Newbie with hose making put put sounds.

General Discussion on any topic relating to CPAP and/or Sleep Apnea.
Heather R
Posts: 15
Joined: Fri Jan 30, 2015 1:58 pm

Newbie with hose making put put sounds.

Post by Heather R » Sun Feb 01, 2015 1:08 pm

Hi I'm a newbie with slight nose bleeds. I put humidity up to 7 on my ResMed AirSense 10 and last night I woke up about 2:30 to sounds something like put- put. Turned machine off but sound continued waking hubby. I then pulled electric connection.

There's still water in chamber. Any suggestions as to what I'm doing wrong? I had been sleeping well.

Heather R

User avatar
OkyDoky
Posts: 2869
Joined: Mon Aug 25, 2014 5:18 pm

Re: Newbie with hose making put put sounds.

Post by OkyDoky » Sun Feb 01, 2015 2:04 pm

What it sounds like is that when you turned up the humidifier the moisture condensed in the tubing due to the room being cooler than the warm air. This can be decreased with a heated hose if you don't already use one, or a hose cozy. You can make your own out of tube socks or flannel or buy one. https://www.cpap.com/productpage/snuggl ... cover.html They come where they slide on with velcro closures at each end or one has a zipper to make it easier to put on.
ResMed Aircurve 10 VAUTO EPAP 11 IPAP 15 / P10 pillows mask / Sleepyhead Software / Back up & travel machine Respironics 760

User avatar
Sir NoddinOff
Posts: 4189
Joined: Mon May 14, 2012 5:30 pm
Location: California

Re: Newbie with hose making put put sounds.

Post by Sir NoddinOff » Sun Feb 01, 2015 2:11 pm

Yep... putt putt sounds are water collecting a loop of your hose, probably where it slumps down between your machine and bed.

_________________
Mask: AirFit™ F10 Full Face Mask with Headgear
Additional Comments: Sleepyhead software v.0.9.8.1 Open GL and Encore Pro v2.2.
I like my ResMed AirFit F10 FFM - reasonably low leaks for my ASV therapy. I'm currently using a PR S1 AutoSV 960P Advanced. I also keep a ResMed S9 Adapt as backup. I use a heated Hibernite hose. Still rockin' with Win 7 by using GWX to stop Win 10.

Heather R
Posts: 15
Joined: Fri Jan 30, 2015 1:58 pm

Re: Newbie with hose making put put sounds.

Post by Heather R » Sun Feb 01, 2015 2:54 pm

Added correct amount of water to humidifier and now it's making gurgling sounds while hose is quiet. Hate to lose another night because of Medicare rules limiting me to only 9 machine free night in 30 days.

Heather R

User avatar
palerider
Posts: 32299
Joined: Wed Dec 16, 2009 5:43 pm
Location: Dallas(ish).

Re: Newbie with hose making put put sounds.

Post by palerider » Sun Feb 01, 2015 3:02 pm

Heather R wrote:Added correct amount of water to humidifier and now it's making gurgling sounds while hose is quiet. Hate to lose another night because of Medicare rules limiting me to only 9 machine free night in 30 days.

Heather R
lay the end of the hose on the floor... see what happens.

_________________
Mask: Bleep DreamPort CPAP Mask Solution
Additional Comments: S9 VPAP Auto
Get OSCAR

Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.

Heather R
Posts: 15
Joined: Fri Jan 30, 2015 1:58 pm

Re: Newbie with hose making put put sounds.

Post by Heather R » Sun Feb 01, 2015 3:09 pm

Pale Rider;

Followed your instruction to lay end of hose on floor ,turned on machine and at 6 humidity setting water came out on floor. Does that mean humidity level too high , water caught in hose or? At least it didn't make a noise.

Heather R

User avatar
Pugsy
Posts: 65301
Joined: Thu May 14, 2009 9:31 am
Location: Missouri, USA

Re: Newbie with hose making put put sounds.

Post by Pugsy » Sun Feb 01, 2015 3:20 pm

Sounds like there was still some water in the hose left over from the condensation that occurred last night. It ran out on the floor...if the hose is now free of water things should quiet down.

Make sure to not over fill the chamber as sometimes when over full water can get blown into the hose from the chamber (at least with the S9 but I haven't seen a new AirSense to see how easy it is to blow back into the hose...I did that a couple of time with my S9 and it made an awful racket)...

If you have a heated hose...turn up the temperature so that condensation won't occur at your higher humidity needs.
If you don't have a heated hose...get one.
Or turn up the house furnace temp so that the bedroom air isn't so cool as to cause the air in the hose to cool and release moisture.

Condensation (we call it rain out) in the mask or hose...it's physics in action.
See my posts here as I go into more detail and explain the options. I think it is my 3rd post that addresses rain out in the hose.
viewtopic/t94035/Pugsys-Pointers-3Deali ... -road.html

_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier
Additional Comments: Mask Bleep Eclipse https://bleepsleep.com/the-eclipse/
I may have to RISE but I refuse to SHINE.

User avatar
palerider
Posts: 32299
Joined: Wed Dec 16, 2009 5:43 pm
Location: Dallas(ish).

Re: Newbie with hose making put put sounds.

Post by palerider » Sun Feb 01, 2015 3:26 pm

Heather R wrote:Followed your instruction to lay end of hose on floor ,turned on machine and at 6 humidity setting water came out on floor. Does that mean humidity level too high , water caught in hose or? At least it didn't make a noise.
you've identified the source of the noise now, if it happens again, lift the hose up, starting at your end, and let the water run back into the humidifier.

anything else I say would just be an echo of the excellent advice that Pugsy has given above

_________________
Mask: Bleep DreamPort CPAP Mask Solution
Additional Comments: S9 VPAP Auto
Get OSCAR

Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.

Heather R
Posts: 15
Joined: Fri Jan 30, 2015 1:58 pm

Re: Newbie with hose making put put sounds.

Post by Heather R » Sun Feb 01, 2015 3:29 pm

Thanks to all; After water was released machine and hose were quiet. I may have to get a heated humidifier although this machine says it has one but there are no controls that I can see for it.

PaleRider I will try that but also ask to try a full face mask covering my nose and mouth to get more air .

Heather R

User avatar
kaiasgram
Posts: 3569
Joined: Sat Jun 02, 2012 2:08 pm
Location: Northern California

Re: Newbie with hose making put put sounds.

Post by kaiasgram » Sun Feb 01, 2015 3:39 pm

Heather R wrote:Thanks to all; After water was released machine and hose were quiet. I may have to get a heated humidifier although this machine says it has one but there are no controls that I can see for it.

PaleRider I will try that but also ask to try a full face mask covering my nose and mouth to get more air .

Heather R
You have a heated humidifier already. A heated hose would help with the rainout: https://www.cpap.com/productpage/resmed ... ve-10.html

When you attach a heated hose you will have more options for temperature and humidification. Those options aren't available when a regular hose is attached.

_________________
Machine: AirSense 10 AutoSet with Heated Humidifer + Aifit N30i Nasal Mask Bundle
Additional Comments: SleepyHead-now-OSCAR software on Mac OSX Ventura

User avatar
Pugsy
Posts: 65301
Joined: Thu May 14, 2009 9:31 am
Location: Missouri, USA

Re: Newbie with hose making put put sounds.

Post by Pugsy » Sun Feb 01, 2015 3:42 pm

If you do have the heated hose there should be a section where the hose actually connects to electric circuit on the back of the machine where the hose attaches to the humidifier.
Non heated hoses have no electrical connections.
Your DME may have locked you out of changing the temperatures. That's easy to remedy.
If you don't have the clinical manual for your machine you can request it here
Just make sure to get the correct model.
Instructions down near the bottom of the page and the manual might help you figure out if you for sure already have the heated hose attached. The machine will recognize the heated hose automatically.
http://www.apneaboard.com/adjust-cpap-p ... tup-manual

_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier
Additional Comments: Mask Bleep Eclipse https://bleepsleep.com/the-eclipse/
I may have to RISE but I refuse to SHINE.

Heather R
Posts: 15
Joined: Fri Jan 30, 2015 1:58 pm

Re: Newbie with hose making put put sounds.

Post by Heather R » Sun Feb 01, 2015 3:47 pm

The heated hose you sent link to looks like mine. I'll have to call local company that I received machine etc. from tomorrow and ask some more questions. In directions it gives more options than are shown on my machine.

Thanks for the link.

Heather R

User avatar
palerider
Posts: 32299
Joined: Wed Dec 16, 2009 5:43 pm
Location: Dallas(ish).

Re: Newbie with hose making put put sounds.

Post by palerider » Sun Feb 01, 2015 3:48 pm

Heather R wrote:Thanks to all; After water was released machine and hose were quiet. I may have to get a heated humidifier although this machine says it has one but there are no controls that I can see for it.
you said earlier "6 humidity setting" <- that's the control for the heated humidifier.

_________________
Mask: Bleep DreamPort CPAP Mask Solution
Additional Comments: S9 VPAP Auto
Get OSCAR

Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.

User avatar
palerider
Posts: 32299
Joined: Wed Dec 16, 2009 5:43 pm
Location: Dallas(ish).

Re: Newbie with hose making put put sounds.

Post by palerider » Sun Feb 01, 2015 3:49 pm

Heather R wrote:The heated hose you sent link to looks like mine. I'll have to call local company that I received machine etc. from tomorrow and ask some more questions. In directions it gives more options than are shown on my machine.
is there a right angle connector on the machine end of the hose?

or does the hose stick straight out the back of the machine.

if there's a right angle connector, then you have a heated hose.

_________________
Mask: Bleep DreamPort CPAP Mask Solution
Additional Comments: S9 VPAP Auto
Get OSCAR

Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.

Heather R
Posts: 15
Joined: Fri Jan 30, 2015 1:58 pm

Re: Newbie with hose making put put sounds.

Post by Heather R » Sun Feb 01, 2015 3:53 pm

Pugsy'

I'll call representative tomorrow since I don't want to mess up machine. There is what looks like an electrical connection when I attached hose to machine but not sure.

Heather R