16$ for a years supply at costco.com, less in the stores.ChicagoGranny wrote: 1. 24-hour generic Zyrtec once in the morning. Shop around and you should find it cheap.
New CPAP user having problems.
Re: New CPAP user having problems.
Get OSCAR
Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.
Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.
- ChicagoGranny
- Posts: 15502
- Joined: Sun Jan 29, 2012 1:43 pm
- Location: USA
Re: New CPAP user having problems.
If you haven't seen him since you were diagnosed with sleep apnea, another visit is in order. He should scope you and examine the size and condition of the nasal turbinates, vocal cords, tonsils, adenoids, soft palate and tongue. Once they know you have obstructive sleep apnea, some ENTs take a different perspective on your case. (The really good ones look for signs and symptoms of OSA on the first damn visit.)Mike208 wrote:I have an ENT and he prescribed nasal spray that really doesn't help very long. I guess it lasts maybe 3 hours or so. He says I have no serious issues with the sinus area. One thing I just realized is that he has never suggested a prescription allergy medicine. All I have ever taken was over-the -counter and it doesn't work well.
"It's not the number of breaths we take, it's the number of moments that take our breath away."
Cuando cuentes cuentos, cuenta cuántas cuentos cuentas.
Cuando cuentes cuentos, cuenta cuántas cuentos cuentas.
Re: New CPAP user having problems.
If you have a ramp of 45 minutes it doesn't sound like you are at therapeutic pressure for very long. I started out with a bilevel machine from the titration study after I had a hard time exhaling and history of GERD. After seeing some comments here about how sometimes you have to fail other machines and be persistent to get one, I consider myself lucky.
With my pressures of 15/11, I quit the ramp after a couple of weeks because it is a comfort feature and I didn't find it helpful, I have had only one episode of GERD since I started 5 months ago.
Seems like a discussion with your ENT about a bilevel machine would be helpful.
With my pressures of 15/11, I quit the ramp after a couple of weeks because it is a comfort feature and I didn't find it helpful, I have had only one episode of GERD since I started 5 months ago.
Seems like a discussion with your ENT about a bilevel machine would be helpful.
_________________
| Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier |
| Mask: AirFit™ P10 Nasal Pillow CPAP Mask with Headgear |
| Additional Comments: Sleepyhead Software |
Last edited by OkyDoky on Tue Jan 06, 2015 11:13 am, edited 1 time in total.
ResMed Aircurve 10 VAUTO EPAP 11 IPAP 15 / P10 pillows mask / Sleepyhead Software / Back up & travel machine Respironics 760
Re: New CPAP user having problems.
That can be true. I learned a long time ago what assume means. LOL Yet, I still do it.palerider wrote:that's a dangerous assumption.Mike208 wrote:I'm not either but I assumed she knew what she was doing. I could tell there was a change in temp and humidity. It seemed to make the congestion worse.tan wrote:1) not sure how ramp can help with air swallowingMike208 wrote:Well, I had a visit with the sleep specialist this morning. She said the high pressure was definitely the cause of the air in the stomach. It was lowered to 11 and the ramp time was changed to 45 min. instead of auto.
_________________
| Machine: ResMed AirSense™ 10 AutoSet™ CPAP Machine with HumidAir™ Heated Humidifier |
| Mask: Quattro™ FX Full Face CPAP Mask with Headgear |
Re: New CPAP user having problems.
He knows about the apnea. He's the one that was pushing for me to get on CPAP. He has run a scope through my nose a couple of times and said nothing stood out that would be aided by surgery.ChicagoGranny wrote:If you haven't seen him since you were diagnosed with sleep apnea, another visit is in order. He should scope you and examine the size and condition of the nasal turbinates, vocal cords, tonsils, adenoids, soft palate and tongue. Once they know you have obstructive sleep apnea, some ENTs take a different perspective on your case. (The really good ones look for signs and symptoms of OSA on the first damn visit.)Mike208 wrote:I have an ENT and he prescribed nasal spray that really doesn't help very long. I guess it lasts maybe 3 hours or so. He says I have no serious issues with the sinus area. One thing I just realized is that he has never suggested a prescription allergy medicine. All I have ever taken was over-the -counter and it doesn't work well.
_________________
| Machine: ResMed AirSense™ 10 AutoSet™ CPAP Machine with HumidAir™ Heated Humidifier |
| Mask: Quattro™ FX Full Face CPAP Mask with Headgear |
Re: New CPAP user having problems.
It sounds like, to me, a biPAP might be the thing I need. I suggested that yesterday and she said she didn't think it was necessary. But, if one of my other doctors agrees that it's necessary I will get it. It's not her call.OkyDoky wrote:If you have a ramp of 45 minutes it doesn't sound like you are at therapeutic pressure for very long. I started out with a bilevel machine from the titration study after I had a hard time exhaling and history of GERD. After seeing some comments here about how sometimes you have fail other machines and be persistant to get one, I consider myself lucky.
With my pressures of 15/11, I quit the ramp after a couple of weeks because it is a comfort feature and I didn't find it helpful, I have had only one episode of GERD since I started 5 months ago.
Seems like a discussion with your ENT about a bilevel machine would be helpful.
_________________
| Machine: ResMed AirSense™ 10 AutoSet™ CPAP Machine with HumidAir™ Heated Humidifier |
| Mask: Quattro™ FX Full Face CPAP Mask with Headgear |
- ChicagoGranny
- Posts: 15502
- Joined: Sun Jan 29, 2012 1:43 pm
- Location: USA
Re: New CPAP user having problems.
Sounds like you have one of the good ones.Mike208 wrote: He's the one that was pushing for me to get on CPAP. He has run a scope through my nose a couple of times and said nothing stood out that would be aided by surgery.
"It's not the number of breaths we take, it's the number of moments that take our breath away."
Cuando cuentes cuentos, cuenta cuántas cuentos cuentas.
Cuando cuentes cuentos, cuenta cuántas cuentos cuentas.
-
Sleeprider
- Posts: 1562
- Joined: Tue May 06, 2008 5:57 pm
- Location: Murrysville, PA
Re: New CPAP user having problems.
Now that is the right attitude! Do YOU know how to access the clinical menus of your device to change settings? Your sense of air starvation early in therapy is from the ramp being too long at too low of a pressure. Also, you can take control of the pressure settings and try to find that threshold where you're not dumping air in your stomach, and if you will download the software already discussed, you can determine if the impact on you therapy is positive or not. Your experience, as you passively wait for the "experts" to get it right, is what drove most of us to take our therapy in hand and do something positive. Someone has to say it, stop being so passive and waiting for someone else to do what you yourself can do today.Mike208 wrote:It sounds like, to me, a biPAP might be the thing I need. I suggested that yesterday and she said she didn't think it was necessary. But, if one of my other doctors agrees that it's necessary I will get it. It's not her call.OkyDoky wrote:If you have a ramp of 45 minutes it doesn't sound like you are at therapeutic pressure for very long. I started out with a bilevel machine from the titration study after I had a hard time exhaling and history of GERD. After seeing some comments here about how sometimes you have fail other machines and be persistant to get one, I consider myself lucky.
With my pressures of 15/11, I quit the ramp after a couple of weeks because it is a comfort feature and I didn't find it helpful, I have had only one episode of GERD since I started 5 months ago.
Seems like a discussion with your ENT about a bilevel machine would be helpful.
_________________
| Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier |
| Mask: AirFit™ P10 Nasal Pillow CPAP Mask with Headgear |
| Additional Comments: Sleepyhead software. Just changed from PRS1 BiPAP Auto DS760TS |
Re: New CPAP user having problems.
Thanks! I'm tired of people screwing around, taking their sweet time, while I sit here suffering. I'm currently going through this with the cervical issues I have. Wait, wait, wait, wait. I do know how to access the clinical menu. I have gone in and reset the ramp, lowered the pressure to 8, and put the temp and humidity back to where it was. I'll try this tonight and see what happens. I refuse to wait any longer.Sleeprider wrote:Now that is the right attitude! Do YOU know how to access the clinical menus of your device to change settings? Your sense of air starvation early in therapy is from the ramp being too long at too low of a pressure. Also, you can take control of the pressure settings and try to find that threshold where you're not dumping air in your stomach, and if you will download the software already discussed, you can determine if the impact on you therapy is positive or not. Your experience, as you passively wait for the "experts" to get it right, is what drove most of us to take our therapy in hand and do something positive. Someone has to say it, stop being so passive and waiting for someone else to do what you yourself can do today.Mike208 wrote:It sounds like, to me, a biPAP might be the thing I need. I suggested that yesterday and she said she didn't think it was necessary. But, if one of my other doctors agrees that it's necessary I will get it. It's not her call.OkyDoky wrote:If you have a ramp of 45 minutes it doesn't sound like you are at therapeutic pressure for very long. I started out with a bilevel machine from the titration study after I had a hard time exhaling and history of GERD. After seeing some comments here about how sometimes you have fail other machines and be persistant to get one, I consider myself lucky.
With my pressures of 15/11, I quit the ramp after a couple of weeks because it is a comfort feature and I didn't find it helpful, I have had only one episode of GERD since I started 5 months ago.
Seems like a discussion with your ENT about a bilevel machine would be helpful.
_________________
| Machine: ResMed AirSense™ 10 AutoSet™ CPAP Machine with HumidAir™ Heated Humidifier |
| Mask: Quattro™ FX Full Face CPAP Mask with Headgear |
-
New guest
Re: New CPAP user having problems.
I had a cervical fusion years ago. I still have nerve damage with my face and neck. I still choke on food and liquid. I have a resumed air sense 10 Autoset. I have the H5 I humidifier, setting on auto which is 81 and 4. I use the P10 mask. My pressures are 12 -18. I have air filling my stomach every time I use the CPAP. I was told by the DME to use the maximum ramp. I start to fill up with air after 30 minutes on ramp. I have tried sitting up but get the same result. Pain in my chest, stomach and intestines. I have seen the gastroenterologist. I took a medication to relax the esophogus but I still have the filling with air problems. I have no help to offer just letting you know I have the neck fusion and nerve damage problem. I am following your posts to see if there are any other ideas to help with filling with air. I am actively seeing doctors to get help too.
-
Sleeprider
- Posts: 1562
- Joined: Tue May 06, 2008 5:57 pm
- Location: Murrysville, PA
Re: New CPAP user having problems.
Way to go Mike!Do you have EPR dialed in at 3?
_________________
| Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier |
| Mask: AirFit™ P10 Nasal Pillow CPAP Mask with Headgear |
| Additional Comments: Sleepyhead software. Just changed from PRS1 BiPAP Auto DS760TS |
Re: New CPAP user having problems.
Thanks! Yes, it is on 3. Tube temp back to 81. Humidity back to 6. Pressure at 8.Sleeprider wrote:Way to go Mike!
Do you have EPR dialed in at 3?
_________________
| Machine: ResMed AirSense™ 10 AutoSet™ CPAP Machine with HumidAir™ Heated Humidifier |
| Mask: Quattro™ FX Full Face CPAP Mask with Headgear |
- ChicagoGranny
- Posts: 15502
- Joined: Sun Jan 29, 2012 1:43 pm
- Location: USA
Re: New CPAP user having problems.
I wonder what you do for a living? You seem to know how to keep a lot of balls in the air.
"It's not the number of breaths we take, it's the number of moments that take our breath away."
Cuando cuentes cuentos, cuenta cuántas cuentos cuentas.
Cuando cuentes cuentos, cuenta cuántas cuentos cuentas.
Re: New CPAP user having problems.
Balls in the air? No, I can't juggle. I tried that with chainsaws once. The results were not impressive. Just kidding.ChicagoGranny wrote:I wonder what you do for a living? You seem to know how to keep a lot of balls in the air.
_________________
| Machine: ResMed AirSense™ 10 AutoSet™ CPAP Machine with HumidAir™ Heated Humidifier |
| Mask: Quattro™ FX Full Face CPAP Mask with Headgear |
Re: New CPAP user having problems.
I'm actually retired, more or less. I went on disability for back injuries suffered in 1992. My body has been steadily falling apart ever since.
_________________
| Machine: ResMed AirSense™ 10 AutoSet™ CPAP Machine with HumidAir™ Heated Humidifier |
| Mask: Quattro™ FX Full Face CPAP Mask with Headgear |