Lots of events at max pressure
Lots of events at max pressure
Hi there, I'm 9 nights into my CPAP therapy and having a great time. Typical nights have been 1-2 AHI, but last night I had an hour-long (515-615am) cluster of hypopneas even though my machine maxed out my prescribed pressure (5-10 cm H2O). Is that an indication that I should start tinkering with the allowed pressure range, or should I stay the course? Please advise if other data are pertinent.


Machine: ResMed AirSense 10 AutoSet, APAP
Mask: Quattro Air full face mask
Mask: Quattro Air full face mask
Re: Lots of events at max pressure
One night is not enough to diagnose or change anything... because any thing could be responsible and you need more time for it to be meaningful.
Re: Lots of events at max pressure
You still are under 5 so I wouldn't change anything just yet.
I wish you well...
I wish you well...
_________________
Machine: DreamStation BiPAP® Auto Machine |
Humidifier: DreamStation Heated Humidifier |
Additional Comments: Settings are IPap 23 EPap 19 |
Re: Lots of events at max pressure
I agree...you said your typical night is 1-2...don't be so anxious to make changes.
_________________
Machine: ResMed AirSense™ 10 AutoSet™ CPAP Machine with HumidAir™ Heated Humidifier |
Mask: Fisher & Paykel Vitera Full Face Mask with Headgear (S, M, or L Cushion) |
Additional Comments: Back up is a new AS10. |
- ChicagoGranny
- Posts: 15314
- Joined: Sun Jan 29, 2012 1:43 pm
- Location: USA
Re: Lots of events at max pressure
If it were me, I would take the minimum to 8 and the max to 20. The machine is running very little of the time under 8 anyway. With the max set at 20, the machine will still not go above 10 unless you are having trouble breathing and need the extra pressure. And I do think you need it at times, probably when you are in REM.zzrrrp wrote:5-10 cm H2O
If you make your screenshots and post your images like this, we can tell better what is going on - viewtopic.php?f=1&t=104876&st=0&sk=t&sd=a&start=30
Bull donkers.Julie wrote:One night is not enough to diagnose or change anything... because any thing could be responsible and you need more time for it to be meaningful.
"It's not the number of breaths we take, it's the number of moments that take our breath away."
Cuando cuentes cuentos, cuenta cuántas cuentos cuentas.
Cuando cuentes cuentos, cuenta cuántas cuentos cuentas.
- Wulfman...
- Posts: 6688
- Joined: Mon Sep 01, 2014 6:41 pm
- Location: Nearest fishing spot
Re: Lots of events at max pressure
It could be "positional" events. Your airway is getting partially closed without actually shutting off your air.
I agree with Julie. Give it more time to observe and see if those situations occur more frequently or if it's just a coincidence.
I would NOT be increasing the maximum pressure.......at this point.
Den
.
I agree with Julie. Give it more time to observe and see if those situations occur more frequently or if it's just a coincidence.
I would NOT be increasing the maximum pressure.......at this point.
Den
.
(5) REMstar Autos w/C-Flex & (6) REMstar Pro 2 CPAPs w/C-Flex - Pressure Setting = 14 cm.
"Passover" Humidification - ResMed Ultra Mirage FF - Encore Pro w/Card Reader & MyEncore software - Chiroflow pillow
User since 05/14/05
"Passover" Humidification - ResMed Ultra Mirage FF - Encore Pro w/Card Reader & MyEncore software - Chiroflow pillow
User since 05/14/05
Re: Lots of events at max pressure
5 is just good enough for doctors... blech.Bill44133 wrote:You still are under 5 so I wouldn't change anything just yet.
I wish you well...
Get OSCAR
Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.
Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.
Re: Lots of events at max pressure
+1ChicagoGranny wrote:If it were me, I would take the minimum to 8 and the max to 20. The machine is running very little of the time under 8 anyway. With the max set at 20, the machine will still not go above 10 unless you are having trouble breathing and need the extra pressure. And I do think you need it at times, probably when you are in REM.zzrrrp wrote:5-10 cm H2O
If you make your screenshots and post your images like this, we can tell better what is going on - viewtopic.php?f=1&t=104876&st=0&sk=t&sd=a&start=30
Get OSCAR
Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.
Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.
- grayghost4
- Posts: 1554
- Joined: Wed Aug 20, 2014 6:52 pm
- Location: Norther Illinois
- Contact:
Re: Lots of events at max pressure
CG "Bull donkers."
Is that a new Nick name
Is that a new Nick name
If you're not part of the solution you're just scumming up the bottom of the beaker!
Get the Clinicians manual here : http://apneaboard.com/adjust-cpap-press ... tup-manual
Get the Clinicians manual here : http://apneaboard.com/adjust-cpap-press ... tup-manual
Re: Lots of events at max pressure
Here's a SleepyHead readout with flow limitation and leak rate. Seems like the prudent thing is to keep watching for clusters like these and slowly increase the pressure range should they recur.
I should have mentioned this before: I did tighten my full face mask last night as a result of a mask fit test. Could the mask pulling my lower jaw back explain some of the extra events?

I should have mentioned this before: I did tighten my full face mask last night as a result of a mask fit test. Could the mask pulling my lower jaw back explain some of the extra events?

Machine: ResMed AirSense 10 AutoSet, APAP
Mask: Quattro Air full face mask
Mask: Quattro Air full face mask
Re: Lots of events at max pressure
+2palerider wrote:+1ChicagoGranny wrote:If it were me, I would take the minimum to 8 and the max to 20. The machine is running very little of the time under 8 anyway. With the max set at 20, the machine will still not go above 10 unless you are having trouble breathing and need the extra pressure. And I do think you need it at times, probably when you are in REM.zzrrrp wrote:5-10 cm H2O
If you make your screenshots and post your images like this, we can tell better what is going on - viewtopic.php?f=1&t=104876&st=0&sk=t&sd=a&start=30
Last edited by VVV on Sat May 16, 2015 3:10 pm, edited 1 time in total.
.....................................V
Re: Lots of events at max pressure
Your mask shouldn't have to fit too tightly - it crushes the thin cushion (which needs to inflate) and can create new smaller leaks... your face changes when lying down of course too, so trying on others needs to be done lying down. I agree your min. pressure should be moved up and also wonder if you're in the habit of ending up with your head down on your chest when sleeping - that will cut off your air (and cpap). A fix for that would be to wear a soft cervical collar (not the big stiff thing for whiplash) that keeps your head up, airway open and mouth closed a lot better... might be worth a try.
Re: Lots of events at max pressure
The pressure is bumping up to 10, not just in the late morning, but almost the entire night. No need to limit the machine when more pressure is needed.Wulfman... wrote:It could be "positional" events. Your airway is getting partially closed without actually shutting off your air.
I agree with Julie. Give it more time to observe and see if those situations occur more frequently or if it's just a coincidence.
I would NOT be increasing the maximum pressure.......at this point.
Den
.
Plus, increasing the minimum from 5 to 8 is prudent because, as Granny said, his machine is running above 8 almost the entire night.
.....................................V
- Wulfman...
- Posts: 6688
- Joined: Mon Sep 01, 2014 6:41 pm
- Location: Nearest fishing spot
Re: Lots of events at max pressure
Do you have a copy of your sleep study? Did it specify an optimal pressure setting?
Was there any mention of Central Apneas in it?
How did you end up with a range of pressures?
Your minimum pressure does seem a bit low to me, but depending on what your sleep study (titration) says, and how you happened to get that pressure range prescribed is what I have questions about.
Den
.
Was there any mention of Central Apneas in it?
How did you end up with a range of pressures?
Your minimum pressure does seem a bit low to me, but depending on what your sleep study (titration) says, and how you happened to get that pressure range prescribed is what I have questions about.
Den
.
(5) REMstar Autos w/C-Flex & (6) REMstar Pro 2 CPAPs w/C-Flex - Pressure Setting = 14 cm.
"Passover" Humidification - ResMed Ultra Mirage FF - Encore Pro w/Card Reader & MyEncore software - Chiroflow pillow
User since 05/14/05
"Passover" Humidification - ResMed Ultra Mirage FF - Encore Pro w/Card Reader & MyEncore software - Chiroflow pillow
User since 05/14/05
- Wulfman...
- Posts: 6688
- Joined: Mon Sep 01, 2014 6:41 pm
- Location: Nearest fishing spot
Re: Lots of events at max pressure
There COULD be.......depending on what was in the sleep study (titration).VVV wrote:The pressure is bumping up to 10, not just in the late morning, but almost the entire night. No need to limit the machine when more pressure is needed.Wulfman... wrote:It could be "positional" events. Your airway is getting partially closed without actually shutting off your air.
I agree with Julie. Give it more time to observe and see if those situations occur more frequently or if it's just a coincidence.
I would NOT be increasing the maximum pressure.......at this point.
Den
.
Plus, increasing the minimum from 5 to 8 is prudent because, as Granny said, his machine is running above 8 almost the entire night.
Read my previous post. I also think it would be prudent to increase the minimum pressure (or use a straight pressure), but I'm curious as to why that pressure range was prescribed.
Edit: The thing that jumps out at me are the Flow Limitations. Increasing the upper pressure may be problematic with them.
Den
.
(5) REMstar Autos w/C-Flex & (6) REMstar Pro 2 CPAPs w/C-Flex - Pressure Setting = 14 cm.
"Passover" Humidification - ResMed Ultra Mirage FF - Encore Pro w/Card Reader & MyEncore software - Chiroflow pillow
User since 05/14/05
"Passover" Humidification - ResMed Ultra Mirage FF - Encore Pro w/Card Reader & MyEncore software - Chiroflow pillow