ResMedAirSense Autoset 10 and on/off function

General Discussion on any topic relating to CPAP and/or Sleep Apnea.
lindalrc
Posts: 77
Joined: Fri Mar 27, 2009 1:34 pm
Location: Maine

ResMedAirSense Autoset 10 and on/off function

Post by lindalrc » Thu May 07, 2015 8:03 am

My husband and I both got these new machines about 6 weeks ago. We have both been on PAP for years and do fine. His pressure is 16 and mine is 10. Both are set for auto-off and auto-on. Mine is just perfect. His often will not turn off or will run for 30 seconds or more after removal. The DME consultant, who has been extremely helpful, says that this happens with ResMeds at higher pressures. Any comments/suggestions?? (Not that just turning the machine off is such a huge deal ) Thanks.

User avatar
cpapernewbie
Posts: 685
Joined: Fri Oct 13, 2006 5:53 pm

Re: ResMedAirSense Autoset 10 and on/off function

Post by cpapernewbie » Thu May 07, 2015 8:32 am

When turning off, the 10 will run for minutes on low pressure. Even if you manually you have push the off/on button. The objective is to dry out the hose if condensation occured.
I do not like this feature and to turn this device off completely, I just push the button longer. The blower will stop completely.

_________________
Machine: DreamStation Auto CPAP Machine
Mask: AirFit™ P10 Nasal Pillow CPAP Mask with Headgear
Humidifier: DreamStation Heated Humidifier
Additional Comments: CPAP history: dumb tank, auto, PR M, PR System 1, PR BIPAP, PR System 1 model 60, Resmed S9, Resmed S10, Dreamstation
Resmed airsense 10

User avatar
palerider
Posts: 32299
Joined: Wed Dec 16, 2009 5:43 pm
Location: Dallas(ish).

Re: ResMedAirSense Autoset 10 and on/off function

Post by palerider » Thu May 07, 2015 9:29 am

lindalrc wrote:My husband and I both got these new machines about 6 weeks ago. We have both been on PAP for years and do fine. His pressure is 16 and mine is 10. Both are set for auto-off and auto-on. Mine is just perfect. His often will not turn off or will run for 30 seconds or more after removal. The DME consultant, who has been extremely helpful, says that this happens with ResMeds at higher pressures. Any comments/suggestions?? (Not that just turning the machine off is such a huge deal ) Thanks.
make sure your hubby's mask type and hose type (if not using a climateline hose) is set correctly. I've seen smartstart problems if the mask type is set wrong.

the DME is, as often happens, wrong. it's documented to sometimes happen at very low pressures.

_________________
Mask: Bleep DreamPort CPAP Mask Solution
Additional Comments: S9 VPAP Auto
Get OSCAR

Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.

User avatar
billbolton
Posts: 2264
Joined: Wed Jun 07, 2006 7:46 pm
Location: Sydney, Australia

Re: ResMedAirSense Autoset 10 and on/off function

Post by billbolton » Thu May 07, 2015 6:59 pm

cpapernewbie wrote: The objective is to dry out the hose if condensation occured.
No, the objective is to prevent any condensation entering the blower from the humidifier/hose, by having a small positive air pressure entering the humidifier and hose. Any drying in the hose is secondary side effect, but is not the objective.

Cheers,
Bill

_________________
MachineMask
Additional Comments: Airmini, Medistrom Pilot 24, CMS 60C Pulse Oximeter, ResScan 6

User avatar
Papit
Posts: 844
Joined: Sun Dec 18, 2011 4:47 pm
Location: USA CA

Re: ResMedAirSense Autoset 10 and on/off function

Post by Papit » Fri May 08, 2015 7:29 pm

billbolton wrote:
cpapernewbie wrote: The objective is to dry out the hose if condensation occured.
No, the objective is to prevent any condensation entering the blower from the humidifier/hose, by having a small positive air pressure entering the humidifier and hose. Any drying in the hose is secondary side effect, but is not the objective.

Cheers,
Bill
Protection for the blower makes sense -- good design. I also value that a lot to dry the air tubes and mask for good sanitation. I use the SmartStart (auto on/off) feature. The slow-blower, that kicks in following the few seconds of hurricane when I remove my mask, lasts for quite a few minutes and I like that. Is this a design feature for all the S9's? In what operating configurations Is the slow-blower supposed to activate? --only when the humidifier is connected? --only when SmartStart is being used? etc. Do you have a handle on the various kick-in modes? I have an AutoSet that doesn't seem to utilize this feature the way my VPAP Adapt does. Might that be a malfunction?

_________________
Mask: DreamWear Nasal CPAP Mask with Headgear
Additional Comments: Machine: AirCurve 10 ASV (37043), Software:ResScan 5.7.0.9477, SleepyHead V1.00BETA2, Oximeter:CMS-50i
Machine: AirCurve 10 ASV, Mask: AirFit N30i
PulseOx Data-transfer to OSCAR4-23-12http://tinyurl.com/nzd64gu
Wireless SD Card Data-transfer to OSCAR 8-14-15http://tiny.cc/z1kv8x

User avatar
Krelvin
Posts: 1977
Joined: Tue Jun 06, 2006 5:23 pm
Location: Metro Phx Area - Dry Heat!

Re: ResMedAirSense Autoset 10 and on/off function

Post by Krelvin » Fri May 08, 2015 7:34 pm

Papit wrote:Is this a design feature for all the S9's? In what operating configurations Is the slow-blower supposed to activate? --only when the humidifier is connected? --only when SmartStart is being used? etc. Do you have a handle on the various kick-in modes? I have an AutoSet that doesn't seem to utilize this feature the way my VPAP Adapt does. Might that be a malfunction?
It is in the manual for the Humidifier...
In order to assist the heater plate in cooling, your S9 device will continue to blow air for up to an hour after treatment has stopped. However, you can unplug the device from the power outlet at any time and allow the heater plate to cool without air flow.
Current Settings PS 4.0 over 10.6-18.0 (cmH2O) - Resmed S9 VPAP Auto w/h5i Humidifier - Quattro Air FFM
TNET Sleep Resource Pages - CPAP Machine Database
Put your equip in your Signature - SleepyHead v1.0.0-beta-1
Kevin... alias Krelvin

Cpappmomma

Re: ResMedAirSense Autoset 10 and on/off function

Post by Cpappmomma » Fri May 08, 2015 8:57 pm

Is the original question referring to the auto on/auto off (smart start) feature of the machine (turns off automatically when you remove mask) or the on/off (hitting the button) feature?

Guest

Re: ResMedAirSense Autoset 10 and on/off function

Post by Guest » Fri May 08, 2015 10:33 pm

First two days I had my Autoset I used the Auto off, and it would run at high pressure for a couple of seconds and shut off. Sleepyhead would report it as a leak and max the leak chart out at 120. I started turning off the machine manually because I didn't like the data getting messed up like that.

User avatar
bobmc
Posts: 49
Joined: Wed Apr 01, 2015 6:11 pm
Location: Hampton Roads, VA

Re: ResMedAirSense Autoset 10 and on/off function

Post by bobmc » Fri May 08, 2015 10:35 pm

Ohhh...kay...wasn't logged in for whatever reason. Oh, well...

User avatar
palerider
Posts: 32299
Joined: Wed Dec 16, 2009 5:43 pm
Location: Dallas(ish).

Re: ResMedAirSense Autoset 10 and on/off function

Post by palerider » Fri May 08, 2015 10:39 pm

BobMC wrote:First two days I had my Autoset I used the Auto off, and it would run at high pressure for a couple of seconds and shut off. Sleepyhead would report it as a leak and max the leak chart out at 120. I started turning off the machine manually because I didn't like the data getting messed up like that.
not that it really matters, but when you take the mask off, the machine runs at very low pressure, but high flow because it's trying to maintain whatever your set pressure is.

to have pressure, you have to have some restriction to the flow, like what the mask vents give you.

now, admittedly, the motor ramps up to full speed to try and generate some pressure, before it says "oh, right, you took your mask off" and shuts off.

_________________
Mask: Bleep DreamPort CPAP Mask Solution
Additional Comments: S9 VPAP Auto
Last edited by palerider on Sat May 09, 2015 2:18 am, edited 1 time in total.
Get OSCAR

Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.

User avatar
Papit
Posts: 844
Joined: Sun Dec 18, 2011 4:47 pm
Location: USA CA

Re: ResMedAirSense Autoset 10 and on/off function

Post by Papit » Sat May 09, 2015 2:07 am

palerider wrote:
Guest wrote:First two days I had my Autoset I used the Auto off, and it would run at high pressure for a couple of seconds and shut off. Sleepyhead would report it as a leak and max the leak chart out at 120. I started turning off the machine manually because I didn't like the data getting messed up like that.
not that it really matters, but when you take the mask off, the machine runs at very low pressure, but high flow because it's trying to maintain whatever your set pressure is.

to have pressure, you have to have some restriction to the flow, like what the mask vents give you.

now, admittedly, the motor ramps up to full speed to try and generate some pressure, before it says "oh, right, you took your mask off" and shuts off.
Yup, that makes sense and, yeah, so what if the leak spikes when you take your mask off. It is a helluva "leak" all of a sudden. But the user knows what that ending leak spike on the graph is.
Help me out on this, though, palerider. I haven't looked at the graphs, but maybe you know in general: Are any of the (10-second) apnea data affected at the tail end of the graphs by the long slow-blowing air? (I think I recall that the recording software is designed to disregard certain 'obvious' anomalies at the beginning and ?at the end of a session.)

_________________
Mask: DreamWear Nasal CPAP Mask with Headgear
Additional Comments: Machine: AirCurve 10 ASV (37043), Software:ResScan 5.7.0.9477, SleepyHead V1.00BETA2, Oximeter:CMS-50i
Machine: AirCurve 10 ASV, Mask: AirFit N30i
PulseOx Data-transfer to OSCAR4-23-12http://tinyurl.com/nzd64gu
Wireless SD Card Data-transfer to OSCAR 8-14-15http://tiny.cc/z1kv8x

User avatar
palerider
Posts: 32299
Joined: Wed Dec 16, 2009 5:43 pm
Location: Dallas(ish).

Re: ResMedAirSense Autoset 10 and on/off function

Post by palerider » Sat May 09, 2015 2:20 am

Papit wrote:Are any of the (10-second) apnea data affected at the tail end of the graphs by the long slow-blowing air? (I think I recall that the recording software is designed to disregard certain 'obvious' anomalies at the beginning and ?at the end of a session.)
no, the machine can tell the difference between a mask off situation and a mask on but not breathing situation.

the software doesn't flag events, the machine does, the software just reports what the machine records.