Newbie with hose making put put sounds.

General Discussion on any topic relating to CPAP and/or Sleep Apnea.
Heather R
Posts: 15
Joined: Fri Jan 30, 2015 1:58 pm

Re: Newbie with hose making put put sounds.

Post by Heather R » Sun Feb 01, 2015 3:57 pm

PaleRider;

It looks like the link Kalisgram posted which is heated but I'll wait until tomorrow.

Thanks,

Heather R

User avatar
archangle
Posts: 9293
Joined: Sun Mar 27, 2011 11:55 am

Re: Newbie with hose making put put sounds.

Post by archangle » Sun Feb 01, 2015 9:02 pm

Rainout fixes.

1) Turn the humidity down.
2) Get a heated hose or a hose cover. A cold hose increases rainout.
3) Route the hose such that it slopes down constantly to the blower so that the water will run back into the humidifier. You can't allow any low spots. (Sort of hard to do.)
4) Route the hose up over something like the headboard so that it slopes up to the top and down to your mask. Any condensation will run back into the tank or down to your mask. Down to your mask sounds bad, but it comes out a drop at a time instead of collecting in a low spot and making noise.
5) Turn up the heat in the room.

_________________
Mask: Swift™ FX Nasal Pillow CPAP Mask with Headgear
Humidifier: S9™ Series H5i™ Heated Humidifier with Climate Control
Additional Comments: Also SleepyHead, PRS1 Auto, Respironics Auto M series, Legacy Auto, and Legacy Plus
Please enter your equipment in your profile so we can help you.
Click here for information on the most common alternative to CPAP.
If it's midnight and a DME tells you it's dark outside, go and check for yourself.

Useful Links.

User avatar
LSAT
Posts: 13350
Joined: Sun Nov 16, 2008 10:11 am
Location: SE Wisconsin

Re: Newbie with hose making put put sounds.

Post by LSAT » Sun Feb 01, 2015 9:30 pm

You could also put your CPAP on the floor. Any water in the hose should run downhill back to the humidifier.