find the best setting using straight pressures in your experiments.

I know you operate off a paradigm of, "straight pressure is best for me, so it must be best for everyone else".

The popularity of APAPs belies your paradigm.
.....................................V

User avatar
VVV
Posts: 532
Joined: Mon Mar 14, 2011 11:28 am

Re: Lots of events at max pressure

Post by VVV » Sat May 16, 2015 3:53 pm

zzrrrp wrote:zzrrrp
BTW, Congratulations on hitting the ground running with your therapy and already posting good screenshots here and understanding a good bit about what is going on.
.....................................V

User avatar
Wulfman...
Posts: 6688
Joined: Mon Sep 01, 2014 6:41 pm
Location: Nearest fishing spot

Re: Lots of events at max pressure

Post by Wulfman... » Sat May 16, 2015 4:07 pm

VVV wrote:
zzrrrp wrote:13 minutes of which were REM
Typically, pressure requirement in REM are higher than in the other sleep stages. 13 minutes is not a lot of time to do a proper titration that accounts for REM periods. This suggests that they titrated your pressure settings too low.

I think if you set your pressure at min 8, max 20, you will start to get a better idea of your pressure needs.
Wulfman... wrote:I would recommend switching to straight pressure
That's exactly bassackwards Wulfman. When you are trying to find an ideal pressure setting at home with your CPAP, you start with a min/max range. Trying to do this with straight pressures will take more, maybe many more, nights of experimenting. You may actually never find the best setting using straight pressures in your experiments.

I know you operate off a paradigm of, "straight pressure is best for me, so it must be best for everyone else".

The popularity of APAPs belies your paradigm.
Considering the OP's breathing patterns, using a range of pressures will be a disaster. And, I've witnessed enough of those on the forum to know what I'm talking about.
Ranges of pressures are not for everybody and optimal pressures can be determined using straight pressures.

AND, the "popularity" of APAPs doesn't have a damn thing to do with it. I ALWAYS recommend a person getting an APAP if possible. The ultimate proof is in "SLEEP QUALITY"!


Den

.
(5) REMstar Autos w/C-Flex & (6) REMstar Pro 2 CPAPs w/C-Flex - Pressure Setting = 14 cm.
"Passover" Humidification - ResMed Ultra Mirage FF - Encore Pro w/Card Reader & MyEncore software - Chiroflow pillow
User since 05/14/05

User avatar
palerider
Posts: 32299
Joined: Wed Dec 16, 2009 5:43 pm
Location: Dallas(ish).

Re: Lots of events at max pressure

Post by palerider » Sat May 16, 2015 5:44 pm

Wulfman... wrote:Edit: The thing that jumps out at me are the Flow Limitations. Increasing the upper pressure may be problematic with them.
yes, it might make them go away and give the OP better sleep.

_________________
Mask: Bleep DreamPort CPAP Mask Solution
Additional Comments: S9 VPAP Auto
Get OSCAR

Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.

User avatar
palerider
Posts: 32299
Joined: Wed Dec 16, 2009 5:43 pm
Location: Dallas(ish).

Re: Lots of events at max pressure

Post by palerider » Sat May 16, 2015 5:47 pm

Wulfman... wrote:Considering the OP's breathing patterns, using a range of pressures will be a disaster.
if, as you *ASSUME* the OP will be bothered by the machine behaving as it's designed, and modifying the pressure to suit the situation.

not everybody is bothered by that. possibly not even a majority.

the OP certainly didn't complain about it.

_________________
Mask: Bleep DreamPort CPAP Mask Solution
Additional Comments: S9 VPAP Auto
Get OSCAR

Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.

User avatar
Wulfman...
Posts: 6688
Joined: Mon Sep 01, 2014 6:41 pm
Location: Nearest fishing spot

Re: Lots of events at max pressure

Post by Wulfman... » Sat May 16, 2015 6:05 pm

palerider wrote:
Wulfman... wrote:Considering the OP's breathing patterns, using a range of pressures will be a disaster.
if, as you *ASSUME* the OP will be bothered by the machine behaving as it's designed, and modifying the pressure to suit the situation.

not everybody is bothered by that. possibly not even a majority.

the OP certainly didn't complain about it.
But, obviously he was concerned about what he saw in his report. SOME things are just obvious and need to be pointed out.

It's his therapy, he can set the machine however he wishes.

One thing I do know is that there have been numerous situations over the years where I've ended up helping users OFF of the forum getting their therapy straightened out after the "pressure range" advocates have gotten them thoroughly messed up.


Den

.
(5) REMstar Autos w/C-Flex & (6) REMstar Pro 2 CPAPs w/C-Flex - Pressure Setting = 14 cm.
"Passover" Humidification - ResMed Ultra Mirage FF - Encore Pro w/Card Reader & MyEncore software - Chiroflow pillow
User since 05/14/05

User avatar
Hang Fire
Posts: 469
Joined: Mon Nov 04, 2013 8:03 pm
Location: Raleigh, NC

Re: Lots of events at max pressure

Post by Hang Fire » Sat May 16, 2015 6:22 pm

Wulfman... wrote:the "pressure range" advocates
The doctors that set the range at 4-20 I would guess.

User avatar
Hang Fire
Posts: 469
Joined: Mon Nov 04, 2013 8:03 pm
Location: Raleigh, NC

Re: Lots of events at max pressure

Post by Hang Fire » Sat May 16, 2015 6:23 pm

palerider wrote:
Wulfman... wrote:Edit: The thing that jumps out at me are the Flow Limitations. Increasing the upper pressure may be problematic with them.
yes, it might make them go away and give the OP better sleep.
+1

User avatar
Wulfman...
Posts: 6688
Joined: Mon Sep 01, 2014 6:41 pm
Location: Nearest fishing spot

Re: Lots of events at max pressure

Post by Wulfman... » Sat May 16, 2015 6:31 pm

palerider wrote:
Wulfman... wrote:Edit: The thing that jumps out at me are the Flow Limitations. Increasing the upper pressure may be problematic with them.
yes, it might make them go away and give the OP better sleep.
Or "worse" sleep.

But, that's why I was asking the OP about the pressure range and whether there may have been any Centrals noted in the sleep lab report......to limit the upper pressure range.


Den

.
(5) REMstar Autos w/C-Flex & (6) REMstar Pro 2 CPAPs w/C-Flex - Pressure Setting = 14 cm.
"Passover" Humidification - ResMed Ultra Mirage FF - Encore Pro w/Card Reader & MyEncore software - Chiroflow pillow
User since 05/14/05

User avatar
Wulfman...
Posts: 6688
Joined: Mon Sep 01, 2014 6:41 pm
Location: Nearest fishing spot

Re: Lots of events at max pressure

Post by Wulfman... » Sat May 16, 2015 6:32 pm

Hang Fire wrote:
Wulfman... wrote:the "pressure range" advocates
The doctors that set the range at 4-20 I would guess.
I was referring to the ones on the forum.


Den

.
(5) REMstar Autos w/C-Flex & (6) REMstar Pro 2 CPAPs w/C-Flex - Pressure Setting = 14 cm.
"Passover" Humidification - ResMed Ultra Mirage FF - Encore Pro w/Card Reader & MyEncore software - Chiroflow pillow
User since 05/14/05

User avatar
Hang Fire
Posts: 469
Joined: Mon Nov 04, 2013 8:03 pm
Location: Raleigh, NC

Re: Lots of events at max pressure

Post by Hang Fire » Sat May 16, 2015 6:35 pm

Wulfman... wrote:I've ended up helping users OFF of the forum getting their therapy straightened out after the "pressure range" advocates have gotten them thoroughly messed up.
Well, I am sure they will soon chime in by the dozens and tell us how you got them on a straight CPAP pressure and their lives thereafter blossomed.

User avatar
Julie
Posts: 20051
Joined: Tue Feb 28, 2006 12:58 pm

Post by Julie » Sat May 16, 2015 7:16 pm

Den - sometimes you can't win for trying!

User avatar
Hang Fire
Posts: 469
Joined: Mon Nov 04, 2013 8:03 pm
Location: Raleigh, NC

Re:

Post by Hang Fire » Sat May 16, 2015 7:21 pm

Julie wrote:Den - sometimes you can't win for trying!
Unfortunately, Den thinks this is about winning.

It is not. It's about helping the OP optimize his therapy.