Sleep Master Wireless XPAP data transfer

General Discussion on any topic relating to CPAP and/or Sleep Apnea.
User avatar
308hjute
Posts: 12
Joined: Thu Mar 12, 2015 5:41 pm
Location: Sydney, Australia

Re: Sleep Master Wireless XPAP data transfer

Post by 308hjute » Mon Apr 20, 2015 1:25 pm

Hi all,

I set my new Flashair card up last night. Tried to transfer data this morning but am getting this error and no data being transferred.

Image

Any ideas on what have I done wrong?

_________________
Machine: ResMed AirSense™ 10 AutoSet™ CPAP Machine with HumidAir™ Heated Humidifier
Mask: Simplus Full Face CPAP Mask with Headgear
Additional Comments: SleepyHead and ResScan 5.3 software

User avatar
Jack_Russel_Fan
Posts: 104
Joined: Wed Apr 09, 2014 2:45 am
Location: Ellijay, GA
Contact:

Re: Sleep Master Wireless XPAP data transfer

Post by Jack_Russel_Fan » Mon Apr 20, 2015 9:48 pm

I would suggest trying Sleep Master data transfer again after another day or two of data is written to the FlashAir by your AS10.

Also, in order to help you further, you can contact me through the sleepmaster.org website and give me more info if this does not work.

Tom

_________________
Mask: TAP PAP Nasal Pillow CPAP Mask with Improved Stability Mouthpiece
Additional Comments: CMS-50F Pulse Oximeter, O2 when sleeping, SleepyHead, Sleep Master
Author of Sleep Master wireless xpap data transfer program.
Machine: PRS1 760P

yaconsult
Posts: 1099
Joined: Fri Nov 14, 2014 10:43 pm
Location: "Silicon Valley", CA

Re: Sleep Master Wireless XPAP data transfer

Post by yaconsult » Mon Apr 20, 2015 11:07 pm

What happens when you browse to the IP address assigned to the card? You should be able to see the folders, like in the picture below. This would verify that your card is connected to your network.
Image

_________________
Mask: AirFit™ P10 Nasal Pillow CPAP Mask with Headgear
Additional Comments: January 2015 Sleep Study Results: Apnea/Hypopnea Index (AHI): 80.2, Sleepyhead

User avatar
308hjute
Posts: 12
Joined: Thu Mar 12, 2015 5:41 pm
Location: Sydney, Australia

Re: Sleep Master Wireless XPAP data transfer

Post by 308hjute » Tue Apr 21, 2015 2:41 am

Hi Tom & yaconsult,

Thanks for the replies. Looks like the connection is ok, I can browse to the card and view folders. Looks like there is data in last night's folder.

Image

I'll give it a couple of more days and see what happens.

Cheers

_________________
Machine: ResMed AirSense™ 10 AutoSet™ CPAP Machine with HumidAir™ Heated Humidifier
Mask: Simplus Full Face CPAP Mask with Headgear
Additional Comments: SleepyHead and ResScan 5.3 software

User avatar
308hjute
Posts: 12
Joined: Thu Mar 12, 2015 5:41 pm
Location: Sydney, Australia

Re: Sleep Master Wireless XPAP data transfer

Post by 308hjute » Wed Apr 22, 2015 2:45 am

All sorted!

Updated to the latest version, which I thought I already had, and tweaked a couple of settings and now it is working perfectly.

Cheers again.

_________________
Machine: ResMed AirSense™ 10 AutoSet™ CPAP Machine with HumidAir™ Heated Humidifier
Mask: Simplus Full Face CPAP Mask with Headgear
Additional Comments: SleepyHead and ResScan 5.3 software

User avatar
JustinD9
Posts: 38
Joined: Tue Feb 17, 2015 8:25 pm

Re: Sleep Master Wireless XPAP data transfer

Post by JustinD9 » Wed Apr 22, 2015 3:32 pm

Just started setup and I am trying to find the config file as per the sleep master request. Any help appreciated . I have connected and downloaded so far. Thanks

_________________
Machine: DreamStation Go Auto CPAP Machine
Mask: AirFit™ P10 Nasal Pillow CPAP Mask with Headgear
Additional Comments: Oscar

User avatar
308hjute
Posts: 12
Joined: Thu Mar 12, 2015 5:41 pm
Location: Sydney, Australia

Re: Sleep Master Wireless XPAP data transfer

Post by 308hjute » Wed Apr 22, 2015 4:02 pm

Mine is created/modified and saved on my Desktop. This is where I run SM from.

_________________
Machine: ResMed AirSense™ 10 AutoSet™ CPAP Machine with HumidAir™ Heated Humidifier
Mask: Simplus Full Face CPAP Mask with Headgear
Additional Comments: SleepyHead and ResScan 5.3 software

User avatar
JustinD9
Posts: 38
Joined: Tue Feb 17, 2015 8:25 pm

Re: Sleep Master Wireless XPAP data transfer

Post by JustinD9 » Wed Apr 22, 2015 4:47 pm

Thanks. Found it!

_________________
Machine: DreamStation Go Auto CPAP Machine
Mask: AirFit™ P10 Nasal Pillow CPAP Mask with Headgear
Additional Comments: Oscar

User avatar
Krelvin
Posts: 1977
Joined: Tue Jun 06, 2006 5:23 pm
Location: Metro Phx Area - Dry Heat!

Re: Sleep Master Wireless XPAP data transfer

Post by Krelvin » Mon May 04, 2015 1:23 pm

Just an FYI... my original card stopped working. Or more specifically the WiFi stopped working. Using the toshiba firmware software I rebuilt the card, etc... and no WiFi still. The SD card part still working just fine, but no WiFi signal out of the card.

I eventually ordered a second card which works just fine.

I just setup an RMA for the first one, that took some looking... you need to use the following URL:
https://support.toshiba.com/support/acclaim/welcome

And in the middle of the page click on the Start button. You need to know all the purchase info etc.. but that would be normal.

BTW the new card I purchased arrived as a W-03 card instead of the W-02 which is what the first card was. It has some new features like the ability to tell the card to keep trying to connect to the AP instead of only trying once. It has been flawless since I got it.

And when I setup the new card I got the same screen:
Image

I then used the machine for about 40 mins and tried again and it worked. I'm using v2.0.1.33.

Not sure how long before I get the replacement card but not in a hurry.
Current Settings PS 4.0 over 10.6-18.0 (cmH2O) - Resmed S9 VPAP Auto w/h5i Humidifier - Quattro Air FFM
TNET Sleep Resource Pages - CPAP Machine Database
Put your equip in your Signature - SleepyHead v1.0.0-beta-1
Kevin... alias Krelvin

User avatar
palerider
Posts: 32304
Joined: Wed Dec 16, 2009 5:43 pm
Location: Dallas(ish).

Re: Sleep Master Wireless XPAP data transfer

Post by palerider » Mon May 04, 2015 1:34 pm

Krelvin wrote:BTW the new card I purchased arrived as a W-03 card instead of the W-02 which is what the first card was. It has some new features like the ability to tell the card to keep trying to connect to the AP instead of only trying once. It has been flawless since I got it..
I'd like some way to make the config files invisible to the s9 machines

_________________
Mask: Bleep DreamPort CPAP Mask Solution
Additional Comments: S9 VPAP Auto
Get OSCAR

Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.

User avatar
Krelvin
Posts: 1977
Joined: Tue Jun 06, 2006 5:23 pm
Location: Metro Phx Area - Dry Heat!

Re: Sleep Master Wireless XPAP data transfer

Post by Krelvin » Mon May 04, 2015 1:44 pm

My S9 doesn't seem to care, though if the power is lost, I need to redo the configuration on the card again. That has really only happened once due to a power failure. I've never seen it say that it is going to reformat the card, any card that I've put into it though.

I really love the ease of not having to pull (and forget) the SDCard to look at the contents. For the week or so that I didn't have it working, it was a pain going back to pulling the card etc..
Current Settings PS 4.0 over 10.6-18.0 (cmH2O) - Resmed S9 VPAP Auto w/h5i Humidifier - Quattro Air FFM
TNET Sleep Resource Pages - CPAP Machine Database
Put your equip in your Signature - SleepyHead v1.0.0-beta-1
Kevin... alias Krelvin

User avatar
Jack_Russel_Fan
Posts: 104
Joined: Wed Apr 09, 2014 2:45 am
Location: Ellijay, GA
Contact:

Re: Sleep Master Wireless XPAP data transfer

Post by Jack_Russel_Fan » Tue May 19, 2015 4:59 am

A Whole lot of changes!!!!!!!!!!!!!!!!!!

After becoming frustrated trying to make Sleep Master work across different OS platforms, I decided to make different versions for different
OS platforms:

v1.0.2.9 for Windows Vista and XP
v2.0.1.34 for Windows 7, 8, and 8.1. This is a new release!
v3.0.0.2 a Java program for Ubuntu 14. This is a new release!

http://www.sleepmaster.org

Tom

_________________
Mask: TAP PAP Nasal Pillow CPAP Mask with Improved Stability Mouthpiece
Additional Comments: CMS-50F Pulse Oximeter, O2 when sleeping, SleepyHead, Sleep Master
Author of Sleep Master wireless xpap data transfer program.
Machine: PRS1 760P

User avatar
Krelvin
Posts: 1977
Joined: Tue Jun 06, 2006 5:23 pm
Location: Metro Phx Area - Dry Heat!

Re: Sleep Master Wireless XPAP data transfer

Post by Krelvin » Tue May 19, 2015 5:11 am

Interesting, will check it out.
Current Settings PS 4.0 over 10.6-18.0 (cmH2O) - Resmed S9 VPAP Auto w/h5i Humidifier - Quattro Air FFM
TNET Sleep Resource Pages - CPAP Machine Database
Put your equip in your Signature - SleepyHead v1.0.0-beta-1
Kevin... alias Krelvin

User avatar
Jack_Russel_Fan
Posts: 104
Joined: Wed Apr 09, 2014 2:45 am
Location: Ellijay, GA
Contact:

Re: Sleep Master Wireless XPAP data transfer

Post by Jack_Russel_Fan » Sun May 24, 2015 3:04 am

A new version of Sleep Master is available for MAC OS X!

Sleep Master Java for MAC v4.0.0.3

http://sleepmaster.org

Tom

_________________
Mask: TAP PAP Nasal Pillow CPAP Mask with Improved Stability Mouthpiece
Additional Comments: CMS-50F Pulse Oximeter, O2 when sleeping, SleepyHead, Sleep Master
Author of Sleep Master wireless xpap data transfer program.
Machine: PRS1 760P

User avatar
Jack_Russel_Fan
Posts: 104
Joined: Wed Apr 09, 2014 2:45 am
Location: Ellijay, GA
Contact:

Re: Sleep Master Wireless XPAP data transfer

Post by Jack_Russel_Fan » Wed May 27, 2015 12:54 am

Just uploaded a new Sleep Master Java for Ubuntu Version v3.0.0.4.

Fixed a few bugs and added a progress indicator to entertain users while long operations are taking place.

http://sleepmaster.org

Tom

_________________
Mask: TAP PAP Nasal Pillow CPAP Mask with Improved Stability Mouthpiece
Additional Comments: CMS-50F Pulse Oximeter, O2 when sleeping, SleepyHead, Sleep Master
Author of Sleep Master wireless xpap data transfer program.
Machine: PRS1 760P